On Sat, 6 Jun 2020 at 12:18, Florian Teply <[email protected]> wrote:

> Funny thing how things work out time-wise: I had a discussion yesterday
> on the very topic during re-audit for ISO 9001.
>

Yes, it is.

In basic terms, verification in metrology is a very slimmed down
> calibration: For a calibration, you essentially check every range of
> your instrument at usually five or more spots within that range in
> order to determine accurracy of your instrument in each range.
> For a verification, you do this only at the spot where you intend to
> measure. So if you were to measure a nominal 7.2V source, you'd compare
> the reading of your meter with your, say, known good 7.5V reference
> instead of doing a full calibration of the meter.
>


> So, in order to determine whether or not your chinese voltage reference
> meets its specs, you'd check your meter against, say, the
> well-characterized LTZ1000A you happen to have in your lab.
>
> Strictly speaking, you still have to do it as carefully as you would do
> a real calibration, taking all known effects into account, but it's
> still much less time-consuming than a full calibration as you check only
> one single point instead of all possible ranges with five points each.
>
> Does this help answer your questions or did I just bring up more
> questions than answers?
>

Yes. I wish VIM was a bit more explicit about it. A single sentence just
does not do it justice.

best regards,
> Florian
>