Hi Michael,

If you have a look at AS2500 page 23 you will see clearly that central line 
insertion is not considered a cardiac type procedure. As for the type of 
procedure being clearly defined ........ essentially it is does electrical 
contact with the ventricle occur? If so - cardiac, if not body. Note it must be 
the ventricle.

From AS2500.....

"Placement of a central venous catheter into the superior vena cava is not a 
cardiac-type
procedure even if the catheter is connected to an electrical device such as a 
pressure
transducer. However, placement of a Swann-Ganz pulmonary catheter through the 
right
ventricle requires cardiac protection.

In procedures in which a subclavian catheter is inserted, e.g. for 
haemodialysis access, with
the intention that the catheter tip will be positioned in the superior vena 
cava, care needs to
be taken that the tip does not enter the heart, since such positioning may 
raise the
classification of subsequent procedures to cardiac-type. Radiographic 
verification of the
positioning of the catheter should be performed.
Measurement by a liquid manometer of right atrial pressures with a 
saline-filled catheter
does not have to be done in a cardiac-or body-protected area as there is no 
electrical
connection.
The use of electrical pressure transducers to measure radial arterial pressure 
is a body-type
procedure and not a cardiac-type procedure."


N.B. This is by no means advice and your institution will need to make its own 
decision.

Regards,
Leigh Baker
Manager Medical Engineering and Physics
[cid:image001.gif@01CA41CE.724148F0]
St Vincent's Hospital
P.O. Box 2900
Fitzroy Victoria 3065
Phone: (03) 9288 4254
Fax: (03) 9288 4347
Email: leigh.ba...@svhm.org.au



From: Sullivan, Michael R [mailto:michael.sulli...@dhhs.tas.gov.au]
Sent: Wednesday, 30 September 2009 12:45 PM
To: Ruff, Megan; Porter, John R; bmelist@bme.asn.au
Subject: RE: [bmelist] Renal Unit Body Protect or Cardiac Protect Wiring

Hi guys,
I'm not a BME, but if clinicians are going to be inserting short to medium term 
central lines, in this unit, then there is a direct path right to the 'source'.
I'm pretty sure if this was the case, (insertions), then cardiac protection 
would be in order.

I reckon that you need the scope of use / practice defined.

M
________________________________
From: Ruff, Megan [mailto:megan.r...@mh.org.au]
Sent: Wednesday, 30 September 2009 11:12 AM
To: Porter, John R; bmelist@bme.asn.au
Subject: RE: [bmelist] Renal Unit Body Protect or Cardiac Protect Wiring

Hi John,

We specify Body Protection.  The flowchart contained in AS/NZS 2500 (section 
and figure 4.5 in the 2004 edition) requires cardiac protection 'if the applied 
part makes contact with the LV or RV of the heart'.  As the fluid lines are 
completely electrically isolated from the machine I'm not sure if they are even 
considered an Applied Part.  However, if you consider them an Applied Part on 
the basis that because they contain conductive fluid they are a potential 
conductor, dialysis falls into the category of a Body procedure under the above 
criteria even if you're dialysing a patient with a permacath that has been 
placed very close to the atrium.  The machines themselves are Type B equipment 
(Gambros are, anyway, I haven't got anything else handy to check), and if you 
have a look at the requirements for home dialysis installations in 2500 and 
3003, Body Protection is specified, not cardiac.

I hope that helps.

Regards,

Megan Ruff
Dialysis Service Technician
Melbourne Health/North West Dialysis Service
ph) 0437 009 708


________________________________
From: Porter, John R [mailto:john.por...@dhhs.tas.gov.au]
Sent: Wednesday, 30 September 2009 8:48 AM
To: bmelist@bme.asn.au
Subject: [bmelist] Renal Unit Body Protect or Cardiac Protect Wiring
We are establishing a new Satellite Renal Unit and my recommendation is that it 
only needs to be wired as Body Protected, outlined in AS/NZS 3003. Other 
individuals believe that is should be Cardiac Protected which I believe is 
unnecessary.

What are the views of others on this issue and what level of protection are 
most Renal Units wired?

John Porter
Team Leader Electrical/Mechanical
Building & Engineering
Launceston General Hospital

Ph: 03 6348 7433
Fx: 03 6348 7430

Email: john.por...@dhhs.tas.gov.au<mailto:john.por...@dhhs.tas.gov.au>


CONFIDENTIALITY NOTICE AND DISCLAIMER
The information in this transmission may be confidential and/or protected by 
legal professional privilege, and is intended only for the person or persons to 
whom it is addressed. If you are not such a person, you are warned that any 
disclosure, copying or dissemination of the information is unauthorised. If you 
have received the transmission in error, please immediately contact this office 
by telephone, fax or email, to inform us of the error and to enable 
arrangements to be made for the destruction of the transmission, or its return 
at our cost. No liability is accepted for any unauthorised use of the 
information contained in this transmission. If the transmission contains 
advice, the advice is based on instructions in relation to, and is provided to 
the addressee in connection with, the matter mentioned above. Responsibility is 
not accepted for reliance upon it by any other person or for any other purpose.
________________________________
WARNING: This message originated from outside the Northern/Melbourne/Western 
Health e-mail network. The sender cannot be validated. Caution is advised. 
Contact IT Services (+61 3 ) 9342 8888 for more information.
CONFIDENTIALITY NOTICE AND DISCLAIMER
The information in this transmission may be confidential and/or protected by 
legal professional privilege, and is intended only for the person or persons to 
whom it is addressed. If you are not such a person, you are warned that any 
disclosure, copying or dissemination of the information is unauthorised. If you 
have received the transmission in error, please immediately contact this office 
by telephone, fax or email, to inform us of the error and to enable 
arrangements to be made for the destruction of the transmission, or its return 
at our cost. No liability is accepted for any unauthorised use of the 
information contained in this transmission. If the transmission contains 
advice, the advice is based on instructions in relation to, and is provided to 
the addressee in connection with, the matter mentioned above. Responsibility is 
not accepted for reliance upon it by any other person or for any other purpose.
______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email
______________________________________________________________________

Disclaimer : The contents of this e-mail including any attachments are intended 
only for the person or entity to which this e-mail is addressed and may contain 
confidential, privileged and/or commercially sensitive material. If you are 
not, or believe you may not be, the intended recipient, please advise the 
sender immediately by return e-mail, delete this e-mail and destroy any copies.
______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

<<inline: image001.gif>>

Reply via email to