Re: [volt-nuts] GPIB Issue with 3458A
Todd, Since the KE5FX NVRAM download software works, that would seem to indicate the GPIB is working, wouldn't it (download completes and checksums OK)? Randy On Sat, Dec 30, 2017 at 9:52 PM, Randy Evanswrote: > Todd, > > The minimal quick test would be greatly appreciated if you have the time. > I know this is above and beyond what I would expect but it would be really > useful to see if there is any hope for the GPIB on this unit. I will do > the service note 13 procedure in the morning. > > Thanks again, > > Randy > > On Sat, Dec 30, 2017 at 8:58 PM, Todd Micallef > wrote: > >> Randy, >> >> Short of swapping out the A5 digital board I have no other ideas. If you >> don't have a NI or Agilent adapter, I am sure I can find a spare for you >> to >> test. >> >> There is a Service Note 3458A-13A that deals with GPIB communication >> failures. There is a test within the document to see if a shorting jumper >> needs to be installed. >> I can try reducing the code to something minimal for a quick test if you >> want. It will take some time in the morning. >> >> Todd >> >> On Sat, Dec 30, 2017 at 11:18 PM, Randy Evans >> wrote: >> >> > Todd, >> > >> > Same as before. Still not working. All it does is print out "HP3458A" >> on >> > the output console and then just sits there. >> > >> > Randy >> > >> > On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallef >> > wrote: >> > >> > > Randy, >> > > >> > > I just remembered that I changed the duration from 30 minutes to 2 >> > minutes >> > > to speed up the test. I forgot to change it back. >> > > >> > > Todd >> > > >> > > On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef >> > > wrote: >> > > >> > > > Randy, >> > > > >> > > > I just sent it. I haven't modified Martin's code yet. That can be >> done >> > > > once you confirm it works for you. >> > > > >> > > > Todd >> > > > >> > > > On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 < >> > > randyevans2...@gmail.com >> > > > > wrote: >> > > > >> > > >> Todd, >> > > >> If you don't mind, just send the code directly to me. >> > > >> Randy >> > > >> >> > > >> Sent from my Verizon, Samsung Galaxy smartphone >> > > >> Original message From: Todd Micallef < >> > > >> tmical...@gmail.com> Date: 12/30/17 6:40 PM (GMT-08:00) To: >> > > Discussion >> > > >> of precise voltage measurement Subject: Re: >> > > >> [volt-nuts] GPIB Issue with 3458A >> > > >> Randy, >> > > >> >> > > >> I have not written anything else for the 3458A using EzGPIB. If >> you >> > > have >> > > >> a >> > > >> National Instruments or Agilent adapters, I have a handful of >> Python >> > > code >> > > >> I >> > > >> use to measure some of my standards. >> > > >> >> > > >> If you are reading the temperature from the HP meter, I would think >> > that >> > > >> the problem lies in how the triggering is configured. The >> triggering >> > has >> > > >> been changed in this script so it may work for you. There was an >> > > >> additional >> > > >> command added for disabling the buffer. >> > > >> The code can be posted here, if possible or I can email it to you >> for >> > > >> testing. >> > > >> >> > > >> Todd >> > > >> >> > > >> On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 < >> > > randyevans2...@gmail.com >> > > >> > >> > > >> wrote: >> > > >> >> > > >> > Todd, thanks for the help. Where will you post the new code? >> > > >> > BTW, do you have any very simple code that is likely to work. I >> am >> > > >> > interested in seeing if anything other than KE5FX's memory dump >> > > program >> > > >> > will work on my HP3458A. >> > > >> > Thanks again, >> > > >> > Randy >> > > >> > >> > > >> > Sent from my Verizon, Samsung Galaxy smartphone >> > > >> > Original message From: Todd Micallef < >> > > >> tmical...@gmail.com> >> > > >> > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise >> > voltage >> > > >> > measurement Subject: Re: [volt-nuts] GPIB >> > Issue >> > > >> with >> > > >> > 3458A >> > > >> > Randy, >> > > >> > >> > > >> > I found the Prologix adapters and have modified the code in the >> > > original >> > > >> > script. I am doing a test run now and will post the script when >> I am >> > > >> > certain I didn't make it worse. >> > > >> > >> > > >> > Todd >> > > >> > >> > > >> > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans < >> > > randyevans2...@gmail.com> >> > > >> > wrote: >> > > >> > >> > > >> > > Todd, >> > > >> > > >> > > >> > > I have swapped cables, moved the Prologix adapter to both >> meters >> > > >> (Agilent >> > > >> > > works, HP doesn't), tried Martins code as well as Xdevs's. I >> have >> > > not >> > > >> > > tried another PC yet, maybe the next step. >> > > >> > > >> > > >> > > Here is the code that Martin modified for two meters, as well >> as >> > the >> > > >> > > modified Xdev's code per your directions (V1.02). >> > > >> > >
Re: [volt-nuts] GPIB Issue with 3458A
Todd, The minimal quick test would be greatly appreciated if you have the time. I know this is above and beyond what I would expect but it would be really useful to see if there is any hope for the GPIB on this unit. I will do the service note 13 procedure in the morning. Thanks again, Randy On Sat, Dec 30, 2017 at 8:58 PM, Todd Micallefwrote: > Randy, > > Short of swapping out the A5 digital board I have no other ideas. If you > don't have a NI or Agilent adapter, I am sure I can find a spare for you to > test. > > There is a Service Note 3458A-13A that deals with GPIB communication > failures. There is a test within the document to see if a shorting jumper > needs to be installed. > I can try reducing the code to something minimal for a quick test if you > want. It will take some time in the morning. > > Todd > > On Sat, Dec 30, 2017 at 11:18 PM, Randy Evans > wrote: > > > Todd, > > > > Same as before. Still not working. All it does is print out "HP3458A" > on > > the output console and then just sits there. > > > > Randy > > > > On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallef > > wrote: > > > > > Randy, > > > > > > I just remembered that I changed the duration from 30 minutes to 2 > > minutes > > > to speed up the test. I forgot to change it back. > > > > > > Todd > > > > > > On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef > > > wrote: > > > > > > > Randy, > > > > > > > > I just sent it. I haven't modified Martin's code yet. That can be > done > > > > once you confirm it works for you. > > > > > > > > Todd > > > > > > > > On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 < > > > randyevans2...@gmail.com > > > > > wrote: > > > > > > > >> Todd, > > > >> If you don't mind, just send the code directly to me. > > > >> Randy > > > >> > > > >> Sent from my Verizon, Samsung Galaxy smartphone > > > >> Original message From: Todd Micallef < > > > >> tmical...@gmail.com> Date: 12/30/17 6:40 PM (GMT-08:00) To: > > > Discussion > > > >> of precise voltage measurement Subject: Re: > > > >> [volt-nuts] GPIB Issue with 3458A > > > >> Randy, > > > >> > > > >> I have not written anything else for the 3458A using EzGPIB. If you > > > have > > > >> a > > > >> National Instruments or Agilent adapters, I have a handful of Python > > > code > > > >> I > > > >> use to measure some of my standards. > > > >> > > > >> If you are reading the temperature from the HP meter, I would think > > that > > > >> the problem lies in how the triggering is configured. The triggering > > has > > > >> been changed in this script so it may work for you. There was an > > > >> additional > > > >> command added for disabling the buffer. > > > >> The code can be posted here, if possible or I can email it to you > for > > > >> testing. > > > >> > > > >> Todd > > > >> > > > >> On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 < > > > randyevans2...@gmail.com > > > >> > > > > >> wrote: > > > >> > > > >> > Todd, thanks for the help. Where will you post the new code? > > > >> > BTW, do you have any very simple code that is likely to work. I > am > > > >> > interested in seeing if anything other than KE5FX's memory dump > > > program > > > >> > will work on my HP3458A. > > > >> > Thanks again, > > > >> > Randy > > > >> > > > > >> > Sent from my Verizon, Samsung Galaxy smartphone > > > >> > Original message From: Todd Micallef < > > > >> tmical...@gmail.com> > > > >> > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise > > voltage > > > >> > measurement Subject: Re: [volt-nuts] GPIB > > Issue > > > >> with > > > >> > 3458A > > > >> > Randy, > > > >> > > > > >> > I found the Prologix adapters and have modified the code in the > > > original > > > >> > script. I am doing a test run now and will post the script when I > am > > > >> > certain I didn't make it worse. > > > >> > > > > >> > Todd > > > >> > > > > >> > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans < > > > randyevans2...@gmail.com> > > > >> > wrote: > > > >> > > > > >> > > Todd, > > > >> > > > > > >> > > I have swapped cables, moved the Prologix adapter to both meters > > > >> (Agilent > > > >> > > works, HP doesn't), tried Martins code as well as Xdevs's. I > have > > > not > > > >> > > tried another PC yet, maybe the next step. > > > >> > > > > > >> > > Here is the code that Martin modified for two meters, as well as > > the > > > >> > > modified Xdev's code per your directions (V1.02). > > > >> > > > > > >> > > Thanks, > > > >> > > > > > >> > > Randy > > > >> > > > > > >> > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef < > > tmical...@gmail.com > > > > > > > >> > > wrote: > > > >> > > > > > >> > > > Randy, > > > >> > > > > > > >> > > > I will continue to look for my adapter. I haven't used it for > > > quite > > > >> a > > > >> > > > while. > > > >> > > > > > > >> > > > I have seen similar issues and
