Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread Kay Diederichs

Am 20:59, schrieb Harry Powell:
...

I think there may be issues with collecting data too finely with a
Pilatus, even in shutterless mode. I don't know where the problems arise
(can't be shutter/rotation axis synchronisation), but it seems to be the
normal thing in XDS (which should have no problems with fine
phi-slicing) to use the PATCH_SHUTTER_PROBLEM=TRUE that Martin
Hallberg suggested, which looks a bit like a fudge to me (but I expect
Kay to correct me on that!).

...

Hi Harry,

no, it is a misunderstanding that the normal thing is to use 
PATCH_SHUTTER_PROBLEM=TRUE in XDS; rather, it may be the last resort to 
try before you abandon a dataset, and if the spindle/shutter 
de-synchronization is so poor that this fudge needs to be used, then the 
BL hardware needs to be fixed before other datasets are collected.


Two other points:
a) one may distinguish weak data resulting in high R-factors from 
problems with hardware or radiation damage resulting in high R-factors 
by looking at (I/sigma)^asymptotic as defined in Acta D66, 733 ( 
http://dx.doi.org/10.1107/S0907444910014836 ).
b) Marcus Müller (SLS) has shown (to me, unambiguously) that the best 
data are obtained from the Pilatus if delta-phi is 1/2 to 1/4 of the 
mosaicity (as reported by XDS). However, for the best data, other 
parameters of the experiment (basically the transmission and the spindle 
speed) have to be optimized as well. The finding of Tassos that 0.5-1 
degree data gave the best result in a particular case might indicate 
that these other factors of his experiment were not optimal (it is 
indeed not trivial to get them right).


best,
Kay



smime.p7s
Description: S/MIME Cryptographic Signature


Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread harry powell

Hi All

Both Martin and Kay ( in a later message) have misinterpreted what I  
wrote - what I meant was that it seems normal in using XDS with  
Pilatus data, not the normal thing to do with data from other detectors.


I had found a number of scripts on the web that deal specifically  
with processing Pilatus data with XDS that seemed to use  
PATCH_SHUTTER_PROBLEM=TRUE - what I didn't notice was that the  
commands had all been commented out, and that FALSE, the default,  
was in operation.


So, in either case I was wrong. Apologies to Kay and Wolfgang, and  
all who write and use scripts for processing Pilatus data with XDS.


On 5 Nov 2010, at 18:33, Martin Hallberg wrote:


Hi Harry,

On Nov 5, 2010, at 5:45 PM, Harry Powell wrote:


I think there may be issues with collecting data too finely with a  
Pilatus, even in shutterless mode. I don't know where the problems  
arise (can't be shutter/rotation axis synchronisation), but it  
seems to be the normal thing in XDS (which should have no problems  
with fine phi-slicing) to use the PATCH_SHUTTER_PROBLEM=TRUE  
that Martin Hallberg suggested, which looks a bit like a fudge to  
me (but I expect Kay to correct me on that!).


It is not the normal thing in XDS but it is perhaps a relatively  
common solution to shutter/spindle synchronisation problems  
discovered afterwards (always process directly at the beam line!).  
The default in XDS is indeed PATCH_SHUTTER_PROBLEM=FALSE


Compensating like this is of course not the best (go and  
recollect!) but still way better than unusable data in the  
meantime. In the case Sergei originally described it would at least  
indicate what the problem may be. Sergei did not say which detector  
was used for the data collection so we don't know if it was a  
Pilatus or a CCD. Maybe Sergei can fill us in on the details?


BTW, any views in the community on crystal lifetime with continuous  
data collection like on the Pilatus (or AXIOM) versus letting the  
crystal rest/cool/die(?) a second between frames on a CCD?


Cheers,

Martin







Harry
--
Dr Harry Powell,
MRC Laboratory of Molecular Biology,
Hills Road,
Cambridge,
CB2 0QH


Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread Guenter Fritz

Dear Ronnie,
we are working with weak diffracting crystals. Many tests (and taking 
into account  Marcus Muellers results) showed that 0.3-0.5 of XDS 
mosaicity, (very) low dose, and high redundancy give the best results. 
Our crystals diffract between 7 and 4 A. At low dose I do not check 
diffraction really by eye. I collect 200 images (fine slicing), process 
and check XDS statistics. (200 images is not much when the dataset 
comprises 9000-12000 images). Especially for anomalous data this 
strategy was pivotal.
We did also comparisons between CCD and PILATUS. We used one long 
crystal and collected on end end a MAD dataset using a  CCD and 
afterwards  at the other end a MAD dataset on PILATUS, same dose per 
degree.  Results were absolutely clear that the fine slicing give much 
better results.

