Richard, (answering message below)

Most safety testing is by inspection of PCB layout, component specs and
component ratings.  For instance, if you use an approved power supply,
the lab usually won't pull any single fault hi-pot tests on your unit.
They'll probably be satisfied with performing a hi-pot test on the
complete unit.  You can rest assured that the manufacturer of your power
supply probably suffered through numerous single fault hi-pot tests.  As
such, they won't be repeated on your design.

However, if there is any doubt, the lab will produce a single fault and
test the circuit for a hazardous condition.  This is especially true if
you design your own power supply, as we have done a few times.   Part of
this testing may be a hi-pot test.  Furthermore, a design may have to
pass a hi-pot test after going through the "blocked fan" single faults
that you mention; perhaps only to verify that the elevated temperatures
didn't compromise any insulation in the unit.  This usually isn't the
case; since most blocked fan tests don't cause temperature rises
sufficient to exceed the temperature ratings of most insulation.

After saying all of that; I have to qualify that it's just my opinion
based upon my own experience and my now failing memory (I cycle through
all three of my girls' names just to get one).  Since I'm a safety lab
"client", I'll defer the verification of my statements to someone who
actually works at a safety lab.

Anybody want to help Richard on this one?

Chris Maxwell
Design Engineer
Nettest


From: Stone, Richard [mailto:[email protected]] 
Sent: Tuesday, February 10, 2004 9:40 AM
To: Chris Maxwell; Chris Wells; emc-pstc
Subject: RE: IEC61010-1 multiple Overvoltage catagories in one
circuit???

Chris,
for #1 below,
are you stating that for a hipot test
the NRTL can do a single fault test,
remove a part or anything else, being the single
fault, then performing a hipot?
I dont agree with this logic,
please educate me if this is the case.
Hipot should be done as a complete unit, intact.
with No faults.....willing to learn from this forum..
Temp test is with one fault for heating purposes.
hipot?
thank you,
Richard...

Richard A. Stone
Excel Switching Corporation
Compliance 
75 Perseverance Way
Hyannis, MA. 02601
508 862 3311 ph.
508 862 3020 fax




This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ieee-pses.org/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to