Re: [volt-nuts] GPIB Issue with 3458A
Randy, Short of swapping out the A5 digital board I have no other ideas. If you don't have a NI or Agilent adapter, I am sure I can find a spare for you to test. There is a Service Note 3458A-13A that deals with GPIB communication failures. There is a test within the document to see if a shorting jumper needs to be installed. I can try reducing the code to something minimal for a quick test if you want. It will take some time in the morning. Todd On Sat, Dec 30, 2017 at 11:18 PM, Randy Evanswrote: > Todd, > > Same as before. Still not working. All it does is print out "HP3458A" on > the output console and then just sits there. > > Randy > > On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallef > wrote: > > > Randy, > > > > I just remembered that I changed the duration from 30 minutes to 2 > minutes > > to speed up the test. I forgot to change it back. > > > > Todd > > > > On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef > > wrote: > > > > > Randy, > > > > > > I just sent it. I haven't modified Martin's code yet. That can be done > > > once you confirm it works for you. > > > > > > Todd > > > > > > On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 < > > randyevans2...@gmail.com > > > > wrote: > > > > > >> Todd, > > >> If you don't mind, just send the code directly to me. > > >> Randy > > >> > > >> Sent from my Verizon, Samsung Galaxy smartphone > > >> Original message From: Todd Micallef < > > >> tmical...@gmail.com> Date: 12/30/17 6:40 PM (GMT-08:00) To: > > Discussion > > >> of precise voltage measurement Subject: Re: > > >> [volt-nuts] GPIB Issue with 3458A > > >> Randy, > > >> > > >> I have not written anything else for the 3458A using EzGPIB. If you > > have > > >> a > > >> National Instruments or Agilent adapters, I have a handful of Python > > code > > >> I > > >> use to measure some of my standards. > > >> > > >> If you are reading the temperature from the HP meter, I would think > that > > >> the problem lies in how the triggering is configured. The triggering > has > > >> been changed in this script so it may work for you. There was an > > >> additional > > >> command added for disabling the buffer. > > >> The code can be posted here, if possible or I can email it to you for > > >> testing. > > >> > > >> Todd > > >> > > >> On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 < > > randyevans2...@gmail.com > > >> > > > >> wrote: > > >> > > >> > Todd, thanks for the help. Where will you post the new code? > > >> > BTW, do you have any very simple code that is likely to work. I am > > >> > interested in seeing if anything other than KE5FX's memory dump > > program > > >> > will work on my HP3458A. > > >> > Thanks again, > > >> > Randy > > >> > > > >> > Sent from my Verizon, Samsung Galaxy smartphone > > >> > Original message From: Todd Micallef < > > >> tmical...@gmail.com> > > >> > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise > voltage > > >> > measurement Subject: Re: [volt-nuts] GPIB > Issue > > >> with > > >> > 3458A > > >> > Randy, > > >> > > > >> > I found the Prologix adapters and have modified the code in the > > original > > >> > script. I am doing a test run now and will post the script when I am > > >> > certain I didn't make it worse. > > >> > > > >> > Todd > > >> > > > >> > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans < > > randyevans2...@gmail.com> > > >> > wrote: > > >> > > > >> > > Todd, > > >> > > > > >> > > I have swapped cables, moved the Prologix adapter to both meters > > >> (Agilent > > >> > > works, HP doesn't), tried Martins code as well as Xdevs's. I have > > not > > >> > > tried another PC yet, maybe the next step. > > >> > > > > >> > > Here is the code that Martin modified for two meters, as well as > the > > >> > > modified Xdev's code per your directions (V1.02). > > >> > > > > >> > > Thanks, > > >> > > > > >> > > Randy > > >> > > > > >> > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef < > tmical...@gmail.com > > > > > >> > > wrote: > > >> > > > > >> > > > Randy, > > >> > > > > > >> > > > I will continue to look for my adapter. I haven't used it for > > quite > > >> a > > >> > > > while. > > >> > > > > > >> > > > I have seen similar issues and swapped cables, moved the adapter > > to > > >> the > > >> > > > other meter, etc... One of my cables was bad from one side of > the > > >> > > connector > > >> > > > to the other (on the same end of the cable) and caused me lots > of > > >> > wasted > > >> > > > time. Weird reading errors were happening. > > >> > > > > > >> > > > Would you email me Martin's code so I can test it here and I > will > > >> see > > >> > if > > >> > > > anything can be done to replicate your issue. > > >> > > > > > >> > > > Todd > > >> > > > > > >> > > > > > >> > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans < > > >> randyevans2...@gmail.com > > >> > > > > >> > > > wrote: > > >> > > >
Re: [volt-nuts] GPIB Issue with 3458A
Todd, Same as before. Still not working. All it does is print out "HP3458A" on the output console and then just sits there. Randy On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallefwrote: > Randy, > > I just remembered that I changed the duration from 30 minutes to 2 minutes > to speed up the test. I forgot to change it back. > > Todd > > On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef > wrote: > > > Randy, > > > > I just sent it. I haven't modified Martin's code yet. That can be done > > once you confirm it works for you. > > > > Todd > > > > On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 < > randyevans2...@gmail.com > > > wrote: > > > >> Todd, > >> If you don't mind, just send the code directly to me. > >> Randy > >> > >> Sent from my Verizon, Samsung Galaxy smartphone > >> Original message From: Todd Micallef < > >> tmical...@gmail.com> Date: 12/30/17 6:40 PM (GMT-08:00) To: > Discussion > >> of precise voltage measurement Subject: Re: > >> [volt-nuts] GPIB Issue with 3458A > >> Randy, > >> > >> I have not written anything else for the 3458A using EzGPIB. If you > have > >> a > >> National Instruments or Agilent adapters, I have a handful of Python > code > >> I > >> use to measure some of my standards. > >> > >> If you are reading the temperature from the HP meter, I would think that > >> the problem lies in how the triggering is configured. The triggering has > >> been changed in this script so it may work for you. There was an > >> additional > >> command added for disabling the buffer. > >> The code can be posted here, if possible or I can email it to you for > >> testing. > >> > >> Todd > >> > >> On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 < > randyevans2...@gmail.com > >> > > >> wrote: > >> > >> > Todd, thanks for the help. Where will you post the new code? > >> > BTW, do you have any very simple code that is likely to work. I am > >> > interested in seeing if anything other than KE5FX's memory dump > program > >> > will work on my HP3458A. > >> > Thanks again, > >> > Randy > >> > > >> > Sent from my Verizon, Samsung Galaxy smartphone > >> > Original message From: Todd Micallef < > >> tmical...@gmail.com> > >> > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise voltage > >> > measurement Subject: Re: [volt-nuts] GPIB Issue > >> with > >> > 3458A > >> > Randy, > >> > > >> > I found the Prologix adapters and have modified the code in the > original > >> > script. I am doing a test run now and will post the script when I am > >> > certain I didn't make it worse. > >> > > >> > Todd > >> > > >> > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans < > randyevans2...