Brian,

You can provide maintenance and servicing info to your customer.
Just be sure your manuals are in order.

To write a any kind of manual, a clear definition audience is necessary.
In your case, you are considering the training and knowledge of the
average laboratory user - about the hazards of performing maintenance
on electrical equipment (best to assume none).  Qualified as a
scientist (even in an electrical field) has no relation to a qualification
to perform maintenance tasks.

Tasks requiring the ability to understand (and avoid) the hazards
present must only be carried out by truly qualified personnel.
You recognize this in the construction of your equipment.
 My main area of difficulty lies in that we rely on the fact
 that our electronic units are located behind locked doors or screwed
 panels to achieve our LVD compliance.

You should clearly establish at least two categories of personnel who
will be interacting with your equipment:

1. Skilled personnel (maintenance staff, specialists).
You must define the general skills/knowledge required
or the necessary qualification(s) of skilled personnel.

You must also specify any specific training/knowledge these
skilled personnel will require to be able to recognize and avoid
any proprietary hazards in your equipment.

2. Users (operators).
Assume:
they cannot recognize any hazards
(they only know what you've told them in the manual)
they will ignore many of your warnings
(which is why some designers use special screws for closing off
non-user areas)

You cannot assume that the user is a skilled person.
EN 61010 does not assume this.

Your duty as the manufacturer is to CLEARLY separate tasks
which can be performed by users from those which must
only be performed by skilled/trained personnel.
Then, you must explain how to carry out these tasks safely.

Regards,

Matt

PS
If you want to talk about the qualifications of scientists - just browse
around
the internet.  Or go directly here:
http://www.rli.com/accident/year_lists/pre-1987.html
#021: 1977: Scientist blinded by pulsed Nd:YAG laser.
A scientist was partially blinded by a reflection from what was called a
relatively weak Nd:YAG The exposure in the eye was approximately 6 mJ.
However, 6 mJ in a 10 nsec exposure time creates enormous peak power
approximately a thousand times greater than the limit allowed into the eye.
Although the laser was thought to be relatively weak, in fact it was many
orders of magnitude above the accepted safe exposure limit.
As a result of the exposure, a vitreous hemorrhage was produced and the
person went into shock. Fortunately, the hemorrhage did not produce a foveal
lesion, and eventually some vision did return.
The accident was due to the fact that although eye protection, was
available, it was not being used. This incident could have been avoided if a
laser safety program had developed a safety awareness so that people wore
protective eyewear.















-
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to majord...@ieee.org
with the single line: unsubscribe emc-pstc (without the
quotes).  For help, send mail to ed.pr...@cubic.com,
jim_bac...@monarch.com, ri...@sdd.hp.com, or
roger.volgst...@compaq.com (the list administrators).



Re: Servicing and repairs

1999-12-22 Thread Brian Kunde at LECO

Hi Brian Harlowe,

I'm not aware of a EU Legislation that says a company can't service or
maintain purchased equipment. Our company is in the same situation as you
are: we manufacture laboratory equipment and sell it world wide. We also
test our products to EN61010 which is not as strict in these areas as
consumer products would be if tested to EN60950. It is assumed that the
operator has the proper training to perform the necessary tasks of their
job.

Many of our customers demand to service and maintain our products
themselves, so we have to provide procedures, schematics, part lists,
assembly diagrams, etc...   This request is so common that we provide all
this information with every unit now.

The manufacturer of a product is not responsible for the technical level of
the operator or the person who maintains or services the instrument at the
customer location.  The customer has the burden to make sure his people are
properly trained for the tasks under their job description.  As a
manufacturer, you have to provide a safe product and provide instructions
for safe operation,  maintenance, and service. As for maintenance and
service, provide the proper warnings in the documentation and on the product
such as Warning: High Voltage. Remove Mains Cord prior to removing panel.
Your job is to identify all hazards whether it is the operator or a service
person who will be performing the task.

You don't have to provide Maintenance and Service information for your
customers. You can make a stand and say that these tasks can only be
performed by your own people. BUT, you will still have to generate
maintenance and service documentation for in-house use. This is a
requirement especially if documentation is required to insure the safety of
the person performing the maintenance or service.

So the question is, what's wrong with providing your customer with
Maintenance and Service information on your product?  Nothing that I am
aware of as long as you do a good job in identifying the hazards.

Our company takes a similar view as our customers. We will not purchase a
piece of equipment, whether a computer, a lift truck, or digital scope
unless we can obtain the schematics, drawings, and service documentation.  I
think in this industry it is for the most part a requirement of doing
business.

Brian Kunde





- Original Message -
From: Brian Harlowe bharl...@vgscientific.com
To: emc-p...@ieee.org
Sent: Tuesday, December 21, 1999 12:04 PM
Subject: Servicing and repairs



Traditionally my company has encouraged it's users to carry out a
limited amount of servicing and repair on our Electronic units.

Under the EU Safety legislation I know this is now a No No.

I am a little bit of a lone voice crying in the wilderness as far as
our management is concerned.

