Dear Doug,

As Secretary of IEC/SC 62D (Subcommittee that published 60601-2-38) and as
the former convener of the hospital bed working group, I would like to
provide some insights on your question about measuring leakage currents for
hospital beds.

Your question regarding note 3 suggests that you've misunderstood the note
or that your bed is extremely unusual in how it connects to the mains. Note
3 addresses equipment that is permanently installed. This means that the
connection to mains and protective earth are "hard wired" into the
institution's mains voltage and protective earth (and protective earth)
distribution system (not connected by a standard power cord). The bed would
have to be connected to mains and protective earth such that a tool would
be
required to disconnect the bed to qualify under Note 3. I have never seen
such an arrangement for a hospital bed. The requirement of clause 52.5.4
(interruption of protective earth) of 60601-1 is performed for all
equipment
connected to the mains through a plug/outlet.

Patient leakage currents (under the single fault condition of interruption
of protective earth and normal condition) are measured from all applied
parts (in essence the surface & siderails of the bed) to earth. Enclosure
leakage is measured from the rest of the bed to earth. Patient auxiliary
current is measured between parts of the applied part. Where the applied
part or enclosure are made of plastic or other non-conductive materials,
those parts are covered by foil and leakage is measured from the foil.

60601-2-38 does not modify the leakage current measurement requirements of
the general standard (other than identifying what is considered the applied
part), so the same requirements apply for beds that apply for any other
medical electrical equipment.

If in fact your bed is hard wired into the hospital mains and protective
earth distribution system, the higher limits for earth leakage current
(leakage current flowing through the protective earth conductor for the
whole device) would apply.

I hope this proves helpful. Feel free to contact me if you have further
questions.

If you do not have a copy of the 60601-2-38, I strongly suggest you acquire
one because there are many requirements that do go beyond the requirements
of the general standard.

Best regards,
Mike Schmidt
Strategic Device Compliance Services
www.devicecompliance.com

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

Reply via email to