@gmail.com> > >> > wrote: > >> > > >> > > Todd, > >> > > > >> > > I have swapped cables, moved the Prologix adapter to both meters > >> (Agilent > >> > > works, HP doesn't), tried Martins code as well as Xdevs's. I have > not > >> > > tried another PC yet, maybe the next step. > >> > > > >> > > Here is the code that Martin modified for two meters, as well as the > >> > > modified Xdev's code per your directions (V1.02). > >> > > > >> > > Thanks, > >> > > > >> > > Randy > >> > > > >> > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef > > >> > > wrote: > >> > > > >> > > > Randy, > >> > > > > >> > > > I will continue to look for my adapter. I haven't used it for > quite > >> a > >> > > > while. > >> > > > > >> > > > I have seen similar issues and swapped cables, moved the adapter > to > >> the > >> > > > other meter, etc... One of my cables was bad from one side of the > >> > > connector > >> > > > to the other (on the same end of the cable) and caused me lots of > >> > wasted > >> > > > time. Weird reading errors were happening. > >> > > > > >> > > > Would you email me Martin's code so I can test it here and I will > >> see > >> > if > >> > > > anything can be done to replicate your issue. > >> > > > > >> > > > Todd > >> > > > > >> > > > > >> > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans < > >> randyevans2...@gmail.com > >> > > > >> > > > wrote: > >> > > > > >> > > > > Todd, > >> > > > > > >> > > > > I tried adding the line but it made no difference. > >> > > > > > >> > > > > Randy > >> > > > > > >> > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans < > >> > > randyevans2...@gmail.com> > >> > > > > wrote: > >> > > > > > >> > > > > > Todd, > >> > > > > > > >> > > > > > Does it matter where I put the line in the Init_Device > >> procedure? > >> > > > > > > >> > > > > > > >> > > > > > procedure Init_Device; > >> > > > > > begin; > >> > > > > >EZGPIB_BusWriteData(gpib_address,GET_ID);// > Get > >> ID > >> > of > >> > > > > meter > >> > > > > >if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) > then > >> > > > > > EZGPIB_ScreenWriteLn(meter_id); > >> > > > > > > >> > > > > >EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);// > Hold > >> > > >
Re: [volt-nuts] GPIB Issue with 3458A
Randy, I just remembered that I changed the duration from 30 minutes to 2 minutes to speed up the test. I forgot to change it back. Todd On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallefwrote: > Randy, > > I just sent it. I haven't modified Martin's code yet. That can be done > once you confirm it works for you. > > Todd > > On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 > wrote: > >> Todd, >> If you don't mind, just send the code directly to me. >> Randy >> >> Sent from my Verizon, Samsung Galaxy smartphone >> Original message From: Todd Micallef < >> tmical...@gmail.com> Date: 12/30/17 6:40 PM (GMT-08:00) To: Discussion >> of precise voltage measurement Subject: Re: >> [volt-nuts] GPIB Issue with 3458A >> Randy, >> >> I have not written anything else for the 3458A using EzGPIB. If you have >> a >> National Instruments or Agilent adapters, I have a handful of Python code >> I >> use to measure some of my standards. >> >> If you are reading the temperature from the HP meter, I would think that >> the problem lies in how the triggering is configured. The triggering has >> been changed in this script so it may work for you. There was an >> additional >> command added for disabling the buffer. >> The code can be posted here, if possible or I can email it to you for >> testing. >> >> Todd >> >> On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 > > >> wrote: >> >> > Todd, thanks for the help. Where will you post the new code? >> > BTW, do you have any very simple code that is likely to work. I am >> > interested in seeing if anything other than KE5FX's memory dump program >> > will work on my HP3458A. >> > Thanks again, >> > Randy >> > >> > Sent from my Verizon, Samsung Galaxy smartphone >> > Original message From: Todd Micallef < >> tmical...@gmail.com> >> > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise voltage >> > measurement Subject: Re: [volt-nuts] GPIB Issue >> with >> > 3458A >> > Randy, >> > >> > I found the Prologix adapters and have modified the code in the original >> > script. I am doing a test run now and will post the script when I am >> > certain I didn't make it worse. >> > >> > Todd >> > >> > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans >> > wrote: >> > >> > > Todd, >> > > >> > > I have swapped cables, moved the Prologix adapter to both meters >> (Agilent >> > > works, HP doesn't), tried Martins code as well as Xdevs's. I have not >> > > tried another PC yet, maybe the next step. >> > > >> > > Here is the code that Martin modified for two meters, as well as the >> > > modified Xdev's code per your directions (V1.02). >> > > >> > > Thanks, >> > > >> > > Randy >> > > >> > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef >> > > wrote: >> > > >> > > > Randy, >> > > > >> > > > I will continue to look for my adapter. I haven't used it for quite >> a >> > > > while. >> > > > >> > > > I have seen similar issues and swapped cables, moved the adapter to >> the >> > > > other meter, etc... One of my cables was bad from one side of the >> > > connector >> > > > to the other (on the same end of the cable) and caused me lots of >> > wasted >> > > > time. Weird reading errors were happening. >> > > > >> > > > Would you email me Martin's code so I can test it here and I will >> see >> > if >> > > > anything can be done to replicate your issue. >> > > > >> > > > Todd >> > > > >> > > > >> > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans < >> randyevans2...@gmail.com >> > > >> > > > wrote: >> > > > >> > > > > Todd, >> > > > > >> > > > > I tried adding the line but it made no difference. >> > > > > >> > > > > Randy >> > > > > >> > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans < >> > > randyevans2...@gmail.com> >> > > > > wrote: >> > > > > >> > > > > > Todd, >> > > > > > >> > > > > > Does it matter where I put the line in the Init_Device >> procedure? >> > > > > > >> > > > > > >> > > > > > procedure Init_Device; >> > > > > > begin; >> > > > > >EZGPIB_BusWriteData(gpib_address,GET_ID);// Get >> ID >> > of >> > > > > meter >> > > > > >if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then >> > > > > > EZGPIB_ScreenWriteLn(meter_id); >> > > > > > >> > > > > >EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);// Hold >> > > > trigger >> > > > > > until meter can be setup >> > > > > >EZGPIB_BusWriteData(gpib_address,'INBUF ON'); >> > > > > >EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); >> > > > > >EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); >> > > > > > end; >> > > > > > >> > > > > > >> > > > > > Randy >> > > > > > >> > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef < >> > tmical...@gmail.com >> > > > >> > > > > > wrote: >> > > > > > >> > > > > >> Randy, >> > > > > >> >> > > > > >> I am the person that wrote that EzGPIB script. It
Re: [volt-nuts] GPIB Issue with 3458A
Randy, I just sent it. I haven't modified Martin's code yet. That can be done once you confirm it works for you. Todd On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688wrote: > Todd, > If you don't mind, just send the code directly to me. > Randy > > Sent from my Verizon, Samsung Galaxy smartphone > Original message From: Todd Micallef > Date: 12/30/17 6:40 PM (GMT-08:00) To: Discussion of precise voltage > measurement Subject: Re: [volt-nuts] GPIB Issue with > 3458A > Randy, > > I have not written anything else for the 3458A using EzGPIB. If you have a > National Instruments or Agilent adapters, I have a handful of Python code I > use to measure some of my standards. > > If you are reading the temperature from the HP meter, I would think that > the problem lies in how the triggering is configured. The triggering has > been changed in this script so it may work for you. There was an additional > command added for disabling the buffer. > The code can be posted here, if possible or I can email it to you for > testing. > > Todd > > On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 > wrote: > > > Todd, thanks for the help. Where will you post the new code? > > BTW, do you have any very simple code that is likely to work. I am > > interested in seeing if anything other than KE5FX's memory dump program > > will work on my HP3458A. > > Thanks again, > > Randy > > > > Sent from my Verizon, Samsung Galaxy smartphone > > Original message From: Todd Micallef < > tmical...@gmail.com> > > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise voltage > > measurement Subject: Re: [volt-nuts] GPIB Issue > with > > 3458A > > Randy, > > > > I found the Prologix adapters and have modified the code in the original > > script. I am doing a test run now and will post the script when I am > > certain I didn't make it worse. > > > > Todd > > > > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans > > wrote: > > > > > Todd, > > > > > > I have swapped cables, moved the Prologix adapter to both meters > (Agilent > > > works, HP doesn't), tried Martins code as well as Xdevs's. I have not > > > tried another PC yet, maybe the next step. > > > > > > Here is the code that Martin modified for two meters, as well as the > > > modified Xdev's code per your directions (V1.02). > > > > > > Thanks, > > > > > > Randy > > > > > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef > > > wrote: > > > > > > > Randy, > > > > > > > > I will continue to look for my adapter. I haven't used it for quite a > > > > while. > > > > > > > > I have seen similar issues and swapped cables, moved the adapter to > the > > > > other meter, etc... One of my cables was bad from one side of the > > > connector > > > > to the other (on the same end of the cable) and caused me lots of > > wasted > > > > time. Weird reading errors were happening. > > > > > > > > Would you email me Martin's code so I can test it here and I will see > > if > > > > anything can be done to replicate your issue. > > > > > > > > Todd > > > > > > > > > > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans < > randyevans2...