Chris,

On Sun, Feb 23, 2014 at 2:52 PM, Chris Fage <cdf...@gmail.com> wrote:

> As suggested, I tried NCS and TLS in phenix.refine, although my R-factors
> did not budge.
> (...)
> Forgive my ignorance, but I am not sure how to check whether the bulk
> solvent model is reasonable.
>


I figure you used phenix.refine for refinement in which case bulk-solvent
modeling and overall anisotropic scaling should be optimal within the
framework of the implemented model used to describe it (for more info see
http://journals.iucr.org/d/issues/2013/04/00/dz5273/dz5273.pdf).

You can check it by looking in refinement log file (this is the table I
always look first thing before doing anything else). There you will see
something like this:

  Resolution Compl Nwork Nfree R_work <Fobs> <Fmodel> kiso  kani kmask
43.996-14.23 96.15    88    12 0.1175 94.939   93.450 1.00 0.148 0.288
14.224-11.37 98.06    94     7 0.1211 84.547   83.471 1.00 0.151 0.295
11.321-9.092 98.90   156    23 0.0967 85.623   85.288 1.00 0.148 0.300
 9.082-7.268 97.97   305    33 0.1400 61.020   59.892 1.00 0.143 0.300
 7.264-5.810 98.96   601    63 0.1495 54.915   53.404 1.00 0.140 0.275
 5.806-4.642 98.67  1134   127 0.1271 71.906   71.017 1.00 0.149 0.274
 4.642-3.711 97.88  2198   203 0.1310 68.399   67.585 1.00 0.156 0.273
 3.710-2.966 97.16  4129   462 0.1611 46.881   46.262 1.00 0.157 0.034
 2.966-2.371 94.27  7737   874 0.1878 25.826   25.084 1.00 0.154 0.000
 2.371-2.200 90.76  3826   418 0.1758 19.757   19.072 1.00 0.158 0.000

Pay attention to:

- completeness (second column). Ideally in all resolution bins it should be
greater than 90-95%. Smaller (especially at low resolution) completeness
may result in corrupted maps.
- kmask (last column) value in the lowest resolution bin: it should be
around 0.2-0.5 (Acta Cryst. (2002). D58, 1387-1392). This characterize
sanity of bulk-solvent model.
- if a resolution bin has outstandingly high r-factor (5th column) this may
indicate a problem.