Forwarding from Dave Osborn. Find below some ESD developments associated with 
IEC 60601-1-2, 3rd and 4th Editions.

Dave, thanks for that info.

Best regards,

Ron

From: Osborn, Dave [mailto:[email protected]]
Sent: Tuesday, January 10, 2012 9:51 AM
To: Ron Pickard; McInturff, Gary
Subject: RE: 61000-4-2 3rd edition

Hi Ron,

Feel free to post to the listserve. I am registered via by ieee alias, so I 
can't post....

3rd edition of IEC 60601-1-2 has:
±2, 4 & 8 kV air
±2, 4 & 6 kV contact

There is a 4th edition under development, the most recent draft has:
± 8 kV contact
± 2 kV, ± 4 kV, ± 8 kV, ± 15 kV air

Best regards,

Dave Osborn
Sr. Manager, International Standards & Regulations
Global Government Affairs, Regulations & Standards
Philips Healthcare

3000 Minuteman Road, Bldg 4, MS0135
Andover, MA 01810-1099  USA
PH-158 3178
Tel: +1 978 659 3178
Mob: +1 781 354 4798
Fax: +1 978 683 6337
E-mail: [email protected]<mailto:[email protected]>

The information contained in this message is confidential and may be legally 
privileged. The message is intended solely for the addressee(s). If you are not 
the intended recipient, you are hereby notified that any use, dissemination, or 
reproduction is strictly prohibited and may be unlawful. If you are not the 
intended recipient, please contact the sender by return e-mail and destroy all 
copies of the original message.



From: [email protected]<mailto:[email protected]> 
[mailto:[email protected]]<mailto:[mailto:[email protected]]> On Behalf Of Ron 
Pickard
Sent: Tuesday, January 10, 2012 11:44 AM
To: McInturff, Gary
Cc: [email protected]<mailto:[email protected]>
Subject: RE: 61000-4-2 3rd edition

Hi Gary,
I cannot find any 3rd Edition of IEC 61000-4-2 as published or in development 
on the IEC website under SC 77B I believe (if anyone knows differently, please 
let us know). As such, you might want to validate the 3rd Edition requirement 
from your customer.

Also, most regulatory-based immunity standards that I am familiar have ESD 
requirements of up to 4kV/8kV, while IEC 60601-1-2, 2nd Edition (latest that I 
have) has ESD requirements of up to 6kV/8kV. And, the much higher ESD threshold 
requirements as you've indicated generally originate from specific industry or 
customer requirements. So, those higher ESD thresholds are quite likely not 
regulatory requirements, but if anyone knows if the ESD requirements from IEC 
60601-1-2, 3rd Edition, please let us know. You might want to verify these 
thresholds from your customer.

IHTH.

Best regards,

Ron

From: McInturff, Gary 
[mailto:[email protected]]<mailto:[mailto:[email protected]]>
Sent: Tuesday, January 10, 2012 8:53 AM
To: [email protected]<mailto:[email protected]>
Subject: [PSES] 61000-4-2 3rd edition

I'm trying to track down a customer requirement for a piece of medical gear and 
apparently I'm a revision behind. They have a requirement for 15 and 30 kV air 
and contact discharge and they reference the 3rd edition standard. My current 
standard doesn't have levels that go that high. Normally I would look at 
60601-2  medical electrical equipment ....EMC compatibility.... But it only 
requires 6 and 8kV.

The standard is going through all the time consuming purchase request channels, 
until then can someone with the 3rd edition help me understand when the 15 and 
30kV is required. The device itself is used in patient monitoring and could be 
used in an operating theater.

We'll meet their requirements I'm just trying to figure out the paper trail 
that would lead them to that requirement, could be nothing more than a desired 
goal but I would like to understand it with your help

Thanks


Gary McInturff
Reliability/Compliance Engineer








Esterline Interface Technologies

Featuring
ADVANCED INPUT, MEMTRON, and LRE MEDICAL products



600 W. Wilbur Avenue
Coeur d'Alene, ID  83815-9496
Office:208-635-8306
Cell:  509 868 2279
Toll Free: 800-444-5923 X 1238
[email protected]<mailto:[email protected]>


www.esterline.com/interfacetechnologies<http://www.esterline.com/advancedinput>

Technology, Innovation, Performance...



-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]<mailto:[email protected]>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]<mailto:[email protected]>>
Mike Cantwell <[email protected]<mailto:[email protected]>>

For policy questions, send mail to:
Jim Bacher <[email protected]<mailto:[email protected]>>
David Heald <[email protected]<mailto:[email protected]>>
-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]<mailto:[email protected]>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]<mailto:[email protected]>>
Mike Cantwell <[email protected]<mailto:[email protected]>>

For policy questions, send mail to:
Jim Bacher <[email protected]<mailto:[email protected]>>
David Heald <[email protected]<mailto:[email protected]>>

________________________________
The information contained in this message may be confidential and legally 
protected under applicable law. The message is intended solely for the 
addressee(s). If you are not the intended recipient, you are hereby notified 
that any use, forwarding, dissemination, or reproduction of this message is 
strictly prohibited and may be unlawful. If you are not the intended recipient, 
please contact the sender by return e-mail and destroy all copies of the 
original message.

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to