@gmail.com > > > > > > > wrote: > > > > > > > > > Todd, > > > > > > > > > > I tried adding the line but it made no difference. > > > > > > > > > > Randy > > > > > > > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans < > > > randyevans2...@gmail.com> > > > > > wrote: > > > > > > > > > > > Todd, > > > > > > > > > > > > Does it matter where I put the line in the Init_Device procedure? > > > > > > > > > > > > > > > > > > procedure Init_Device; > > > > > > begin; > > > > > >EZGPIB_BusWriteData(gpib_address,GET_ID);// Get > ID > > of > > > > > meter > > > > > >if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then > > > > > > EZGPIB_ScreenWriteLn(meter_id); > > > > > > > > > > > >EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);// Hold > > > > trigger > > > > > > until meter can be setup > > > > > >EZGPIB_BusWriteData(gpib_address,'INBUF ON'); > > > > > >EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); > > > > > >EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); > > > > > > end; > > > > > > > > > > > > > > > > > > Randy > > > > > > > > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef < > > tmical...@gmail.com > > > > > > > > > > wrote: > > > > > > > > > > > >> Randy, > > > > > >> > > > > > >> I am the person that wrote that EzGPIB script. It was my first > > > before > > > > I > > > > > >> moved on to Python and I have only written a couple for the > > > different > > > > > >> meters I have on hand. Version 1.01 is the most current version > I > > > have > > > > > for > > > > > >> the 3458A. > > > > > >> I see that I should have added at least one more command to the > > > meter > > > > > >> initialization. The additional init
Re: [volt-nuts] GPIB Issue with 3458A
Todd, If you don't mind, just send the code directly to me. Randy Sent from my Verizon, Samsung Galaxy smartphone Original message From: Todd MicallefDate: 12/30/17 6:40 PM (GMT-08:00) To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] GPIB Issue with 3458A Randy, I have not written anything else for the 3458A using EzGPIB. If you have a National Instruments or Agilent adapters, I have a handful of Python code I use to measure some of my standards. If you are reading the temperature from the HP meter, I would think that the problem lies in how the triggering is configured. The triggering has been changed in this script so it may work for you. There was an additional command added for disabling the buffer. The code can be posted here, if possible or I can email it to you for testing. Todd On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 wrote: > Todd, thanks for the help. Where will you post the new code? > BTW, do you have any very simple code that is likely to work. I am > interested in seeing if anything other than KE5FX's memory dump program > will work on my HP3458A. > Thanks again, > Randy > > Sent from my Verizon, Samsung Galaxy smartphone > Original message From: Todd Micallef > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise voltage > measurement Subject: Re: [volt-nuts] GPIB Issue with > 3458A > Randy, > > I found the Prologix adapters and have modified the code in the original > script. I am doing a test run now and will post the script when I am > certain I didn't make it worse. > > Todd > > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans > wrote: > > > Todd, > > > > I have swapped cables, moved the Prologix adapter to both meters (Agilent > > works, HP doesn't), tried Martins code as well as Xdevs's. I have not > > tried another PC yet, maybe the next step. > > > > Here is the code that Martin modified for two meters, as well as the > > modified Xdev's code per your directions (V1.02). > > > > Thanks, > > > > Randy > > > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef > > wrote: > > > > > Randy, > > > > > > I will continue to look for my adapter. I haven't used it for quite a > > > while. > > > > > > I have seen similar issues and swapped cables, moved the adapter to the > > > other meter, etc... One of my cables was bad from one side of the > > connector > > > to the other (on the same end of the cable) and caused me lots of > wasted > > > time. Weird reading errors were happening. > > > > > > Would you email me Martin's code so I can test it here and I will see > if > > > anything can be done to replicate your issue. > > > > > > Todd > > > > > > > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans > > > > wrote: > > > > > > > Todd, > > > > > > > > I tried adding the line but it made no difference. > > > > > > > > Randy > > > > > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans < > > randyevans2...@gmail.com> > > > > wrote: > > > > > > > > > Todd, > > > > > > > > > > Does it matter where I put the line in the Init_Device procedure? > > > > > > > > > > > > > > > procedure Init_Device; > > > > > begin; > > > > > EZGPIB_BusWriteData(gpib_address,GET_ID); // Get ID > of > > > > meter > > > > > if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then > > > > > EZGPIB_ScreenWriteLn(meter_id); > > > > > > > > > > EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE); // Hold > > > trigger > > > > > until meter can be setup > > > > > EZGPIB_BusWriteData(gpib_address,'INBUF ON'); > > > > > EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); > > > > > EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); > > > > > end; > > > > > > > > > > > > > > > Randy > > > > > > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef < > tmical...@gmail.com > > > > > > > > wrote: > > > > > > > > > >> Randy, > > > > >> > > > > >> I am the person that wrote that EzGPIB script. It was my first > > before > > > I > > > > >> moved on to Python and I have only written a couple for the > > different > > > > >> meters I have on hand. Version 1.01 is the most current version I > > have > > > > for > > > > >> the 3458A. > > > > >> I see that I should have added at least one more command to the > > meter > > > > >> initialization. The additional init command is used in the > scripts > > on > > > > >> xDevs. > > > > >> > > > > >> Add the following line in the Init_Device procedure > > > > >> EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); > > > > >> > > > > >> This tells the meter to send an EOI on every read. > > > > >> > > > > >> I am looking for one of my Prologix adapters or I would test this. > > > When > > > > I > > > > >> find it, I will also test it. I want to make sure there are no > > further > > > > >> issues. > > > > >>
Re: [volt-nuts] GPIB Issue with 3458A
Randy, I have not written anything else for the 3458A using EzGPIB. If you have a National Instruments or Agilent adapters, I have a handful of Python code I use to measure some of my standards. If you are reading the temperature from the HP meter, I would think that the problem lies in how the triggering is configured. The triggering has been changed in this script so it may work for you. There was an additional command added for disabling the buffer. The code can be posted here, if possible or I can email it to you for testing. Todd On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688wrote: > Todd, thanks for the help. Where will you post the new code? > BTW, do you have any very simple code that is likely to work. I am > interested in seeing if anything other than KE5FX's memory dump program > will work on my HP3458A. > Thanks again, > Randy > > Sent from my Verizon, Samsung Galaxy smartphone > Original message From: Todd Micallef > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise voltage > measurement Subject: Re: [volt-nuts] GPIB Issue with > 3458A > Randy, > > I found the Prologix adapters and have modified the code in the original > script. I am doing a test run now and will post the script when I am > certain I didn't make it worse. > > Todd > > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans > wrote: > > > Todd, > > > > I have swapped cables, moved the Prologix adapter to both meters (Agilent > > works, HP doesn't), tried Martins code as well as Xdevs's. I have not > > tried another PC yet, maybe the next step. > > > > Here is the code that Martin modified for two meters, as well as the > > modified Xdev's code per your directions (V1.02). > > > > Thanks, > > > > Randy > > > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef > > wrote: > > > > > Randy, > > > > > > I will continue to look for my adapter. I haven't used it for quite a > > > while. > > > > > > I have seen similar issues and swapped cables, moved the adapter to the > > > other meter, etc... One of my cables was bad from one side of the > > connector > > > to the other (on the same end of the cable) and caused me lots of > wasted > > > time. Weird reading errors were happening. > > > > > > Would you email me Martin's code so I can test it here and I will see > if > > > anything can be done to replicate your issue. > > > > > > Todd > > > > > > > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans > > > > wrote: > > > > > > > Todd, > > > > > > > > I tried adding the line but it made no difference. > > > > > > > > Randy > > > > > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans < > > randyevans2...