Doug,

60601-1 is not the easiest standard to interpret, but I think that in this
instance you are interpreting it incorrectly. Patient leakage current is
that current which flows between an Applied Part and Earth via the patient.
The accessible parts of the frame are not considered as Applied Parts. It is
acceptable for protection of the patient from electric shock risk arising
>from contact with an accessible part to be provided by earthing. As you
rightly point out, the potential of the accessible part when the earth
impedance is low will be negligible. You must consider single fault
conditions, so it would be wise to include interruption of the earth
conductor in this case. However, perhaps you could have one earth connection
via the power cord and one separate connection to a permanent earth
conductor, thus providing a redundant protective earth.
This is my opinion only, and I stand to be corrected by anybody more
familiar with this standard than I.

Best regards

Neil R. Barker CEng MIEE FSEE MIEEE
Manager
Compliance Engineering
e2v technologies (uk) ltd
106 Waterhouse Lane
Chelmsford
Essex CM1 2QU
UK

Tel: (+44) 1245 453616
Fax: (+44) 1245 453410
Mob: (+44) 7801 723735



From: Doug Beckwith [mailto:[email protected]]
Sent: 11 October 2005 16:33
To: EMC-PSTC
Subject: 60601-1 and patient leakage current


Hi All,
Question time again, this time concerning patient leakage
current and permanently connected Class 1 equipment with a
permanently connected earth.

I have a product that exceeds the patient leakage current
between the mains and the accessible metal parts of the frame (a
motorised bed). The accessible metal part is pemanently
connected to PE. The measured leakage current from the power
supply exceeds the patient leakage requirements (0.1mA). I
searched 60601-1 and I cannot find any exception that allows for
increased leakge current limits. 
If you have a metal part that is connected to earth with a
resistance of <0.1 ohm, the potential on the accessible part
shoul closely approximate 0V. I can understand the leakage
requirement if the part is not earthed, but there should be an
exception if the part is at earth potential.

Help, I have a feeling I am missing something. All assistance
gratefully received.

Regards

Doug

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

Reply via email to