Ray,

I agree with Opinion 1.  The required warning must have the following three 
components:

1) WARNING 

2) Risk of Electric Shock.

3) To avoid .........Do xxxxxx.


On Fri, 20 Sep 1996 09:24:37 -0700 Ray Russell wrote:

> 
> 
>      Dear Friends,
>      
>      I've received some contradicting information in relation to the need 
>      for interlocks, and would like to bring it up to the group for 
>      discussion.  We have an instrument that has a removable side panel. A 
>      tool is required to remove the panel and there is hazardous voltages 
>      behind the panel. The user is instructed to remove the panel for 
>      installing options, and maintenance. 
>      
>      Opinion 1 - Access is restricted by a tool, so no interlock is 
>      required. EN61010-1 sec 6.1.1 states:
>      
>      The following Hazardous live parts are permitted to be accessible to 
>      the operator during normal use:
>      
>      - parts intended to be replaced by the operator (e.g. batteries) and 
>      which may be Hazardous live during the replacement or other operator 
>      action, but only if they are accessible only by means of a tool and 
>      have a warning marking..
>      
>      Opinion 2 - When an operator is instructed to remove the panel, an 
>      interlock is required. An agency told me that once an operator is 
>      instructed to remove a panel, it is treated as if no tool was required 
>      to remove it. 
>      
>      Also, if anyone has information on high integrity components, such as 
>      interlock switches and relays, I would appreciate some information.
>      
>      Thank you for your thoughts.
>      
>      Ray
>      
>      [email protected]
> 