Back to a few of the original issues. A big questionh would be just how accurate the average volt nut needs the 732A to be? I once owned two 732A, one working normally, and kept hot 24/7, and one that needed some work with the charging circuits and new batteries. My 3458A being newly cal'd at Agilent in Colorado Springs, I matched the two 732As to the 3458A, then powered off the one with problems, worked on it off and on over several weeks, then powered it back up. What I found was that the one I worked on (after 48 hours warm-up) was within 0.2ppm of the hot one and the 3458A. Over the next year, the drift got a little larger among the three, but not by much -- a spread of about another 0.2ppm if I remember right.

That's when I decided that since I wasn't going to try to be a standards lab, shipping a 732A cold for cal was much easier and effective than trying to do it hot, and I live about 30 air miles from Fluke in Everett.

Can you live with 1ppm uncertainty in your 732A? Some here obviously can't, but I think many of us can and would be happy with that level of accuracy -- certainly good enough to cal the various 6-1/2 digit meters around, including the 3457 mentioned originally. And given my experience of 732A stability, especially the good old ones that come on ebay and the like, that 1ppm uncertainty is also a good check on the state of cal of even the 8-1/2 digit boxes, which despite their resolution, are not spec'd long-term for even 1ppm at 10V.