>     On 19 January 2018 at 00:57 "David C. Partridge" 
> <[email protected]> wrote:
> 
>     Starting with a warmed up Prema 6048, if I power it off and then on, 
> select
>     20V DC range, 4 second integration and connect to the 10V output of a 
> Fluke
>     732A, I get a reading of:
> 
>     +10.000047
> 
>     If I reverse the polarity of the cables and allow the reading to settle 
> down
>     for a minute, I get a reading of:
> 
>     -10.000006
> 
>     If I short the inputs of the meter, I get a reading of:
> 
>     +0.000105
> 
>     which is different from the +0.000020 which I would have expected from the
>     readings of the 732A output.
> 
>     Proceeding on the assumption that the meter has stored a zero offset of
>     -105uV
> 
>     I then press the ZERO button to null out the offset and then reconnect the
>     732A in normal polarity I get an incorrect reading of:
> 
>     +10.000153
> 
>     This is not what I would expect to see - I would expect to see a reading 
> of
>     +10.000047 - 0.000105 = +09.999942 This calculation
> 
>     I then reverse the inputs and get a reading of:
> 
>     -10.000113
> 
>     which is close to the expected reading of -10.000006 - 0.000105 = 
> -10.000111
> 
>     I have mailed Prema about this problem, and will let you know what they 
> have
>     to say.
> 
>     Cheers
>     David Partridge
> 
>     _______________________________________________
>     volt-nuts mailing list -- [email protected]
>     To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>     and follow the instructions there.
> 
_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to