Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-04-15 Thread Chavas Leo

Dear Joe --

Does anyone have information about how long it takes to set up a 96- 
well tray for the crystallization robots available?  Besides cost  
per tray and maintenance cost, another important feature we  
consider is the time for setting up a 96-well tray.  It is an  
important factor since we are talking about sub-microliter drops.


The japanese protein-crystallization system (PXS) can handle 7680  
drops / hour, from sample and solution disposal to sealing and  
incubator storage. You can have a look at the movie:


http://pfweis.kek.jp/system/PXS.html

or at the paper (PubMed 16929107).

HTH.

Kind regards.

Leo

Chavas Leonard, Ph.D.
Research Associate

Faculty of Life Sciences
The University of Manchester
The Michael Smith Building
Oxford Road
Manchester Lancashire
M13 9PT

Tel: +44(0)161-275-1586
e-mail: [EMAIL PROTECTED]




Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-04-15 Thread Frank von Delft
Actually, 3.5mins plate-to-plate for our 3-drop protocols (50:100, 
75:75, 100:50ul) on our mosquito. (Not sure what protocol Artem is 
using, must have lots of tip changes.)


Janet's reply reads slightly misleadingly: if you're concerned with 
plates/hour (which we are, since we have users show up with piles of 
20-40 trays on Friday evenings), then mosquito seems(*) to be quickest 
by quite a bit, as it does not have a wash cycle: as soon as one plate 
is done, you wang on the next one.


If you only do a plate at a time(**), then the wash cycle won't bother 
you, and phoenix is indeed as quick as mosquito.


phx.


(*) I must confess, though, I've not done extensive research recently. 
When I did poke around a year ago, the plate-to-plate timing was 
exceedingly hard to get out of sales-persons -- frustratingly.


(**) Mind you, I'm not sure why you'd only do a plate at a time: protein 
variation is far more useful to explore than chemical variation.





Artem Evdokimov wrote:


About 2 – 2.5 minutes on Mosquito for single-drop protocol, scaled 
according to the number of drops per well (7 minutes for 3-drop 
trays). About 25 minutes on a Tecan-derived platform (numbers vary 
greatly depending on the particular configuration).


Artem



*From:* CCP4 bulletin board [mailto:[EMAIL PROTECTED] *On Behalf 
Of *JOE CRYSTAL

*Sent:* Monday, April 14, 2008 5:10 PM
*To:* CCP4BB@JISCMAIL.AC.UK
*Subject:* Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

Hi,


Does anyone have information about how long it takes to set up a 
96-well tray for the crystallization robots available? Besides cost 
per tray and maintenance cost, another important feature we consider 
is the time for setting up a 96-well tray. It is an important factor 
since we are talking about sub-microliter drops.



Best,


Joe

On Fri, Jan 18, 2008 at 12:28 PM, Lisa A Nagy [EMAIL PROTECTED] 
mailto:[EMAIL PROTECTED] wrote:


Al's Oil on the plates:
What a nightmare!!!
The oil creeps up the plate and over the sides. It dissolves adhesives.
It makes me say bad words in multiple languages.
Bigger drops + no oil = fewer bad words.

Lisa

--
Lisa A. Nagy, Ph.D.
University of Alabama-Birmingham
[EMAIL PROTECTED] mailto:[EMAIL PROTECTED]

-Original Message-
From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK 
mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of


Patrick Shaw Stewart
Sent: Friday, January 18, 2008 2:20 AM
To: CCP4BB@JISCMAIL.AC.UK mailto:CCP4BB@JISCMAIL.AC.UK

Subject: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

One thing that people often overlook is that quite a lot of protein
can be lost by denaturation on the surface of the drop. This is more
significant for smaller drops. Two suggestions: (1) increase the
proportion of protein in the - technical term - teeny drop to say two
thirds and (2) cover the drops with oil eg Al's oils
(silicone/paraffin). You still get vapor diffusion though the oil ,
and you'd like to slow up equilibration. of course (2) slows up the
robotics a little, but both should be trivial to set up..



Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-04-15 Thread Yusuke Yamada

Thanks Leo for your introduction.

However, the link is broken.
Please look at the following page.
http://pfweis.kek.jp/protein/Robot/PXS/index-e.html

Best regards,

Yusuke

---
Yusuke Yamada, Ph.D.
Photon Factory
High Energy Accelerator Research Organization
[EMAIL PROTECTED]
---


Chavas Leo wrote:

Dear Joe --

Does anyone have information about how long it takes to set up a 
96-well tray for the crystallization robots available?  Besides cost 
per tray and maintenance cost, another important feature we consider 
is the time for setting up a 96-well tray.  It is an important factor 
since we are talking about sub-microliter drops.


The japanese protein-crystallization system (PXS) can handle 7680 drops 
/ hour, from sample and solution disposal to sealing and incubator 
storage. You can have a look at the movie:


http://pfweis.kek.jp/system/PXS.html

or at the paper (PubMed 16929107).

HTH.

Kind regards.

Leo

Chavas Leonard, Ph.D.
Research Associate

Faculty of Life Sciences
The University of Manchester
The Michael Smith Building
Oxford Road
Manchester Lancashire
M13 9PT

Tel: +44(0)161-275-1586
e-mail: [EMAIL PROTECTED]





Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-04-15 Thread Tom Walter
We use Cartesian Honeybee X8 machines (8 tips). They take about 10 minutes to 
set a 96-drop plate including the washes of the tips. 3 or 4 drops per 
condition wouldnt take much longer. Optimisation and additive/detergent screens 
take a little less time.  

The plates are pipetted under a close-fitting cover to (virtually) eliminate 
evaporation, which IMO is better than a humidity chamber. Consumable costs 
extend to isopropanol and water, with the occasional replacement valve or tip.

Since people here also tend to turn up at beer o'clock on a Friday evening 
(must be an Oxford thing...) we have two machines (and another one imminent) to 
increase throughput.

HTH
Tom

**  Tom Walter B.Sc. M.Res.   **
** Oxford Protein Production FacilityTel: +44 (0)1865 287747  **
** Wellcome Trust Centre for Human Genetics  Fax: +44 (0)1865 287547  **
** Roosevelt Drive   [EMAIL PROTECTED]   **
** Headington, Oxford OX3 7BNhttp://www.oppf.ox.ac.uk **


 Original message 
Date: Mon, 14 Apr 2008 17:10:26 -0400
From: JOE CRYSTAL [EMAIL PROTECTED]  
Subject: Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot  
To: CCP4BB@JISCMAIL.AC.UK

   Hi,

   Does anyone have information about how long it takes to set up
   a 96-well tray for the crystallization robots available? 
   Besides cost per tray and maintenance cost, another important
   feature we consider is the time for setting up a 96-well
   tray.  It is an important factor since we are talking about
   sub-microliter drops.

   Best,

   Joe

   On Fri, Jan 18, 2008 at 12:28 PM, Lisa A Nagy
   [EMAIL PROTECTED] wrote:

 Al's Oil on the plates:
 What a nightmare!!!
 The oil creeps up the plate and over the sides. It dissolves
 adhesives.
 It makes me say bad words in multiple languages.
 Bigger drops + no oil = fewer bad words.
 Lisa
 --
 Lisa A. Nagy, Ph.D.
 University of Alabama-Birmingham
 [EMAIL PROTECTED]

 -Original Message-
 From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On
 Behalf Of
 Patrick Shaw Stewart
 Sent: Friday, January 18, 2008 2:20 AM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

 One thing that people often overlook is that quite a lot of
 protein
 can be lost by denaturation on the surface of the drop.
  This is more
 significant for smaller drops.  Two suggestions: (1)
 increase the
 proportion of protein in the - technical term - teeny drop
 to say two
 thirds and (2) cover the drops with oil eg Al's oils
 (silicone/paraffin).  You still get vapor diffusion though
 the oil ,
 and you'd like to slow up equilibration.  of course (2)
 slows up the
 robotics a little, but both should be trivial to set up..


Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-04-15 Thread Jochen Muller-Dieckmann


Off the top of my head:

Phoenix Robot: ~2min (incl. reformatting)
Hydra2plusOne: ~5min (incl. reformatting and washing of needles)
Mosquito: ~2min (no reformatting, needles are disposable)
Cartesian: ~20min (incl. washing)

At 11:10 PM 4/14/2008, JOE CRYSTAL wrote:

Hi,


Does anyone have information about how long it takes to set up a 
96-well tray for the crystallization robots available?  Besides cost 
per tray and maintenance cost, another important feature we consider 
is the time for setting up a 96-well tray.  It is an important 
factor since we are talking about sub-microliter drops.



