Brian,

You wrote:
> I have recent experience where a CPS with a 60601 
> report, and tested for patient contact, was installed in ITE. The 
> box could no longer pass all safety requirements for a C.O. device.

'C.O. device' -- 'Central Office'?  Does this means the medical power 
supply was subjected to Bellcore/Verizon-style telecomm system tests? 

Pat Lawler
EMC Engineer
SL Power Electronics Corp.

"Brian O'Connell" <[email protected]> wrote on 09/15/2010 08:56:52 
AM:
> My employer designs and manufactures, among other things, component 
> power supplies for medical, IT, instrument, industrial, and other weird 
stuff.
>
> Unlike other safety-critical components such as X/Y caps, VDRs, 
> switches, connectors, etc; there are no 'accepted' safety standards 
> for component power supplies. No need to tell me about IEC61204, 
> UL1310, UL1012, IEC61558, etc; as these are 'boutique' standards. 
> Each major product safety standard has some specific requirements 
> for power sources - of which most requirements overlap. Other than 
> an indicated conformity to Articles 600 or 725 of NEC, it is very 
> unusual to be able to drop in a component power supply from 
> different sources even when the ratings are identical.
>
> Until there is an industry-standard component power supply product 
> safety standard that has been harmonized by all national codes, and 
> is cited in the Big Four (60950/60065/60335/60601), there will be no
> drop-in replacements without re-test with an alternate construction 
> amendment to your CB report.
>
> Final note - I have recent experience where a CPS with a 60601 
> report, and tested for patient contact, was installed in ITE. The 
> box could no longer pass all safety requirements for a C.O. device. 
> This has been discussed in previous PSES threads - a medical power 
> supply is not necessarily 'safe' in a consumer or ITE box.
>
> Brian

> > -----Original Message-----
> > From: [email protected] [mailto:[email protected]]On Behalf Of Amund
> > Westin
> > Sent: Wednesday, September 15, 2010 12:19 AM
> > To: [email protected]
> > Subject: SV: [PSES] IEC 60950-1 vs IEC 60601-1
> >
> > The end product is an ITE (DVB) and in the previous models,
> > we have made use
> > of an open frame AC/DC power supply unit. This PSU had CB certificate
> > according to IEC60950-1. Then we went to a third party body
> > for testing the
> > complete unit according to IEC60950-1. Since we had a CB
> > certificate, the
> > scope of work for testing the ITE was limited a lot, since
> > no testing /
> > examination on the PSU itself was necessary.
> >
> > Now, we want to change PSU to another open frame AC/DC power
> > supply unit.
> > This PSU has a CB cert acc to IEC 60601-1, while the end
> > product still shall
> > conform according to IEC60950-1.
> >
> > Since the new PSU shall be placed inside the ITE, some
> > safety test and
> > evaluations must be done.
> > But we would also assume that a 601-1 PSU should be more
> > "safe" than an
> > 950-1 PSU, seen from an application point of view.
> >
> > We accept that some safety testing must be done on the end
> > product. But we
> > find it a bit strange if the introduction of this medical
> > PSU involves more
> > testing than if we had made use of a 950-1 PSU.
> >
> > #Amund
> >
> >
> > -----Opprinnelig melding-----
> > Fra: Brian O'Connell [mailto:[email protected]]
> > Sendt: 14. september 2010 18:56
> > Til: [email protected]
> > Emne: Re: [PSES] IEC 60950-1 vs IEC 60601-1
> >
> > WARNING - Fractious dichotomous isochroous nondeterministic
> > reply follows.
> >
> > I am not certain of Mr. Westin's use of "PSU". IEC60950-1
> > has no definition
> > for the phrase 'power supply unit', and contextual meaning is not
> > consistent. IEC60601-1 uses the phrase 'power supply unit' only when
> > referring to a separate external power source.
> >
> > For a COMPONENT power supply, where the final determination
> > is the end-use
> > installation - the 'PSU' CB report can use a TRF based on
> > almost any product
> > safety standard that would scope a 'PSU'.
> >
> > For a power supply that has been assessed as an end-product, neither
> > standard provides complete coverage for the requirements of
> > the other for
> > many uses, and for some uses either standard could provide
> > coverage for a
> > superset of the requirements. Anyone wanna see a De Morgan
> > reduction of that
> > logic ?
> >
> > The coming (????) of the UL/EN/IEC62368-1 standard may
> > provide an unintended
> > formal method to 'bridge' various requirements for component
> > power supplies.
> >
> > Most I.T. CB reports for component power supplies require
> > the repeat of some
> > Type Tests in the end-use installation. So for a medical
> > 'PSU', the pudding
> > would not get cooked until it was subject to tests in your I.T. box.
> > Conversely, the 3d edition of 60601-1 specifically addresses
> > the use of I.T.
> > component power supplies for some medical equipment.
> >
> > The CB report could be considered valid; or not.
> > Additional tests could be required; or not.
> >
> > Non-related Note: the '#' at the signature threw some
> > interesting errors for
> > my auto-mail program - thanks for the programmatic work-out.
> >
> > Brian
> >
> >  > -----Original Message-----
> >  > From: [email protected] [mailto:[email protected]]On
> > Behalf Of Amund
> >  > Westin
> >  > Sent: Tuesday, September 14, 2010 1:05 AM
> >  > To: [email protected]
> >  > Subject: IEC 60950-1 vs IEC 60601-1
> >  >
> >  > Is it possible to make use of a CB certificated (IEC
> >  > 60601-1) PSU, into a
> >  > ITE which shall be approved according to IEC60950-1?
> >  >
> >  > Well, I assume that's OK, but should we expect additional
> >  > tests on the PSU /
> >  > ITE since it's not CB acc to IEC60950-1?
> >  >
> >  > Thanks.
> >  > #Amund

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to