Can any one out there quote me any instances or cases that I can use
to drive home the point with our management

A happy Christmas to you fellow compliance people and if we survive
the Y2K business.  Good luck in the new century and may this
newsgroup continue to prosper

Best Regards

Brian Harlowe
* opinions expressed here are personal and in no way reflect the position of
VG Scientific

-
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to majord...@ieee.org
with the single line: unsubscribe emc-pstc (without the
quotes).  For help, send mail to ed.pr...@cubic.com,
jim_bac...@monarch.com, ri...@sdd.hp.com, or
roger.volgst...@compaq.com (the list administrators).





-
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to majord...@ieee.org
with the single line: unsubscribe emc-pstc (without the
quotes).  For help, send mail to ed.pr...@cubic.com,
jim_bac...@monarch.com, ri...@sdd.hp.com, or
roger.volgst...@compaq.com (the list administrators).



RE: Servicing and Repairs

1999-12-22 Thread John Allen

Brian

Your comments also make interesting reading.

Unfortunately I don't have time to comment in detail as I am finishing for 
XMAS in about 2 hrs!

However, in respect of your issue about access of untrained people to 
potentially dangerous situation should be prevented by appropriate 
inter-locking, labelling and instructions.

If EN61010 does not address the issue in sufficient detail then look at 
EN60950 Safety of Information Technology Equipment - especially Cl 2.8 
Safety Interlocks - as this standard is aimed squarely at equipment for 
use by untrainned personnel.

Additionally, you might want to look at the access and interlock 
requirements standards called up under the Machinery Directive 98/37/EC and 
the guidance information put out by authorities such as the UK Health  
Safety Executive.

Regards

John Allen.

--
From:   Brian Harlowe[SMTP:bharl...@vgscientific.com]
Sent:   22 December 1999 11:54
To: emc-p...@ieee.org
Cc: john.al...@rdel.co.uk
Subject:Servicing and Repairs


Hi Folks
John Allen has opened my can of worms even further with
his very useful comments to my e.mail. Obviously the military/defence
situation is somewhat different as he states but maybe there is some
guidance there.

The answers to his comments are his follows:-

Our products are large Scientific instruments and the EMC, LVD
and Product liability directives apply

The standards we have applied are EN 61326-1 and EN 61010-1

Servicing would be to a Minor level although in some instances access
is required to units to carry out conditioning procedures

The target customer is generally a qualified Scientist but not
necessarily from an Electrical/Electronic Background.

Manuals to an appropriate level could be produced but are currently
only available  to our service organisation.

My main area of difficulty lies in that we rely on the fact
that our electronic units are located behind locked doors or screwed
panels to achieve our LVD compliance.

It then seems at odds to encourage the user to undo these panels to
carry out routine service or repair tasks.

Also I am very uncomfortable on the product liability front if
someone was hurt or killed because we had suggested he opened panels
behind which were dangerous voltages.

It has been suggested that we should obtain a declaration that the
user has qualified trained staff who are capable of doing this work.
Could staff be regarded as qualified if they had not been trained by
us in the use of proprietry units? Many of our instruments go to
academic users who usually consider themselves competant in every
area. I would be interested in the groups comments in this area.

Finally as John states very specific instructions would have to be
given on reassembling units in order to maintain the units/systems
emc signature

These are the main reasons I feel these operations should only be
carried out by our own Staff. Although I can understand a customers
position when he has a piece of capital equipment worth up to stlg750k
he is unable to use due to the failure of a simple component

Regards

Brian Harlowe
* opinions expressed here are personal and in no way reflect the position 
of VG Scientific

-
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to majord...@ieee.org
with the single line: unsubscribe emc-pstc (without the
quotes).  For help, send mail to ed.pr...@cubic.com,
jim_bac...@monarch.com, ri...@sdd.hp.com, or
roger.volgst...@compaq.com (the list administrators).



-
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to majord...@ieee.org
with the single line: unsubscribe emc-pstc (without the
quotes).  For help, send mail to ed.pr...@cubic.com,
jim_bac...@monarch.com, ri...@sdd.hp.com, or
roger.volgst...@compaq.com (the list administrators).



RE: Servicing and repairs

1999-12-22 Thread John Allen

Hi folks

Brian raises an interesting issue - but I would be grateful if he or 
someone else can identify the EU legislation to which he refers.

His comments are particularly to companies like ours who supply mainly to a 
military market which almost demands to do its own servicing because it 
locates and uses the equipment in places where normal manufacturer's 
service cover is unavailable (certainly we don't have a service department 
in Bosnia or Kosovo!!). However, the customers' requirements' of this 
nature are stated in contract and typically includes full maintenance and 
repair training and documentation.

I think our situation digresses from the issue raised, but it does have a 
bearing on my comments, notably that each case should be considered on its 
merits,

Notably, the following appear to be relevant:

a) The specific items of EU legislation which apply to the product.

b) The type of product and the safety/EMC/telecomms standards to which it 
has been assessed will be relevant.

c) The level of servicing which will/could be required.

d) The typical target customer market, as it will affect the level of skill 
required to correctly maintain/repair the product'.

e) The availability to the customer of the maintenance and repair 
instructions, the depth of detail included, and the warnings/cautions 
provided on what must - and must not be - done, including any tests to be 
applied at the end of repair/maintenance.

f) The availability and assembly level of the correct routine spares, and 
of replacement or re-conditioned exchange units.

