Well, meanwhile I think your 3458A might have a bug in its 10k range, if 1k and 
100 were measured stable in their corresponding ranges.\
Then the HW,  e.g. the current source of the 3458A might be affected.

How does you 3458A measure the 1k in its 10k range, btw.?

I also read through the whole Service Note 12-B, which describes a lot of 
Firmware bugs between your 4.6 and the latest, 8 or 9.

I found one which is related to OCOMP and Delay > 0.85sec, or so, and another 
one which deals with slow implicit reading of the buffer, or so, where the 
3458A delivers only 1 valid reading. I could not match those directly with your 
problem, but anyhow there might be another hidden error which might explain.

At RWTH Aachen in 1989, I used one of the first 3458As, also measuring 10k and 
100k reference resistors with OCOMP. \
I never used the DELAY, as I was not aware of these polarization problems yet, 
so I did not encounter a behavior as yours… and usually I find every failure on 
new instruments. Like I did then: That was the calibration constant issue of 
the PERIOD function, which was resolved very late with Rev. 8. 

Frank