Heather,
The Sherlock Tip Location System IFU is your best source
for information as the IFU has been approved by the FDA and is therefore covered
by extensive testing. Here are the two statements from the IFU that you
should know:
"The Sherlock detector can be used on pateints
who have implanted cardiac rhythm devices such as pacemakers and
defribillators. When a cardiac rhythm device is present, it is recommended
that the Sherlock tip location system stylet be placed on the contralateral
side."
"Temporary disruption of the cardiac rhythm
device may occur is the Sherlock tip location stylet passes within 1cm of the
cardiac rhythm device. Use care if placing the Sherlock tip location
stylet on the same side as the cardiac rhythm
device."
Keep in mind, we are speaking of the actual pacer or defib
box, not the pacer wires. So first choice is to place the PICC on the
contralateral side. If this can't be avoided, check to see where the pacer
box is. If it is within 1cm of the subclavian, you should consult the
patient's cardiologist before placing a PICC while using Sherlock. It is
also a good idea to have this patient on an EKG monitor during the
procedure. Bottom line, if the Sherlock does affect the pacer--and that's
a big if because the magnetic signal put out by Sherlock is so minor that if it
does affect the pacer, it will only be a temporary disruption that will
clear once the catheter has moved away from the pacer
box.
Please let me know if you have additional questions.
Thanks, Paul
![]()
Confidentiality Notice: This e-mail and any attachments are intended only for the use of those to whom it is addressed and may contain information that is confidential and prohibited from further disclosure under law. If you have received this e-mail in error, its review, use, retention and/or distribution is strictly prohibited. If you are not the intended recipient, please contact the sender by reply e-mail and destroy all copies of the original message and any attachments.[v1.0]
From: [EMAIL PROTECTED]
[mailto:[EMAIL PROTECTED] On Behalf Of Heather
Nichols
Sent: Thursday, August 10, 2006 6:00 AM To: [EMAIL PROTECTED] Subject: Info needed Question for any of you who have used the Sherlock with Bard. I
was told that the Sherlock was NOT contraindicated for use when a patient has a
pace maker or electrical device of some kind in their chest. I have
recently been told that you must be careful not to get the device (the Sherlock)
anywhere near the implant, and as long as you do not, you are ok to use
it. Does anyone have the actual manufacturer recommendation? I hate
to give out incorrect information. Paul? Are you out there?
Kathy? Do you know?
Heather Nichols RN BSN CRNI
Infusion Services University of Louisville Trauma Institute 530 S. Jackson St. Lou. Ky. 40202 (502)562-3530
This message, including any attachments, is
confidential, intended only for the named recipient(s) and may contain
information that is privileged or exempt from disclosure under applicable law,
including PHI (Protected Health Information) covered under the Health Insurance
Portability and Accountability Act (HIPAA) of 1996. If you are not the intended
recipient(s), you are notified that the dissemination, distribution, or copying
of this message is strictly prohibited.
If you receive this message in error, or are not the named recipient(s),
please notify the sender or contact the University of Louisville Health
Care I.S. helpdesk at 502.562.3637 to report an inadvertently received
message. ----------------------------------------------------- |
- Info needed Heather Nichols
- RE: Info needed Blackburn, Paul