Doug,

    Altho I am not completely familiar with 60601-1 since I haven't
worked with it for a number of years, I will comment on leakage current
requirements for earthed equipment in general.  

    By and large it is recognized that earth/ground is unreliable for
cord connected equipment around the world.  (I know that US medical
facilities work harder than most to provide a proper earth, but that doesn't
negate the comment).  For this reason safety standards specify a leakage
current (Touch Current) limit for the normal, earthed case - 0.1mA as you
have mentioned for your case.  So, for this product - when the earth is not
present - the patient will not get more than the 0.1mA allowed.  

    I would be surprised if there were an exception in 60601 to allow
anything higher than this.  

    You could, of course, provide for a permanent, hard wired electrical
connection for the equipment which will provide a permanent, secure
earth/ground.  Conversely you could provide an isolating transformer for the
product which would isolate the product from ground so that there would be
no leakage current to ground other than the leakage current from the
transformer (a good double insulated one).  I'm sure that neither of these
seems attractive.  Surely you can get a proper motor with low leakage
current for this application,  

:>)     br,     Pete
 
Peter E Perkins, PE
Principal Product Safety Engineer
PO Box 23427
Tigard, ORe  97281-3427
 
503/452-1201     fone/fax
[email protected]
 

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

Reply via email to