Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-07 Thread James Holton


Jacob,

The main reason why it is not common practice to saturate every crystal 
with every heavy metal under the sun is radiation damage.  X-ray 
absorption increases very rapidly with atomic number (third power), so 
on the order of 100 mM of heavy atom is usually enough to cut your 
crystal's useful life in half.  It doesn't matter if the heavy atom is 
in the lattice or not.  At 12 keV the photoelectron cascade crashes 
through about 3 microns worth of organic matter before finally coming to 
rest (Cole, 1969).


The program RADDOSE (Murray et al. 2004, 2005, Paithankar et al 2009, 
2010) was written to calculate expected crystal lifetime given your 
buffer concentrations, etc.  For example, with a 50% solvent crystal, 
250 mM iodide in the solvent channels is dose doubling (life 
halving).  That is, the number of photons you will get scattered into 
spots before the crystal is dead will be half that which you would get 
with no iodine.  This is because the iodine-soaked crystal is absorbing 
twice as much energy per incident photon, but scattering at pretty much 
the same rate with or without the iodine.  500 mM iodine will cut the 
useful life to 1/3 of what it would be with no iodine, and 1 M will 
cut it to 1/5th.  (250/([iodine]+250)).  For Cs, 200 mM is the 
dose-doubling buffer concentration.  For Rb it is 1 M.  For Cl, it is 
2.4 M, and for fluoride there is no concentration that doubles the dose 
because water absorbs more than fouoride.  This is why I recommend using 
low-Z buffers in crystallography whenever possible.


I have listed dose-doubling concentrations of a few common elements on 
the last page of this document:

http://bl831.als.lbl.gov/damage_rates.pdf
and a fairly comprehensive look up table is here:
http://bl831.als.lbl.gov/~jamesh/pickup/dose_doubler_solc50.txt

The dose doubling concentration does depend on the wavelength.  The 
worst one at 1 A is tungsten (57 mM), but this is not all that 
different from tantalum (62 mM) or mercury (74 mM), or gold (76 mM), or 
uranium (105 mM).  So, by and large ~100 mM concentration of anything 
heavy will double the dose per scattered photon.  Note that this is 
the concentration of the atoms, not molecules.  The dose-doubling 
concentration of Ta6Br12 clusters is 9.5 mM at 1 A.


As for the original poster's question, 20 mM iodide disrupting your 
crystallization is actually a good sign that it's binding, but the bad 
news is that it appears your crystals don't like that.  Nevertheless it 
could be worth a shot, you can estimate your expected Bijvoet ratio 
assuming one site per molecule using this little web app I wrote:

http://bl831.als.lbl.gov/xtalsize.html
 In general, a Bijvoet ratio of 3% or so is needed to solve a structure 
(the current world record is 0.5% and lots of multiplicity).  The above 
web page will also tell you how many crystals you need if you type in 
their size in all three dimensions. but this estimate assumes that you 
don't have high concentrations of heavy metals in your solution!  So, if 
it says you can get away with one crystal but you know your have a 
dose-doubling concentration of something, then you're going to need to 
average data from two crystals, etc.


Good luck!

-James Holton
MAD Scientist

On 5/3/2012 8:29 AM, Jacob Keller wrote:
I have wondered for a long time now why it is not standard practice 
for all crystallization protein stocks to contain either Br- or I- 
ions instead of Cl-, even for cationic buffers like TRIS, which could 
be titrated with HBr or HI to get in the 10+ mM range. Also, one could 
use Cs or Rb for the cations (and titrate anionic buffers with the 
respective hydroxides). What's there to lose? The gain is obviously 
the possible anomalous signal (always helpful), and one might pick up 
additional interesting and possibly physiologically-relevant halide or 
alkali metal sites. Seems that structural genomics people might 
standardize this into the pipeline as well, and thereby potentially 
cut out SeMet protein production in many if not most cases.


JPK

On Thu, May 3, 2012 at 9:05 AM, Jan Abendroth jan.abendr...@gmail.com 
mailto:jan.abendr...@gmail.com wrote:


Hi Rajesh,
it can be a bit all over the place:
For quick soaks, we typically use 500mM-1000mM. A good starting
point might be to simply replace the NaCl concentration in your
protein buffer. By some serendipity we also managed to solve a
structure by I/S SAD after a 1mM NaI soak. One iodide had found
its way into a nice binding pocket.
For co-crystallization, mostly 200mM should be fine.
Another approach could be to supplement your cryo buffer with
iodide, replacing NaCl. NaI is highly soluble in ethylene glycol.