Best,
Guenter

Dear Tassos,

I'm interested in your third point. Do you have any explanation for 
why 0.5-1 degrees oscillation gave better data? Purely due to the fact 
that the crystals survived longer and thus yielded higher redundancy 
data, or also other parameters?
Also do anyone know where the threshold lies for when /not/ to use 
fine phi slicing on the PILATUS? ie, at what level of diffraction 
would one need to increase the exposure (and oscillation in order to 
still get redundant data)?


We'll be in a similar position in the coming weeks with data 
collection using PILATUS detectors, and would like to maximize the 
potential data quality from our weak diffracting crystals. Any input 
on this would be greatly appreciated!


Cheers,
Ronnie Berntsson



On Nov 5, 2010, at 16:16, Anastassis Perrakis wrote:


three additional points:

1.


OTOH, if The diffraction is quite weak, one may be limited by counting
statistics.  This also cannot be overcome by processing.


As JIm suggests above then, maybe you should look if the 15% Rmerge 
is almost reasonable given the specific I/sigI at low resolution?



2. If there is one thing I do not like in XDS, is that there is no 
(or I have failed to find) statistics of I/sigI and Rmerge as 
function of image.

Have a look at the SCALA output. Maybe some images are bad?

3. making too fine slices of too weak diffraction images ends up with 
either too weak counting statistics or inability to 'lock' the 
refinement.
we did that for one crystal form, collecting 0.1, 0.2, 0.35, 0.5, 
0.7, 1.0 from various crystals (with the same dose per degree, at SLS 
using a PILATUS, mosaicity 0.4-0.6) in an attempt to get better Se 
signal. We miserably failed to get any useful signal at the end, but 
learned that for these very weak diffracting plates (submicron) 
collecting 0.5-1.0 degrees was actually giving at the end better data.


A.





--
***

Priv.Doz.Dr. Guenter Fritz
Fachbereich Biologie
Universitaet Konstanz
http://www.biologie.uni-konstanz.de/fritz

e-mail: guenter.fr...@uni-konstanz.de

e-mail: guenter.fr...@uniklinik-freiburg.de
http://www.uniklinik-freiburg.de/neuropathologie/live/forschung/ag-g-fritz.html
Tel.: +49 761 270 5078
Fax.: +49 761 270 5050



Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread Sergei Strelkov

Dear All,

first of all, I would like to thank the many
good people who have responded to my query.
Yet another truly interesting discussion on this BB!

As a partial summary, two points:

1. Few people suggested that our high Rmerge problem
could be caused by experimental troubles
like phi angle imprecision rather than the crystal itself.
Thus far we could not find indications for that but maybe
we did not look carefully enough. However, the
frames ARE weak, and I tend to think
that this is the main cause of the problem.

2.  Few people argued that data processing programs
(XDS in particular) 'should' handle thin frames
as efficiently as thicker frames...
After digesting these (very useful!) arguments,
I still tend to think that the best proof would be in the pudding,
i.e. trying to pool thin frames into thicker frames
(but I do not have an immediate means of doing this... )

Sergei




Dear All,

I am processing a dataset collected (not by me) with 0.1 degree
oscillations.
The diffraction is quite weak even though there is a clean diffraction
pattern to about 3A.

Either Mosflm or XDS processes the data readily with +/- default settings
but both yield a high overall Rmerge of about 0.23 in the expected symmetry.
Processing in P1 yields an overall Rmerge of ~0.18, but what is
especially disappointing
is that Rmerge is as high as 0.15 at ~5A resolution already.

The question is, how can we process the data so that the merging statistics
becomes more reasonable?

Apparent mosaicity turns out to be ~0.5A. My naive way of thinking is
to try treating each five consecutive frames as a single 0.5 degree frame.
Does anyone have experience with this?

Many thanks in advance,
Sergei





--
Prof. Sergei V. Strelkov
Laboratory for Biocrystallography
Department of Pharmaceutical Sciences, Katholieke Universiteit Leuven
ON2, Campus Gasthuisberg, Herestraat 49 bus 822, 3000 Leuven, Belgium
Work phone: +32 16 330845  Fax: +32 16 323469 OR +32 16 323460
Mobile: +32 486 294132
Lab pages: http://pharm.kuleuven.be/anafar


Re: [ccp4bb] High Rmerge with thin frames

2010-11-06 Thread Kay Diederichs

Am 20:59, schrieb Sergei Strelkov:
...

