If there are any other members of the group with a 6048, I'd be grateful if you 
would attempt to recreate the scenario I describe to determine if this is just 
my meter or impacts other 6048 users.

Thanks
Dave

-----Original Message-----
From: volt-nuts [mailto:[email protected]] On Behalf Of David C. 
Partridge
Sent: 18 January 2018 11:57
To: 'Discussion of precise voltage measurement'
Subject: [volt-nuts] Prema 6048 zero offset problems

Starting with a warmed up Prema 6048, if I power it off and then on, select 20V 
DC range, 4 second integration and connect to the 10V output of a Fluke 732A, I 
get a reading of:

 +10.000047

If I reverse the polarity of the cables and allow the reading to settle down 
for a minute, I get a reading of:

 -10.000006

If I short the inputs of the meter, I get a reading of:

 +0.000105

which is different from the  +0.000020 which I would have expected from the 
readings of the 732A output.

Proceeding on the assumption that the meter has stored a zero offset of -105uV

I then press the ZERO button to null out the offset and then reconnect the 732A 
in normal polarity I get an incorrect reading of:

 +10.000153

This is not what I would expect to see - I would expect to see a reading of
+10.000047 - 0.000105 = +09.999942  This calculation

I then reverse the inputs and get a reading of:

 -10.000113

which is close to the expected reading of -10.000006 - 0.000105 = -10.000111

I have mailed Prema about this problem, and will let you know what they have to 
say.

Cheers
David Partridge

_______________________________________________
volt-nuts mailing list -- [email protected] To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.