If one does a ratiometric measurement comparing the voltage drop across the RTD 
with the voltage drop across a stable low Tc resistor connected in series with 
the RTD the excitation source only needs to be quiet with good short term 
stability.

Bruce

> On 29 September 2020 at 07:48 "John Moran, Scawby Design" 
> <j...@scawbydesign.co.uk> wrote:
> 
> 
> Thanks for going easy on me Bob ... a case of more haste, less speed! I 
> focussed on low long-term drift specs without realising I had turned up a 
> voltage reference, sorry.
> 
> However, I have found some YSI glass encased thermistors that have long-term 
> drift specs of <10mK at 25C and 75C over a period of 100 months. They are in 
> the YSI 46000 series - data sheet attached.
> 
> http://www.farnell.com/datasheets/169207.pdf
> 
> There is an interesting paper by NIST on achieving the International 
> Temperature scale - link attached (it is 196 pages and 10MB) that seems to 
> indicate platinum sensors are the most stable at less than 1mK and, of 
> course, to be able to measure these resistors accurately, you need an equally 
> low-drift voltage/current source. :-)
> 
> https://nvlpubs.nist.gov/nistpubs/Legacy/TN/nbstechnicalnote1265.pdf
> 
> This reference appeared from an EEVblog where I think some Volt-nuts were 
> discussing temperature. One of them confirmed that the most economical way 
> was to have a group of lower-cost sensors and characterise them.
> 
> https://www.eevblog.com/forum/metrology/long-term-stability-of-temperature-sensors/
> 
> John 
> 
> 
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to 
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.

_______________________________________________