I still tend to think that the best proof would be in the pudding,
i.e. trying to pool thin frames into thicker frames
(but I do not have an immediate means of doing this... )


Sergei,

try to locate 2pck and 2pck.man which were part of former XDS 
distributions (there's a link to a Linux binary in article 
http://strucbio.biologie.uni-konstanz.de/xdswiki/index.php/2pck ).


2pck can add several images, and writes them out as one new image. After 
the conversion, you may only have to change the filetype, the 
oscillation range and the number of images in your XDS.INP . Pls tell us 
what you find!


HTH,

Kay




smime.p7s
Description: S/MIME Cryptographic Signature


[ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Sergei Strelkov

Dear All,

I am processing a dataset collected (not by me) with 0.1 degree 
oscillations.
The diffraction is quite weak even though there is a clean diffraction 
pattern to about 3A.


Either Mosflm or XDS processes the data readily with +/- default settings
but both yield a high overall Rmerge of about 0.23 in the expected symmetry.
Processing in P1 yields an overall Rmerge of ~0.18, but what is 
especially disappointing

is that Rmerge is as high as 0.15 at ~5A resolution already.

The question is, how can we process the data so that the merging statistics
becomes more reasonable?

Apparent mosaicity turns out to be ~0.5A. My naive way of thinking is
to try treating each five consecutive frames as a single 0.5 degree frame.
Does anyone have experience with this?

Many thanks in advance,
Sergei


Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Tim Gruene
Dear Sergei,

with only 3A data and 0.1 deg frame width my first guess would be radiation
damage.
In that case there is little you can do - the Rmerge might just be realistic.
XDS has not problem dealing with thin frames (on the contrary!) and it won't
help pooling frames together.

Check out the statistics and plots from scala, they should tell you more about
whether e.g. the scaling per frame increases with frame number, which is quite a
sure sign of radiation damage.

I would suggest to pick a new crystal and collect data with 0.5deg/frame.

Tim

On Fri, Nov 05, 2010 at 09:40:57AM +0100, Sergei Strelkov wrote:
 Dear All,

 I am processing a dataset collected (not by me) with 0.1 degree  
 oscillations.
 The diffraction is quite weak even though there is a clean diffraction  
 pattern to about 3A.

 Either Mosflm or XDS processes the data readily with +/- default settings
 but both yield a high overall Rmerge of about 0.23 in the expected symmetry.
 Processing in P1 yields an overall Rmerge of ~0.18, but what is  
 especially disappointing
 is that Rmerge is as high as 0.15 at ~5A resolution already.

 The question is, how can we process the data so that the merging statistics
 becomes more reasonable?

 Apparent mosaicity turns out to be ~0.5A. My naive way of thinking is
 to try treating each five consecutive frames as a single 0.5 degree frame.
 Does anyone have experience with this?

 Many thanks in advance,
 Sergei

-- 
--
Tim Gruene
Institut fuer anorganische Chemie
Tammannstr. 4
D-37077 Goettingen

phone: +49 (0)551 39 22149

GPG Key ID = A46BEE1A



signature.asc
Description: Digital signature


Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Dirk Kostrewa

Dear Sergei,

since your Rmerge is high at low resolution even in P1, my guess is, 
that there was a problem either with the crystal or with the data 
collection. Fine slicing should improve the data quality, because your 
get a better description of the reference profiles and reduce the 
background. Different merging of fine-sliced images into 
coarser-sliced images was already systematically done by Wolfgang 
Kabsch, with the result, that finer slices gave better results, although 
slices much finer than half the full mosaicity angle of a reflection 
didn't improve the results much further (I've heard a talk given by 
Wolfgang, but never found a publication about these results). There is a 
publication in 1999 about fine-slicing by Jim Pflugrath, that you might 
want to read about this (Acta Cryst D55, 1718-1725). So, merging 
fine-sliced images into coarse-sliced probably won't help.


Anyway, good luck!

Dirk.

Am 05.11.10 09:40, schrieb Sergei Strelkov:

Dear All,

I am processing a dataset collected (not by me) with 0.1 degree 
oscillations.
The diffraction is quite weak even though there is a clean diffraction 
pattern to about 3A.