Also see here: http://www.ncbi.nlm.nih.gov/pubmed/21359836

Good luck!
Cheers,
Jan
--
Jan Abendroth
Emerald Bio
Seattle / Bainbridge Island WA, USA
home: Jan.Abendroth_at_gmail.com http://Jan.Abendroth_at_gmail.com
work: 

Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-07 Thread Pete Meyer

James Holton wrote:


 In general, a Bijvoet ratio of 3% or so is needed to solve a structure (the 
current world record is 0.5% and lots of multiplicity).  The above web page 
will also tell you how many crystals you need if you type in their size in all 
three dimensions. but this estimate assumes that you don't have high 
concentrations of heavy metals in your solution!  So, if it says you can get 
away with one crystal but you know your have a dose-doubling concentration of 
something, then you're going to need to average data from two crystals, etc.



I don't know of any systematic studies, but in my experience if you've 
got weak anomalous signal you're better off with multi-crystal phasing 
than multi-crystal merging.


Pete


[ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Rajesh Kumar

Dear All,
I have very thin crystals but diffracting. I was not able to handle them easily 
for iodide soak. I always lost the crystals during manipulation and other big 
crystals obtained after seeding doesn't even give any diffraction. I tried for 
co-crystallizing with NaI. The crystals appear only up to 20 mM in 1:2 (3ul 
drop of 1 ul protein and 2ul reservoir).Is this concentration of iodide is 
enough for SAD data  ( if it had good incorporation) ?I appreciate your help.
ThanksRajesh  

Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Jan Abendroth
Hi Rajesh,
it can be a bit all over the place:
For quick soaks, we typically use 500mM-1000mM. A good starting point might be 
to simply replace the NaCl concentration in your protein buffer. By some 
serendipity we also managed to solve a structure by I/S SAD after a 1mM NaI 
soak. One iodide had found its way into a nice binding pocket. 
For co-crystallization, mostly 200mM should be fine.
Another approach could be to supplement your cryo buffer with iodide, replacing 
NaCl. NaI is highly soluble in ethylene glycol.

Also see here: http://www.ncbi.nlm.nih.gov/pubmed/21359836

Good luck!
Cheers,
Jan
--
Jan Abendroth
Emerald Bio
Seattle / Bainbridge Island WA, USA
home: Jan.Abendroth_at_gmail.com
work: JAbendroth_at_embios.com
http://www.emeraldbiostructures.com

On May 3, 2012, at 6:51 AM, Rajesh Kumar wrote:

 Dear All,
 
 I have very thin crystals but diffracting. I was not able to handle them 
 easily for iodide soak. I always lost the crystals during manipulation and 
 other big crystals obtained after seeding doesn't even give any diffraction. 
 I tried for co-crystallizing with NaI. The crystals appear only up to 20 mM 
 in 1:2 (3ul drop of 1 ul protein and 2ul reservoir).
 Is this concentration of iodide is enough for SAD data  ( if it had good 
 incorporation) ?
 I appreciate your help.
 
 Thanks
 Rajesh



Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Jim Pflugrath
Iodide is a fantastic derivative.  One does not need a lot with modern X-ray 
equipment, careful data collection, and great software.


From: CCP4 bulletin board [CCP4BB@JISCMAIL.AC.UK] on behalf of Rajesh Kumar 
[ccp4...@hotmail.com]
Sent: Thursday, May 03, 2012 8:51 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] effective iodide conc. for SAD data

Dear All,

I have very thin crystals but diffracting. I was not able to handle them easily 
for iodide soak. I always lost the crystals during manipulation and other big 
crystals obtained after seeding doesn't even give any diffraction. I tried for 
co-crystallizing with NaI. The crystals appear only up to 20 mM in 1:2 (3ul 
drop of 1 ul protein and 2ul reservoir).
Is this concentration of iodide is enough for SAD data  ( if it had good 
incorporation) ?
I appreciate your help.