Best,


Joe

On Fri, Jan 18, 2008 at 12:28 PM, Lisa A Nagy 
mailto:[EMAIL PROTECTED][EMAIL PROTECTED] wrote:

Al's Oil on the plates:
What a nightmare!!!
The oil creeps up the plate and over the sides. It dissolves adhesives.
It makes me say bad words in multiple languages.
Bigger drops + no oil = fewer bad words.

Lisa
--
Lisa A. Nagy, Ph.D.
University of Alabama-Birmingham
mailto:[EMAIL PROTECTED][EMAIL PROTECTED]

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Patrick Shaw Stewart
Sent: Friday, January 18, 2008 2:20 AM
To: mailto:CCP4BB@JISCMAIL.AC.UKCCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

One thing that people often overlook is that quite a lot of protein
can be lost by denaturation on the surface of the drop.  This is more
significant for smaller drops.  Two suggestions: (1) increase the
proportion of protein in the - technical term - teeny drop to say two
thirds and (2) cover the drops with oil eg Al's oils
(silicone/paraffin).  You still get vapor diffusion though the oil ,
and you'd like to slow up equilibration.  of course (2) slows up the
robotics a little, but both should be trivial to set up..





___

 Jochen Muller-Dieckmann, Ph.D.

 Team Leader

 EMBL Hamburg Outstation
 c/o DESY, Notkestr. 85
 D-22603 Hamburg
 Germany

 Ph: +49-40-89902-201FAX: +49-40-89902-215
___


http://www.embl-hamburg.de/services/crystallisation 

Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-04-14 Thread JOE CRYSTAL
Hi,


Does anyone have information about how long it takes to set up a 96-well
tray for the crystallization robots available?  Besides cost per tray and
maintenance cost, another important feature we consider is the time for
setting up a 96-well tray.  It is an important factor since we are talking
about sub-microliter drops.


Best,


Joe

On Fri, Jan 18, 2008 at 12:28 PM, Lisa A Nagy [EMAIL PROTECTED] wrote:

 Al's Oil on the plates:
 What a nightmare!!!
 The oil creeps up the plate and over the sides. It dissolves adhesives.
 It makes me say bad words in multiple languages.
 Bigger drops + no oil = fewer bad words.

 Lisa
 --
 Lisa A. Nagy, Ph.D.
 University of Alabama-Birmingham
 [EMAIL PROTECTED]

 -Original Message-
 From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
 Patrick Shaw Stewart
 Sent: Friday, January 18, 2008 2:20 AM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

 One thing that people often overlook is that quite a lot of protein
 can be lost by denaturation on the surface of the drop.  This is more
 significant for smaller drops.  Two suggestions: (1) increase the
 proportion of protein in the - technical term - teeny drop to say two
 thirds and (2) cover the drops with oil eg Al's oils
 (silicone/paraffin).  You still get vapor diffusion though the oil ,
 and you'd like to slow up equilibration.  of course (2) slows up the
 robotics a little, but both should be trivial to set up..



Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-04-14 Thread William Scott

Does this include the customary grieving period?


On Apr 14, 2008, at 2:10 PM, JOE CRYSTAL wrote:


Hi,


Does anyone have information about how long it takes to set up a 96- 
well
tray for the crystallization robots available?  Besides cost per  
tray and
maintenance cost, another important feature we consider is the time  
for
setting up a 96-well tray.  It is an important factor since we are  
talking

about sub-microliter drops.


Best,


Joe

On Fri, Jan 18, 2008 at 12:28 PM, Lisa A Nagy [EMAIL PROTECTED]  
wrote:



Al's Oil on the plates:
What a nightmare!!!
The oil creeps up the plate and over the sides. It dissolves  
adhesives.

It makes me say bad words in multiple languages.
Bigger drops + no oil = fewer bad words.

Lisa
--
Lisa A. Nagy, Ph.D.
University of Alabama-Birmingham
[EMAIL PROTECTED]

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Patrick Shaw Stewart
Sent: Friday, January 18, 2008 2:20 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

One thing that people often overlook is that quite a lot of protein
can be lost by denaturation on the surface of the drop.  This is more
significant for smaller drops.  Two suggestions: (1) increase the
proportion of protein in the - technical term - teeny drop to say two
thirds and (2) cover the drops with oil eg Al's oils
(silicone/paraffin).  You still get vapor diffusion though the oil ,
and you'd like to slow up equilibration.  of course (2) slows up the
robotics a little, but both should be trivial to set up..



Re: [ccp4bb] crystallisation robot

2008-04-14 Thread Bart Hazes

Hi Joe,

We have a 32-head Honeybee robot which sets up a 96-well plate with a 
single drop per well in ~6 minutes and 3 drops per well at ~9 minutes. A 
96-head phoenix or hummingbird-like system is likely going to be faster, 
but not by that much. Our robot has its own humidified cabinet and as 
long as you keep evaporation under control I don't think set-up speed, 
within reason, is that important.


Bart

JOE CRYSTAL wrote:

Hi,


Does anyone have information about how long it takes to set up a 96-well 
tray for the crystallization robots available?  Besides cost per tray 
and maintenance cost, another important feature we consider is the time 
for setting up a 96-well tray.  It is an important factor since we are 
talking about sub-microliter drops.



Best,


Joe

On Fri, Jan 18, 2008 at 12:28 PM, Lisa A Nagy [EMAIL PROTECTED] 
mailto:[EMAIL PROTECTED] wrote:


Al's Oil on the plates:
What a nightmare!!!
The oil creeps up the plate and over the sides. It dissolves adhesives.
It makes me say bad words in multiple languages.
Bigger drops + no oil = fewer bad words.

Lisa
--
Lisa A. Nagy, Ph.D.
University of Alabama-Birmingham
[EMAIL PROTECTED] mailto:[EMAIL PROTECTED]

-Original Message-
From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK
mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of
Patrick Shaw Stewart
Sent: Friday, January 18, 2008 2:20 AM
To: CCP4BB@JISCMAIL.AC.UK mailto:CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

One thing that people often overlook is that quite a lot of protein
can be lost by denaturation on the surface of the drop.  This is more
significant for smaller drops.  Two suggestions: (1) increase the
proportion of protein in the - technical term - teeny drop to say two
thirds and (2) cover the drops with oil eg Al's oils
(silicone/paraffin).  You still get vapor diffusion though the oil ,
and you'd like to slow up equilibration.  of course (2) slows up the
robotics a little, but both should be trivial to set up..





--

==

Bart Hazes (Assistant Professor)
Dept. of Medical Microbiology  Immunology
University of Alberta
1-15 Medical Sciences Building
Edmonton, Alberta
Canada, T6G 2H7
phone:  1-780-492-0042
fax:1-780-492-7521

==


Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-04-14 Thread Artem Evdokimov
About 2 - 2.5 minutes on Mosquito for single-drop protocol, scaled according
to the number of drops per well (7 minutes for 3-drop trays). About 25
minutes on a Tecan-derived platform (numbers vary greatly depending on the
particular configuration).

 

Artem

 

  _  

From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of JOE
CRYSTAL
Sent: Monday, April 14, 2008 5:10 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

 

Hi,


Does anyone have information about how long it takes to set up a 96-well
tray for the crystallization robots available?  Besides cost per tray and
maintenance cost, another important feature we consider is the time for
setting up a 96-well tray.  It is an important factor since we are talking
about sub-microliter drops.


Best,


Joe

On Fri, Jan 18, 2008 at 12:28 PM, Lisa A Nagy [EMAIL PROTECTED] wrote:

Al's Oil on the plates:
What a nightmare!!!
The oil creeps up the plate and over the sides. It dissolves adhesives.
It makes me say bad words in multiple languages.
Bigger drops + no oil = fewer bad words.

Lisa

--
Lisa A. Nagy, Ph.D.
University of Alabama-Birmingham
[EMAIL PROTECTED]

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of

Patrick Shaw Stewart
Sent: Friday, January 18, 2008 2:20 AM
To: CCP4BB@JISCMAIL.AC.UK

Subject: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

One thing that people often overlook is that quite a lot of protein
can be lost by denaturation on the surface of the drop.  This is more
significant for smaller drops.  Two suggestions: (1) increase the
proportion of protein in the - technical term - teeny drop to say two
thirds and (2) cover the drops with oil eg Al's oils
(silicone/paraffin).  You still get vapor diffusion though the oil ,
and you'd like to slow up equilibration.  of course (2) slows up the
robotics a little, but both should be trivial to set up..