Either Mosflm or XDS processes the data readily with +/- default settings
but both yield a high overall Rmerge of about 0.23 in the expected 
symmetry.
Processing in P1 yields an overall Rmerge of ~0.18, but what is 
especially disappointing

is that Rmerge is as high as 0.15 at ~5A resolution already.

The question is, how can we process the data so that the merging 
statistics

becomes more reasonable?

Apparent mosaicity turns out to be ~0.5A. My naive way of thinking is
to try treating each five consecutive frames as a single 0.5 degree 
frame.

Does anyone have experience with this?

Many thanks in advance,
Sergei


--

***
Dirk Kostrewa
Gene Center Munich, A5.07
Department of Biochemistry
Ludwig-Maximilians-Universität München
Feodor-Lynen-Str. 25
D-81377 Munich
Germany
Phone:  +49-89-2180-76845
Fax:+49-89-2180-76999
E-mail: kostr...@genzentrum.lmu.de
WWW:www.genzentrum.lmu.de
***


Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Savvas Savvides
Dear Sergei
how much do the refined unit cell parameters (given a fixed detector distance) 
vary as a function of frame number? We have been using such initial diagnostic 
approach to trace radiation damage issues (among other problems) for a number 
of crystal forms that maximally diffracted in the 3.8-4.5 angstron range. 

I copy/paste an excerpt from the XDS wiki that reinforces this approach:

REFINE(INTEGRATE)
The defaults (REFINE(INTEGRATE)=DISTANCE BEAM ORIENTATION CELL) could be 
modified by omitting DISTANCE, because one should assume that the distance is 
constant. This is particularly recommended if SPACE_GROUP_NUMBER=0 or 1. 
Furthermore, by fixing the distance one can better see from the results of the 
refinement whether the cell parameters are stable, or whether they change due 
to radiation damage. There are situations when one wants to reduce the number 
of parameters to be refined even more, see Optimization.

Best regards
Savvas


Savvas Savvides
Unit for Structural Biology @ L-ProBE
Ghent University
K.L. Ledeganckstraat 35, 9000 Ghent, Belgium
Ph. +32  (0)472 928 519 http://www.LProBE.ugent.be/xray.html



On 05 Nov 2010, at 09:40, Sergei Strelkov wrote:

 Dear All,
 
 I am processing a dataset collected (not by me) with 0.1 degree oscillations.
 The diffraction is quite weak even though there is a clean diffraction 
 pattern to about 3A.
 
 Either Mosflm or XDS processes the data readily with +/- default settings
 but both yield a high overall Rmerge of about 0.23 in the expected symmetry.
 Processing in P1 yields an overall Rmerge of ~0.18, but what is especially 
 disappointing
 is that Rmerge is as high as 0.15 at ~5A resolution already.
 
 The question is, how can we process the data so that the merging statistics
 becomes more reasonable?
 
 Apparent mosaicity turns out to be ~0.5A. My naive way of thinking is
 to try treating each five consecutive frames as a single 0.5 degree frame.
 Does anyone have experience with this?
 
 Many thanks in advance,
 Sergei



Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Martin Hallberg
Dear Sergei,
Did you check the mean intensity as function of spindle position statistics 
in the CORRECT.LP file?
Any (even minute) shutter problems will affect these thin frames significantly. 
If this is indeed the problem, you could then try to set:

PATCH_SHUTTER_PROBLEM=TRUE

for the CORRECT stage in XDS.

Cheers,

Martin






On Nov 5, 2010, at 9:40 AM, Sergei Strelkov wrote:

 Dear All,
 
 I am processing a dataset collected (not by me) with 0.1 degree oscillations.
 The diffraction is quite weak even though there is a clean diffraction 
 pattern to about 3A.
 
 Either Mosflm or XDS processes the data readily with +/- default settings
 but both yield a high overall Rmerge of about 0.23 in the expected symmetry.
 Processing in P1 yields an overall Rmerge of ~0.18, but what is especially 
 disappointing
 is that Rmerge is as high as 0.15 at ~5A resolution already.
 
 The question is, how can we process the data so that the merging statistics
 becomes more reasonable?
 
 Apparent mosaicity turns out to be ~0.5A. My naive way of thinking is
 to try treating each five consecutive frames as a single 0.5 degree frame.
 Does anyone have experience with this?
 
 Many thanks in advance,
 Sergei


[ccp4bb] RE : [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread SHEPARD William
Dear Sergei, 
 
It is difficult to say without looking at XDS or MOSFLM logfiles, but this sort 
of problem (high  flat Rsym in all resolution bins) sounds like the crystal 
vibrating wildly in the cryostream. You could ask the data collector the 
following questions: 1) Was the cryostream misaligned or too far from the 
sample? 2) Is the crystal mounted in a long  wobbly loop? 3) Was the incident 
beam varying (in XDS check the average background count for each image)?
 
