On 8/25/2014 9:54 PM, Don@True-Cal wrote:
Why?

Let me count the ways.You can never count on any Seebeck voltage to
be immediately offset, there are far too many variables.
...
> A set of
5440A-7002 (banana plug) cables comes with this calibrator
(5440A-7003 spade lugs for 5720A)

For those leads, Fluke says they have the advantage of "minimizing the emf caused by temperature difference between the ends of the leads." I don't have much argument with that - if there's a temperature difference between the ends, you won't have an equal thermocouple offset, so minimizing such effects makes sense.

But, since those leads use pure copper wiring ("Belden 8262 or 8719 or equivalent"), how does having only the contact itself made of tellurium copper (TC) make a difference? Shouldn't they be TC end-to-end, with gas tight crimped connections? Does the tinning on that copper cable have an effect? What's the Seebeck coefficient for a copper to tellurium copper connection?

You said that "Tellurium Copper is usually not used for a device's terminal posts but used as the lead wire..." But, Fluke does exactly the opposite - those Fluke leads use TC connectors and are characterized "while engaged in a five-way binding post of Tellurium Copper Alloy 145, half hard." Since they use pure copper cabling, I'd think they'd work even better with a pure copper binding post, since the offsetting thermocouples would be in close proximity, and not at different ends of the cable.

You earlier said "Any type of Silver or Gold plating on the terminal or wire will introduce the undesirable dissimilar metal properties, both at the plating junction and at the plating metal to DUT terminal." Now you emphasize that by saying "If someone used a set of gold-plated interconnects for this procedure, they would be laughed out of the lab and the calibrator would be useless until recalibrated properly..."

But, the specs for the Fluke 5440A-7003 which you say are appropriate for the 5720A state: "Connector materials: Copper space lug with gold plating over silver plating." (Both platings!!) Is someone to be laughed out of the lab for using the cables they're supposed to use? The 5440A-7002 specs aren't clear, but if they're not plated, are you required to clean them of copper oxide for every use? It seems that would be necessary, since a gold-copper thermocouple is ~0.5 uV/C, while a copper-copper oxide one is ~1000 uV/C.

You seem to be stating one thing, but then giving evidence which contradicts your claim. It seems that Fluke uses TC where there's thermal mass, and pure copper where there's little, and uses plating in addition.

Please clarify, this isn't making sense to me.