I would add that there are some issues with air, you have to be careful with 
nanodrop that the path is ok, and also if concentrations are low, <1 mg/ml for 
instance, i am not sure one can trust it - compare 50 ul 1 cm path results with 
nano at 0.5-1 mg/ml... i get inconsistency there.. its good for concentrated 
samples and fast. and handy, we use it a lot, but
for dilute samples, i always use 1 cm light path. just my feel on it.

Tommi

On Jun 16, 2011, at 11:20 PM, aaleshin wrote:

> Filip,
> 25% accuracy is observed only for very diluted (OD280< 0.1) or concentrated 
> samples. But those sample a rarely used for ITC or CD. The concentrated 
> samples require dilution but a regular spec does it too. Since the light 
> passway is very short in Nanodrop it is accurate with more concentrated 
> samples, which we crystallographers use, so Nanodrop is ideal instrument for 
> our trade.
> 
> If the drop is within recommended volume like 1-2 ul for our model, its size 
> has a very small influence on the measurement. 
> 
>> Cuvettes will give a better accuracy provided you clean them properly. 
> I hated those times when I had to measure a concentration because of a need 
> to wash a cuvette. In a biological lab they are always dirty. We switched to 
> plastic disposable cuvettes for that reason...
> 
> Alex
> 
> On Jun 16, 2011, at 1:06 PM, Filip Van Petegem wrote:
> 
>> 25% is not acceptable for ITC or CD experiments though...
>> 
>> I was just sharing our bad experience with a demo nanodrop we had. Even if 
>> evaporation is not an issue, one has to take pipetting errors into account 
>> when dealing with small volumes.  The relative error on 1-2ul is a lot 
>> bigger than on 50ul. Unless you want to pre-mix 50ul and use a small 
>> quantity of that, which defeats the purpose of miniaturization...  It all 
>> depends on your applications and sample availability, but if you want a very 
>> accurate measurement, miniaturized volumes just won't get you the same 
>> accuracy.
>> 
>> Cuvettes will give a better accuracy provided you clean them properly. Just 
>> some water or EtOH is *not* enough...
>> 
>> Filip Van Petegem
>> 
>> 
>> 
>> On Thu, Jun 16, 2011 at 12:52 PM, aaleshin <aales...@burnham.org> wrote:
>> I also like our Nanodrop, but I do not recommend using it for Bradford 
>> measurements.
>> 
>> The 25% accuracy mentioned by Flip is pretty good for biological samples.  
>> Using 50 ul cuvette in a traditional spectrophotometer will not give this 
>> accuracy because cleanness of the cuvette will be a big issue...
>> 
>> Alex
>> 
>> On Jun 16, 2011, at 12:43 PM, Oganesyan, Vaheh wrote:
>> 
>>> I completely disagree with Filip’s assessment. I’ve been using nanodrop 
>>> nearly 5 years and never had inconsistency issues. If you work at 
>>> reasonable speed (if you put a drop there then lower the lever and click 
>>> measure before you do anything else) there will be no issues. At very high 
>>> concentrations the accuracy and therefore consistency may become lower. 
>>> Concentrations between 5 and 10 mg/ml should be fine. The instrument is 
>>> pricey though.
>>>  
>>>      Vaheh 
>>>  
>>>  
>>>  
>>> From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Filip 
>>> Van Petegem
>>> Sent: Thursday, June 16, 2011 3:34 PM
>>> To: CCP4BB@JISCMAIL.AC.UK
>>> Subject: Re: [ccp4bb] Nanodrop versus Nanophotomter Pearl versus good old 
>>> Bradford.
>>>  
>>> Dear Arnon,
>>>  
>>> the Bradford method is not recommended for accurate measurements.  The 
>>> readings are strongly dependent on the amino acid composition.  A much 
>>> better method is using the absorption at 280nm under denaturing conditions 
>>> (6M Guanidine), and using calculated extinction coefficients based on the 