I wrote a web page to troubleshoot this sort of problem when I was beamline 
scientist on ID29 at the ESRF. Unfortunately the pages are no longer available, 
but I might have a backup if you are interested.
 
Cheers, 
Bill 
 



De: CCP4 bulletin board de la part de Sergei Strelkov
Date: ven. 05/11/2010 09:40
À: CCP4BB@JISCMAIL.AC.UK
Objet : [ccp4bb] High Rmerge with thin frames



Dear All,

I am processing a dataset collected (not by me) with 0.1 degree
oscillations.
The diffraction is quite weak even though there is a clean diffraction
pattern to about 3A.

Either Mosflm or XDS processes the data readily with +/- default settings
but both yield a high overall Rmerge of about 0.23 in the expected symmetry.
Processing in P1 yields an overall Rmerge of ~0.18, but what is
especially disappointing
is that Rmerge is as high as 0.15 at ~5A resolution already.

The question is, how can we process the data so that the merging statistics
becomes more reasonable?

Apparent mosaicity turns out to be ~0.5A. My naive way of thinking is
to try treating each five consecutive frames as a single 0.5 degree frame.
Does anyone have experience with this?

Many thanks in advance,
Sergei


Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Peter Cherepanov
Hi Sergei, 

such fine slicing during data collection would suggest a large cell. How many 
reflections are you merging? And what is the redundancy (in the expected 
symmetry)? Rmerge tends to go up with more reflections added. 

Peter

On 5 Nov 2010, at 08:40, Sergei Strelkov wrote:

 Dear All,
 
 I am processing a dataset collected (not by me) with 0.1 degree 
 oscillations.
 The diffraction is quite weak even though there is a clean diffraction 
 pattern to about 3A.
 
 Either Mosflm or XDS processes the data readily with +/- default settings
 but both yield a high overall Rmerge of about 0.23 in the expected symmetry.
 Processing in P1 yields an overall Rmerge of ~0.18, but what is 
 especially disappointing
 is that Rmerge is as high as 0.15 at ~5A resolution already.
 
 The question is, how can we process the data so that the merging statistics
 becomes more reasonable?
 
 Apparent mosaicity turns out to be ~0.5A. My naive way of thinking is
 to try treating each five consecutive frames as a single 0.5 degree frame.
 Does anyone have experience with this?
 
 Many thanks in advance,
 Sergei


Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Jim Pflugrath
In general, if the Rmeas or Rmerge is high in the low resolution shells,
then something is not optimal with the data collection.

Bill Shepard has already mentioned the loop vibrating or moving in the
cryogenic gas flow.  Other problems could be the goniometer head was loose,
the magnet was loose, the pin was loose, etc.  There could be excessive
shutter shutter.  The shutter and the crystal rotation could be poorly
synchronized.  There could be some other vibration in the system which could
cause the X-ray flux to vary quite a bit.  It could be as simple as the
cycling of the cooling for the monochromator, the room or hutch, or the
X-ray source itself.  All of these would affect non-thin-images as wells.

As already mentioned, combining thin-images into wider images will not
overcome these problems.

If Rmerge or Rmeas was 40% or more in the low resolution shells, then the
diffraction pattern is probably mis-indexed, but that kind of high value was
not reported here.

OTOH, if The diffraction is quite weak, one may be limited by counting
statistics.  This also cannot be overcome by processing.

Jim

-Original Message-
From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of Sergei
Strelkov
Sent: Friday, November 05, 2010 3:41 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] High Rmerge with thin frames

Dear All,

I am processing a dataset collected (not by me) with 0.1 degree
oscillations.
The diffraction is quite weak even though there is a clean diffraction
pattern to about 3A.
...


Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Anastassis Perrakis

three additional points:

1.

OTOH, if The diffraction is quite weak, one may be limited by  
counting

statistics.  This also cannot be overcome by processing.


As JIm suggests above then, maybe you should look if the 15% Rmerge is  
almost reasonable given the specific I/sigI at low resolution?



2. If there is one thing I do not like in XDS, is that there is no (or  
I have failed to find) statistics of I/sigI and Rmerge as function of  
image.

Have a look at the SCALA output. Maybe some images are bad?

