Ray -
You asked about info on interlock switches and relays. Switches I am not 
familiar with - but a special type relay is often required for interlocks on 
safety gates for machine tools, robots, presses, etc. - don't know about 
your instrument. The German machine tool assoc has a standard ZH1/457 which 
requires "forcibly guided contacts" maintaining at least 0.5mm gap over 
life. No contact may change state out of syncronization with another even in 
the event of welding, coil failure, etc..
Both Siemens and Hengstler in Germany make such a device. I am not aware of 
a US mfr.. If interested, you may contact me for a Siemens datasheet. They 
are not in the same Siemens group we are, but I can provide contacts at 
either company if you are interested.

Best regards,

Harold Leipold
Siemens EC
Princeton, IN
Tel. (812)386-2161
Fax. (812)386-2616
E-mail   [email protected]
 ----------
From: Ray Russell
To: emc-pstc
Subject: Interlocks
List-Post: [email protected]
Date: Friday, September 20, 1996 9:24AM

     Dear Friends,

     I've received some contradicting information in relation to the need
     for interlocks, and would like to bring it up to the group for
     discussion.  We have an instrument that has a removable side panel. A
     tool is required to remove the panel and there is hazardous voltages
     behind the panel. The user is instructed to remove the panel for
     installing options, and maintenance.

     Opinion 1 - Access is restricted by a tool, so no interlock is
     required. EN61010-1 sec 6.1.1 states:

     The following Hazardous live parts are permitted to be accessible to
     the operator during normal use:

     - parts intended to be replaced by the operator (e.g. batteries) and
     which may be Hazardous live during the replacement or other operator
     action, but only if they are accessible only by means of a tool and
     have a warning marking..

     Opinion 2 - When an operator is instructed to remove the panel, an
     interlock is required. An agency told me that once an operator is
     instructed to remove a panel, it is treated as if no tool was required
     to remove it.

     Also, if anyone has information on high integrity components, such as
     interlock switches and relays, I would appreciate some information.

     Thank you for your thoughts.

     Ray

     [email protected]

Reply via email to