Hi

The “gotcha” with compensated devices is that there can be interesting 
breakpoints or decoupling to the sensor. Those show up more if the compensation
is working very hard to get the job done. Put another way - if you start at 
1x10^-9/C 
and compensate to 1x10^-11/C that is very different than starting at 3x10^-11/C.
A pretty good Rb or DOCXO would be a 1x10^-10 -30 to +70 spec. That would get 
you
into the 1x10^-12/C range. A 1x10^-9 part over the same range would be a good 
single
oven or a lower spec Rb. One always needs to be careful with those sort of 
translations. 
They may be optimistic by a factor or 3 or 6 or ….

Bob 