Thanks
Rajesh


Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Roger Rowlett

Rajesh,

Why not try a soak-in-place by adding iodide and/or cryoprotectant 
directly to the crystallization drop? Then you only have to fish out the 
crystal once, minimizing handling. I usually do this by preparing an 
artificial mother liquor solution that has 125% of the final desired 
concentration of cryoprotectant and/or ligand. Add 4 volumes of this 
solution to your crystallization drop all at once or in stages. (I 
usually add 0.25, 0.25, 0.5, 1.0, and 2.0 drop volumes a few minutes 
apart). The drop can be incubated over its reservoir well during this 
procedure between additions to minimize evaporation. This often works 
well for fragile crystals or crystals that are sensitive to evaporation 
of the drop.


Cheers,

___
Roger S. Rowlett
Gordon  Dorothy Kline Professor
Department of Chemistry
Colgate University
13 Oak Drive
Hamilton, NY 13346

tel: (315)-228-7245
ofc: (315)-228-7395
fax: (315)-228-7935
email: rrowl...@colgate.edu

On 5/3/2012 9:51 AM, Rajesh Kumar wrote:

Dear All,

I have very thin crystals but diffracting. I was not able to handle 
them easily for iodide soak. I always lost the crystals during 
manipulation and other big crystals obtained after 
seeding doesn't even give any diffraction. I tried for 
co-crystallizing with NaI. The crystals appear only up to 20 mM in 1:2 
(3ul drop of 1 ul protein and 2ul reservoir).
Is this concentration of iodide is enough for SAD data ( if it had 
good incorporation) ?

I appreciate your help.

Thanks
Rajesh


Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Jacob Keller
I have wondered for a long time now why it is not standard practice for all
crystallization protein stocks to contain either Br- or I- ions instead of
Cl-, even for cationic buffers like TRIS, which could be titrated with HBr
or HI to get in the 10+ mM range. Also, one could use Cs or Rb for the
cations (and titrate anionic buffers with the respective hydroxides).
What's there to lose? The gain is obviously the possible anomalous signal
(always helpful), and one might pick up additional interesting and possibly
physiologically-relevant halide or alkali metal sites. Seems that
structural genomics people might standardize this into the pipeline as
well, and thereby potentially cut out SeMet protein production in many if
not most cases.

JPK

On Thu, May 3, 2012 at 9:05 AM, Jan Abendroth jan.abendr...@gmail.comwrote:

 Hi Rajesh,
 it can be a bit all over the place:
 For quick soaks, we typically use 500mM-1000mM. A good starting point
 might be to simply replace the NaCl concentration in your protein buffer.
 By some serendipity we also managed to solve a structure by I/S SAD after a
 1mM NaI soak. One iodide had found its way into a nice binding pocket.
 For co-crystallization, mostly 200mM should be fine.
 Another approach could be to supplement your cryo buffer with iodide,
 replacing NaCl. NaI is highly soluble in ethylene glycol.

 Also see here: http://www.ncbi.nlm.nih.gov/pubmed/21359836

 Good luck!
 Cheers,
 Jan
  --
 Jan Abendroth
 Emerald Bio
 Seattle / Bainbridge Island WA, USA
 home: Jan.Abendroth_at_gmail.com
 work: JAbendroth_at_embios.com
 http://www.emeraldbiostructures.com

 On May 3, 2012, at 6:51 AM, Rajesh Kumar wrote:

 Dear All,

 I have very thin crystals but diffracting. I was not able to handle them
 easily for iodide soak. I always lost the crystals during manipulation
 and other big crystals obtained after seeding doesn't even give any
 diffraction. I tried for co-crystallizing with NaI. The crystals appear
 only up to 20 mM in 1:2 (3ul drop of 1 ul protein and 2ul reservoir).
 Is this concentration of iodide is enough for SAD data  ( if it had good
 incorporation) ?
 I appreciate your help.

 Thanks
 Rajesh





-- 
***
Jacob Pearson Keller
Northwestern University
Medical Scientist Training Program
email: j-kell...@northwestern.edu
***


Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Roger Rowlett
Interesting idea. The only caveat that springs to mind is that the more 
useful anions (e.g., iodide and bromide) are on the chaotropic end of 
the Hofmeister series and may potentially destabilize protein structure 
or protein-protein interactions, which might complicate 
co-crystallization starting from known conditions, especially at higher 
concentrations of anion. Alternate cations may be less problematic.


