Gentlemen 

An isolation xfmr does NOT necessarily reduce leakage current through the
ground connector. Most iso xfmrs connect the ground pin, and float line and
neutral to the end-use equipment.

If the power supply has components bridging the pri/sec boundary that allow
current flow, and depending on if the secondary outputs are ground-referenced,
then human (ground-referenced) touch current under some conditions can be very
high.

If the installed power supply bridges the pri/sec boundary using only galvanic
isolation, then almost all human touch current is through the power supply's
y-caps that are between primary and ground. And there is no such thing as
"medical" y-caps that have low leakage current. 

The magnitude of current thru the y-caps is simple to calculate; it is the
voltage across the cap divided by the cap's reactance. It is important to note
that the capacitance for some y-caps can be voltage-dependent. 

If your empirical measurements are not close to calculated 50/60Hz leakage
current values, then it is important to note the quality of the input power,
as high crest-factor voltage waveforms will cause the y-caps to "work harder"
and thus have more current flowing from the mains input back through ground.

Do NOT ever modify components in a certified/recognized component power
supply. A slight change in some component values can have profound effects on
the performance of the unit under some SFCs and/or transient load conditions.
The people that make component power supplies have tested the stability of
their units under wide input and output conditions; and if the unit is not
stable it will probably not be safe.

(CONFLICT OF INTEREST ALERT: my employer makes component power supplies) It is
important to use component power supplies with supporting Type Test data that
have published leakage currents and indicate test conditions. The OP noted
that the test voltage for Touch Current was 264V, so I assume that this is a
1010 or 65-based installation. If the test standard for the component power
supply is 60950, then the Touch Current test voltage was probably 254V. You
may be using a component outside of its ratings. You must observe the
conditions of acceptability for safety-critical components. Please note that
the unit's power supply is probably the most safety-critical component in your
box.

luck, and let's not hurt anybody, 
Brian 


-----Original Message----- 
From: FastWave [ mailto:[email protected]] 
Sent: Wednesday, November 17, 2004 6:23 AM 
To: '[email protected]'; 'EMC-PSTC List' 
Subject: RE: Multiple Leakage/Touch Current Sources 


Ron, 

I have conducted hundreds of leakage current tests but never personally
redesigned a product to reduce leakage, so here is my "no previous experience"
input:

My guess is that your leakage current is predominantly line frequency (you may
wish to confirm by looking at the leakage current frequency composition). I
also guess that the switching power supplies have additional filtering in the
front end with capacitors to ground (in addition to the 2 separate filters).
This is making for a lot of capacitance to ground which is the source of your
leakage current. If these assumptions are correct, the only way to reduce
leakage current would be to reduce capacitance to ground (do you really need
all those filters). Or, isolate everything behind an isolation transformer
(i.e. the medical way).

If my presumption that filter capacitance is the cause is correct, it would be
additive. 
Regards, 
Bill Bisenius, N.C.E. 
E.D.& D. 
www.productsafeT.com 

  
-----Original Message----- 
From: Ronald R. Wellman [ mailto:[email protected]] 
Sent: Wednesday, November 17, 2004 8:26 AM 
To: EMC-PSTC List 
Subject: Multiple Leakage/Touch Current Sources 

Howdy all, 

I have come across a problem where a product is exceeding 3.5 ma of touch
current. The product in question has 2 line filters and 4 Switching Mode Power
Supplies. Because of the addition of the fourth power supply the overall touch
current can be as much as 4.0 ma. Ideally, we do not want to make the product
so it has to be hard-wired to the building mains. 

Because of the high touch current, does anyone know what the relationships are
between different touch current sources? I've heard that touch current is not
linear when you add up the different sources and phase shifts between L-G and
N-G might be a problem. However, I'm looking for some detailed information on
this issue so I can provide ways to investigate the problem and hopefully
reduce the overall touch current.

BTW, the product is not medical equipment and all touch current tests are done
with the product operating at 264 V AC, 50 Hz.

Best regards, 
Ron Wellman 

---------------------------------------------------------------- This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Ron Pickard: [email protected] Scott Douglas [email protected] 


For policy questions, send mail to: 


Richard Nute: [email protected] Jim Bacher: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc


Reply via email to