 



[ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-01-18 Thread Patrick Shaw Stewart
One thing that people often overlook is that quite a lot of protein
can be lost by denaturation on the surface of the drop.  This is more
significant for smaller drops.  Two suggestions: (1) increase the
proportion of protein in the - technical term - teeny drop to say two
thirds and (2) cover the drops with oil eg Al's oils
(silicone/paraffin).  You still get vapor diffusion though the oil ,
and you'd like to slow up equilibration.  of course (2) slows up the
robotics a little, but both should be trivial to set up..

On 1/17/08, Oganesyan, Vaheh [EMAIL PROTECTED] wrote:




 Mark,



 What was the state of the larger drops when tiny counterparts had crystals?
 My guess - they all precipitated.

 I'm trying to understand why some proteins or some conditions require change
 in protein concentration while others do not when migrating from smaller
 drops to larger ones. If it is protein dependent then I'm afraid there might
 be no one answer; if it is not then there should be a trend and explanation
 of phenomena.






 Vaheh

  


 From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
 [EMAIL PROTECTED]
  Sent: Wednesday, January 16, 2008 8:31 PM
  To: CCP4BB@JISCMAIL.AC.UK
  Subject: Re: [ccp4bb] crystallisation robot





 Once upon a time I worked in a group that was interested in developing
 crystallization in microfluidics. This was before the time that Fluidigm
 existed and we had not heard of crystallization with the aid of
 microfluidics at the time. We had good reason to try to make a system that
 was as small and light as possible - it had something to do with the cost of
 shipping proteins and precipitants - less was better. And we also wanted all
 protein drops to be fully enclosed, out of safety considerations.

  Like Tassos, we were very worried what would happen if you scaled back
 drops along the lines of this discussion - several uL downto tens of
 nanoliters. If the stochastic process had a major influence over this
 process, we thought that we would never get any crystals. So we set up
 side-by-side experiments at larger volumes and smaller volumes - basically
 scanning several orders of magnitude - expecting a decrease of the number of
 crystals when volumes decrease. To our great surprise the outcome was that
 smaller volumes almost always gave MORE (I almost want to say 'dramatically
 more') crystals, more nucleation, and indeed in various cases the crystals
 grew much faster also. Indeed, it was trivial to observe that the
 surface-to-volume ratio was the primary driver for the nucleation process.
 We had control over geometry to some extent and were able to observe
 surfaces while crystals grow. The crystals would most commonly nucleate on a
 surface.

  So although there probably is something to stochastic aspects, it is clear
 that other aspects can be more important and overrule the stochastic
 considerations.
  The somewhat unpleasant consquence is of course that results acquired in
 very small volumes (with larger surface-to-volume ratio) cannot necessarily
 be repeated in larger volumes (smaller surface-to-volume ratio).

  This is not a flame, even if heat might be a good thing on a night with
 temperatures predicted far below 0F.

   :-)

  Mark







 -Original Message-
  From: Anastassis Perrakis [EMAIL PROTECTED]
  To: CCP4BB@JISCMAIL.AC.UK
  Sent: Wed, 16 Jan 2008 6:17 am
  Subject: Re: [ccp4bb] crystallisation robot


  Oryxnano 50+50 nL
  
   Demetres
  

  Which, indirectly, brings up an interesting (but not relevant to the Oryx)
 question.

  Nucleation is a process that does have a stochastic aspect.

  Thus, one could argue that compromising to 200-300 nl might be better than
 either extremes of 50nl (too small volume and less chance for nucleation) or
 1000 nl (too much sample).

  any comments ? (let the flames begin).

  A.

  PS1
  another interesting issue that has has been hardly touched in these emails
 is the real sample loss: left in wells and not easy to recover, lost because
 of contamination with system liquid, etc ...


  PS2
  I see lots of people with new robots. please do have a look at the
 www.BIOXHIT.org page and if you have a few minutes to assemble a table we
 will be happy to add your specs to our pages. it can be a nice resource and
 it has already enough things and already one response to my last email ;-)
 To make life easier to potential contributors we can provide an Excel sheet
 to fill up with your specs - just ask.

  On Jan 16, 2008, at 12:46, Demetres D. Leonidas wrote:

  
   David Briggs wrote:
   I'll defend the honour of the phoenix... (again)
  
   Bernhard Rupp 100+100 nl
   Dave Briggs (and all users at Univ of Manchester, UK) 100+100nl
   Others..
  
   Only time we have ANY problems is when the nano dispensing tip  gets
 clogged. Often a good wash whilst still on the machine will  clear the
 blockage.
  
   Dave
  
  
  
  
   -- 
   David C

Re: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

2008-01-18 Thread Lisa A Nagy
Al's Oil on the plates:
What a nightmare!!!
The oil creeps up the plate and over the sides. It dissolves adhesives. 
It makes me say bad words in multiple languages. 
Bigger drops + no oil = fewer bad words.

Lisa
--
Lisa A. Nagy, Ph.D.
University of Alabama-Birmingham
[EMAIL PROTECTED]

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Patrick Shaw Stewart
Sent: Friday, January 18, 2008 2:20 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] Fwd: [ccp4bb] crystallisation robot

One thing that people often overlook is that quite a lot of protein
can be lost by denaturation on the surface of the drop.  This is more
significant for smaller drops.  Two suggestions: (1) increase the
proportion of protein in the - technical term - teeny drop to say two
thirds and (2) cover the drops with oil eg Al's oils
(silicone/paraffin).  You still get vapor diffusion though the oil ,
and you'd like to slow up equilibration.  of course (2) slows up the
robotics a little, but both should be trivial to set up..


Re: [ccp4bb] crystallisation robot

2008-01-17 Thread Flip Hoedemaeker
This illustrates nicely that the number of crystals per volume unit
ultimately depends on the number of nucleants per volume unit. If the
surface of your vessel has the power to nucleate it dramatically changes
the effect of going to smaller volumes. I guess everybody is right here,
depending on the vessel they use.

Flip


[EMAIL PROTECTED] schreef:

  Once upon a time I worked in a group that was interested in developing
 crystallization in microfluidics. This was before the time that Fluidigm
 existed and we had not heard of crystallization with the aid of
 microfluidics at the time. We had good reason to try to make a system
 that was as small and light as possible - it had something to do with the
 cost of shipping proteins and precipitants - less was better. And we also
 wanted all protein drops to be fully enclosed, out of safety
 considerations.

 Like Tassos, we were very worried what would happen if you scaled back
 drops along the lines of this discussion - several uL downto tens of
 nanoliters. If the stochastic process had a major influence over this
 process, we thought that we would never get any crystals. So we set up
 side-by-side experiments at larger volumes and smaller volumes - basically
 scanning several orders of magnitude - expecting a decrease of the number
 of crystals when volumes decrease. To our great surprise the outcome was
 that smaller volumes almost always gave MORE (I almost want to say
 'dramatically more') crystals, more nucleation, and indeed in various
 cases the crystals grew much faster also. Indeed, it was trivial to
 observe that the surface-to-volume ratio was the primary driver for the
 nucleation process. We had control over geometry to some extent and were
 able to observe surfaces while crystals grow. The crystals would most
 commonly nucleate on a surface.

 So although there probably is something to stochastic aspects, it is clear
 that other aspects can be more important and overrule the stochastic
 considerations.
 The somewhat unpleasant consquence is of course that results acquired in
 very small volumes (with larger surface-to-volume ratio) cannot
 necessarily be repeated in larger volumes (smaller surface-to-volume
 ratio).

 This is not a flame, even if heat might be a good thing on a night with
 temperatures predicted far below 0F.

 ?:-)

 Mark







 -Original Message-
 From: Anastassis Perrakis [EMAIL PROTECTED]
 To: CCP4BB@JISCMAIL.AC.UK
 Sent: Wed, 16 Jan 2008 6:17 am
 Subject: Re: [ccp4bb] crystallisation robot









 Oryxnano 50+50 nL?

?

 Demetres?

?
 ?

 Which, indirectly, brings up an interesting (but not relevant to the Oryx)
 question.?
 ?

 Nucleation is a process that does have a stochastic aspect.?
 ?

 Thus, one could argue that compromising to 200-300 nl might be better than
 either extremes of 50nl (too small volume and less chance for nucleation)
 or 1000 nl (too much sample).?
 ?

 any comments ? (let the flames begin).?
 ?

 A.?
 ?

 PS1?

 another interesting issue that has has been hardly touched in these emails
 is the real sample loss: left in wells and not easy to recover, lost
 because of contamination with system liquid, etc ...?
 ?

 PS2?

