well, I do that with my references if they cannot be adjusted reasonably such 
as the 732a. but for meters and calibrators I do not. 
achievieng high precison with such an instrument then means you still need to 
characterize it (i.e. calibrate, and not adjust it, doesnt save money 
essentially)
and secondly you always need to apply a transfer factor, i am too lazy for 
that, i rather get it adjusted when i cal it.
also, the reference is stable, true, but other things do drift, and we talk 10+ 
yeras here (last cal sticker i saw). it would certainly be interesting to know 
where its different ranges went inbetween (if it was known that it was adjusted 
to within spec back then, which probably is not known)


> Gesendet: Mittwoch, 08. Oktober 2014 um 10:32 Uhr
> Von: "Mike S" <mi...@flatsurface.com>
> An: volt-nuts@febo.com
> Betreff: Re: [volt-nuts] Old HP3458A - SN: 2823A 03939
>
> On 10/8/2014 4:23 AM, acb...@gmx.de wrote:
> > the EPROMs are in sockets, no soldering needed.
> > but again, buying a precision instrument but reprogramming cal data that is 
> > years old does not make any sense.
> > unless of course if you are just a collector and do not use its accuracy.
> 
> It makes perfect sense, for the same reason that HP doesn't touch the 
> cal if it's in spec - for tracking/characterization. By keeping the same 
> cal constants, if and when he does send it in for calibration he'll be 
> able to know how much it drifted since it was last cal'd (25 years ago?).
> 
> -- 
> Mike
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
> 
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.