@gmail.com> > > > > wrote: > > > > > > > > > Todd, > > > > > > > > > > Does it matter where I put the line in the Init_Device procedure? > > > > > > > > > > > > > > > procedure Init_Device; > > > > > begin; > > > > >EZGPIB_BusWriteData(gpib_address,GET_ID);// Get ID > of > > > > meter > > > > >if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then > > > > > EZGPIB_ScreenWriteLn(meter_id); > > > > > > > > > >EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);// Hold > > > trigger > > > > > until meter can be setup > > > > >EZGPIB_BusWriteData(gpib_address,'INBUF ON'); > > > > >EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); > > > > >EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); > > > > > end; > > > > > > > > > > > > > > > Randy > > > > > > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef < > tmical...@gmail.com > > > > > > > > wrote: > > > > > > > > > >> Randy, > > > > >> > > > > >> I am the person that wrote that EzGPIB script. It was my first > > before > > > I > > > > >> moved on to Python and I have only written a couple for the > > different > > > > >> meters I have on hand. Version 1.01 is the most current version I > > have > > > > for > > > > >> the 3458A. > > > > >> I see that I should have added at least one more command to the > > meter > > > > >> initialization. The additional init command is used in the > scripts > > on > > > > >> xDevs. > > > > >> > > > > >> Add the following line in the Init_Device procedure > > > > >>EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); > > > > >> > > > > >> This tells the meter to send an EOI on every read. > > > > >> > > > > >> I am looking for one of my Prologix adapters or I would test this. > > > When > > > > I > > > > >> find it, I will also test it. I want to make sure there are no > > further > > > > >> issues. > > > > >> Also, do both meters run the same firmware versions? I wonder if > the > > > HP > > > > >> meter is using a much older version. > > > > >> > > > > >> Todd > > > > >> > > > > >> On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans < > > > randyevans2...@gmail.com > > > > > > > > > >> wrote: > > > > >> > > > > >> > Todd, > > > > >> > > > > > >> > I have
Re: [volt-nuts] GPIB Issue with 3458A
Todd, thanks for the help. Where will you post the new code? BTW, do you have any very simple code that is likely to work. I am interested in seeing if anything other than KE5FX's memory dump program will work on my HP3458A. Thanks again, Randy Sent from my Verizon, Samsung Galaxy smartphone Original message From: Todd MicallefDate: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] GPIB Issue with 3458A Randy, I found the Prologix adapters and have modified the code in the original script. I am doing a test run now and will post the script when I am certain I didn't make it worse. Todd On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans wrote: > Todd, > > I have swapped cables, moved the Prologix adapter to both meters (Agilent > works, HP doesn't), tried Martins code as well as Xdevs's. I have not > tried another PC yet, maybe the next step. > > Here is the code that Martin modified for two meters, as well as the > modified Xdev's code per your directions (V1.02). > > Thanks, > > Randy > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef > wrote: > > > Randy, > > > > I will continue to look for my adapter. I haven't used it for quite a > > while. > > > > I have seen similar issues and swapped cables, moved the adapter to the > > other meter, etc... One of my cables was bad from one side of the > connector > > to the other (on the same end of the cable) and caused me lots of wasted > > time. Weird reading errors were happening. > > > > Would you email me Martin's code so I can test it here and I will see if > > anything can be done to replicate your issue. > > > > Todd > > > > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans > > wrote: > > > > > Todd, > > > > > > I tried adding the line but it made no difference. > > > > > > Randy > > > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans < > randyevans2...@gmail.com> > > > wrote: > > > > > > > Todd, > > > > > > > > Does it matter where I put the line in the Init_Device procedure? > > > > > > > > > > > > procedure Init_Device; > > > > begin; > > > > EZGPIB_BusWriteData(gpib_address,GET_ID); // Get ID of > > > meter > > > > if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then > > > > EZGPIB_ScreenWriteLn(meter_id); > > > > > > > > EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE); // Hold > > trigger > > > > until meter can be setup > > > > EZGPIB_BusWriteData(gpib_address,'INBUF ON'); > > > > EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); > > > > EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); > > > > end; > > > > > > > > > > > > Randy > > > > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef > > > > > wrote: > > > > > > > >> Randy, > > > >> > > > >> I am the person that wrote that EzGPIB script. It was my first > before > > I > > > >> moved on to Python and I have only written a couple for the > different > > > >> meters I have on hand. Version 1.01 is the most current version I > have > > > for > > > >> the 3458A. > > > >> I see that I should have added at least one more command to the > meter > > > >> initialization. The additional init command is used in the scripts > on > > > >> xDevs. > > > >> > > > >> Add the following line in the Init_Device procedure > > > >> EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); > > > >> > > > >> This tells the meter to send an EOI on every read. > > > >> > > > >> I am looking for one of my Prologix adapters or I would test this. > > When > > > I > > > >> find it, I will also test it. I want to make sure there are no > further > > > >> issues. > > > >> Also, do both meters run the same firmware versions? I wonder if the > > HP > > > >> meter is using a much older version. > > > >> > > > >> Todd > > > >> > > > >> On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans < > > randyevans2...@gmail.com > > > > > > > >> wrote: > > > >> > > > >> > Todd, > > > >> > > > > >> > I have version 1.01. Yes, please e-mail me your modified version. > > > >> Maybe it > > > >> > will help. > > > >> > > > > >> > Thanks, > > > >> > > > > >> > Randy > > > >> > > > > >> > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef < > tmical...@gmail.com > > > > > > >> > wrote: > > > >> > > > > >> > > Randy, > > > >> > > > > > >> > > Is there a version number at the top of the file? I had two > > versions > > > >> > (1.00 > > > >> > > and 1.01). I can't remember all that changed between the two > since > > > it > > > >> has > > > >> > > been a while but I did find an additional timeout code that > > limited > > > >> the > > > >> > > wait time for reading data to 1 minute. I am guessing I had a > > > problem > > > >> > with > > > >> > > mine and added it. > > > >> > > > > > >> > > I can email it to you if you want to try it. If it works for > you, > > I > > > >> will > > > >> > > have Illya update it on
Re: [volt-nuts] GPIB Issue with 3458A
