Dave,

Thats the question. It depends on the quality of the cells where the AC calibration info is stored.
I'm not aware of a checksum function on these EAROM's.

If it is the AC part, I've some spare parts if necessary... Just let me know.

Kind regards, Wim.

Wim de Jong
[email protected]

Op 7-12-2012 10:25, David C. Partridge schreef:
Wim

I don't think that is likely, as Phil clearly says that all the DC ranges are 
correct,  it is just the AC ranges that are a problem.  If the EAROMs were bad, 
you'd lose all the calibration constants.

Dave
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf 
Of Wim de Jong
Sent: 07 December 2012 08:04
To: [email protected]
Subject: [volt-nuts] Fwd: Solartron 7150plus - Chasing AC gremlins

Phil,
I suspect it to be the EAROM(s) IC 105/6 of this meter. Losing its calibration 
constants.
Regards, Wim.

Wim de Jong
[email protected]


_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.