 I see lots of people with new robots. please do have a look at the
 www.BIOXHIT.org page and if you have a few minutes to assemble a table we
 will be happy to add your specs to our pages. it can be a nice resource
 and it has already enough things and already one response to my last email
 ;-) To make life easier to potential contributors we can provide an Excel
 sheet to fill up with your specs - just ask.?
 ?

 On Jan 16, 2008, at 12:46, Demetres D. Leonidas wrote:?
 ?

?

 David Briggs wrote:?

 I'll defend the honour of the phoenix... (again)?

?

 Bernhard Rupp 100+100 nl?

 Dave Briggs (and all users at Univ of Manchester, UK) 100+100nl?

 Others..?

?

 Only time we have ANY problems is when the nano dispensing tip  gets
 clogged. Often a good wash whilst still on the machine will  clear
 the blockage.?

?

 Dave?

?

?

?

?

 -- ?

 David C. Briggs PhD?

 Father  Crystallographer?

 http://www.dbriggs.talktalk.net http://www.dbriggs.talktalk.net?

 AIM ID: dbassophile?

 ?

?

 -- Demetres D. Leonidas, Ph.D.?

 Structural Biology  Chemistry Group?

 Institute of Organic and Pharmaceutical Chemistry?

 The National Hellenic Research Foundation?

 48, Vassileos Constantinou Avenue?

 Athens 116 35, Greece?

 ==?

 Tel. +30 210 7273841 (office)?

 +30 210 7273895 (lab) Fax. +30 210 7273831?

 E-mail: [EMAIL PROTECTED]

 URL: http://athena.eie.gr?

 ==?






 
 More new features than ever.  Check out the new AIM(R) Mail ! -
 http://webmail.aim.com



---
Flip Hoedemaeker Ph.D, CEO
Key Drug Prototyping BV

Re: [ccp4bb] crystallisation robot

2008-01-17 Thread Oganesyan, Vaheh
Mark,

 

What was the state of the larger drops when tiny counterparts had
crystals? My guess - they all precipitated.

I'm trying to understand why some proteins or some conditions require
change in protein concentration while others do not when migrating from
smaller drops to larger ones. If it is protein dependent then I'm afraid
there might be no one answer; if it is not then there should be a trend
and explanation of phenomena.

 

 

Vaheh 



From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
[EMAIL PROTECTED]
Sent: Wednesday, January 16, 2008 8:31 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] crystallisation robot

 

Once upon a time I worked in a group that was interested in developing
crystallization in microfluidics. This was before the time that Fluidigm
existed and we had not heard of crystallization with the aid of
microfluidics at the time. We had good reason to try to make a system
that was as small and light as possible - it had something to do with
the cost of shipping proteins and precipitants - less was better. And we
also wanted all protein drops to be fully enclosed, out of safety
considerations.

Like Tassos, we were very worried what would happen if you scaled back
drops along the lines of this discussion - several uL downto tens of
nanoliters. If the stochastic process had a major influence over this
process, we thought that we would never get any crystals. So we set up
side-by-side experiments at larger volumes and smaller volumes -
basically scanning several orders of magnitude - expecting a decrease of
the number of crystals when volumes decrease. To our great surprise the
outcome was that smaller volumes almost always gave MORE (I almost want
to say 'dramatically more') crystals, more nucleation, and indeed in
various cases the crystals grew much faster also. Indeed, it was trivial
to observe that the surface-to-volume ratio was the primary driver for
the nucleation process. We had control over geometry to some extent and
were able to observe surfaces while crystals grow. The crystals would
most commonly nucleate on a surface. 

So although there probably is something to stochastic aspects, it is
clear that other aspects can be more important and overrule the
stochastic considerations.
The somewhat unpleasant consquence is of course that results acquired in
very small volumes (with larger surface-to-volume ratio) cannot
necessarily be repeated in larger volumes (smaller surface-to-volume
ratio).

This is not a flame, even if heat might be a good thing on a night with
temperatures predicted far below 0F.

 :-)

Mark

 

 

-Original Message-
From: Anastassis Perrakis [EMAIL PROTECTED]
To: CCP4BB@JISCMAIL.AC.UK
Sent: Wed, 16 Jan 2008 6:17 am
Subject: Re: [ccp4bb] crystallisation robot

 Oryxnano 50+50 nL 
 
 Demetres 
 
 
Which, indirectly, brings up an interesting (but not relevant to the
Oryx) question. 
 
Nucleation is a process that does have a stochastic aspect. 
 
Thus, one could argue that compromising to 200-300 nl might be better
than either extremes of 50nl (too small volume and less chance for
nucleation) or 1000 nl (too much sample). 
 
any comments ? (let the flames begin). 
 
A. 
 
PS1 
another interesting issue that has has been hardly touched in these
emails is the real sample loss: left in wells and not easy to recover,
lost because of contamination with system liquid, etc ... 
 
PS2 
I see lots of people with new robots. please do have a look at the
www.BIOXHIT.org page and if you have a few minutes to assemble a table
we will be happy to add your specs to our pages. it can be a nice
resource and it has already enough things and already one response to my
last email ;-) To make life easier to potential contributors we can
provide an Excel sheet to fill up with your specs - just ask. 
 
On Jan 16, 2008, at 12:46, Demetres D. Leonidas wrote: 
 
 
 David Briggs wrote: 
 I'll defend the honour of the phoenix... (again) 
 
 Bernhard Rupp 100+100 nl 
 Dave Briggs (and all users at Univ of Manchester, UK) 100+100nl 
 Others.. 
 
 Only time we have ANY problems is when the nano dispensing tip 
gets clogged. Often a good wash whilst still on the machine will 
clear the blockage. 
 
 Dave 
 
 
 
 
 --  
 David C. Briggs PhD 
 Father  Crystallographer 
 http://www.dbriggs.talktalk.net http://www.dbriggs.talktalk.net 
 AIM ID: dbassophile 
  
 
 -- Demetres D. Leonidas, Ph.D. 
 Structural Biology  Chemistry Group 
 Institute of Organic and Pharmaceutical Chemistry 
 The National Hellenic Research Foundation 
 48, Vassileos Constantinou Avenue 
 Athens 116 35, Greece 
 == 
 Tel. +30 210 7273841 (office) 
 +30 210 7273895 (lab) Fax. +30 210 7273831 
 E-mail: [EMAIL PROTECTED] 
 URL: http://athena.eie.gr 
 == 



size=2 width=100

Re: [ccp4bb] crystallisation robot

2008-01-17 Thread mjvdwoerd

 Vaheh,

I don't recall precipitation at all, but I do remember that we were prepared to 
change the crystallization recipes (i.e. adjust the recipes from previous 
'large volume' crystallization to make more nuclei). For example, our first 
tests were with lysozyme (sorry, hardly a representative protein, I know) and 
it is known (paper by RA Judge et al) that buffer pH is a major determinant in 
nucleation (for that protein). Our initial concern was: is it possible to grow 
crystals in exceedingly small volumes? So we changed the pH to increase the 
nucleation rate. I think I recall that in case of lysozyme the nucleation rate 
is roughly inversely proportional to acid concentration, i.e. increase pH by 1 
unit (10X less acid), gives ~10 times more nuclei. In general (for other 
proteins) we did make some minor adjustments in crystallization, but generally 
only in protein/precipitant concentrations. 'Minor' means adjustments of about 
10% or so - going from 10% precipitant to 9% or 11%.

When you go to very small volumes there is another consideration that you have 
to think about: mixing. If you assume that mixing takes place by diffusion (no 
stirring, it is very difficult to stir a very small volume reliably, i.e. to 
the same extent every time, unless you use microfluidic flow, in which case 
complete mixing to homogeneous mixtures is trivial - note: the Fluidigm system 
is 'microstatic', not microfluidic), so - if you assume mixing to take place by 
diffusion only, then the diffusion length in 100nL drops is very small and you 
can calculate in the worst case scenario how long it will take to accomplish 
full mixing by diffusion only. The time scale will be in the order of seconds. 
If you do the same calculation for a large drop (1 uL), the time scale is MUCH 
longer. So in case of batch crystallization, the 'end point' is reached very 
quickly in small volumes, while it takes much longer in large drops. If the end 
point causes protein precipitation, then this will happen very quickly after 
starting the experiment. In the larger scale experiment it will take much 
longer to reach precipitation and your system may go slowly through a process 
of enabling crystal growth before you reach drastic insolubility that causes 
precipitation. 

So as someone else said, the kinetics are completely different, the end point 
is not, and this can significantly affect the outcome of the experiment.

