Re: Production Line Test Requirements - Medical Devices

2001-05-21 Thread Rich Nute





Hi John:


   Because continuity at low current does not ensure that the protective
   circuit will carry a large fault current - it might be 'hanging on by
   one strand'.

Yes, for one strand.  No, for five strands.

Some years ago, I did some experiments on what
problems the 25-amp test would detect.  I 
simulated broken strands by cutting them one
at a time.  With five strands intact, the circuit 
passed the 25-amp, 2-minute test.  It failed at 
4 strands and 1 minute.

(The tested wire was 18 AWG comprised of 36 
strands of 34 AWG.)

The ability of a few strands to carry the 25-amp
current depends on the free length of the 
those few strands, which in turn determines the
heat-sinking provided to those strands.  The 
free length was on the order of 3 mm.  

My experiment assumed the problem was caused by 
an incorrectly set wire stripper, that cut a 
number of strands.  So there was a very small 
free length of strands.

I published this study in the Product Safety
Newsletter, Vol. 10, No.1, January-March, 1997.


Best regards,
Rich





---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Environmental test labs

2001-05-21 Thread David Heald

Greetings all,
  I am looking for recommendations for a NEBS environmental test lab
that has an operational temp/humidity chamber of significant size and
with tremendous cooling capacity (8kW or more).  I have a large
(6+'wide, 7+'tall) telecom product that consumes about 8kW of energy
during operation and I am having trouble finding environmental labs with
chambers that can handle this without a liquid nitrogen feed.  I am
based in the Boston area and am talking to a few local labs, but my next
product will be bigger and use more power.  I am willing to expand the
search outside of the Boston area since no one I know will likely be
able to handle it.
  All you lab sales guys out there - I am actually asking for you to
contact me on this one (OFF LIST, private email, of course).  

Thanks in advance,

Dave Heald

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




RE: Cooking induction apparatus and FCC

2001-05-21 Thread Gary McInturff

Sorry, once again ham fisted typing lead to C64.5 rather than the proper
C63.4 - my error hope it didn't confuse to many folks. Thanks for
straightening that out.
Gary

-Original Message-
From: rehel...@mmm.com [mailto:rehel...@mmm.com]
Sent: Monday, May 21, 2001 12:39 PM
To: Gary McInturff
Cc: emc-p...@majordomo.ieee.org
Subject: RE: Cooking induction apparatus and FCC



Part 18 still uses MP-5 but it is not mandatory. In cases of dispute, MP-5
will be used by the FCC. You can't go wrong, however,  using ANSI C63.4 as
the test set-up. Or you can use the CISPR limits as Gary described. I do
not know what C64.5 is.

You do not have to submit your test results through a TCB.


Bob Heller
3M Product Safety, 76-1-01
St. Paul, MN 55107-1208
Tel:  651- 778-6336
Fax:  651-778-6252





 

Gary McInturff

Gary.McInturff@worldwidepa  To: 'Pierre
SELVA' pierre.se...@worldonline.fr
ckets.com Forum Safety-emc
emc-p...@majordomo.ieee.org  
 cc: (bcc: Robert E.
Heller/US-Corporate/3M/US)
05/21/2001 10:59 AM  Subject: RE:
Cooking induction apparatus and FCC  
Please respond to Gary

McInturff

 

 






Bon Jour (That's about it for my French)
Pierre, MP-5 isn't relevant any more. I believe you  want ANSI C64.5.
and I think you'll find mostly similarities. I can't think of any
dissimilarities of the top of my head, Even FCC part 15 allows use of the
CISPR limits for radiated and conducted emissions. The only caveat (ah,
more
French!) is that if you accept the radiated limits during test you also
have
to accept the conducted limits. Some folks trip up on that one. It means
that the conducted limits are really 150Khz even at US line voltage, not
the
normal FCC start of 450 Khz.
Gary

-Original

-Original Message-
From: Pierre SELVA [mailto:pierre.se...@worldonline.fr]
Sent: Monday, May 21, 2001 6:09 AM
To: Forum Safety-emc
Cc: Pierre SELVA
Subject: Cooking induction apparatus and FCC


Dear colleagues,

I need to know your opinion on the following :
I have to perform EMC testing on an induction cooking table and I would
like
to obtain the FCC certification.

In Europe, this kind of apparatus is subjected to the EN55011 (CISPR11) and
the test conditions are clearly stated, mainly for the EUT configuration.

I understand for the USA, the product has to be tested according to Part 18
requirements. This part explains that the technical requirements for the
measurement are described in the MP-5 document. As this one is a little bit
old (1986), do you believe I can use the CISPR11 to demonstrate compliance,
or should I use the MP5 ?

If I have to use the MP-5, do you know if somewhere the EUT configuration
is
described, or is the worst case at the manufacturer discretion ?

And, at last, do you know if I have to submit my test results and EUT
description to FCC directly (electronic submittal), or should I have to go
to a TCB ?

