Here’s the core of my program, that configuration is everything you need:

..      cmd:= 'OHMF 1000;APER 1;OCOMP ON;TRIG HOLD;DELAY 0.1;';  // for 1kOhm\
        else cmd:= 'OHMF 10000;APER 1;OCOMP ON;TRIG HOLD;DELAY 1;';  // OCOMP 
requires settling time, 1s for VHP for 0.1ppm\
        sendstr(HP3458);

        for N_Samp:= 1 to 16 do begin\
                ..\
                TRIGHP34401;\
                cmd:= 'TRIG SGL;’;\
                sendstr(HP3458);\
                Temperature:= NTC_2k8(fetch(HP34401))-273.16;\
                ..\
                Resistance:= fetch(HP3458); // ASCII read\
                ..\
                Resistance_25:= Resistance \* (1 + (25.0-Temperature)\* 
TC_Factor);\
                Do_Statistics;\
        end;\
        cmd:= 'TRIG AUTO;';\
        sendstr(HP3458);\
        Goto_Local(HP34401);\
        Goto_Local(HP3458);\
        EndCEC;\
        ..

You are using AZERO ONCE, that is not necessary at all with OCOMP as the DMM in 
this mode substracts e.m.f. and offsets anyhow.

Then you are using TARM, I think,  that’s also not necessary, maybe even 
counter productive.

I have found a comment in my DCV acquisition program, that TRIG SGL and TARM 
creates many ppm of offset, so I skipped that, and let the DMM running free on 
TRIG AUTO. \
For Ohm, TRIG SGL seems to work for my FW 9, maybe that is still a problem on 
your old FW version.

Background is, that in the DCV specification there is a footnote, that the 
first sample might have 1ppm deviation, and that seems to apply for halted 
(armed & triggered) measurements in particular. 

So maybe you find some further hints to eliminate your measurement problem.

Frank