3. making too fine slices of too weak diffraction images ends up with  
either too weak counting statistics or inability to 'lock' the  
refinement.
we did that for one crystal form, collecting 0.1, 0.2, 0.35, 0.5, 0.7,  
1.0 from various crystals (with the same dose per degree, at SLS using  
a PILATUS, mosaicity 0.4-0.6) in an attempt to get better Se signal.  
We miserably failed to get any useful signal at the end, but learned  
that for these very weak diffracting plates (submicron) collecting  
0.5-1.0 degrees was actually giving at the end better data.


A.


Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Jacob Keller
3. making too fine slices of too weak diffraction images ends up with 
either too weak counting statistics or inability to 'lock' the 
refinement.
we did that for one crystal form, collecting 0.1, 0.2, 0.35, 0.5, 0.7, 
1.0 from various crystals (with the same dose per degree, at SLS using  a 
PILATUS, mosaicity 0.4-0.6) in an attempt to get better Se signal.  We 
miserably failed to get any useful signal at the end, but learned  that 
for these very weak diffracting plates (submicron) collecting  0.5-1.0 
degrees was actually giving at the end better data.



Perhaps the reason for the better data was an instance of the 
redundancy-vs-long-exposure dilemma. Given, say, 1deg exposures: should one 
collect 500x1s exposures or 250x2s exposures? I think this has been 
examined, and while it does depend on the details of the parameters, it is 
often better to collect 250x2s exposures because there is a flat rate per 
frame noise level in the detector. I am wondering whether you just 
increased the number of flat rates in your data sets by increasing the 
number of frames while keeping the exposure/degree equal?


JPK 


Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Anastassis Perrakis


On Nov 5, 2010, at 16:57, Ronnie Berntsson wrote:


Dear Tassos,

I'm interested in your third point. Do you have any explanation for  
why 0.5-1 degrees oscillation gave better data? Purely due to the  
fact that the crystals survived longer and thus yielded higher  
redundancy data, or also other parameters?


No, x-ray beam survival had nothing to do with this. The dose was the  
same per degree, so the damage was the same in all, at least on  
principle, and from what I see in practice.


I simply think that the very low partiality of all reflections ends up  
with counts that are just above the noise (even for a well set up  
experiment and an excellent detector). The diffracting volume of these  
crystals in some orientation was really small! 1x40x70 microns, only,  
so the signals were very low. I think that the end integration goes  
wrong because images cannot be well refined with so low-count data. As  
soon as the reflections were stronger, things 'catch up'.


Also do anyone know where the threshold lies for when not to use  
fine phi slicing on the PILATUS? ie, at what level of diffraction  
would one need to increase the exposure (and oscillation in order to  
still get redundant data)?


In general, the slicing of the Pilatus works great for us. My only  
negative experience was really really small crystals.


We'll be in a similar position in the coming weeks with data  
collection using PILATUS detectors, and would like to maximize the  
potential data quality from our weak diffracting crystals. Any input  
on this would be greatly appreciated!


I would aim to be able to see nice spots, to at least 3.5 A: these  
would be enough to 'lock' the orientations, and I would expect either  
XDS or MOSFLM to integrate even very low signals.


A.



Cheers,
Ronnie Berntsson



On Nov 5, 2010, at 16:16, Anastassis Perrakis wrote:


three additional points:

1.

OTOH, if The diffraction is quite weak, one may be limited by  
counting

statistics.  This also cannot be overcome by processing.


As JIm suggests above then, maybe you should look if the 15% Rmerge  
is almost reasonable given the specific I/sigI at low resolution?



2. If there is one thing I do not like in XDS, is that there is no  
(or I have failed to find) statistics of I/sigI and Rmerge as  
function of image.

Have a look at the SCALA output. Maybe some images are bad?

3. making too fine slices of too weak diffraction images ends up  
with either too weak counting statistics or inability to 'lock' the  
refinement.
we did that for one crystal form, collecting 0.1, 0.2, 0.35, 0.5,  
0.7, 1.0 from various crystals (with the same dose per degree, at  
SLS using a PILATUS, mosaicity 0.4-0.6) in an attempt to get better  
Se signal. We miserably failed to get any useful signal at the end,  
but learned that for these very weak diffracting plates (submicron)  
collecting 0.5-1.0 degrees was actually giving at the end better  
data.


A.