In advance, I thank you a lot for your answers,

Best regards from France


Pierre SELVA
2 route de la Grobelle
73000 JACOB BELLECOMBETTE - France
Tel : 33 (0)6 60 52 04 96
Fax : 33 (0)6 61 37 87 48
e-mail :  mailto:pierre.se...@worldonline.fr pierre.se...@worldonline.fr
  mailto:pierrese...@onetelnet.fr pierrese...@onetelnet.fr







 File att1.htm not included with reply 




---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




RE: Cooking induction apparatus and FCC

2001-05-21 Thread reheller


Part 18 still uses MP-5 but it is not mandatory. In cases of dispute, MP-5
will be used by the FCC. You can't go wrong, however,  using ANSI C63.4 as
the test set-up. Or you can use the CISPR limits as Gary described. I do
not know what C64.5 is.

You do not have to submit your test results through a TCB.


Bob Heller
3M Product Safety, 76-1-01
St. Paul, MN 55107-1208
Tel:  651- 778-6336
Fax:  651-778-6252





   
Gary McInturff  
   
Gary.McInturff@worldwidepa  To: 'Pierre SELVA' 
pierre.se...@worldonline.fr
ckets.com Forum Safety-emc 
emc-p...@majordomo.ieee.org  
 cc: (bcc: Robert E. 
Heller/US-Corporate/3M/US)
05/21/2001 10:59 AM  Subject: RE: Cooking 
induction apparatus and FCC  
Please respond to Gary  
   
McInturff   
   

   

   





Bon Jour (That's about it for my French)
Pierre, MP-5 isn't relevant any more. I believe you  want ANSI C64.5.
and I think you'll find mostly similarities. I can't think of any
dissimilarities of the top of my head, Even FCC part 15 allows use of the
CISPR limits for radiated and conducted emissions. The only caveat (ah,
more
French!) is that if you accept the radiated limits during test you also
have
to accept the conducted limits. Some folks trip up on that one. It means
that the conducted limits are really 150Khz even at US line voltage, not
the
normal FCC start of 450 Khz.
Gary

-Original

-Original Message-
From: Pierre SELVA [mailto:pierre.se...@worldonline.fr]
Sent: Monday, May 21, 2001 6:09 AM
To: Forum Safety-emc
Cc: Pierre SELVA
Subject: Cooking induction apparatus and FCC


Dear colleagues,

I need to know your opinion on the following :
I have to perform EMC testing on an induction cooking table and I would
like
to obtain the FCC certification.

In Europe, this kind of apparatus is subjected to the EN55011 (CISPR11) and
the test conditions are clearly stated, mainly for the EUT configuration.

I understand for the USA, the product has to be tested according to Part 18
requirements. This part explains that the technical requirements for the
measurement are described in the MP-5 document. As this one is a little bit
old (1986), do you believe I can use the CISPR11 to demonstrate compliance,
or should I use the MP5 ?

If I have to use the MP-5, do you know if somewhere the EUT configuration
is
described, or is the worst case at the manufacturer discretion ?

And, at last, do you know if I have to submit my test results and EUT
description to FCC directly (electronic submittal), or should I have to go
to a TCB ?

In advance, I thank you a lot for your answers,

Best regards from France


Pierre SELVA
2 route de la Grobelle
73000 JACOB BELLECOMBETTE - France
Tel : 33 (0)6 60 52 04 96
Fax : 33 (0)6 61 37 87 48
e-mail :  mailto:pierre.se...@worldonline.fr pierre.se...@worldonline.fr
  mailto:pierrese...@onetelnet.fr pierrese...@onetelnet.fr







 File att1.htm not included with reply 





---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Re: Public Health and Safety Signs - Tomfoolery so delete if you

2001-05-21 Thread John Woodgate

200105211432.kaa24...@interlock2.lexmark.com, oover...@lexmark.com
inimitably wrote:
I once read a safety article (tongue-in-cheek I believe) that said that the
safety industry was weaking the species by allowing the weak and feable to
continue exist and procreate.  I believe that this was directed toward the
mentally weak and feable (read stoopid).

By warning them about things that generally intellegient people would know were
hazardous, this special group would reproduce and create even more special
people.  Ultimately, the special people would out number the others to such 
an
extent that homosapiens would face extinction.

One of the assumptions was that these special people, through their routine
actions, would eliminate most of the other people through collateral damage.
(e.g.  drunk drivers, cell phone drivers, equipment operators, greedy managers
with only the bottom line in mind (otherwise known as bottom feeders))

Well, it has gotten to the point, as Tania has so elegantly pointed out, that
the safety profession only facilitates this; but the legal and political system
now rewards being special (again read stoopid).  Not only do we protect them
and allow them to procreate, we pay them large sums of money to those special
people for being so special.

This is a distasteful posting, and even if you wrote it in innocence you
may find you get attention from extreme right-wing elements. 