Randy, I found the Prologix adapters and have modified the code in the original script. I am doing a test run now and will post the script when I am certain I didn't make it worse. Todd On Sat, Dec 30, 2017 at 7:15 PM, Randy Evanswrote: > Todd, > > I have swapped cables, moved the Prologix adapter to both meters (Agilent > works, HP doesn't), tried Martins code as well as Xdevs's. I have not > tried another PC yet, maybe the next step. > > Here is the code that Martin modified for two meters, as well as the > modified Xdev's code per your directions (V1.02). > > Thanks, > > Randy > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef > wrote: > > > Randy, > > > > I will continue to look for my adapter. I haven't used it for quite a > > while. > > > > I have seen similar issues and swapped cables, moved the adapter to the > > other meter, etc... One of my cables was bad from one side of the > connector > > to the other (on the same end of the cable) and caused me lots of wasted > > time. Weird reading errors were happening. > > > > Would you email me Martin's code so I can test it here and I will see if > > anything can be done to replicate your issue. > > > > Todd > > > > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans > > wrote: > > > > > Todd, > > > > > > I tried adding the line but it made no difference. > > > > > > Randy > > > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans < > randyevans2...@gmail.com> > > > wrote: > > > > > > > Todd, > > > > > > > > Does it matter where I put the line in the Init_Device procedure? > > > > > > > > > > > > procedure Init_Device; > > > > begin; > > > >EZGPIB_BusWriteData(gpib_address,GET_ID);// Get ID of > > > meter > > > >if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then > > > > EZGPIB_ScreenWriteLn(meter_id); > > > > > > > >EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);// Hold > > trigger > > > > until meter can be setup > > > >EZGPIB_BusWriteData(gpib_address,'INBUF ON'); > > > >EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); > > > >EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); > > > > end; > > > > > > > > > > > > Randy > > > > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef > > > > > wrote: > > > > > > > >> Randy, > > > >> > > > >> I am the person that wrote that EzGPIB script. It was my first > before > > I > > > >> moved on to Python and I have only written a couple for the > different > > > >> meters I have on hand. Version 1.01 is the most current version I > have > > > for > > > >> the 3458A. > > > >> I see that I should have added at least one more command to the > meter > > > >> initialization. The additional init command is used in the scripts > on > > > >> xDevs. > > > >> > > > >> Add the following line in the Init_Device procedure > > > >>EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); > > > >> > > > >> This tells the meter to send an EOI on every read. > > > >> > > > >> I am looking for one of my Prologix adapters or I would test this. > > When > > > I > > > >> find it, I will also test it. I want to make sure there are no > further > > > >> issues. > > > >> Also, do both meters run the same firmware versions? I wonder if the > > HP > > > >> meter is using a much older version. > > > >> > > > >> Todd > > > >> > > > >> On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans < > > randyevans2...@gmail.com > > > > > > > >> wrote: > > > >> > > > >> > Todd, > > > >> > > > > >> > I have version 1.01. Yes, please e-mail me your modified version. > > > >> Maybe it > > > >> > will help. > > > >> > > > > >> > Thanks, > > > >> > > > > >> > Randy > > > >> > > > > >> > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef < > tmical...@gmail.com > > > > > > >> > wrote: > > > >> > > > > >> > > Randy, > > > >> > > > > > >> > > Is there a version number at the top of the file? I had two > > versions > > > >> > (1.00 > > > >> > > and 1.01). I can't remember all that changed between the two > since > > > it > > > >> has > > > >> > > been a while but I did find an additional timeout code that > > limited > > > >> the > > > >> > > wait time for reading data to 1 minute. I am guessing I had a > > > problem > > > >> > with > > > >> > > mine and added it. > > > >> > > > > > >> > > I can email it to you if you want to try it. If it works for > you, > > I > > > >> will > > > >> > > have Illya update it on xDevs. > > > >> > > > > > >> > > Todd > > > >> > > > > > >> > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans < > > > >> randyevans2...@gmail.com> > > > >> > > wrote: > > > >> > > > > > >> > > > I am still unable to get my HP-3458A to work with the xdevs > > > program > > > >> > > > "HP3458A_Measure_DMM_Noise.488" ( > https://xdevs.com/article/dmm > > > >> _noise/ > > > >> > ). > > > >> > > > It > > > >> > > > works fine with my Agilent 3458A and with a friend's HP-3458A > > and > > > >> his > > > >> > > > Agilent 3458A
Re: [volt-nuts] GPIB Issue with 3458A
Todd, I have swapped cables, moved the Prologix adapter to both meters (Agilent works, HP doesn't), tried Martins code as well as Xdevs's. I have not tried another PC yet, maybe the next step. Here is the code that Martin modified for two meters, as well as the modified Xdev's code per your directions (V1.02). Thanks, Randy On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallefwrote: > Randy, > > I will continue to look for my adapter. I haven't used it for quite a > while. > > I have seen similar issues and swapped cables, moved the adapter to the > other meter, etc... One of my cables was bad from one side of the connector > to the other (on the same end of the cable) and caused me lots of wasted > time. Weird reading errors were happening. > > Would you email me Martin's code so I can test it here and I will see if > anything can be done to replicate your issue. > > Todd > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans > wrote: > > > Todd, > > > > I tried adding the line but it made no difference. > > > > Randy > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans > > wrote: > > > > > Todd, > > > > > > Does it matter where I put the line in the Init_Device procedure? > > > > > > > > > procedure Init_Device; > > > begin; > > >EZGPIB_BusWriteData(gpib_address,GET_ID);// Get ID of > > meter > > >if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then > > > EZGPIB_ScreenWriteLn(meter_id); > > > > > >EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);// Hold > trigger > > > until meter can be setup > > >EZGPIB_BusWriteData(gpib_address,'INBUF ON'); > > >EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); > > >EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); > > > end; > > > > > > > > > Randy > > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef > > > wrote: > > > > > >> Randy, > > >> > > >> I am the person that wrote that EzGPIB script. It was my first before > I > > >> moved on to Python and I have only written a couple for the different > > >> meters I have on hand. Version 1.01 is the most current version I have > > for > > >> the 3458A. > > >> I see that I should have added at least one more command to the meter > > >> initialization. The additional init command is used in the scripts on > > >> xDevs. > > >> > > >> Add the following line in the Init_Device procedure > > >>EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); > > >> > > >> This tells the meter to send an EOI on every read. > > >> > > >> I am looking for one of my Prologix adapters or I would test this. > When > > I > > >> find it, I will also test it. I want to make sure there are no further > > >> issues. > > >> Also, do both meters run the same firmware versions? I wonder if the > HP > > >> meter is using a much older version. > > >> > > >> Todd > > >> > > >> On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans < > randyevans2...@gmail.com > > > > > >> wrote: > > >> > > >> > Todd, > > >> > > > >> > I have version 1.01. Yes, please e-mail me your modified version. > > >> Maybe it > > >> > will help. > > >> > > > >> > Thanks, > > >> > > > >> > Randy > > >> > > > >> > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef > > > >> > wrote: > > >> > > > >> > > Randy, > > >> > > > > >> > > Is there a version number at the top of the file? I had two > versions > > >> > (1.00 > > >> > > and 1.01). I can't remember all that changed between the two since > > it > > >> has > > >> > > been a while but I did find an additional timeout code that > limited > > >> the > > >> > > wait time for reading data to 1 minute. I am guessing I had a > > problem > > >> > with > > >> > > mine and added it. > > >> > > > > >> > > I can email it to you if you want to try it. If it works for you, > I > > >> will > > >> > > have Illya update it on xDevs. > > >> > > > > >> > > Todd > > >> > > > > >> > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans < > > >> randyevans2...@gmail.com> > > >> > > wrote: > > >> > > > > >> > > > I am still unable to get my HP-3458A to work with the xdevs > > program > > >> > > > "HP3458A_Measure_DMM_Noise.488" (https://xdevs.com/article/dmm > > >> _noise/ > > >> > ). > > >> > > > It > > >> > > > works fine with my Agilent 3458A and with a friend's HP-3458A > and > > >> his > > >> > > > Agilent 3458A using the EZGPIB program. The SW versions are > > >> identical > > >> > > with > > >> > > > my friend's DMMs, the HP units are (8.2) and with the Agilent > > units > > >> are > > >> > > > (9.2). However, both of my units work fine with the KE5FX GPIB > > >> Toolkit > > >> > > > HP3458.EXE program to download the NVRAM CAL data. This would > > seem > > >> to > > >> > > > indicate that my HP-3458 GPIB works at least partially. > > >> > > > > > >> > > > Does anyone have a EZGPIB program that I could try to run (that > is > > >> > > > different from the Xdev program)? It would be nice to see if I >
Re: [volt-nuts] GPIB Issue with 3458A