Cheers,

___
Roger S. Rowlett
Gordon  Dorothy Kline Professor
Department of Chemistry
Colgate University
13 Oak Drive
Hamilton, NY 13346

tel: (315)-228-7245
ofc: (315)-228-7395
fax: (315)-228-7935
email: rrowl...@colgate.edu

On 5/3/2012 11:29 AM, Jacob Keller wrote:
I have wondered for a long time now why it is not standard practice 
for all crystallization protein stocks to contain either Br- or I- 
ions instead of Cl-, even for cationic buffers like TRIS, which could 
be titrated with HBr or HI to get in the 10+ mM range. Also, one could 
use Cs or Rb for the cations (and titrate anionic buffers with the 
respective hydroxides). What's there to lose? The gain is obviously 
the possible anomalous signal (always helpful), and one might pick up 
additional interesting and possibly physiologically-relevant halide or 
alkali metal sites. Seems that structural genomics people might 
standardize this into the pipeline as well, and thereby potentially 
cut out SeMet protein production in many if not most cases.


JPK

On Thu, May 3, 2012 at 9:05 AM, Jan Abendroth jan.abendr...@gmail.com 
mailto:jan.abendr...@gmail.com wrote:


Hi Rajesh,
it can be a bit all over the place:
For quick soaks, we typically use 500mM-1000mM. A good starting
point might be to simply replace the NaCl concentration in your
protein buffer. By some serendipity we also managed to solve a
structure by I/S SAD after a 1mM NaI soak. One iodide had found
its way into a nice binding pocket.
For co-crystallization, mostly 200mM should be fine.
Another approach could be to supplement your cryo buffer with
iodide, replacing NaCl. NaI is highly soluble in ethylene glycol.

Also see here: http://www.ncbi.nlm.nih.gov/pubmed/21359836

Good luck!
Cheers,
Jan
--
Jan Abendroth
Emerald Bio
Seattle / Bainbridge Island WA, USA
home: Jan.Abendroth_at_gmail.com http://Jan.Abendroth_at_gmail.com
work: JAbendroth_at_embios.com
http://www.emeraldbiostructures.com

On May 3, 2012, at 6:51 AM, Rajesh Kumar wrote:


Dear All,

I have very thin crystals but diffracting. I was not able to
handle them easily for iodide soak. I always lost the crystals
during manipulation and other big crystals obtained after
seeding doesn't even give any diffraction. I tried for
co-crystallizing with NaI. The crystals appear only up to 20 mM
in 1:2 (3ul drop of 1 ul protein and 2ul reservoir).
Is this concentration of iodide is enough for SAD data ( if it
had good incorporation) ?
I appreciate your help.

Thanks
Rajesh





--
***
Jacob Pearson Keller
Northwestern University
Medical Scientist Training Program
email: j-kell...@northwestern.edu mailto:j-kell...@northwestern.edu
***


Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Jim Pflugrath
Please see my poster at the ACA 2012 meeting.  See also:
(1) Dauter, Z., Dauter, M.  Rajashankar, K.R. (2000) Acta Cryst. D56, 232-237.
(2) Nagem, R.A.P, Dauter, Z.  Polikarpov, I. (2001) Acta Cryst. D57, 996-1002.

:)


From: CCP4 bulletin board [CCP4BB@JISCMAIL.AC.UK] on behalf of Roger Rowlett 
[rrowl...@colgate.edu]
Sent: Thursday, May 03, 2012 11:00 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] effective iodide conc. for SAD data

Interesting idea. The only caveat that springs to mind is that the more useful 
anions (e.g., iodide and bromide) are on the chaotropic end of the Hofmeister 
series and may potentially destabilize protein structure or protein-protein 
interactions, which might complicate co-crystallization starting from known 
conditions, especially at higher concentrations of anion. Alternate cations may 
be less problematic.