The idea that low intelligence is a dominant inherited characteristic is
simply not true. There is a heuristic observation of an 'equalization'
principle in human and other animal inheritance; unusually tall parents
have tall but not so unusually tall children. Many highly intelligent
parents have children whose intelligence is not so outstanding. Some
simple parents have genius-level children.
-- 
Regards, John Woodgate, OOO - Own Opinions Only. Phone +44 (0)1268 747839
Fax +44 (0)1268 777124. http://www.jmwa.demon.co.uk Why not call a vertically-
applied manulo-pedally-operated quasi-planar chernozem-penetrating and 
excavating implement a SPADE?

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Re: Production Line Test Requirements - Medical Devices

2001-05-21 Thread John Woodgate

36BDBCA75E0FD411A80100104B93ABF202C2C079@MGCMAIL, Dick Grobner
dick.grob...@medgraph.com inimitably wrote:
Why
wouldn't a simple continuity tests as defined by UL be appropriate from
equipment coming off of the production line (assuring that the protective
earth circuit in intake)? 

Because continuity at low current does not ensure that the protective
circuit will carry a large fault current - it might be 'hanging on by
one strand'.
-- 
Regards, John Woodgate, OOO - Own Opinions Only. Phone +44 (0)1268 747839
Fax +44 (0)1268 777124. http://www.jmwa.demon.co.uk Why not call a vertically-
applied manulo-pedally-operated quasi-planar chernozem-penetrating and 
excavating implement a SPADE?

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Re: UL P.A.G.

2001-05-21 Thread John Woodgate

lobbjbnlajjdfilpcaancehncjaa.peperkin...@cs.com, Pete Perkins
peperkin...@cs.com inimitably wrote:
Some standards committees themselves work to clarify the requirements; for
instance TC74/IEC 60950 has a chairman's advisory panel that will answer
questions from IEC national committees (not from individuals) that are
phrased in a manner that they can be answered by a simple yes/no response.
On occasion the process of answering these questions has prompted a change
in the standard to provide additional clarification to the issue.  These
answers are provided back to the IEC national committees for distribution.

IEC TC92 (responsible for IEC/EN60065) has a similar arrangement, but
will respond to questions requiring a quite complex response.

Within the CB Scheme the CTL (Committee of Testing Laboratories) has an
internal method for polling laboratories as to interpretation of
requirements or testing practice and, after discussion, providing an
agreed-upon interpretation for all labs to use.  This is intended to provide
uniform interpretation and application of requirements and testing
throughout all of the participating labs on a worldwide basis. I don’t
believe that these interpretations are generally available outside the labs.

TC92 now sends its interpretations to CTL and to CENELEC OSM (a similar
body), also to all IEC National Committees, who are expected to make the
interpretations publicly available (but not necessarily free, AIUI).
-- 
Regards, John Woodgate, OOO - Own Opinions Only. Phone +44 (0)1268 747839
Fax +44 (0)1268 777124. http://www.jmwa.demon.co.uk Why not call a vertically-
applied manulo-pedally-operated quasi-planar chernozem-penetrating and 
excavating implement a SPADE?

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




RE: Production Line Test Requirements - Medical Devices

2001-05-21 Thread Ned Devine

Hi Dick,

The requirement for using 10 to 25 Amps comes from the test house.  It is
their mark.  If you want to use it, they can require you to do anything they
want.  Is it fair, no.  Does the requirement make engineering sense, no (I
believe that Rich Nute did an article on how this requirement did not
identify anymore failures than the simple ohm meter did).  Do you still have
to do it, yes!

If you are just declaring conformity for the MDD, use your risk analysis
(i.e. EN 1441) to show that the risk of using a ohm meter is an acceptable
one.  Your NB might review it, but you should be able to switch to the ohm
meter.

Ned


Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548

616 248 9671 Phone
616 574 9752 Fax
ndev...@entela.com  e-mail




---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Test Equipment

2001-05-21 Thread Courtland Thomas

Hello group,

I am interested in purchasing some new/used test equipment. I am looking
specifically for a 'Digital Balance Analyzer' and an 'AC Stress/Hazardous
Voltage/Hazardous Current Analyzer'. Any suggestions would be appreciated.

Thanks,

Courtland Thomas
Patton Electronics


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




RE: UL P.A.G.

2001-05-21 Thread Fee John

Hi
You mentioned the CTL Decisions. While I understand that not all the
decisions have been placed on the web yet, you will find many of them
at www.iecee.org. Try under CTL or CTL Decisions
John Fee
National Electronics Technology Centre
Enterprise Ireland
Phone +353-1-8082214
Fax +353-1-8370705
email f...@netc.ie 

-Original Message-
From: Pete Perkins [mailto:peperkin...@cs.com]
Sent: 21 May 2001 16:12
To: PSNet; Brian O'Connell
Subject: RE: UL P.A.G.