Randy, I will continue to look for my adapter. I haven't used it for quite a while. I have seen similar issues and swapped cables, moved the adapter to the other meter, etc... One of my cables was bad from one side of the connector to the other (on the same end of the cable) and caused me lots of wasted time. Weird reading errors were happening. Would you email me Martin's code so I can test it here and I will see if anything can be done to replicate your issue. Todd On Sat, Dec 30, 2017 at 6:42 PM, Randy Evanswrote: > Todd, > > I tried adding the line but it made no difference. > > Randy > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans > wrote: > > > Todd, > > > > Does it matter where I put the line in the Init_Device procedure? > > > > > > procedure Init_Device; > > begin; > >EZGPIB_BusWriteData(gpib_address,GET_ID);// Get ID of > meter > >if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then > > EZGPIB_ScreenWriteLn(meter_id); > > > >EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);// Hold trigger > > until meter can be setup > >EZGPIB_BusWriteData(gpib_address,'INBUF ON'); > >EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); > >EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); > > end; > > > > > > Randy > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef > > wrote: > > > >> Randy, > >> > >> I am the person that wrote that EzGPIB script. It was my first before I > >> moved on to Python and I have only written a couple for the different > >> meters I have on hand. Version 1.01 is the most current version I have > for > >> the 3458A. > >> I see that I should have added at least one more command to the meter > >> initialization. The additional init command is used in the scripts on > >> xDevs. > >> > >> Add the following line in the Init_Device procedure > >>EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); > >> > >> This tells the meter to send an EOI on every read. > >> > >> I am looking for one of my Prologix adapters or I would test this. When > I > >> find it, I will also test it. I want to make sure there are no further > >> issues. > >> Also, do both meters run the same firmware versions? I wonder if the HP > >> meter is using a much older version. > >> > >> Todd > >> > >> On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans > > >> wrote: > >> > >> > Todd, > >> > > >> > I have version 1.01. Yes, please e-mail me your modified version. > >> Maybe it > >> > will help. > >> > > >> > Thanks, > >> > > >> > Randy > >> > > >> > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef > >> > wrote: > >> > > >> > > Randy, > >> > > > >> > > Is there a version number at the top of the file? I had two versions > >> > (1.00 > >> > > and 1.01). I can't remember all that changed between the two since > it > >> has > >> > > been a while but I did find an additional timeout code that limited > >> the > >> > > wait time for reading data to 1 minute. I am guessing I had a > problem > >> > with > >> > > mine and added it. > >> > > > >> > > I can email it to you if you want to try it. If it works for you, I > >> will > >> > > have Illya update it on xDevs. > >> > > > >> > > Todd > >> > > > >> > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans < > >> randyevans2...@gmail.com> > >> > > wrote: > >> > > > >> > > > I am still unable to get my HP-3458A to work with the xdevs > program > >> > > > "HP3458A_Measure_DMM_Noise.488" (https://xdevs.com/article/dmm > >> _noise/ > >> > ). > >> > > > It > >> > > > works fine with my Agilent 3458A and with a friend's HP-3458A and > >> his > >> > > > Agilent 3458A using the EZGPIB program. The SW versions are > >> identical > >> > > with > >> > > > my friend's DMMs, the HP units are (8.2) and with the Agilent > units > >> are > >> > > > (9.2). However, both of my units work fine with the KE5FX GPIB > >> Toolkit > >> > > > HP3458.EXE program to download the NVRAM CAL data. This would > seem > >> to > >> > > > indicate that my HP-3458 GPIB works at least partially. > >> > > > > >> > > > Does anyone have a EZGPIB program that I could try to run (that is > >> > > > different from the Xdev program)? It would be nice to see if I > >> could > >> > get > >> > > > something to work other than the HP3458.EXE program. Or if any > one > >> has > >> > > > some ideas on what to look for. > >> > > > > >> > > > Thanks, > >> > > > > >> > > > Randy Evans > >> > > > > >> > > > On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans < > >> randyevans2...@gmail.com > >> > > > >> > > > wrote: > >> > > > > >> > > > > I am hoping some of you experts might be able to assist with a > >> > strange > >> > > > > problem i am having trying to make automated measurements using > >> > EZGPIB > >> > > > with > >> > > > > an Agilent and an HP 3458A. I have a program given to me by > >> Martin > >> > > > > Reynolds, which is a variant on the Xdevs noise
Re: [volt-nuts] GPIB Issue with 3458A
Todd, I tried adding the line but it made no difference. Randy On Sat, Dec 30, 2017 at 12:19 PM, Randy Evanswrote: > Todd, > > Does it matter where I put the line in the Init_Device procedure? > > > procedure Init_Device; > begin; >EZGPIB_BusWriteData(gpib_address,GET_ID);// Get ID of meter >if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then > EZGPIB_ScreenWriteLn(meter_id); > >EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);// Hold trigger > until meter can be setup >EZGPIB_BusWriteData(gpib_address,'INBUF ON'); >EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); >EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); > end; > > > Randy > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef > wrote: > >> Randy, >> >> I am the person that wrote that EzGPIB script. It was my first before I >> moved on to Python and I have only written a couple for the different >> meters I have on hand. Version 1.01 is the most current version I have for >> the 3458A. >> I see that I should have added at least one more command to the meter >> initialization. The additional init command is used in the scripts on >> xDevs. >> >> Add the following line in the Init_Device procedure >>EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); >> >> This tells the meter to send an EOI on every read. >> >> I am looking for one of my Prologix adapters or I would test this. When I >> find it, I will also test it. I want to make sure there are no further >> issues. >> Also, do both meters run the same firmware versions? I wonder if the HP >> meter is using a much older version. >> >> Todd >> >> On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans >> wrote: >> >> > Todd, >> > >> > I have version 1.01. Yes, please e-mail me your modified version. >> Maybe it >> > will help. >> > >> > Thanks, >> > >> > Randy >> > >> > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef >> > wrote: >> > >> > > Randy, >> > > >> > > Is there a version number at the top of the file? I had two versions >> > (1.00 >> > > and 1.01). I can't remember all that changed between the two since it >> has >> > > been a while but I did find an additional timeout code that limited >> the >> > > wait time for reading data to 1 minute. I am guessing I had a problem >> > with >> > > mine and added it. >> > > >> > > I can email it to you if you want to try it. If it works for you, I >> will >> > > have Illya update it on xDevs. >> > > >> > > Todd >> > > >> > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans < >> randyevans2...@gmail.com> >> > > wrote: >> > > >> > > > I am still unable to get my HP-3458A to work with the xdevs program >> > > > "HP3458A_Measure_DMM_Noise.488" (https://xdevs.com/article/dmm >> _noise/ >> > ). >> > > > It >> > > > works fine with my Agilent 3458A and with a friend's HP-3458A and >> his >> > > > Agilent 3458A using the EZGPIB program. The SW versions are >> identical >> > > with >> > > > my friend's DMMs, the HP units are (8.2) and with the Agilent units >> are >> > > > (9.2). However, both of my units work fine with the KE5FX GPIB >> Toolkit >> > > > HP3458.EXE program to download the NVRAM CAL data. This would seem >> to >> > > > indicate that my HP-3458 GPIB works at least partially. >> > > > >> > > > Does anyone have a EZGPIB program that I could try to run (that is >> > > > different from the Xdev program)? It would be nice to see if I >> could >> > get >> > > > something to work other than the HP3458.EXE program. Or if any one >> has >> > > > some ideas on what to look for. >> > > > >> > > > Thanks, >> > > > >> > > > Randy Evans >> > > > >> > > > On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans < >> randyevans2...@gmail.com >> > > >> > > > wrote: >> > > > >> > > > > I am hoping some of you experts might be able to assist with a >> > strange >> > > > > problem i am having trying to make automated measurements using >> > EZGPIB >> > > > with >> > > > > an Agilent and an HP 3458A. I have a program given to me by >> Martin >> > > > > Reynolds, which is a variant on the Xdevs noise measurement >> program. >> > > > > Martin's program continuously reads the voltage measurements >> > > sequentially >> > > > > from the two meters and then stores then in a measurements file, >> > along >> > > > with >> > > > > the temperature measurements from both meters. The program works >> > fine >> > > on >> > > > > his Agilent and HP 3458As. >> > > > > >> > > > > On my two 3458As, it takes the temperature measurements just fine >> but >> > > > only >> > > > > takes a voltage measurement from the Agilent unit. The HP unit >> will >> > > not >> > > > > make any voltage measurements although I can see both the 3458As >> in >> > > talk >> > > > > states and, of course, it does sent the temperature measurements. >> > > > However, >> > > > > if I turn the HP unit off and then on again, the HP will start to >> > take >> > > > > voltage and