P please don't print this e-mail unless you really need to
Anastassis (Tassos) Perrakis, Principal Investigator / Staff Member
Department of Biochemistry (B8)
Netherlands Cancer Institute,
Dept. B8, 1066 CX Amsterdam, The Netherlands
Tel: +31 20 512 1951 Fax: +31 20 512 1954 Mobile / SMS: +31 6 28 597791






Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Anastassis Perrakis
In the Pilatus mode these are open-shutter experiment, where the  
Pilatus integrates over different times
- all these exposure times are slower than the frequency of the  
detector, as far as I understand the setup.


So, the crystal gets the 'full blast' in all cases, and the blast is  
the same for the same rotation, as we did the experiments.


btw, I have the suspicion, that for our system, it was better to do  
'100 sec' in the pilatus, than 100x1 sec and read for 3-4 secs
with a CCD in between, but I have no rigorous proof for that. Its just  
an observation that should be treated with caution.


Tassos

On Nov 5, 2010, at 17:11, Jacob Keller wrote:


3. making too fine slices of too weak diffraction images ends up with
either too weak counting statistics or inability to 'lock' the
refinement.
we did that for one crystal form, collecting 0.1, 0.2, 0.35, 0.5,  
0.7,
1.0 from various crystals (with the same dose per degree, at SLS  
using  a
PILATUS, mosaicity 0.4-0.6) in an attempt to get better Se signal.   
We
miserably failed to get any useful signal at the end, but learned   
that
for these very weak diffracting plates (submicron) collecting   
0.5-1.0

degrees was actually giving at the end better data.



Perhaps the reason for the better data was an instance of the
redundancy-vs-long-exposure dilemma. Given, say, 1deg exposures:  
should one

collect 500x1s exposures or 250x2s exposures? I think this has been
examined, and while it does depend on the details of the parameters,  
it is
often better to collect 250x2s exposures because there is a flat  
rate per

frame noise level in the detector. I am wondering whether you just
increased the number of flat rates in your data sets by increasing  
the

number of frames while keeping the exposure/degree equal?

JPK



P please don't print this e-mail unless you really need to
Anastassis (Tassos) Perrakis, Principal Investigator / Staff Member
Department of Biochemistry (B8)
Netherlands Cancer Institute,
Dept. B8, 1066 CX Amsterdam, The Netherlands
Tel: +31 20 512 1951 Fax: +31 20 512 1954 Mobile / SMS: +31 6 28 597791






Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Harry Powell
Hi

I'd read Jim Pflugrath's 1999 paper in Acta D for a discussion on fine phi 
slicing - in general, if memory serves me correctly, he suggested using an 
oscillation angle ~0.5x the mosaic spread.

I think there may be issues with collecting data too finely with a Pilatus, 
even in shutterless mode. I don't know where the problems arise (can't be 
shutter/rotation axis synchronisation), but it seems to be the normal thing in 
XDS (which should have no problems with fine phi-slicing) to use the 
PATCH_SHUTTER_PROBLEM=TRUE that Martin Hallberg suggested, which looks a bit 
like a fudge to me (but I expect Kay to correct me on that!).

The important thing is to treat each data collection as a scientific 
experiment, and to treat each crystal as an individual. Use the strategy tools 
(e.g. BEST, DNA, Edna) to work out how to get the best data from each crystal - 
remember that the data collection itself is the last experimental step.

Any of the issues mentioned earlier in this thread will be exacerbated by very 
short exposure times - so it may be an idea to attenuate the beam and spend 
five minutes on a data collection rather than one

On 5 Nov 2010, at 15:57, Ronnie Berntsson wrote:

 Dear Tassos,
 
 I'm interested in your third point. Do you have any explanation for why 0.5-1 
 degrees oscillation gave better data? Purely due to the fact that the 
 crystals survived longer and thus yielded higher redundancy data, or also 
 other parameters?
 Also do anyone know where the threshold lies for when not to use fine phi 
 slicing on the PILATUS? ie, at what level of diffraction would one need to 
 increase the exposure (and oscillation in order to still get redundant data)?
 
 We'll be in a similar position in the coming weeks with data collection using 
 PILATUS detectors, and would like to maximize the potential data quality from 
 our weak diffracting crystals. Any input on this would be greatly appreciated!
 
 Cheers,
 Ronnie Berntsson
 
 
 
 On Nov 5, 2010, at 16:16, Anastassis Perrakis wrote:
 
 three additional points:
 
 1.
 
 OTOH, if The diffraction is quite weak, one may be limited by counting
 statistics.  This also cannot be overcome by processing.
 