PSNet

UL Practical application guide discussion

Several good points have been made in the discussion of the practical
application guide supplied by UL; I’ll add my U$ two-bits worth here

Let's start with a story: the question of interpretation has been with us
for a long time.  An old timer, a German trained engineer who later worked
for a test house, described his indoctrination into the world of safety
testing.  As a beginning engineer, one day his boss came to him and said
when they go up to VDE to have the product examined he should remember to
take along his copy of the standard and make notes in the margin as to the
interpretation of requirements as they were described.  Later, as a project
engineer, when he took his product to VDE his boss reminded him to take
along the new engineer with his copy of the standard to do the same thing.
This was obviously training by apprenticeship.

Having worked in standards development arena for 25 years or so, I found
that is not easy to make the standard absolutely clear – there is an ongoing
pressure for conciseness and compromise (since it is written by a
committee - you know all the old jokes: an elephant is a mouse designed to
MIL specs; a camel is a donkey designed by a committee).  This means that
there also is ongoing effort to get clarification of requirements and to
clarify the written standard itself.

We know that in large companies there is considerable discussion among
designers and safety engineering professionals as to the intent and
requirements of the standards.  In some cases it is organized, but much is
not.

Much of the discussion that we have here on the PSNet is along the same
lines - to provide the experience of others to help clarify the requirements
or the intent of the requirements in a standard.  As with the company
efforts, the advent of email has gone a long way to get things written down,
but not always organized in any recoverable way (RCIC not withstanding).  My
experience is that UL engineers are generally not familiar with the PSNet
discussions.

There are some formal, traditional efforts to do this clarification 
codification.

Some standards committees themselves work to clarify the requirements; for
instance TC74/IEC 60950 has a chairman's advisory panel that will answer
questions from IEC national committees (not from individuals) that are
phrased in a manner that they can be answered by a simple yes/no response.
On occasion the process of answering these questions has prompted a change
in the standard to provide additional clarification to the issue.  These
answers are provided back to the IEC national committees for distribution.

Within the CB Scheme the CTL (Committee of Testing Laboratories) has an
internal method for polling laboratories as to interpretation of
requirements or testing practice and, after discussion, providing an
agreed-upon interpretation for all labs to use.  This is intended to provide
uniform interpretation and application of requirements and testing
throughout all of the participating labs on a worldwide basis. I don’t
believe that these interpretations are generally available outside the labs.

Finally, UL has its P. A. G. that we've been discussing.  This guide was
developed so that UL engineers would provide the same interpretation of
requirements from any office.  It's really quite convenient that they offer
this outside of the lab for use but there's no obligation to do so.
Charging a nominal amount to support making this available does not, in my
opinion, seem unreasonable.  This could also easily be done on-line where
you would be charged on a page-by-page basis (how many peeks per U$ dollar
at those Russian sneak-a-peek sites - once you enter your credit card
number?).

Whether or not these are more than guidelines - it seems that trying to work
around them is like trying to push aside the preponderance of the evidence
in legal proceedings.  In each case they are considered opinions from a
group of knowledgeable participants in each process.

It would be most convenient to find some way to combine all the interpretive
information into the standard such as is done with the US NEC (National
Electric Code) Handbook.  In that book each clause is followed by
interpretive information to further explain the application of the
requirements.  (Hmmm – on reflection this looks like it would be a good
retirement project for me if I could get my hands on the interpretations.
What would you be willing to pay?)

I hope that this little 

RE: Cooking induction apparatus and FCC

2001-05-21 Thread Gary McInturff
Bon Jour (That's about it for my French)
Pierre, MP-5 isn't relevant any more. I believe you  want ANSI C64.5.
and I think you'll find mostly similarities. I can't think of any
dissimilarities of the top of my head, Even FCC part 15 allows use of the
CISPR limits for radiated and conducted emissions. The only caveat (ah, more
French!) is that if you accept the radiated limits during test you also have
to accept the conducted limits. Some folks trip up on that one. It means
that the conducted limits are really 150Khz even at US line voltage, not the
normal FCC start of 450 Khz.
Gary

-Original 

-Original Message-
From: Pierre SELVA [mailto:pierre.se...@worldonline.fr]
Sent: Monday, May 21, 2001 6:09 AM
To: Forum Safety-emc
Cc: Pierre SELVA
Subject: Cooking induction apparatus and FCC


Dear colleagues,
 
I need to know your opinion on the following :
I have to perform EMC testing on an induction cooking table and I would like
to obtain the FCC certification.
 
In Europe, this kind of apparatus is subjected to the EN55011 (CISPR11) and
the test conditions are clearly stated, mainly for the EUT configuration.
 
I understand for the USA, the product has to be tested according to Part 18
requirements. This part explains that the technical requirements for the
measurement are described in the MP-5 document. As this one is a little bit
old (1986), do you believe I can use the CISPR11 to demonstrate compliance,
or should I use the MP5 ?
 
