Hello Moshe:


Virtually all leakage current arises from the stray and 
real capacitances between line (live) and ground (earth).

To reduce leakage current, you have two choices:

1.  Reduce the capacitance.

2.  Install a 1:1 isolating transformer between the mains
    and the load.  

2a. If the transformer is built into the equipment, the 
    leakage current changes to that of the primary of the 
    isolating transformer.  

    (This technique is commonly used for medical equipment
    systems with high leakage current.)

2b. If the transformer is external to the equipment, then 
    the secondary of the transformer must be floating 
    (i.e., no neutral).

Here are some choices that don't reduce the leakage current,
but prevent the leakage current from becoming accessible.

3.  Change to all-insulated construction so that there are
    no accessible conductive parts, including low-voltage
    secondary circuits.  This doesn't change the leakage 
    current in the green/yellow wire, but it does make the
    leakage current inaccessible.  

4.  Redundant ground connection.  Some standards waive the
    leakage current requirement if the equipment is installed
    with a second, separate ground wire, whose connection is
    permanently connected.  This, likewise, assures that the
    leakage current does not become accessible.


Best regards,
Rich


Reply via email to