I did a simple test run. driving the 1PPS input with a Tbolt..  over 2 hours 
the reported PRS case temperature varied 0.8 C.  The TT time tag varied 60 ns 
(75 ns / degree).   The FC frequency control word varied the frequency by  
0.231 ppt (0.300 ppt / degree).   All my high res counters are tide up doing 
other things, so I could not measure the actual freq.

It looks like the PRS10 time tag processor / interpolator is rather crappy.   I 
need to feed it the 1PPS from a 5071A to eliminate any temperature effects on 
the Tbolt.

I did modify the message polling code to read the case temperature every 
second.  That increased the time needed to get all the data values from the 
PRS10 to 35 seconds... a 10 second increase,   You can only get around 5 values 
per second from the PRS10 and there are around 50 values it can report.  If 
three of those 5 values per second are TT, FC, and AD10 that leaves room for 
getting 2 of those 50 values per second.

-----------

> Have you done a least-square fit to determine the sensitivity?