Cheers,

___
Roger S. Rowlett
Gordon  Dorothy Kline Professor
Department of Chemistry
Colgate University
13 Oak Drive
Hamilton, NY 13346

tel: (315)-228-7245
ofc: (315)-228-7395
fax: (315)-228-7935
email: rrowl...@colgate.edumailto:rrowl...@colgate.edu

On 5/3/2012 11:29 AM, Jacob Keller wrote:
I have wondered for a long time now why it is not standard practice for all 
crystallization protein stocks to contain either Br- or I- ions instead of Cl-, 
even for cationic buffers like TRIS, which could be titrated with HBr or HI to 
get in the 10+ mM range. Also, one could use Cs or Rb for the cations (and 
titrate anionic buffers with the respective hydroxides). What's there to lose? 
The gain is obviously the possible anomalous signal (always helpful), and one 
might pick up additional interesting and possibly physiologically-relevant 
halide or alkali metal sites. Seems that structural genomics people might 
standardize this into the pipeline as well, and thereby potentially cut out 
SeMet protein production in many if not most cases.

JPK

On Thu, May 3, 2012 at 9:05 AM, Jan Abendroth 
jan.abendr...@gmail.commailto:jan.abendr...@gmail.com wrote:
Hi Rajesh,
it can be a bit all over the place:
For quick soaks, we typically use 500mM-1000mM. A good starting point might be 
to simply replace the NaCl concentration in your protein buffer. By some 
serendipity we also managed to solve a structure by I/S SAD after a 1mM NaI 
soak. One iodide had found its way into a nice binding pocket.
For co-crystallization, mostly 200mM should be fine.
Another approach could be to supplement your cryo buffer with iodide, replacing 
NaCl. NaI is highly soluble in ethylene glycol.

Also see here: http://www.ncbi.nlm.nih.gov/pubmed/21359836

Good luck!
Cheers,
Jan
--
Jan Abendroth
Emerald Bio
Seattle / Bainbridge Island WA, USA
home: Jan.Abendroth_at_gmail.comhttp://Jan.Abendroth_at_gmail.com
work: JAbendroth_at_embios.com
http://www.emeraldbiostructures.com

On May 3, 2012, at 6:51 AM, Rajesh Kumar wrote:

Dear All,

I have very thin crystals but diffracting. I was not able to handle them easily 
for iodide soak. I always lost the crystals during manipulation and other big 
crystals obtained after seeding doesn't even give any diffraction. I tried for 
co-crystallizing with NaI. The crystals appear only up to 20 mM in 1:2 (3ul 
drop of 1 ul protein and 2ul reservoir).
Is this concentration of iodide is enough for SAD data  ( if it had good 
incorporation) ?
I appreciate your help.

Thanks
Rajesh




--
***
Jacob Pearson Keller
Northwestern University
Medical Scientist Training Program
email: j-kell...@northwestern.edumailto:j-kell...@northwestern.edu
***


Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Florian Sauer
Dear Rajesh,

another method to incorporate Iodine into your crystal is by simply
placing a drop of KI/I2 solution next to the crystallization drop.

Have a look here:

Acta Cryst. (2006). D62, 280-289
New methods to prepare iodinated derivatives by vaporizing iodine
labelling (VIL) and hydrogen peroxide VIL (HYPER-VIL)

Worked for me with salt concentration sensitive crystal and a
derivatization time 10 min.

Good luck,

Florian




Am 03.05.12 15:51, schrieb Rajesh Kumar:
 Dear All,

 I have very thin crystals but diffracting. I was not able to handle
 them easily for iodide soak. I always lost the crystals during
 manipulation and other big crystals obtained after
 seeding doesn't even give any diffraction. I tried for
 co-crystallizing with NaI. The crystals appear only up to 20 mM in 1:2
 (3ul drop of 1 ul protein and 2ul reservoir).
 Is this concentration of iodide is enough for SAD data  ( if it had
 good incorporation) ?
 I appreciate your help.

 Thanks
 Rajesh



Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Edward A. Berry

Did you locate the Iodine(s)? Did you have iodotyrosine, or I^- bound
at anion binding sites? There are two distinct methods based on I2/I-.

Florian Sauer wrote:

  Dear Rajesh,

another method to incorporate Iodine into your crystal is by simply
placing a drop of KI/I2 solution next to the crystallization drop.

Have a look here:

Acta Cryst. (2006). D62, 280-289
New methods to prepare iodinated derivatives by vaporizing iodine
labelling (VIL) and hydrogen peroxide VIL (HYPER-VIL)

Worked for me with salt concentration sensitive crystal and a
derivatization time 10 min.

