Actually it's a 3.6V (SL-750/T) one. It still has ~3.68V, and no NVRAM checksum errors are reported.

It's a no opts unit (only DCV), so the errors are not related to resistance measurement. As Hubert (Dankeschön) pointed out, it's related to the int cal src drifting outside certain ranges. As Laurence (thank you) observed, it probably would be fixed with the int src cal characterization after ext cal, if nothing is physically wrong.

The manual specifies:
N.B.
Full Selftest cannot be selected unless a successful
'Internal Source Calibration' has been carried out
since the most-recent External Calibration.
Caution
The success of Full Selftest can be inhibited by:
• temperature not in the range: 13°C to 33°C;
• more than 1 year since Internal Source
Calibration executed;
• temperature more than 10degC different from
Internal Source Calibration; or
• presence of excessive RFI or Line noise.

2273    1V Range - Attenuated +10V Ref Magnitude Drift
        Digital comparison of the present magnitude against that
        recorded at the most-recent Internal Source Cal.
        +100mV Signal Mag. Drift   1-10ppm < drift < 1+10ppm

2283    100mV Range - Attenuated +10V Ref Magnitude Drift
        Digital comparison of the present magnitude against that
        recorded at the most-recent Internal Source Cal.
        +100mV Signal Mag. Drift   1-20ppm < drift < 1+20ppm


How would that drift be determined, without an ext cal?

Regards,
Gabi

On 1/20/2015 5:42 PM, Todd Micallef wrote:
I have not seen inside a 1281, but my 4920 has a 3V battery for the cal
constants. I am guessing that the unit would have flagged error codes
2110-2113.

Looking at the specs, it looks like the meter uses the 1V and 100mV ranges
for measuring resistance based on the selected range and whether or not it
is in LoI mode. The error messages appear that it is a gross measurement
error of over 100mV on both flagged ranges. I would think the 5522A would
be good enough for a quick sanity check of all possible inputs.

Todd

_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to