If I have to use the MP-5, do you know if somewhere the EUT configuration is
described, or is the worst case at the manufacturer discretion ?
 
And, at last, do you know if I have to submit my test results and EUT
description to FCC directly (electronic submittal), or should I have to go
to a TCB ?
 
In advance, I thank you a lot for your answers,
 
Best regards from France


Pierre SELVA
2 route de la Grobelle
73000 JACOB BELLECOMBETTE - France
Tel : 33 (0)6 60 52 04 96
Fax : 33 (0)6 61 37 87 48
e-mail :  mailto:pierre.se...@worldonline.fr pierre.se...@worldonline.fr
  mailto:pierrese...@onetelnet.fr pierrese...@onetelnet.fr
 

 



RE: Cooking induction apparatus and FCC

2001-05-21 Thread Gary McInturff
 

-Original Message-
From: Pierre SELVA [mailto:pierre.se...@worldonline.fr]
Sent: Monday, May 21, 2001 6:09 AM
To: Forum Safety-emc
Cc: Pierre SELVA
Subject: Cooking induction apparatus and FCC


Dear colleagues,
 
I need to know your opinion on the following :
I have to perform EMC testing on an induction cooking table and I would like
to obtain the FCC certification.
 
In Europe, this kind of apparatus is subjected to the EN55011 (CISPR11) and
the test conditions are clearly stated, mainly for the EUT configuration.
 
I understand for the USA, the product has to be tested according to Part 18
requirements. This part explains that the technical requirements for the
measurement are described in the MP-5 document. As this one is a little bit
old (1986), do you believe I can use the CISPR11 to demonstrate compliance,
or should I use the MP5 ?
 
If I have to use the MP-5, do you know if somewhere the EUT configuration is
described, or is the worst case at the manufacturer discretion ?
 
And, at last, do you know if I have to submit my test results and EUT
description to FCC directly (electronic submittal), or should I have to go
to a TCB ?
 
In advance, I thank you a lot for your answers,
 
Best regards from France


Pierre SELVA
2 route de la Grobelle
73000 JACOB BELLECOMBETTE - France
Tel : 33 (0)6 60 52 04 96
Fax : 33 (0)6 61 37 87 48
e-mail :  mailto:pierre.se...@worldonline.fr pierre.se...@worldonline.fr
  mailto:pierrese...@onetelnet.fr pierrese...@onetelnet.fr
 

 



RE: Production Line Test Requirements - Medical Devices

2001-05-21 Thread Dick Grobner

Jon
Thx for the info, however, I re-checked the two known US standards (UL2601,
Appendix D - Manufacturer's Responsibilities, Construction Considerations
and Requirements for Factory Tests) (NFPA 99, Grounding Circuit Continuity -
Measurement of Resistance). UL defines Production Line Grounding Continuity
Test Equipment as: Any suitable continuity indication device (such as an
ohmmeter, a battery and buzzer combination, or the like) may be used to
determine compliance with Grounding Continuity Test requirements. The NFPA
std is even less definitive on this matter.
EN60601-1, Appendix B - Testing During Manufacture and/or Installation,
Not Used. See rational to sub-clause 4.1. Which states - Tests described in
this standard are type tests.

I checked all of the amendments including EN60601-1 for medical systems, and
I do not see anything changing this.

My questions remains, where does it state to use the 10-25Amp, 50 or 60Hz @
6Volt for 5 seconds test? Is this a carry over from another EN standard? Why
wouldn't a simple continuity tests as defined by UL be appropriate from
equipment coming off of the production line (assuring that the protective
earth circuit in intake)? 

Appreciate your feedback Jon!  thanks again.
 
Does anyone else on this forum have any input - would appreciate it.


-Original Message-
From: Jon Griver [mailto:jo...@medson.com]
Sent: Monday, May 21, 2001 9:16 AM
To: Dick Grobner
Subject: RE: Production Line Test Requirements - Medical Devices


Dick,

The requirements for production tests of medical equipment generally are:

1. Hipot at 1500AC - between live and neutral connected together and earth.
2. Earth continuity at 25A between earth pin on plug and enclosure (choose a
point that may be problematic if there are manufacturing problems)
3. Earth and/or patient leakage. Whether you do both, one or neither of
these tests depends on the product.

As the production tests are intended to find production problems (as opposed
to type tests which are intended to find design problems), choose the tests
that you think make sense from the production engineering point of view.
Document the reasoning behind your choice.

IT equipment only requires Hipot and earth continuity tests.

