Hello Luciano,

On 22.08.20 14:35, [email protected] wrote:

    Thanks to all those who have also answered me directly.(spade Lugs)

    At a flea market I bought two references, one for voltage and other for 
current, Digistand 6406 and 6426 respectively. I wrote to the manufacturer for 
the manuals and he sent me the one for 6426 but he told me that he did not find 
the one that interested me most , ie voltage reference 6406. I repaired both 
power supplies that had shorted electrolytic capacitors and for one of them I 
had to replace the custom power transformer.

    Now they both work but the voltage calibrator needs to be calibrated. 
Inside are a handful of trimmers but no indication of their function. 
Calibration is not possible without manual. Extreme ratio I want to ask you if 
any of you have this manual.

    Thank you,

    Luciano


in 2016, quarks from the EEVBlog forum has offered exactly those two models.
Perhaps he can help you with a manual.

https://www.eevblog.com/forum/buysellwanted/fs-burster-calibrator-(location-germany)/

HTH

Best regards,

Mounty

_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Reply via email to