I have an HP5065A Rubidium Frequency Standard which is not working.

All of the oven temperatures look OK.
The power supply voltages look OK.
The ~60 MHz going into the RVFS package looks OK.
The 5.000 MHz and the ~5.315 MHz signals are within a few Hz.  (The unit is 
running open loop only.)
The A7 AC amp looks to be working OK.
The A8 137/234 Hz signals look OK.
There is 20 VDC feeding the RVFS lamp oscillator.

But, there is no current coming from the photodiode.  (Maybe a few picoamps of 
dark current, certainly not the expected ~50 microamps of DC, and no 137/234 Hz 
signals).

It seems the most likely problem is that the Rb lamp is not coming on.  Maybe 
possible that the photodiode is faulty, or a broken wire, etc.

Any other possibilities I should check before setting off into the innards of 
the RVFS package?

Thanks in advance!

Cheers,

Scott