Hope this helps,


Jon


 -Original Message-
 From: Dick Grobner [mailto:dick.grob...@medgraph.com]
 Sent: Monday, May 21, 2001 16:36
 To: 'Jon Griver'
 Subject: RE: Production Line Test Requirements - Medical Devices


 Jon
 Thanks, you have confirmed what I thought I knew. We just had our
 annual ISO
 9001 audit (2 weeks ago). One issue that one of the two auditors
 raised was
 why are you not doing the ground integrity test (25 Amp test) on 100% of
 your production units? I asked do you mean ground continuity
 (Ohm test) and
 he said no, the ground integrity test. That's when I started digging into
 the standards. UL (and the former ETL) states high potential and ground
 continuity tests on all production units. EN60601-1 states that all tests
 within this document are type tests. The auditor gave me no reference to
 any EN, etc. when I asked. So, I have a suspicion that this is
 his wish and
 not stated in any EN (at least that I know of so far). SO - I
 will continue
 to pursue with other outside sources (but not this one auditor!)
 Thanks for the reply back!
 PS - does ITE equipment require this test (ground integrity ((25
 Amp)) test?


 -Original Message-
 From: Jon Griver [mailto:jo...@medson.com]
 Sent: Sunday, May 20, 2001 1:34 AM
 To: emc-p...@majordomo.ieee.org
 Subject: FW: Production Line Test Requirements - Medical Devices



 Dick,

 To the best of my knowledge there is no EN standard or guidance on
 production tests for medical equipment (There is a standard, EN 50116, for
 IT equipment).
 Your best bet is to discuss this with your Notified Body and to come to a
 mutually acceptable set of tests.

 Regards,

 Jon Griver
 Medson Ltd.


  Good Day Everyone
  Question I have - Does anyone out there know if an EN standard
 or guidance
  document exists that deals with production line test requirements of
  finished medical devices.
  Reading in EN60601-1 it states that the test (ground resistance, high
  potential, etc.) are type test only (See appendix B, and then paragraph
  4.1). Thus - they are not identified as production line tests.
 I know that
  the NRTL's in the USA (UL/ETL) specifically call out what tests
 are to be
  performed after the production build of equipment.
  Any input would be appreciated.
  Thx
 
  Dick Grobner
  Medical Graphics Corporation
  350 Oak Grove Parkway
  St Paul MN 55127
  651-766-3395
  651-766-3389 (fax)
  dick.grob...@medgraph.com
 
 
 
  ---
  This message is from the IEEE EMC Society Product Safety
  Technical Committee emc-pstc discussion list.
 
  Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
 
  To cancel your subscription, send mail to:
   majord...@ieee.org
  with the 

RE: Public Health and Safety Signs - Tomfoolery so delete if you don't have the time.

2001-05-21 Thread Nerad, Daren HS-SNS

Gary,
We learned about this one almost 13 years ago with the first baby...
My wife inadvertently went through about a pot of coffee while on the phone
(long distance) with a rarely seen friend.  In a way, the results of that
were much worse than passing on a glass of wine or two!!  Since neither one
was used to that much coffee, in that short a period, it was unpleasant, but
educational.  From an engineering perspective: it was a very conclusive
test!

Daren A. Nerad
EMC Engineer
815.226.6123


-Original Message-
From: Gary McInturff [mailto:gary.mcintu...@worldwidepackets.com]
Sent: Friday, May 18, 2001 4:40 PM
To: 'Michael Mertinooke'; wo...@sensormatic.com;
emc-p...@majordomo.ieee.org
Subject: RE: Public Health and Safety Signs - Tomfoolery so delete if
you don't have the time.



Here in the US, awhile back,  a woman was suing the liquor industry
because she gave birth to a fetal alcohol syndrome child. Apparently, nobody
in their right mind would assume that consuming a fifth of whiskey a day
could be harmful to a developing fetus making the liquor industry patiently
and damnably negligent in not putting warning labels on the bottles. (We got
them now thank God!)
During the coverage of the trial, and I don't remember the context,
but the issue of passing nastiness to infants who were being breast fed was
also brought up. While I didn't hear the end of this I often have wondered
that if that was true, and this woman's case had merit (her lawyer took it
up didn't he?) then the logical extension would be that mothers milk should
come with a warning. 

Soo Just what the heck will this label look like, and even
more importantly, just where are they going to put it so that people, can
easily read it!

Gary

-Original Message-
From: Michael Mertinooke [mailto:mertino...@skyskan.com]
Sent: Friday, May 18, 2001 12:37 PM
To: wo...@sensormatic.com; emc-p...@majordomo.ieee.org
Subject: RE: Public Health and Safety Signs




signs at work. Is there a similar Directive for health and safety signs for
the general public?

Whoo! The mind boggles! You mean with like people with exclamation
point in triangle tattoos on various portions of the anatomy? Or
biohazard labels on the door of the kids' rooms? Judging
from some of the ANSI Z535 safety labels I see in the catalogs, the
Human Warning Labels would be interesting indeed.   =]

Cheers!
Mike


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




RE: UL P.A.G.

