Thomas wrote:

You may find a local lab with less accreditation charging half then price that is fully capable of of calibrating to the limit of the 732A but cannot document to the level of a primary standards lab.

Very, very doubtful. Very few cal labs have a 732A or equivalent, much less anything better. The only labs with *better* uncertainty than a properly working 732A are those with JJAs. If you look at the NIST NVLAP accreditation list and run down it, looking at each lab's "Scope of Accreditation," you will only find 4 or 5 labs on the list with better uncertainty than a properly working 732A (I'm not sure you will find *any* on the A2LA accreditation list, but I haven't run down it lately). The Fluke cal lab and the Los Alamos and Sandia standards labs are three of those four or five (plus, of course, NIST itself). Boeing (Seattle) is another. Interestingly, you will find many labs that are rigorously accredited to only .003% or so (30 ppm), because the best voltage standard they own is an HP 34401A DMM. Even the HP Houston cal lab is certified to only 0.0007%, or 7 ppm (using a Fluke 5700A calibrator).

Yes a 1-2PPM Cals is not as sexy as a .1PPM Cal but in the real world the results when used in you home lab my be the same.

To get a calibration with an uncertainty of 1 or 2 ppm, the lab would need, at a minimum, a 732A or 732B to compare with (as well as a 720A Kelvin-Varley bridge, or equivalent, and a null meter that can reliably be read to 0.1uV, if you want the calibration certified to 1 or 2 ppm at voltages other than 10v). I don't think there are even ten labs on the NVLAP list that claim to have a 732A or B (the equipment used is often listed in the "remarks" column).

It does not take long to run down the whole list -- it's a short list and the "Scope of Accreditation" documents load fast. I recommend the exercise, to get a feel for what's out there. Same with the A2LA list, but it is longer and not as well organized and it usually takes 2 or 3 steps (running off to the lab's site) to get to the "Scope of Accreditation." (If you look at A2LA labs, pay attention to the lab class and only look at "open" commercial labs -- the non-commercial ones do not take in third-party calibration work.)

A list of NVLAP-accredited labs can be found here:

<http://ts.nist.gov/standards/scopes/dclow.htm>

There seems to be this myth of cal labs that can do just as good a job as the expensive, accredited labs, but don't bother with accreditation so they are much cheaper. First, note that to do a job as good as an expensive, accredited lab, any lab would have to do the same documentation as the accredited lab. If there is no documentation, there can be no claim as to the calibration's uncertainty. Having done the documentation, which is the time-consuming (thus, expensive) part, no commercial cal lab is going to do without the accreditation (which is nothing but an audit of the lab's procedures and documentation). I stress again -- if there is no documentation, there can be no claim as to the uncertainty of a lab's work. And since the documentation is the part that contributes most to the cost, there simply are not any commercial labs that can claim to have uncertainties on par with accredited cal labs, but are not themselves accredited.