However the energy is transferred to the body tissues, and RF currents and 
voltages are induced. At frequencies that begin to exceed 100 kHz, the 
effect is energy absorption in the tissues (heating). At these frequencies, 
it is considered that there is no neuro or muscular stimulation. The effect 
is energy transfer. That's why localized SAR measurements would likely be 
appropriate.

The medical implant manufacturers set limits up to about 10 MHz but that is 
for EMI concerns (interfering with the electronics and mimicing the heart 
signals). My understanding is that the frequencies, that you question, are 
too high for neuromechanisms, such a those that pace the heart, to respond 
to.

dB


On Feb 8 2007, Paolo Peruzzi wrote:

> My original question had nothing to do with SAR, because I'm talking 
> about devices which deliver rf power "galvanically" to the patient skin, 
> by means of an electrode.
>
> Is it reasonable to assume that for frequencies above 100 kHz, a direct 
> application of rf current to the human skin is very unlikely (or unable) 
> to cause ventricular fibrillation?
> 
>Paolo Peruzzi
>
>
>---------- Original Message -----------
>From: "Garnier, David S (GE Healthcare)" <[email protected]>
>To: "Paolo Peruzzi" <[email protected]>
>Sent: Thu, 8 Feb 2007 09:47:52 -0500
>Subject: RE: effects of high frequency currents on human body
>
>> What kind of device are we talking about?  I know the folks
>> in MRI are concerned about SAR... There was concern about
>> generating SAR numbers for wireless ECG telemetry...
>> 
>> Regards,
>> 
>> dave garnier
>> 
>>                      David Garnier 
>>                      e GE HealthCare 
>>                      ___________________________________________
>> 
>>                      David S. Garnier 
>>                      Quality Assurance Technician 
>>                      Quality-DI RA-EMC 
>>                      3000 N. Grandview Ave - M/S W-1250 
>>                      Waukesha, Wi. 53188 
>>                      Tel: 262.312.7246 
>>                      Cel:  414.899.7580
>> 
>> -----Original Message-----
>> From: [email protected] [mailto:[email protected]] On Behalf Of Paolo
>> Peruzzi
>> Sent: Thursday, February 08, 2007 3:50 AM
>> To: [email protected]; E-mail
>> Subject: effects of high frequency currents on human body
>> 
>> Dear all,
>> I'm looking for a guidance on the effects of currents on the human body
>> for frequencies above 100 kHz and up to tens of MHz. 
>> I've read IEC 60479-2 and ICNIRP guidelines, but none of them seems 
>> to give satisfactory information to set a reasonably safe 
>> current/voltage/power limit for an equipment intended to delivery rf 
>> or shortwave current to a patient by means of a bipolar or unipolar 
>electrode.
>> A lot of systems recently have appeared on the market for medical and
>> aesthetic applications, but it looks like high frequency electrical
>> features are up to the manufacturer, with no real regulatory constraint.
>> 
>> Regards,
>> 
>> Paolo Peruzzi
>> Regulatory Medical R&D
>> El.En. S.p.A.
>> Via Baldanzese, 17
>> 50041 Calenzano (FI)
>> Italy
>> 
>> Tel. +39 055 8826807
>> FAX  +39 055 8832884
>> http://www.elengroup.com/
>> 
>> -
>> ----------------------------------------------------------------
>> This message is from the IEEE Product Safety Engineering Society
>> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>> 
>> To post a message to the list, send your e-mail to [email protected]
>> 
>> Instructions:  http://listserv.ieee.org/request/user-guide.html
>> 
>> List rules: http://www.ieee-pses.org/listrules.html
>> 
>> For help, send mail to the list administrators:
>> 
>>      Scott Douglas           [email protected]
>>      Mike Cantwell           [email protected]
>> 
>> For policy questions, send mail to:
>> 
>>      Jim Bacher:             [email protected]
>>      David Heald:            [email protected]
>> 
>> All emc-pstc postings are archived and searchable on the web at:
>> 
>>     http://www.ieeecommunities.org/emc-pstc
>------- End of Original Message -------
>
>-
>----------------------------------------------------------------
>This message is from the IEEE Product Safety Engineering Society
>emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>
>To post a message to the list, send your e-mail to [email protected]
>
>Instructions:  http://listserv.ieee.org/request/user-guide.html
>
>List rules: http://www.ieee-pses.org/listrules.html
>
>For help, send mail to the list administrators:
>
>     Scott Douglas           [email protected]
>     Mike Cantwell           [email protected]
>
>For policy questions, send mail to:
>
>     Jim Bacher:             [email protected]
>     David Heald:            [email protected]
>
>All emc-pstc postings are archived and searchable on the web at:
>
>    http://www.ieeecommunities.org/emc-pstc
>

-- 
dBaron
dB-emf
(512)917-8346
[email protected]

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

Reply via email to