Mark


 


 

-Original Message-
From: Oganesyan, Vaheh [EMAIL PROTECTED]
To: CCP4BB@JISCMAIL.AC.UK
Sent: Thu, 17 Jan 2008 7:40 am
Subject: Re: [ccp4bb] crystallisation robot



































Mark,



 



What was the state of the larger drops when
tiny counterparts had crystals? My guess - they all precipitated.



I’m trying to understand why some
proteins or some conditions require change in protein concentration while others
do not when migrating from smaller drops to larger ones. If it is protein
dependent then I’m afraid there might be no one answer; if it is not then
there should be a trend and explanation of phenomena.



 



 






Vaheh




















From: CCP4 bulletin
board [mailto:[EMAIL PROTECTED] On
Behalf Of [EMAIL PROTECTED]

Sent: Wednesday, January 16, 2008
8:31 PM

To: CCP4BB@JISCMAIL.AC.UK

Subject: Re: [ccp4bb]
crystallisation robot






 






Once upon a time I worked in a group that was interested in
developing crystallization in microfluidics. This was before the time that
Fluidigm existed and we had not heard of crystallization with the aid of
microfluidics at the time. We had good reason to try to make a system that was
as small and light as possible - it had something to do with the cost of
shipping proteins and precipitants - less was better. And we also wanted all
protein drops to be fully enclosed, out of safety considerations.



Like Tassos, we were very worried what would happen if you scaled back drops
along the lines of this discussion - several uL downto tens of nanoliters. If
the stochastic process had a major influence over this process, we thought that
we would never get any crystals. So we set up side-by-side experiments at
larger volumes and smaller volumes - basically scanning several orders of
magnitude - expecting a decrease of the number of crystals when volumes
decrease. To our great surprise the outcome was that smaller volumes almost
always gave MORE (I almost want to say 'dramatically more') crystals, more
nucleation, and indeed in various cases the crystals grew much faster also.
Indeed, it was trivial to observe that the surface-to-volume ratio was the
primary driver for the nucleation process. We had control over geometry to some
extent and were able to observe surfaces while crystals grow. The crystals
would most commonly nucleate on a surface. 



So although there probably is something to stochastic aspects, it is clear that
other aspects can be more important and overrule the stochastic
considerations.

The somewhat

Re: [ccp4bb] crystallisation robot

2008-01-16 Thread David Briggs
I'll defend the honour of the phoenix... (again)

Bernhard Rupp 100+100 nl
Dave Briggs (and all users at Univ of Manchester, UK) 100+100nl

Others..
Only time we have ANY problems is when the nano dispensing tip gets clogged.
Often a good wash whilst still on the machine will clear the blockage.

Dave




-- 

David C. Briggs PhD
Father  Crystallographer
http://www.dbriggs.talktalk.net
AIM ID: dbassophile



Re: [ccp4bb] crystallisation robot

2008-01-16 Thread Alexey Rak
Dear Bernard,

I do not have long experience using Phoenix. I tried the machine during a 
couple of demonstrations I had chance to participate and also learned about 
other people experience who used the robot (various MPIs and universities in 
Germany and a couple of users from industry in Europe), I hope they can comment 
further. 

The crash of Phoenix needles that dissolved a myth on non-breakable Phoenix 
needles we witnessed during Art Robbins instruments presentation in PSDI 
meeting, 28th - 30th October 2007 Autrans, France. The needles were bended by 
rep during demonstration in the first day of the conference so that the robot 
head was moving without needles for the rest of the conference, it was funny.

For the maintenance, all labs I know those using Phoenix have someone looking 
after the machine. In some lab setups the situation exaggerated even further: 
lab members do not have direct access to the delicate machine and all the 
crystallization by Phoenix is carried out by person who is trained as the User 
and dedicated to. As being quite experienced user of Cartesian nano dispenser 
(1 year), I can understand the problems of cleanness for the robots with not 
dispensable and therefore requiring washing steps needles/tips. As I know, the 
situation is significantly improved in Innovadyne screenmaker, for those who 
prefer flexibilities of 96 channel system over speed, simplicity, no cross 
contamination, accuracy of pipetting high PEGs Mosquito system (4 years 
experience).

The electrostatic effect I mentioned was observed on the plates performed by 
Phoenix, while setting the same plates (Greiner triple, Corning 3550) by 
Mosquito did not cause similar problems (same drops volumes, same screening 
solutions), I hope colleagues may detail further. I would also suggest 
experienced Phoenix users to let us know what drops volume is routinely set in 
their labs.

Bernhard Rupp 100+100 nl
Others??? 
Others..

All the best,
Alexey.

 

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Bernhard Rupp
Sent: Wednesday, January 16, 2008 1:53 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] crystallisation robot

Dear Alexey,

being involved in the development of the 'fixed needles + a few' 
robots and 96-well plates early on, I wonder about your bad experiences. 

You seem to say that the Phoenix requires more maintenance than the Mosquito - 
how long have you had that Phoenix or what model is giving you the trouble?

For the electrostatics issue: it’s the plates that are electrostatic not the 
robot. This is a common lament for plastic plate users, and the Intelliplate 
designed to go with the Phoenix comes in anti-static bags as used in electronic 
component packaging. Other plates have anti-static coating, and there may be 
other tricks as well to avoid electrostatic charge 
(comments anyone?)
  
Also, the limitation of the 200+200 you claim for the Phoenix seems extreme, 
users have set up 200+200 already with the old hydra+1 contraptions, and 
100+100 seems to work reliable with the Phoenix (maybe some users can 
comment?). Could it be that yours is not set up correctly? I am sure ARI will 
be happy to send a European rep by to check.

Best regards, BR 

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Alexey Rak
Sent: Monday, January 14, 2008 9:36 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] crystallisation robot

we have recently tried the new lid mentioned below. It works very well, 
Mosquito makes 3 hanging drops 96 well plate in 3-4 minutes! The film is very 
easy to handle and it is very transparent. The price for the lid is 
significantly reduced: from 13€ before to 4€ now.

Comparing to Cartesian and Phoenix robots the Mosquito is REALLY maintenance 
free machine, especially the third generation insect where many 1st and 2nd 
Mosquito generation users advices have been implicated.

Alexey Rak
Structural Biology, Chemical Sciences
Sanofi-Aventis
Centre de Recherche Paris


Re: [ccp4bb] crystallisation robot

2008-01-16 Thread Demetres D. Leonidas

Oryxnano 50+50 nL

Demetres


David Briggs wrote:

I'll defend the honour of the phoenix... (again)

Bernhard Rupp 100+100 nl
Dave Briggs (and all users at Univ of Manchester, UK) 100+100nl   


Others..

Only time we have ANY problems is when the nano dispensing tip gets 
clogged. Often a good wash whilst still on the machine will clear the 
blockage.


Dave




--

David C. Briggs PhD
Father  Crystallographer
http://www.dbriggs.talktalk.net http://www.dbriggs.talktalk.net
AIM ID: dbassophile
 


--
Demetres D. Leonidas, Ph.D.
Structural Biology  Chemistry Group
Institute of Organic and Pharmaceutical Chemistry
The National Hellenic Research Foundation
48, Vassileos Constantinou Avenue
Athens 116 35, Greece
==
Tel. +30 210 7273841 (office)
+30 210 7273895 (lab) 
Fax. +30 210 7273831

E-mail: [EMAIL PROTECTED]
URL: http://athena.eie.gr
==


Re: [ccp4bb] crystallisation robot

2008-01-16 Thread Santarsiero, Bernard D.
In our original work on a prototype, we used the Cartesian technology. We
were able to dispense 10nL+10nL drops with a large range of viscosities
without difficulty. The main issue was to wash the tips after use to
prevent clogging. That was with a system that could dispense 1nL droplets.
The present systems are sub-nL now. We didn't have a problem with
electrostatics.

More recently, I've looked at all of the crystallization robot vendors.
For single lab users, all of the systems work well. Systems like the Hydra
or Mosquito are less automatic, but provide the basic functions for
crystallization trial setup. For more of a user facility with a large
number of users from several groups, you want more automation to avoid
protocols that can damage the components, like alignment or breakage of
the needles. You also want to consider the total annual cost of
expendables and maintenance.

