Hi
> I had processed the images using iMOSFLM. The option of “automatic ice and 
> powder ring exclusion” was toggled ON when I processed the data. It is only 
> now I realize that this is not the way to get rid of ice rings. 
> 
This is due to the exclusion limits being set too conservatively for the ice 
rings; you might consider it a bug, because this should be the way to get rid 
of the ice rings!
> The latest paper on the use of iMOSFLM (Powell. H. R et al, Nature Protocols, 
> 2017) suggests excluding data within specific resolution shells to get rid of 
> the ice ring problem. I observe that if I set the limits 3.62-3.68, 
> 2.23-2.26, 1.90-1.93 Å in “excluded resolution ranges” option of iMOSFLM, 
> only the spots upto 3.6 Å are found and also predicted. Moreover all high 
> resolution data is lost.  Somehow I am not able to get this strategy working 
> in iMOSFLM.
> 

This is due to a bug in the iMosflm code; it will be fixed in the next release 
(I've told the current developer about it...).

I could send you a fix so that this option works if you like.

Harry
--
Dr Harry Powell
Chairman of International Union of Crystallography Commission on 
Crystallographic Computing
Chairman of European Crystallographic Association SIG9 (Crystallographic 
Computing) 



On 9 Aug 2017, at 13:17, Satvik Kumar wrote:

> Dear All,
> 
> 
> 
> Thank you all for your inputs.
> 
> 
> 
> You are all correct. The diffraction images have ice rings at 3.67, 2.24 and 
> 1.9 Å. The intensity of these ice rings decrease with increasing resolution. 
> In the Wilson plot, I clearly observe the spikes in intensity corresponding 
> to these resolutions.
> 
> 
> 
> 
> 
> The latest paper on the use of iMOSFLM (Powell. H. R et al, Nature Protocols, 
> 2017) suggests excluding data within specific resolution shells to get rid of 
> the ice ring problem. I observe that if I set the limits 3.62-3.68, 
> 2.23-2.26, 1.90-1.93 Å in “excluded resolution ranges” option of iMOSFLM, 
> only the spots upto 3.6 Å are found and also predicted. Moreover all high 
> resolution data is lost.  Somehow I am not able to get this strategy working 
> in iMOSFLM.
> 
> 
> 
> The other suggestion was to deice using AUSPEX or DEICE. The information 
> available on the internet suggests AUSPEX is a diagnostic tool. Is it 
> possible to use it to deice? I will be trying to get DEICE working shortly.
> 
> 
> 
> Please share your thoughts as to how I should proceed.
> 
>  
> Thanks,
> 
> Satvik
> 
> 
> On Tue, Aug 8, 2017 at 11:47 PM, Eleanor Dodson <eleanor.dod...@york.ac.uk> 
> wrote:
> You have some horrible ice rings - some data processing software may be able 
> to cut them out.. how are you processing it?
> Eleanor
> 
> On 8 August 2017 at 15:43, Christian Roth <christianroth...@gmail.com> wrote:
> Your plots look strangely different to the old Scala output you posted 
> before, but never mind.
> 
> Paul is right that a negative intensity is not desired and your crystal has 
> some issues with ice. 
> 
> That one icering around 2.26 must be massive taken into account how haywire 
> your curve goes there. 
> 
> Have you had a look at the images? There should be something visible in that 
> area. 
> 
> Christian
> 
> 
> Am 08.08.2017 um 15:17 schrieb Paul Emsley:
>> On 08/08/2017 15:07, Satvik Kumar wrote: 
>>> Dear Prof. Powell and Prof. Dodson, 
>>> 
>>> Thanks for your reply and advise. 
>>> 
>>> As per your suggestion, I have re-scaled the intensities using Aimless at 
>>> 1.861 A. 
>>> 
>>> I observe that the I/sigI has dropped to -0.8
>> 
>> That's not good. 
>> 
>> > and the behaviour of CC_1/2 is still anomalous. 
>> 
>> That made me laugh out loud. Perhaps not the best choice of adjective. 
>> 
>>> 
>>> Also, when I inspect the Wilson plot (Fig. 1), I observe that the curve 
>>> does not fall smoothly with respect to the reference curve (blue). Even 
>>> with respect to one more Wilson plot from CCP4 website (Fig. 2), the curve 
>>> from my aimless output is different and discontinuous. 
>> 
>> Icy! 
>> 
>> /me wonders if CCP4 are distributing auspex yet... 
>> 
>>> 
>>> The second moment of I is constant only up to a resolution of 2.4 Å at a 
>>> value of 3 (Fig. 3). I was not able to get some other plot to compare 
>>> against mine. 
>>> 
>>> Please tell me if I can still go ahead and refine at 1.861 A. 
>> 
>> No you can't. 
>> 
>> Maybe with some chopping you can rescue some reflections beyond 2.1. 
>> 
>> Paul 
>> 
> 
> 
> 
> <090817.pdf>