Robbin,

The syringe and the cap are each independent issues.

re the syringe
Basically what happens there is that the silicone rubber tip on the syringe plunger (the black part) can get compressed if you push hard on the plunger at the end of the flush. Then when you release the pressure on the plunger, that rubber tip bounces back. That bounce back can cause blood reflux.

BD's Posi-Flush is a syringe that is designed to minimize the compression/bounce. 
However, if you push really hard, you can still make it happen. 
Other syringe brands compress quite easily and almost inevitably cause bounce back / blood reflux.

The solution is to never fully compress the plunger or at least never push hard at the completion of the flush.

Unfortunately, there are folks who were taught  long ago to press hard on the plunger as they pulled the syringe and cannula out of rubber diaphragm / split septum type caps and who do not understand that the technique causes problems with valve type caps.

The positive pressure effect of the CLC2000 and other positive pressure caps happens only during the process of disconnecting the syringe from the valve. Positive pressure caps essentially store some of the flush fluid, which is then released and pushed down stream as the syringe is unscrewed from the cap and the valve closes.

Nurses can defeat the positive pressure caps by:

Clamping before disconnection - prevents the positive pressure effect from occurring and in some products, prevents the valve from moving outward to the closed position.

Placing those red/blue dead end caps on the positive pressure cap when the IV is not in use. They do this with the good intention of keeping the cap clean. However, when you connect TO a positive pressure cap there is always blood reflux into the catheter tip. So the lumen clots as it is sitting there with that reflux until the next dose is due.

/Martha

On Feb 25, 2006, at 4:47 PM, [EMAIL PROTECTED] wrote:

Since I started this thread I am very interested in your questions--can you be brand specific--Are you speaking about the BD posi-flush saline syringe?--We are currently using this product in conjunction with the CLC which is a positive displacement mechanical valve and have had countless withdrawl occlusion issues--Can the physcists or mechanical engineer types out there please explain if these products are working against each other?????????
 
--
Robbin K. George RN
Vascular Access Resource
Alexandria Hospital Virginia
 
-------------- Original message --------------
From: [EMAIL PROTECTED]
Hi group:
A question for those of you using this device:
 
Do you know for sure the rebound reflux that can be caused by certain syringes when they are 'bottomed out' at the end of the flush cycle is mitigated by this product design?
 
If you DO know that, how was it proven to you?
 
thanks
 
-------------- Original message --------------
From: "Martha Pike" <[EMAIL PROTECTED]>

> We are having great success with the InVision-Plus.
> One of the advantages is that it is not technique dependent.
> It does not matter when or if you clamp.
> We are using a saline flush protocol for everything except implanted
> ports and having very few problems.
> My colleague in CT has some pre and post statistics re occlusion that
> she may be will ing to share when she returns from vacation
> It also seems to have a greatly improved barrier to prevent bacteria
> from entering the fluid pathway compared to other products. The bug
> glow test is very impressive. (However, there are no completed
> studies r/t BSI as yet.)
> /Martha
>
> On Feb 21, 2006, at 10:54 PM, [EMAIL PROTECTED] wrote:
>
> > Have been reviewing some old emails I copied for personal archive--
> > In early December Martha Pike wrote a lengthy and glowing
> > evaluation of the Rymed Invision-Plus Neutral injection cap--If she
> > could please give us an update and allow us to use her comments for
> > benchmarking purposes we would be most grateful--We are having
> > tremendous issues with the CLC in our ICU and would like to
> > recommend a change--I have visited the Web site and received an
> > information packet from the company but could use some first hand
> > professional support and end user input--Anyone else out there
> > using this product please contribute your thoughts to the thread--
> > Thanks to everyone in advance
> >
> > --
> > Robbin K. George RN
> > Vascular Access Resource
> > Alexandria Hospital Virginia
>
>

Reply via email to