Hi John

 

I agree with Megan and have been fighting this battle for a long time
usually with hospital planners who seem to be nearly always nurses. This is
particularly problematic in EDs where they always want to put CF everywhere
and it's not necessary.

 

Bruce.

 

From: Ruff, Megan [mailto:megan.r...@mh.org.au] 
Sent: Wednesday, 30 September, 2009 10:12 AM
To: Porter, John R; bmelist@bme.asn.au
Subject: RE: [bmelist] Renal Unit Body Protect or Cardiac Protect Wiring

 

Hi John,

 

We specify Body Protection.  The flowchart contained in AS/NZS 2500 (section
and figure 4.5 in the 2004 edition) requires cardiac protection 'if the
applied part makes contact with the LV or RV of the heart'.  As the fluid
lines are completely electrically isolated from the machine I'm not sure if
they are even considered an Applied Part.  However, if you consider them an
Applied Part on the basis that because they contain conductive fluid they
are a potential conductor, dialysis falls into the category of a Body
procedure under the above criteria even if you're dialysing a patient with a
permacath that has been placed very close to the atrium.  The machines
themselves are Type B equipment (Gambros are, anyway, I haven't got anything
else handy to check), and if you have a look at the requirements for home
dialysis installations in 2500 and 3003, Body Protection is specified, not
cardiac.  

 

I hope that helps.

 

Regards,

 

Megan Ruff

Dialysis Service Technician

Melbourne Health/North West Dialysis Service

ph) 0437 009 708

 

 

  _____  

From: Porter, John R [mailto:john.por...@dhhs.tas.gov.au] 
Sent: Wednesday, 30 September 2009 8:48 AM
To: bmelist@bme.asn.au
Subject: [bmelist] Renal Unit Body Protect or Cardiac Protect Wiring

We are establishing a new Satellite Renal Unit and my recommendation is that
it only needs to be wired as Body Protected, outlined in AS/NZS 3003. Other
individuals believe that is should be Cardiac Protected which I believe is
unnecessary.

 

What are the views of others on this issue and what level of protection are
most Renal Units wired? 

 

John Porter

Team Leader Electrical/Mechanical

Building & Engineering

Launceston General Hospital

 

Ph: 03 6348 7433

Fx: 03 6348 7430

 

Email: john.por...@dhhs.tas.gov.au

 

 

CONFIDENTIALITY NOTICE AND DISCLAIMER 

The information in this transmission may be confidential and/or protected by
legal professional privilege, and is intended only for the person or persons
to whom it is addressed. If you are not such a person, you are warned that
any disclosure, copying or dissemination of the information is unauthorised.
If you have received the transmission in error, please immediately contact
this office by telephone, fax or email, to inform us of the error and to
enable arrangements to be made for the destruction of the transmission, or
its return at our cost. No liability is accepted for any unauthorised use of
the information contained in this transmission. If the transmission contains
advice, the advice is based on instructions in relation to, and is provided
to the addressee in connection with, the matter mentioned above.
Responsibility is not accepted for reliance upon it by any other person or
for any other purpose. 

  _____  

WARNING: This message originated from outside the Northern/Melbourne/Western
Health e-mail network. The sender cannot be validated. Caution is advised.
Contact IT Services (+61 3 ) 9342 8888 for more information. 

No virus found in this incoming message.
Checked by AVG - www.avg.com
Version: 8.5.409 / Virus Database: 270.13.113/2400 - Release Date: 09/28/09
05:51:00

Reply via email to