Bernie Santarsiero

On Wed, January 16, 2008 5:46 am, Demetres D. Leonidas wrote:
 Oryxnano 50+50 nL

 Demetres


 David Briggs wrote:
 I'll defend the honour of the phoenix... (again)
 Bernhard Rupp 100+100 nl
 Dave Briggs (and all users at Univ of Manchester, UK) 100+100nl Others..
 Only time we have ANY problems is when the nano dispensing tip gets
clogged. Often a good wash whilst still on the machine will clear the
blockage.
 Dave
 --
 
 David C. Briggs PhD
 Father  Crystallographer
 http://www.dbriggs.talktalk.net http://www.dbriggs.talktalk.net AIM
ID: dbassophile
 

 --
 Demetres D. Leonidas, Ph.D.
 Structural Biology  Chemistry Group
 Institute of Organic and Pharmaceutical Chemistry
 The National Hellenic Research Foundation
 48, Vassileos Constantinou Avenue
 Athens 116 35, Greece
 ==
 Tel. +30 210 7273841 (office)
  +30 210 7273895 (lab)
 Fax. +30 210 7273831
 E-mail: [EMAIL PROTECTED]
 URL: http://athena.eie.gr
 ==



Re: [ccp4bb] crystallisation robot

2008-01-16 Thread Anastassis Perrakis
More recently, I've looked at all of the crystallization robot  
vendors.
For single lab users, all of the systems work well. Systems like  
the Hydra

or Mosquito are less automatic, but provide the basic functions for
crystallization trial setup. For more of a user facility with a large
number of users from several groups, you want more automation to avoid
protocols that can damage the components, like alignment or  
breakage of

the needles. You also want to consider the total annual cost of
expendables and maintenance.


There are two major kinds of  a user facility with a large number of  
users from several groups


a. One with an operator: then buy what the operator likes ;-)  
basically all machines do work
b. NO operator: buy something that is hard to break and needs minimal  
maintenance


We are in b. and happy with our Mosquito for 4 years or more now,  
although I am sure the

other choices mentioned are also clearly fine.

A.


Re: [ccp4bb] crystallisation robot - volume?

2008-01-16 Thread cdekker
Regarding stochastic processes: to increase the chances of nucleation 
one would like to have protein in the right nucleation zone for a 
certain period of time.


If the drop becomes too small, the ratio evaporation surface/drop 
volume changes. So, the nucleation zone might be reached within the 
drop but either too fast  ('racing through the phase diagram too 
quickly') or in the smaller droplet turbulence caused by evaporation 
might be more pronounced and interfere with nucleation.


At some stage I looked at this myself, although not for tiny volumes 
(0.3, 0.5 and 0.8ul droplets). With smaller volume I saw more and 
faster nucleation, but less crystal growth. There's got to be an 
optimum in terms of drop volume when using vapour diffusion. Any 
comments?


Carien

On 16 Jan 2008, at 13:17, Anastassis Perrakis wrote:


Oryxnano 50+50 nL

Demetres



Which, indirectly, brings up an interesting (but not relevant to the 
Oryx) question.


Nucleation is a process that does have a stochastic aspect.

Thus, one could argue that compromising to 200-300 nl might be better 
than either extremes of 50nl (too small volume and less chance for 
nucleation) or 1000 nl (too much sample).


any comments ? (let the flames begin).

A.

PS1
another interesting issue that has has been hardly touched in these 
emails is the real sample loss: left in wells and not easy to recover, 
lost because of contamination with system liquid, etc ...


PS2
I see lots of people with new robots. please do have a look at the 
www.BIOXHIT.org page and if you have a few minutes to assemble a table 
we will be happy to add your specs to our pages. it can be a nice 
resource and it has already enough things and already one response to 
my last email ;-) To make life easier to potential contributors we can 
provide an Excel sheet to fill up with your specs - just ask.


On Jan 16, 2008, at 12:46, Demetres D. Leonidas wrote:



David Briggs wrote:

I'll defend the honour of the phoenix... (again)

Bernhard Rupp 100+100 nl
Dave Briggs (and all users at Univ of Manchester, UK) 100+100nl
Others..

Only time we have ANY problems is when the nano dispensing tip gets 
clogged. Often a good wash whilst still on the machine will clear 
the blockage.


Dave




--

David C. Briggs PhD
Father  Crystallographer
http://www.dbriggs.talktalk.net http://www.dbriggs.talktalk.net
AIM ID: dbassophile



--
Demetres D. Leonidas, Ph.D.
Structural Biology  Chemistry Group
Institute of Organic and Pharmaceutical Chemistry
The National Hellenic Research Foundation
48, Vassileos Constantinou Avenue
Athens 116 35, Greece
==
Tel. +30 210 7273841 (office)
+30 210 7273895 (lab) Fax. +30 210 7273831
E-mail: [EMAIL PROTECTED]
URL: http://athena.eie.gr
==



The Institute of Cancer Research: Royal Cancer Hospital, a charitable Company 
Limited by Guarantee, Registered in England under Company No. 534147 with its 
Registered Office at 123 Old Brompton Road, London SW7 3RP.

This e-mail message is confidential and for use by the addressee only.  If the 
message is received by anyone other than the addressee, please return the 
message to the sender by replying to it and then delete the message from your 
computer and network.


Re: [ccp4bb] crystallisation robot

2008-01-16 Thread Oganesyan, Vaheh
I have been using Phoenix for more than two years and so far there were
no issues with maintenance. Wash the needles and nano-dispenser before
and after the runs and you are good to go.
Electrostatic effects have been seen in a way of drops being positioned
on the side of the flat bottom plate, but the drops do not climb too
high in my case probably because I'm using fairly large drops (200 +
200 nl).
No matter how robust the system is, there will be always a way to break
it.

Once in a while I get large crystal that can be used for data
collection, but in most of the cases optimization in hanging 1+1 uL
drops is required.
What I found quite difficult to do is to choose the appropriate protein
concentration when moving from 200+200nL to 1+1 uL. Sometimes protein
should be diluted 3-4 times, sometimes it shouldn't. How others are
approaching this issue?

Vaheh

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Anastassis Perrakis
Sent: Wednesday, January 16, 2008 8:23 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] crystallisation robot

 More recently, I've looked at all of the crystallization robot  
 vendors.
 For single lab users, all of the systems work well. Systems like  
 the Hydra
 or Mosquito are less automatic, but provide the basic functions for
 crystallization trial setup. For more of a user facility with a large
 number of users from several groups, you want more automation to avoid
 protocols that can damage the components, like alignment or  
 breakage of
 the needles. You also want to consider the total annual cost of
 expendables and maintenance.

There are two major kinds of  a user facility with a large number of  
users from several groups

a. One with an operator: then buy what the operator likes ;-)  
basically all machines do work
b. NO operator: buy something that is hard to break and needs minimal  
maintenance

We are in b. and happy with our Mosquito for 4 years or more now,  
although I am sure the
other choices mentioned are also clearly fine.

A.


Re: [ccp4bb] crystallisation robot

2008-01-16 Thread Shirley Roberts

Vaheh,
If I'm optimising in 24 well trays after a hit in 96 I generally use 
sitting drop trays (cryschem) as a first step and try different ratios 
of protein:mother liquor first rather than changing the protein 
concentration. I know that sometimes the crystals stick to the plastic 
but I use the hypodermic trick mentioned on the BB a few days ago.


On the subject of drop size - I was at the RAMC meeting where this was 
discussed and several attendees had looked at the issue,  300+300nl or 
400+400nl were mentioned as more optimal drop sizes for obtaining 
crystal hits during screening.See summary on:

http://www.hamptonresearch.com/stuff/RAMC/RAMC2007Notes.aspx

At York we use 150+150nl  routinely for screening using the mosquito. 
I've moved up to 200+200 since the conference!


Shirley Roberts
ps I think I was responsible for the z to s change in the title
begin:vcard
fn:Shirley M.  Roberts
n:Roberts;Shirley M. 
org:York Structural Biology Laboratory
adr:University of York;;Chemistry Department;York;;YO10 5YW;UK
email;internet:[EMAIL PROTECTED]
tel;work:01904 328268
tel;fax:01904 328266
url:http://www.ysbl.york.ac.uk/people/5.htm
version:2.1
end:vcard



Re: [ccp4bb] crystallisation robot

2008-01-16 Thread Lisa A Nagy
You optimize that variable as well in scale-up.
That is the problem with nano-drop screening- sometimes it is near
impossible to scale up the drops. You have to re-screen around the
conditions (response surface methods can conserve experiments) because
the kinetics change so much between nano and big drops.

The smaller they are, the harder it is to scale up, so we compromise
between protein consumption and scalability. I think with most of these
machines, you _can_ tweak it to dispense in the 20-50nl range
accurately. To do that, you have to avoid using an air gap- either by
drawing (and diluting) your protein against a column of water, or by
using some other liquid (oil) between your protein and the dispensing
fluid. Many times these tiny drops yield crystals much faster because
they have more surface area to volume. But once you have them- well-
that's where I've pulled out my hair. Scale-up is hard, and for delicate
proteins (1% crystalline hits), near impossible from the 20 nl drops.
That's the main reason we use bigger drops. 

