Hi Danny and Dan,

Thanks for your advices!
The main cause was indeed the quantization and the FFT bitwidth.
I did increase FFT bitwidth to 25 and increase the quant block output
to ufix_64_25
To do so I had to also increase the vacc bitwidth and store the output
in 64bit (two 32bit bram)
After all this bit growing I got a nice flat noise floor 75dB below
the test tone I put (which was at the ADC full scale = 0dbm).
the response is totally linear all the way!
The optimal shift seems to still be FFF.

Now the problem will be to put two of this spectrometers in a ROACH
and compile for high speed (1.0 - 1.5GHz) !
I am getting into this now !

Cheers,

r




On Thu, Jul 26, 2012 at 9:40 PM, Danny Price <[email protected]> wrote:
> Hi Ricardo
>
> I think along with the FFT shift, the quant block is a likely culprit. I'd
> recommend increasing the quant block from 6 bits to somewhere above 8 bits
> and seeing if the problem persists. (You'll have to modify the vacc block
> too).
>
> Also, you might want to consider digitizers with more bits if dynamic range
> is critical. The iADC and 083000 have 8-bits (255 levels), so the dynamic
> range is 20*log10(1/255) = 48dBm, a tad lower than the 50dB figure you
> mentioned.
>
> Regards
> Danny
>
> On 26/07/2012 21:01, Ricardo Finger wrote:
>>
>> Hello Casper community,
>>
>> I am working with the tutorial 3 spectrometer (slightly modified to work
>> with the iADC083000 board).
>> I need to analyze strong CW tones with a dynamic range of at least 50dB. I
>> am using the standard libraries and running at 500MHz ADC clock.
>> I started reducing the digital gain of the 'quant' block, and adding noise
>> to the tone to reduce the digital artifacts/harmonics.
>> So far so good, I got a strong line 75dB above the noise floor. The
>> problem is that those 75dBs are not the effective dynamic range because
>> there is a strange nonlinearity in the response of the spectrometer at the
>> lower power end. Please see the attachment "200MHz tone with 0 to 40dB
>> attenuation.pdf"
>>
>> When I reduce the tone power in 10dB steps I can see the peak going down
>> accordingly for the first two steps (20dB) but with the next 10dB of
>> attenuation, it drops more than 20dB, and with the following 10dB of
>> attenuation (40dB in total) it does completely disappear, dropping more than
>> 35 additional dBs.
>> Have any of you seen this 'low end' nonlinearity before?
>> Is it part of the ADC response or an artifact of the digital processing?
>> is it maybe related to the re-quantization before accumulation?
>>
>> The nonlinear response also applies to the noise floor as it can bee seen
>> in the attachment "200MHz tone with 5 steps digital gain.pdf", where for a
>> gain of '0F 00 00 00'  the noise floor has a slope of around 10dB, but for a
>> gain of ' 00 80 00 00" the slope is more than 50dB.
>>
>> Regards,
>>
>> Ricardo.
>>
>
>



-- 
Ricardo Finger Camus
Electrical Engineer
Astronomy Department
University of Chile
Of: 56(2)9771122
Casilla 36-D, Santiago.
http://www.das.uchile.cl/lab_mwl/