>>> composition of mostly Tyrosine and Tryptophan residues (+ disulfide bonds). 
>>>  This method is also old (Edelhoch, 1967), but very reliable.
>>>  
>>> One thing about the nanodrop: smaller volume = more evaporation.  On the 
>>> demo we've had, I was so unimpressed with the precision (>25% variability 
>>> between two consecutive measurement) that we didn't consider this 
>>> instrument at all.  So unless you just want a 'rough' estimate, I wouldn't 
>>> recommend it at all. But most respectable spectrophotometers will take 
>>> cuvettes with 50ul volumes - a big step up from 1ml volumes...
>>>  
>>> Filip Van Petegem
>>>  
>>>  
>>>   
>>> On Thu, Jun 16, 2011 at 12:15 PM, Arnon Lavie <la...@uic.edu> wrote:
>>> Dear fellow crystallographers - a question about spectrophotometers for 
>>> protein concentration determination.
>>> 
>>> We are so last millennium - using Bradford reagent/ 1 ml cuvette for 
>>> protein conc. determination.
>>> 
>>> We have been considering buying a Nanodrop machine (small volume, no 
>>> dilution needed, fast, easy).
>>> However, while testing our samples using a colleague's machine, we have 
>>> gotten readings up to 100% different to our Bradford assay (all fully 
>>> purified proteins). For example, Bradford says 6 mg/ml, Nanodrop 3 mg/ml. 
>>> So while it is fun/easy to use the Nanodrop, I am not sure how reliable are 
>>> the measurements (your thoughts?).
>>> 
>>> So QUESTION 1: What are people's experience regarding the correlation 
>>> between Nanodrop and Bradford?
>>> 
>>> While researching the Nanodrop machine, I heard about the Implen 
>>> NanoPhotmeter Pearl.
>>> So Question 2: Is the Pearl better/worse/same as the Nanodrop for our 
>>> purpose?
>>> 
>>> Thank you for helping us to advance to the next millennium, even if it is 
>>> nearly a dozen years late.
>>> 
>>> Arnon
>>> 
>>> -- 
>>> ***********************************************************
>>> Arnon Lavie, Professor
>>> Dept. of Biochemistry and Molecular Genetics
>>> University of Illinois at Chicago
>>> 900 S. Ashland Ave.
>>> Molecular Biology Research Building, Room 1108 (M/C 669)
>>> Chicago, IL 60607
>>> U.S.A.
>>>                             Tel:        (312) 355-5029
>>>                             Fax:        (312) 355-4535
>>>                             E-mail:     la...@uic.edu
>>>                             http://www.uic.edu/labs/lavie/
>>> ***********************************************************
>>> 
>>> 
>>> 
>>> -- 
>>> Filip Van Petegem, PhD
>>> Assistant Professor
>>> The University of British Columbia
>>> Dept. of Biochemistry and Molecular Biology
>>> 2350 Health Sciences Mall - Rm 2.356
>>> Vancouver, V6T 1Z3
>>> 
>>> phone: +1 604 827 4267
>>> email: filip.vanpete...@gmail.com
>>> http://crg.ubc.ca/VanPetegem/
>>> To the extent this electronic communication or any of its attachments 
>>> contain information that is not in the public domain, such information is 
>>> considered by MedImmune to be confidential and proprietary. This 
>>> communication is expected to be read and/or used only by the individual(s) 
>>> for whom it is intended. If you have received this electronic communication 
>>> in error, please reply to the sender advising of the error in transmission 
>>> and delete the original message and any accompanying documents from your 
>>> system immediately, without copying, reviewing or otherwise using them for 
>>> any purpose. Thank you for your cooperation.
>> 
>> 
>> 
>> 
>> -- 
>> Filip Van Petegem, PhD
>> Assistant Professor
>> The University of British Columbia
>> Dept. of Biochemistry and Molecular Biology
>> 2350 Health Sciences Mall - Rm 2.356
>> Vancouver, V6T 1Z3
>> 
>> phone: +1 604 827 4267
>> email: filip.vanpete...@gmail.com
>> http://crg.ubc.ca/VanPetegem/
> 