2001-05-21 Thread Pete Perkins


PSNet

UL Practical application guide discussion

Several good points have been made in the discussion of the practical
application guide supplied by UL; I’ll add my U$ two-bits worth here

Let's start with a story: the question of interpretation has been with us
for a long time.  An old timer, a German trained engineer who later worked
for a test house, described his indoctrination into the world of safety
testing.  As a beginning engineer, one day his boss came to him and said
when they go up to VDE to have the product examined he should remember to
take along his copy of the standard and make notes in the margin as to the
interpretation of requirements as they were described.  Later, as a project
engineer, when he took his product to VDE his boss reminded him to take
along the new engineer with his copy of the standard to do the same thing.
This was obviously training by apprenticeship.

Having worked in standards development arena for 25 years or so, I found
that is not easy to make the standard absolutely clear – there is an ongoing
pressure for conciseness and compromise (since it is written by a
committee - you know all the old jokes: an elephant is a mouse designed to
MIL specs; a camel is a donkey designed by a committee).  This means that
there also is ongoing effort to get clarification of requirements and to
clarify the written standard itself.

We know that in large companies there is considerable discussion among
designers and safety engineering professionals as to the intent and
requirements of the standards.  In some cases it is organized, but much is
not.

Much of the discussion that we have here on the PSNet is along the same
lines - to provide the experience of others to help clarify the requirements
or the intent of the requirements in a standard.  As with the company
efforts, the advent of email has gone a long way to get things written down,
but not always organized in any recoverable way (RCIC not withstanding).  My
experience is that UL engineers are generally not familiar with the PSNet
discussions.

There are some formal, traditional efforts to do this clarification 
codification.

Some standards committees themselves work to clarify the requirements; for
instance TC74/IEC 60950 has a chairman's advisory panel that will answer
questions from IEC national committees (not from individuals) that are
phrased in a manner that they can be answered by a simple yes/no response.
On occasion the process of answering these questions has prompted a change
in the standard to provide additional clarification to the issue.  These
answers are provided back to the IEC national committees for distribution.

Within the CB Scheme the CTL (Committee of Testing Laboratories) has an
internal method for polling laboratories as to interpretation of
requirements or testing practice and, after discussion, providing an
agreed-upon interpretation for all labs to use.  This is intended to provide
uniform interpretation and application of requirements and testing
throughout all of the participating labs on a worldwide basis. I don’t
believe that these interpretations are generally available outside the labs.

Finally, UL has its P. A. G. that we've been discussing.  This guide was
developed so that UL engineers would provide the same interpretation of
requirements from any office.  It's really quite convenient that they offer
this outside of the lab for use but there's no obligation to do so.
Charging a nominal amount to support making this available does not, in my
opinion, seem unreasonable.  This could also easily be done on-line where
you would be charged on a page-by-page basis (how many peeks per U$ dollar
at those Russian sneak-a-peek sites - once you enter your credit card
number?).

Whether or not these are more than guidelines - it seems that trying to work
around them is like trying to push aside the preponderance of the evidence
in legal proceedings.  In each case they are considered opinions from a
group of knowledgeable participants in each process.

It would be most convenient to find some way to combine all the interpretive
information into the standard such as is done with the US NEC (National
Electric Code) Handbook.  In that book each clause is followed by
interpretive information to further explain the application of the
requirements.  (Hmmm – on reflection this looks like it would be a good
retirement project for me if I could get my hands on the interpretations.
What would you be willing to pay?)

I hope that this little summary is useful.

  br, Pete

  Peter E Perkins, PE
  Principal Product Safety Consultant
  Tigard, ORe 97281-3427
  503/452-1201 fone/fax
  p.perk...@ieee.org



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 

RE: Li-ion Batteries

2001-05-21 Thread Michael Mertinooke

Is there a European standard for Li-ion batteries?
Is there a battery directive? If yes, what standards is applicable?
CE mark required?
--
As far as I know, the relevant EU directive is 98/101/EC, 22 December
1998, which adapts 91/157/EEC on batteries and accumulators. This
only relates to the chemicals used in the battery.

I asked Moltech Power Systems (Energizer) what directives applied
to EU, and got the reply attached below.

Mike Mertinooke


--
Michael:

Are you joking with this response?  Are you kidding me?

Regards,

Joe DeCarlo
Moltech Power Systems
Ph: 860-257-1020
Fx: 860-257-0165



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Re: Public Health and Safety Signs - Tomfoolery so delete if you

2001-05-21 Thread ooverton



I once read a safety article (tongue-in-cheek I believe) that said that the
safety industry was weaking the species by allowing the weak and feable to
continue exist and procreate.  I believe that this was directed toward the
mentally weak and feable (read stoopid).

By warning them about things that generally intellegient people would know were
hazardous, this special group would reproduce and create even more special
people.  Ultimately, the special people would out number the others to such an
extent that homosapiens would face extinction.

One of the assumptions was that these special people, through their routine
actions, would eliminate most of the other people through collateral damage.
(e.g.  drunk drivers, cell phone drivers, equipment operators, greedy managers
with only the bottom line in mind (otherwise known as bottom feeders))

Well, it has gotten to the point, as Tania has so elegantly pointed out, that
the safety profession only facilitates this; but the legal and political system
now rewards being special (again read stoopid).  Not only do we protect them
and allow them to procreate, we pay them large sums of money to those special
people for being so special.







