I thought is was closer to 800 RBC's

24ga. (arrow) lumen = 0.022 inches

RBC's = (6) to 8 angstroms = 0.000031 inch

 

Peter Marino RN BSN

(LIJ IV Team)

http://www.arrowintl.com/products/critical_care/faqs2/MISCdefault.asp

http://www.answers.com/topic/angstrom-1

 

---- Original Message -----

From: "Bartholomew , Jerry A (SPO)" <[EMAIL PROTECTED]>

Date: Thursday, December 22, 2005 2:01 pm

Subject: RE: BLOOD ADMINISTRATION

> I have personally transfused several units through a 24 gauge
> catheter.I did some research on the relative diameters of red
> blood cells, and
> the 24 gauge catheter.  The catheter diameter is equal to the
> width of
> 60-80 red blood cells.  Under normal pressures, this will not cause
> hemolysis.  The limiting factor is the speed of the transfusion.  If
> you're working the ER and you need to slam the unit into a trauma
> victim, then the #24 is not for you.  But if you are satisfied with
> rates of 200-400ml per hour, you shouldn't have any problems,
> because a
> #24 will accommodate up to 1200ml/hour.  Just as an aside, I also
> infusequite a bit of IVIG, and Prolastin, and use the #24s almost
> exclusively.You can also find support for this in the manual put
> out by the American
> Association of Blood Banking.
>
>
> Jerry Bartholomew RN BSN CRNI
> Vascular Access Specialist
> Veterans Affairs Medical Center
> Spokane, Washington