I have a 'friend" (actually a person who I have never met, but is a pain in the ***).
He sold a Agilent 34401A multimeter which the customer said is out of specification and can not be adjusted. Quickly scanning the measurent results, the meter is not showing results with any huge errors (say 1%), but which (if any) range it is out of specification I don't know. Unlike the Keysight calibration certificates I have seen, the measurement results doesn't show the upper and lower limits for the specification of the meter, so one can not tell from a quick glance if the meter is out of specification. One would need to study the detailed specification. *IF* I understand correctly, (and the data is in Spanish which I don't speak), with an input of 9,99990 V, the meter reads 9,99899 which the data indicates is an error of -91 uV/V with an uncertainty of 7.7 uV/V. The lab said the meter can not be adjusted. I don't know if they mean 1) This lab has no ability to adjust the meter. OR 2) It is too far off to adjust. Does anybody know if a 34401A can be adjusted by a competent 3rd party lab, or are the details of how to adjust the meter not available outside of Keysight? Dave _______________________________________________ volt-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.