Hi Rich

Indeed that is interesting. Could you try adjusting the log-level to error or 
trace in tcpborphserver with ?log-level error or ?log-level trace and see if 
any additional messages are outputted when the sensors are queried?

Adam

On May 28, 2556 BE, at 20:41, Rich Lacasse <[email protected]> wrote:

> Hi Adam,
> 
> With respect to the LED, I found that the alarm was caused by the external 
> fans being disconnected.  I was able to probe R74 through R77 to narrow it 
> down to fans.  I am operating with the board externally cooled, not in a 
> chassis, so the chassis fans are not connected.  Interestingly enough, when I 
> connected to tborphserver3 interface and queried the individual sensors as 
> you recommended, it reported 0's for three of the fans and the fault went 
> away!  Does that make any sense?
> 
> Thanks for the help,
> Rich
> 
> On 5/28/2013 1:42 PM, [email protected] wrote:
>> 
>> Fault LED