I connected a Tbolt to the 1PPS input of a Symmetricom SA22.c rubidium 
oscillator and ran Heather's software discipling code on it.   For some strange 
reason, you do get  better performance using a clean 1PPS rather than the 
crappiest one you can find ;-)

Anyway some interesting results came out.  White traces are the DDS frequency 
offset Heather is using to steer the SA22.c output freq,  yellow are 
temperature, magenta is the difference between the 1PPS in and 1PPS out (16.666 
ns res of the SA22.c TIC).   The straight white and yellow traces are linear 
regression trend lines of those parameters.   A little math shows a clear 
dependence of the SA22 output freq on temperature of around 2.664E-11 parts per 
degree C.  This sort of agrees with a plot in the SA22.c data sheet that shows 
the TC vs ambient temperature... if "ambient temperature" means the baseplate 
temp reported by the unit.