In the early stages of a project, there is often a great deal of batch
to batch variability in the protein samples. Minimizing this, and
getting consistent polydispersity measurements can go a long way toward
making scale-ups easier.
All this screening is for naught if you need to optimize using a second
batch of protein which behaves differently from the first, or if your
first/only batch is degraded by the time you have hits to optimize. If
you are fortunate to be producing your own protein, you can personally
make sure you are producing consistent batches. The worst thing you can
do is screen on the hearts, then optimize/scaleup with the tails (of the
protein peak in the purification). 

Personally, I think having a liquid handling robot is almost as
important as a crystallization robot.

Lisa

P.S. Confidential to various sales people:
1. If you had visited us, you would have known we were going to buy
another robot. 
2. If you had visited us, you would have known I was in the group.
3. If a member of our group didn't contact you and request information
for purchasing your fabulous robot based on your website/mailings/ads,
then your marketing people (not us) have the problem.
4. Sole source justification. Features and Price. 20K is a significant
difference, and we didn't need the extra features on your machine. Or we
needed features yours didn't have. 
5. Sending snarky email messages about me to my coworkers makes all of
us less inclined to view your company favorably for future purchases. 

--
Lisa A. Nagy, Ph.D.
University of Alabama-Birmingham
[EMAIL PROTECTED]



-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Oganesyan, Vaheh
Sent: Wednesday, January 16, 2008 9:06 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] crystallisation robot
...
Once in a while I get large crystal that can be used for data
collection, but in most of the cases optimization in hanging 1+1 uL
drops is required.
What I found quite difficult to do is to choose the appropriate protein
concentration when moving from 200+200nL to 1+1 uL. Sometimes protein
should be diluted 3-4 times, sometimes it shouldn't. How others are
approaching this issue?

Vaheh


Re: [ccp4bb] crystallisation robot

2008-01-16 Thread mjvdwoerd

 Once upon a time I worked in a group that was interested in developing 
crystallization in microfluidics. This was before the time that Fluidigm 
existed and we had not heard of crystallization with the aid of microfluidics 
at the time. We had good reason to try to make a system that was as small and 
light as possible - it had something to do with the cost of shipping proteins 
and precipitants - less was better. And we also wanted all protein drops to be 
fully enclosed, out of safety considerations.

Like Tassos, we were very worried what would happen if you scaled back drops 
along the lines of this discussion - several uL downto tens of nanoliters. If 
the stochastic process had a major influence over this process, we thought that 
we would never get any crystals. So we set up side-by-side experiments at 
larger volumes and smaller volumes - basically scanning several orders of 
magnitude - expecting a decrease of the number of crystals when volumes 
decrease. To our great surprise the outcome was that smaller volumes almost 
always gave MORE (I almost want to say 'dramatically more') crystals, more 
nucleation, and indeed in various cases the crystals grew much faster also. 
Indeed, it was trivial to observe that the surface-to-volume ratio was the 
primary driver for the nucleation process. We had control over geometry to some 
extent and were able to observe surfaces while crystals grow. The crystals 
would most commonly nucleate on a surface. 

So although there probably is something to stochastic aspects, it is clear that 
other aspects can be more important and overrule the stochastic 
considerations.
The somewhat unpleasant consquence is of course that results acquired in very 
small volumes (with larger surface-to-volume ratio) cannot necessarily be 
repeated in larger volumes (smaller surface-to-volume ratio).

This is not a flame, even if heat might be a good thing on a night with 
temperatures predicted far below 0F.

?:-)

Mark


 


 

-Original Message-
From: Anastassis Perrakis [EMAIL PROTECTED]
To: CCP4BB@JISCMAIL.AC.UK
Sent: Wed, 16 Jan 2008 6:17 am
Subject: Re: [ccp4bb] crystallisation robot









 Oryxnano 50+50 nL?

?

 Demetres?

?
?

Which, indirectly, brings up an interesting (but not relevant to the Oryx) 
question.?
?

Nucleation is a process that does have a stochastic aspect.?
?

Thus, one could argue that compromising to 200-300 nl might be better than 
either extremes of 50nl (too small volume and less chance for nucleation) or 
1000 nl (too much sample).?
?

any comments ? (let the flames begin).?
?

A.?
?

PS1?

another interesting issue that has has been hardly touched in these emails is 
the real sample loss: left in wells and not easy to recover, lost because of 
contamination with system liquid, etc ...?
?

PS2?

I see lots of people with new robots. please do have a look at the 
www.BIOXHIT.org page and if you have a few minutes to assemble a table we will 
be happy to add your specs to our pages. it can be a nice resource and it has 
already enough things and already one response to my last email ;-) To make 
life easier to potential contributors we can provide an Excel sheet to fill up 
with your specs - just ask.?
?

On Jan 16, 2008, at 12:46, Demetres D. Leonidas wrote:?
?

?

 David Briggs wrote:?

 I'll defend the honour of the phoenix... (again)?

?

 Bernhard Rupp 100+100 nl?

 Dave Briggs (and all users at Univ of Manchester, UK) 100+100nl?

 Others..?

?

 Only time we have ANY problems is when the nano dispensing tip  gets 
 clogged. Often a good wash whilst still on the machine will  clear the 
 blockage.?

?

 Dave?

?

?

?

?

 -- ?

 David C. Briggs PhD?

 Father  Crystallographer?

 http://www.dbriggs.talktalk.net http://www.dbriggs.talktalk.net?

 AIM ID: dbassophile?

 ?

?

 -- Demetres D. Leonidas, Ph.D.?

 Structural Biology  Chemistry Group?

 Institute of Organic and Pharmaceutical Chemistry?

 The National Hellenic Research Foundation?

 48, Vassileos Constantinou Avenue?

 Athens 116 35, Greece?

 ==?

 Tel. +30 210 7273841 (office)?

 +30 210 7273895 (lab) Fax. +30 210 7273831?

 E-mail: [EMAIL PROTECTED]

 URL: http://athena.eie.gr?

 ==?



 



More new features than ever.  Check out the new AIM(R) Mail ! - 
http://webmail.aim.com


Re: [ccp4bb] crystallisation robot

2008-01-16 Thread mjvdwoerd

 Once upon a time I worked in a group that was interested in developing 
crystallization in microfluidics. This was before the time that Fluidigm 
existed and we had not heard of crystallization with the aid of microfluidics 
at the time. We had good reason to try to make a system that was as small and 
light as possible - it had something to do with the cost of shipping proteins 
and precipitants - less was better. And we also wanted all protein drops to be 
fully enclosed, out of safety considerations.

Like Tassos, we were very worried what would happen if you scaled back drops 
along the lines of this discussion - several uL downto tens of nanoliters. If 
the stochastic process had a major influence over this process, we thought that 
we would never get any crystals. So we set up side-by-side experiments at 
larger volumes and smaller volumes - basically scanning several orders of 
magnitude - expecting a decrease of the number of crystals when volumes 
decrease. To our great surprise the outcome was that smaller volumes almost 
always gave MORE (I almost want to say 'dramatically more') crystals, more 
nucleation, and indeed in various cases the crystals grew much faster also. 
Indeed, it was trivial to observe that the surface-to-volume ratio was the 
primary driver for the nucleation process. We had control over geometry to some 
extent and were able to observe surfaces while crystals grow. The crystals 
would most commonly nucleate on a surface. 

So although there probably is something to stochastic aspects, it is clear that 
other aspects can be more important and overrule the stochastic 
considerations.
The somewhat unpleasant consquence is of course that results acquired in very 
small volumes (with larger surface-to-volume ratio) cannot necessarily be 
repeated in larger volumes (smaller surface-to-volume ratio).

This is not a flame, even if heat might be a good thing on a night with 
temperatures predicted far below 0F.

?:-)

Mark


 


 

-Original Message-
From: Anastassis Perrakis [EMAIL PROTECTED]
To: CCP4BB@JISCMAIL.AC.UK
Sent: Wed, 16 Jan 2008 6:17 am
Subject: Re: [ccp4bb] crystallisation robot









 Oryxnano 50+50 nL?

?

 Demetres?

?
?

Which, indirectly, brings up an interesting (but not relevant to the Oryx) 
question.?
?

Nucleation is a process that does have a stochastic aspect.?
?

Thus, one could argue that compromising to 200-300 nl might be better than 
either extremes of 50nl (too small volume and less chance for nucleation) or 
1000 nl (too much sample).?
?

any comments ? (let the flames begin).?
?