 As JIm suggests above then, maybe you should look if the 15% Rmerge is 
 almost reasonable given the specific I/sigI at low resolution?
 
 
 2. If there is one thing I do not like in XDS, is that there is no (or I 
 have failed to find) statistics of I/sigI and Rmerge as function of image.
 Have a look at the SCALA output. Maybe some images are bad?
 
 3. making too fine slices of too weak diffraction images ends up with either 
 too weak counting statistics or inability to 'lock' the refinement.
 we did that for one crystal form, collecting 0.1, 0.2, 0.35, 0.5, 0.7, 1.0 
 from various crystals (with the same dose per degree, at SLS using a 
 PILATUS, mosaicity 0.4-0.6) in an attempt to get better Se signal. We 
 miserably failed to get any useful signal at the end, but learned that for 
 these very weak diffracting plates (submicron) collecting 0.5-1.0 degrees 
 was actually giving at the end better data.
 
 A.
 

Harry
--
Dr Harry Powell, MRC Laboratory of Molecular Biology, MRC Centre, Hills Road, 
Cambridge, CB2 0QH



Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Tim Gruene
Hello Tassos,

the data you are missing are available from XDS_ASCII.HKL and can e.g. be
generated with Kay Diederichs xdsstat, see
http://strucbio.biologie.uni-konstanz.de/xdswiki/index.php/XDSSTAT
According to the web site, it prints Rmeas instead of Rmerge, but since Rmeas
should anyhow be promoted and Rmerge abandoned, that's a good thing.


In case you have access to it, sadabs also generate some useful graphs from
XDS_ASCII.HKL

Cheers, Tim

On Fri, Nov 05, 2010 at 04:16:02PM +0100, Anastassis Perrakis wrote:
 2. If there is one thing I do not like in XDS, is that there is no (or I 
 have failed to find) statistics of I/sigI and Rmerge as function of  
 image.
 Have a look at the SCALA output. Maybe some images are bad?

 A.

-- 
--
Tim Gruene
Institut fuer anorganische Chemie
Tammannstr. 4
D-37077 Goettingen

phone: +49 (0)551 39 22149

GPG Key ID = A46BEE1A



signature.asc
Description: Digital signature


Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Martin Hallberg
Hi Harry,

On Nov 5, 2010, at 5:45 PM, Harry Powell wrote:
 
 I think there may be issues with collecting data too finely with a Pilatus, 
 even in shutterless mode. I don't know where the problems arise (can't be 
 shutter/rotation axis synchronisation), but it seems to be the normal thing 
 in XDS (which should have no problems with fine phi-slicing) to use the 
 PATCH_SHUTTER_PROBLEM=TRUE that Martin Hallberg suggested, which looks a 
 bit like a fudge to me (but I expect Kay to correct me on that!).

It is not the normal thing in XDS but it is perhaps a relatively common 
solution to shutter/spindle synchronisation problems discovered afterwards 
(always process directly at the beam line!). The default in XDS is indeed 
PATCH_SHUTTER_PROBLEM=FALSE 

Compensating like this is of course not the best (go and recollect!) but still 
way better than unusable data in the meantime. In the case Sergei originally 
described it would at least indicate what the problem may be. Sergei did not 
say which detector was used for the data collection so we don't know if it was 
a Pilatus or a CCD. Maybe Sergei can fill us in on the details?

BTW, any views in the community on crystal lifetime with continuous data 
collection like on the Pilatus (or AXIOM) versus letting the crystal 
rest/cool/die(?) a second between frames on a CCD?

Cheers,

Martin







Re: [ccp4bb] High Rmerge with thin frames

2010-11-05 Thread Chris Ulens

Compensating like this is of course not the best (go and recollect!)
but still way better than unusable data in the meantime. In the case
Sergei originally described it would at least indicate what the
problem may be. Sergei did not say which detector was used for the
data collection so we don't know if it was a Pilatus or a CCD. Maybe
Sergei can fill us in on the details?



The data that Sergei described were collected by yours truly at
beamline ID14-2 of the ESRF, so detector is a Quantum CCD. I collected
data for another protein crystal a few hours before with 0.5 degree
oscillations - no abnormal statistics here, so I don't suspect cryo or
loop instabilities are the underlying cause as suggested before.
The choice of 0.1 degree oscillations appears somewhat unfortunate in
retrospect, but this is what several programs (Best/mosflm/DNA)
suggested to reduce overlaps. The data were collected for the only
crystal that showed decent diffraction to ~3A out of more than 30-40
screened.

-Chris