Frankly - whilst I fully sympathise with Brian's viewpoint - I think it is 
very difficult to make a blanket statement that Mr Customer, thou shalt 
not service/repair this equipment, particularly in respect of routine 
maintenance functions, e.g. changing filters, topping up fluids, cleaning, 
etc.

Many of us do these tasks on our own cars - and would object violently to 
having to go back to the manufacturer (not an agent, according to Brian) to 
get these tasks done at the cost of a lot of time and money.

However, when it get to more complex functions, then the definition of what 
can/cannot be done is more difficult.

Nevertheless, that is what you should do in detail, and you should then 
clearly specify:
- The maintenance documentation (by reference number etc.) in the operating 
instructions.
- Which tasks can/cannot be done, how often, and what the required skill 
levels are.
- All the Warnings and Cautions for the safety of both the repairer and the 
user, and to prevent the equipment going out of compliance with the 
safety/emc/etc., standards to which it was designed and manufactured.
- Which spares shall be used.
- What checks and tests must be performed at the end of the tasks.
- What records should be kept of the maintenance and repairs performed 
(what, when, by whom).

If you - the manufacturer - do the above, then I can see no real reason for 
not permitting professional customers to service their equipment.

However, finally, I do accept that if you supply equipment intended for use 
by the general public - particularly in their own homes - you may have to 
clearly specify on both the equipment and in the operating instructions 
that no-user servicing is permitted. You might even have to make access 
inside the equipment very difficult by sealing the construction or by using 
special security fixings that cannot be opened with normal handtools.

Regards  Happy Xmas

John Allen
Racal Defence Electronics Ltd


--
From:   Brian Harlowe[SMTP:bharl...@vgscientific.com]
Sent:   21 December 1999 17:04
To: emc-p...@ieee.org
Subject:Servicing and repairs


Traditionally my company has encouraged it's users to carry out a
limited amount of servicing and repair on our Electronic units.

Under the EU Safety legislation I know this is now a No No.

I am a little bit of a lone voice crying in the wilderness as far as
our management is concerned.

Can any one out there quote me any instances or cases that I can use
to drive home the point with our management

A happy Christmas to you fellow compliance people and if we survive
the Y2K business.  Good luck in the new century and may this
newsgroup continue to prosper

Best Regards

Brian Harlowe
* opinions expressed here are personal and in no way reflect the position 
of VG Scientific

-
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to majord...@ieee.org
with the single line: unsubscribe emc-pstc (without the
quotes).  For help, send mail to ed.pr...@cubic.com,
jim_bac...@monarch.com, ri...@sdd.hp.com, or
roger.volgst...@compaq.com (the list administrators).



-
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to majord...@ieee.org
with the single line: unsubscribe emc-pstc (without the
quotes).  For help, send mail to ed.pr...@cubic.com,
jim_bac...@monarch.com, 

Re: Servicing and repairs

1999-12-21 Thread Ralph Cameron

Brian:

As a comment I would say that the liability to meet EU approval ends when
the product is sold and ownership is transferred. That's the way it happens
with consumer goods in Canada, in spite of the fact there is CSA approval
required on the original product.

Compliments of the Season

Ralph Cameron
EMC Consultant and Suppression of Consumer Electronic Equipment
(After Sale).

- Original Message -
From: Brian Harlowe bharl...@vgscientific.com
To: emc-p...@ieee.org
Sent: Tuesday, December 21, 1999 9:04 AM
Subject: Servicing and repairs



 Traditionally my company has encouraged it's users to carry out a
 limited amount of servicing and repair on our Electronic units.

 Under the EU Safety legislation I know this is now a No No.

 I am a little bit of a lone voice crying in the wilderness as far as
 our management is concerned.

 Can any one out there quote me any instances or cases that I can use
 to drive home the point with our management

 A happy Christmas to you fellow compliance people and if we survive
 the Y2K business.  Good luck in the new century and may this
 newsgroup continue to prosper

 Best Regards

 Brian Harlowe
 * opinions expressed here are personal and in no way reflect the position
of VG Scientific

 -
 This message is coming from the emc-pstc discussion list.
 To cancel your subscription, send mail to majord...@ieee.org
 with the single line: unsubscribe emc-pstc (without the
 quotes).  For help, send mail to ed.pr...@cubic.com,
 jim_bac...@monarch.com, ri...@sdd.hp.com, or
 roger.volgst...@compaq.com (the list administrators).





-
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to majord...@ieee.org
with the single line: unsubscribe emc-pstc (without the
quotes).  For help, send mail to ed.pr...@cubic.com,
jim_bac...@monarch.com, ri...@sdd.hp.com, or
roger.volgst...@compaq.com (the list administrators).