Tania Grant taniagrant%msn@interlock.lexmark.com on 05/18/2001 09:11:49
PM

Please respond to Tania Grant taniagrant%msn@interlock.lexmark.com

To:   Gary McInturff
  gary.mcinturff%worldwidepackets@interlock.lexmark.com, 'Michael
  Mertinooke' mertinooke%skyskan@interlock.lexmark.com,
  woods%sensormatic@interlock.lexmark.com,
  emc-pstc%majordomo.ieee@interlock.lexmark.com
cc:(bcc: Oscar Overton/Lex/Lexmark)
Subject:  Re: Public Health and Safety Signs - Tomfoolery so delete if you



Gary,

Kudos to you for trying;--  but I think you are confusing your metaphors.   And
you can't equate logic with brainless juries or judges.   And you know better
than to rely on warning labels for safety protection!

1.  It is assumed that intelligent people will make intelligent choices when
opening a bottle with Warning labels that state the content can do you in.

2.  It is assumed that a nursing baby, while potentially intelligent, does not
have the capability to make any choices whatsoever as to what she is consuming.

3.  It is assumed that a lactating mother has a direct effect on the well being
of her baby.

4.  Therefore, the mother is the responsible party for any adverse effects the
baby may suffer due to conditions such as described in 1. above.

5.  Therefore, we can assume that any decision rendered otherwise by judge
and/or jury is brainless.

For technological widgets we bend over backwards to make them safe and we don't
rely on labels to protect the general user (trained service persons, however,
can be protected by labels in certain cases).  However, our
legislators/politicians think they can affect and protect our behavior by labels
and warning statements.This does not really work;--warning labels are for
others, never for ourselves!

So, what is the answer? --  Education for intelligent people; and safe design
(in case of a single fault, and a subsequent fault)  for dumb widgets.

Plastering warning labels on breasts will no more eliminate alcoholic babies
than plastering warning labels on men's ...  will eliminate HIV.

Tania Grant
taniagr...@msn.com



- Original Message -
From: Gary McInturff
Sent: Friday, May 18, 2001 5:33 PM
To: 'Michael Mertinooke'; wo...@sensormatic.com; emc-p...@majordomo.ieee.org
Subject: RE: Public Health and Safety Signs - Tomfoolery so delete if you



Here in the US, awhile back,  a woman was suing the liquor industry
because she gave birth to a fetal alcohol syndrome child. Apparently, nobody
in their right mind would assume that consuming a fifth of whiskey a day
could be harmful to a developing fetus making the liquor industry patiently
and damnably negligent in not putting warning labels on the bottles. (We got
them now thank God!)
During the coverage of the trial, and I don't remember the context,
but the issue of passing nastiness to infants who were being breast fed was
also brought up. While I didn't hear the end of this I often have wondered
that if that was true, and this woman's case had merit (her lawyer took it
up didn't he?) then the logical extension would be that mothers milk should
come with a warning.

Soo Just what the heck will this label look like, and even
more importantly, just where are they going to put it so that people, can
easily read it!

Gary

-Original Message-
From: Michael Mertinooke [mailto:mertino...@skyskan.com]
Sent: Friday, May 18, 2001 12:37 PM
To: wo...@sensormatic.com; emc-p...@majordomo.ieee.org
Subject: RE: Public Health and Safety Signs




signs at work. Is there a similar Directive for health and safety signs for
the general public?

Whoo! The mind boggles! You mean with like people with exclamation
point in triangle tattoos on various portions of the anatomy? Or

Cooking induction apparatus and FCC

2001-05-21 Thread Pierre SELVA
Dear colleagues,

I need to know your opinion on the following :
I have to perform EMC testing on an induction cooking table and I would like
to obtain the FCC certification.

In Europe, this kind of apparatus is subjected to the EN55011 (CISPR11) and
the test conditions are clearly stated, mainly for the EUT configuration.

I understand for the USA, the product has to be tested according to Part 18
requirements. This part explains that the technical requirements for the
measurement are described in the MP-5 document. As this one is a little bit
old (1986), do you believe I can use the CISPR11 to demonstrate compliance,
or should I use the MP5 ?

If I have to use the MP-5, do you know if somewhere the EUT configuration is
described, or is the worst case at the manufacturer discretion ?

And, at last, do you know if I have to submit my test results and EUT
description to FCC directly (electronic submittal), or should I have to go
to a TCB ?

In advance, I thank you a lot for your answers,

Best regards from France

Pierre SELVA
2 route de la Grobelle
73000 JACOB BELLECOMBETTE - France
Tel : 33 (0)6 60 52 04 96
Fax : 33 (0)6 61 37 87 48
e-mail : pierre.se...@worldonline.fr