Good luck,

Florian




Am 03.05.12 15:51, schrieb Rajesh Kumar:

Dear All,

I have very thin crystals but diffracting. I was not able to handle
them easily for iodide soak. I always lost the crystals during
manipulation and other big crystals obtained after seeding doesn't
even give any diffraction. I tried for co-crystallizing with NaI. The
crystals appear only up to 20 mM in 1:2 (3ul drop of 1 ul protein and
2ul reservoir).
Is this concentration of iodide is enough for SAD data ( if it had
good incorporation) ?
I appreciate your help.

Thanks
Rajesh




Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Florian Sauer
Yes I did and could solve the structure by SAD.

In total 8 I atoms found. 6 bound at anion binding sites. 2 as I2.
None covalently bound to tyrosine.


Florian

Am 03.05.12 19:42, schrieb Edward A. Berry:
 Did you locate the Iodine(s)? Did you have iodotyrosine, or I^- bound
 at anion binding sites? There are two distinct methods based on I2/I-.

 Florian Sauer wrote:
 Dear Rajesh,

 another method to incorporate Iodine into your crystal is by simply
 placing a drop of KI/I2 solution next to the crystallization drop.

 Have a look here:

 Acta Cryst. (2006). D62, 280-289
 New methods to prepare iodinated derivatives by vaporizing iodine
 labelling (VIL) and hydrogen peroxide VIL (HYPER-VIL)

 Worked for me with salt concentration sensitive crystal and a
 derivatization time 10 min.

 Good luck,

 Florian




 Am 03.05.12 15:51, schrieb Rajesh Kumar:
 Dear All,

 I have very thin crystals but diffracting. I was not able to handle
 them easily for iodide soak. I always lost the crystals during
 manipulation and other big crystals obtained after seeding doesn't
 even give any diffraction. I tried for co-crystallizing with NaI. The
 crystals appear only up to 20 mM in 1:2 (3ul drop of 1 ul protein and
 2ul reservoir).
 Is this concentration of iodide is enough for SAD data ( if it had
 good incorporation) ?
 I appreciate your help.

 Thanks
 Rajesh






Re: [ccp4bb] effective iodide conc. for SAD data

2012-05-03 Thread Rajesh Kumar

Dear All,

I really thank everyone who gave me excellent suggestions.
I will try them to see which one works better for me and my crystals.
Again thanks a lot for your help.

Regards,
Rajesh

Date: Thu, 3 May 2012 20:56:21 +0200
From: florian.sa...@embl-hamburg.de
Subject: Re: [ccp4bb] effective iodide conc. for SAD data
To: CCP4BB@JISCMAIL.AC.UK


  

  
  
Yes I did and could solve the structure by SAD.



In total 8 I atoms found. 6 bound at anion binding sites. 2 as I2.

None covalently bound to tyrosine.





Florian



Am 03.05.12 19:42, schrieb Edward A. Berry:

 Did you locate the Iodine(s)?
  Did you have iodotyrosine, or I^- bound

   at anion binding sites? There are two distinct methods based
  on I2/I-.

  

   Florian Sauer wrote:

   Dear Rajesh,

  

   another method to incorporate Iodine into your crystal is
  by simply

   placing a drop of KI/I2 solution next to the
  crystallization drop.

  

   Have a look here:

  

   Acta Cryst. (2006). D62, 280-289

   New methods to prepare iodinated derivatives by
  vaporizing iodine

   labelling (VIL) and hydrogen peroxide VIL (HYPER-VIL)

  

   Worked for me with salt concentration sensitive crystal
  and a

   derivatization time 10 min.

  

   Good luck,

  

   Florian

  

  

  

  

   Am 03.05.12 15:51, schrieb Rajesh Kumar:

   Dear All,

  

   I have very thin crystals but diffracting. I was not
  able to handle

   them easily for iodide soak. I always lost the
  crystals during

   manipulation and other big crystals obtained after
  seeding doesn't

   even give any diffraction. I tried for
  co-crystallizing with NaI. The

   crystals appear only up to 20 mM in 1:2 (3ul drop of
  1 ul protein and

   2ul reservoir).

   Is this concentration of iodide is enough for SAD
  data ( if it had

   good incorporation) ?

   I appreciate your help.

  

   Thanks

   Rajesh