A.?
?

PS1?

another interesting issue that has has been hardly touched in these emails is 
the real sample loss: left in wells and not easy to recover, lost because of 
contamination with system liquid, etc ...?
?

PS2?

I see lots of people with new robots. please do have a look at the 
www.BIOXHIT.org page and if you have a few minutes to assemble a table we will 
be happy to add your specs to our pages. it can be a nice resource and it has 
already enough things and already one response to my last email ;-) To make 
life easier to potential contributors we can provide an Excel sheet to fill up 
with your specs - just ask.?
?

On Jan 16, 2008, at 12:46, Demetres D. Leonidas wrote:?
?

?

 David Briggs wrote:?

 I'll defend the honour of the phoenix... (again)?

?

 Bernhard Rupp 100+100 nl?

 Dave Briggs (and all users at Univ of Manchester, UK) 100+100nl?

 Others..?

?

 Only time we have ANY problems is when the nano dispensing tip  gets 
 clogged. Often a good wash whilst still on the machine will  clear the 
 blockage.?

?

 Dave?

?

?

?

?

 -- ?

 David C. Briggs PhD?

 Father  Crystallographer?

 http://www.dbriggs.talktalk.net http://www.dbriggs.talktalk.net?

 AIM ID: dbassophile?

 ?

?

 -- Demetres D. Leonidas, Ph.D.?

 Structural Biology  Chemistry Group?

 Institute of Organic and Pharmaceutical Chemistry?

 The National Hellenic Research Foundation?

 48, Vassileos Constantinou Avenue?

 Athens 116 35, Greece?

 ==?

 Tel. +30 210 7273841 (office)?

 +30 210 7273895 (lab) Fax. +30 210 7273831?

 E-mail: [EMAIL PROTECTED]

 URL: http://athena.eie.gr?

 ==?



 



More new features than ever.  Check out the new AIM(R) Mail ! - 
http://webmail.aim.com


Re: [ccp4bb] crystallisation robot

2008-01-15 Thread Bernhard Rupp
Dear Alexey,

being involved in the development of the 'fixed needles + a few' 
robots and 96-well plates early on, I wonder about your bad experiences. 

You seem to say that the Phoenix requires more maintenance than the 
Mosquito - how long have you had that Phoenix or what model is giving 
you the trouble?

For the electrostatics issue: it’s the plates that are electrostatic 
not the robot. This is a common lament for plastic plate users, and 
the Intelliplate designed to go with the Phoenix comes in anti-static 
bags as used in electronic component packaging. Other plates have anti-static 
coating, and there may be other tricks as well to avoid electrostatic charge 
(comments anyone?)
  
Also, the limitation of the 200+200 you claim for the Phoenix seems extreme, 
users have set up 200+200 already with the old hydra+1 contraptions,
and 100+100 seems to work reliable with the Phoenix (maybe some 
users can comment?). Could it be that yours is not set up correctly? I am 
sure ARI will be happy to send a European rep by to check.

Best regards, BR 

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Alexey Rak
Sent: Monday, January 14, 2008 9:36 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] crystallisation robot

we have recently tried the new lid mentioned below. It works very well, 
Mosquito makes 3 hanging drops 96 well plate in 3-4 minutes! The film is very 
easy to handle and it is very transparent. The price for the lid is 
significantly reduced: from 13€ before to 4€ now.

Comparing to Cartesian and Phoenix robots the Mosquito is REALLY maintenance 
free machine, especially the third generation insect where many 1st and 2nd 
Mosquito generation users advices have been implicated.

Alexey Rak
Structural Biology, Chemical Sciences
Sanofi-Aventis
Centre de Recherche Paris


Re: [ccp4bb] crystallisation robot

2008-01-14 Thread Alexey Rak
we have recently tried the new lid mentioned below. It works very well, 
Mosquito makes 3 hanging drops 96 well plate in 3-4 minutes! The film is very 
easy to handle and it is very transparent. The price for the lid is 
significantly reduced: from 13€ before to 4€ now.

Comparing to Cartesian and Phoenix robots the Mosquito is REALLY maintenance 
free machine, especially the third generation insect where many 1st and 2nd 
Mosquito generation users advices have been implicated.

Alexey Rak
Structural Biology, Chemical Sciences
Sanofi-Aventis
Centre de Recherche Paris



-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Shirley Roberts
Sent: Friday, January 11, 2008 7:25 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] crystallisation robot

I sent this personal reply on wednesday to Madhavi Nalam's original question.
In the light of recent comments I'm posting it to the BB to support TTP labtech 
and Mosquito!
Some of the points have been made by others now, excuse the repeats!

We at YSBL have been very happy with our mosquito. We use it to setup sitting 
drops with the MRC Wilden plate, this can be for screening or optimisation (not 
always necessary to go to 24-well format). You can also setup hanging drops, 
this can work better if detergents are involved. A new lid has just been 
released  from TTP that costs £2.50 and takes 3 drops per well but we havn't 
tried this yet. You will need another robot (or multichannel pipette) for 
transfering the screen solutions.We use the hydra from alpha-helix ltd with 
flexible needles so hard to break.
The phoenix robot does both jobs and seems popular. The Douglas robot has been 
mentioned, the Cartesian is the other main robot in use but I don't have any 
personal experience of it, I've heard it is quite high maintenance.
Good luck with your purchase!
Shirley Roberts


Re: [ccp4bb] crystallisation robot

2008-01-11 Thread Shirley Roberts
I sent this personal reply on wednesday to Madhavi Nalam's original 
question.
In the light of recent comments I'm posting it to the BB to support TTP 
labtech and Mosquito!

Some of the points have been made by others now, excuse the repeats!

We at YSBL have been very happy with our mosquito. We use it to setup 
sitting drops with the MRC Wilden plate, this can be for screening or 
optimisation (not always necessary to go to 24-well format). You can 
also setup hanging drops, this can work better if detergents are 
involved. A new lid has just been released  from TTP that costs £2.50 
and takes 3 drops per well but we havn't tried this yet. You will need 
another robot (or multichannel pipette) for transfering the screen 
solutions.We use the hydra from alpha-helix ltd with flexible needles so 
hard to break.
The phoenix robot does both jobs and seems popular. The Douglas robot 
has been mentioned, the Cartesian is the other main robot in use but I 
don't have any personal experience of it, I've heard it is quite high 
maintenance.

Good luck with your purchase!
Shirley Roberts
begin:vcard
fn:Shirley M.  Roberts
n:Roberts;Shirley M. 
org:York Structural Biology Laboratory
adr:University of York;;Chemistry Department;York;;YO10 5YW;UK
email;internet:[EMAIL PROTECTED]
tel;work:01904 328268
tel;fax:01904 328266
url:http://www.ysbl.york.ac.uk/people/5.htm
version:2.1
end:vcard



Re: [ccp4bb] crystallisation robot

2008-01-11 Thread Nalam, Madhavi
Thanks to all who replied to my query. 
Madhavi


Re: [ccp4bb] crystallisation robot

2008-01-11 Thread Anastassis Perrakis

Dear all -

Under the BIOXHIT home page,

http://www.bioxhit.org

you can navigate to Section 1 or HTP Crystallization,
or simply follow the link below:

http://icarus.embl-hamburg.de/bioxhit/bioXHITSection1.jsp

It has quite a lot of data on what people use, available facilities,  
and testing of some popular robots.


If any lab/facility wants to contribute data to that resource please  
get in contact with me.


My best regards, Tassos

On 11 Jan 2008, at 19:25, Shirley Roberts wrote:

I sent this personal reply on wednesday to Madhavi Nalam's original  
question.
In the light of recent comments I'm posting it to the BB to support  
TTP labtech and Mosquito!

Some of the points have been made by others now, excuse the repeats!

We at YSBL have been very happy with our mosquito. We use it to  
setup sitting drops with the MRC Wilden plate, this can be for  
screening or optimisation (not always necessary to go to 24-well  
format). You can also setup hanging drops, this can work better if  
detergents are involved. A new lid has just been released  from TTP  
that costs £2.50 and takes 3 drops per well but we havn't tried  
this yet. You will need another robot (or multichannel pipette) for  
transfering the screen solutions.We use the hydra from alpha-helix  
ltd with flexible needles so hard to break.
The phoenix robot does both jobs and seems popular. The Douglas  
robot has been mentioned, the Cartesian is the other main robot in  
use but I don't have any personal experience of it, I've heard it  
is quite high maintenance.

Good luck with your purchase!
Shirley Robertsroberts.vcf