Re: [volt-nuts] GPIB Issue with 3458A
Randy, You can add it after the last line with the OUTPUT_FORMAT Todd On Sat, Dec 30, 2017 at 3:19 PM, Randy Evanswrote: > Todd, > > Does it matter where I put the line in the Init_Device procedure? > > > procedure Init_Device; > begin; >EZGPIB_BusWriteData(gpib_address,GET_ID);// Get ID of meter >if EZGPIB_BusWaitForData(gpib_address,meter_id,Timeout) then > EZGPIB_ScreenWriteLn(meter_id); > >EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);// Hold trigger > until meter can be setup >EZGPIB_BusWriteData(gpib_address,'INBUF ON'); >EZGPIB_BusWriteData(gpib_address,AUTO_ZERO); >EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT); > end; > > > Randy > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef > wrote: > > > Randy, > > > > I am the person that wrote that EzGPIB script. It was my first before I > > moved on to Python and I have only written a couple for the different > > meters I have on hand. Version 1.01 is the most current version I have > for > > the 3458A. > > I see that I should have added at least one more command to the meter > > initialization. The additional init command is used in the scripts on > > xDevs. > > > > Add the following line in the Init_Device procedure > >EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); > > > > This tells the meter to send an EOI on every read. > > > > I am looking for one of my Prologix adapters or I would test this. When I > > find it, I will also test it. I want to make sure there are no further > > issues. > > Also, do both meters run the same firmware versions? I wonder if the HP > > meter is using a much older version. > > > > Todd > > > > On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans > > wrote: > > > > > Todd, > > > > > > I have version 1.01. Yes, please e-mail me your modified version. > Maybe > > it > > > will help. > > > > > > Thanks, > > > > > > Randy > > > > > > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef > > > wrote: > > > > > > > Randy, > > > > > > > > Is there a version number at the top of the file? I had two versions > > > (1.00 > > > > and 1.01). I can't remember all that changed between the two since it > > has > > > > been a while but I did find an additional timeout code that limited > the > > > > wait time for reading data to 1 minute. I am guessing I had a problem > > > with > > > > mine and added it. > > > > > > > > I can email it to you if you want to try it. If it works for you, I > > will > > > > have Illya update it on xDevs. > > > > > > > > Todd > > > > > > > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans < > > randyevans2...@gmail.com> > > > > wrote: > > > > > > > > > I am still unable to get my HP-3458A to work with the xdevs program > > > > > "HP3458A_Measure_DMM_Noise.488" (https://xdevs.com/article/ > > dmm_noise/ > > > ). > > > > > It > > > > > works fine with my Agilent 3458A and with a friend's HP-3458A and > his > > > > > Agilent 3458A using the EZGPIB program. The SW versions are > > identical > > > > with > > > > > my friend's DMMs, the HP units are (8.2) and with the Agilent units > > are > > > > > (9.2). However, both of my units work fine with the KE5FX GPIB > > Toolkit > > > > > HP3458.EXE program to download the NVRAM CAL data. This would seem > > to > > > > > indicate that my HP-3458 GPIB works at least partially. > > > > > > > > > > Does anyone have a EZGPIB program that I could try to run (that is > > > > > different from the Xdev program)? It would be nice to see if I > could > > > get > > > > > something to work other than the HP3458.EXE program. Or if any one > > has > > > > > some ideas on what to look for. > > > > > > > > > > Thanks, > > > > > > > > > > Randy Evans > > > > > > > > > > On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans < > > randyevans2...@gmail.com > > > > > > > > > wrote: > > > > > > > > > > > I am hoping some of you experts might be able to assist with a > > > strange > > > > > > problem i am having trying to make automated measurements using > > > EZGPIB > > > > > with > > > > > > an Agilent and an HP 3458A. I have a program given to me by > Martin > > > > > > Reynolds, which is a variant on the Xdevs noise measurement > > program. > > > > > > Martin's program continuously reads the voltage measurements > > > > sequentially > > > > > > from the two meters and then stores then in a measurements file, > > > along > > > > > with > > > > > > the temperature measurements from both meters. The program works > > > fine > > > > on > > > > > > his Agilent and HP 3458As. > > > > > > > > > > > > On my two 3458As, it takes the temperature measurements just fine > > but > > > > > only > > > > > > takes a voltage measurement from the Agilent unit. The HP unit > > will > > > > not > > > > > > make any voltage measurements although I can see both the 3458As > in > > > > talk > > > > > > states and, of course, it does sent the temperature measurements. > > > >
Re: [volt-nuts] GPIB Issue with 3458A
Randy, I am the person that wrote that EzGPIB script. It was my first before I moved on to Python and I have only written a couple for the different meters I have on hand. Version 1.01 is the most current version I have for the 3458A. I see that I should have added at least one more command to the meter initialization. The additional init command is used in the scripts on xDevs. Add the following line in the Init_Device procedure EZGPIB_BusWriteData(gpib_address,'END ALWAYS'); This tells the meter to send an EOI on every read. I am looking for one of my Prologix adapters or I would test this. When I find it, I will also test it. I want to make sure there are no further issues. Also, do both meters run the same firmware versions? I wonder if the HP meter is using a much older version. Todd On Sat, Dec 30, 2017 at 10:42 AM, Randy Evanswrote: > Todd, > > I have version 1.01. Yes, please e-mail me your modified version. Maybe it > will help. > > Thanks, > > Randy > > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef > wrote: > > > Randy, > > > > Is there a version number at the top of the file? I had two versions > (1.00 > > and 1.01). I can't remember all that changed between the two since it has > > been a while but I did find an additional timeout code that limited the > > wait time for reading data to 1 minute. I am guessing I had a problem > with > > mine and added it. > > > > I can email it to you if you want to try it. If it works for you, I will > > have Illya update it on xDevs. > > > > Todd > > > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans > > wrote: > > > > > I am still unable to get my HP-3458A to work with the xdevs program > > > "HP3458A_Measure_DMM_Noise.488" (https://xdevs.com/article/dmm_noise/ > ). > > > It > > > works fine with my Agilent 3458A and with a friend's HP-3458A and his > > > Agilent 3458A using the EZGPIB program. The SW versions are identical > > with > > > my friend's DMMs, the HP units are (8.2) and with the Agilent units are > > > (9.2). However, both of my units work fine with the KE5FX GPIB Toolkit > > > HP3458.EXE program to download the NVRAM CAL data. This would seem to > > > indicate that my HP-3458 GPIB works at least partially. > > > > > > Does anyone have a EZGPIB program that I could try to run (that is > > > different from the Xdev program)? It would be nice to see if I could > get > > > something to work other than the HP3458.EXE program. Or if any one has > > > some ideas on what to look for. > > > > > > Thanks, > > > > > > Randy Evans > > > > > > On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans > > > > wrote: > > > > > > > I am hoping some of you experts might be able to assist with a > strange > > > > problem i am having trying to make automated measurements using > EZGPIB > > > with > > > > an Agilent and an HP 3458A. I have a program given to me by Martin > > > > Reynolds, which is a variant on the Xdevs noise measurement program. > > > > Martin's program continuously reads the voltage measurements > > sequentially > > > > from the two meters and then stores then in a measurements file, > along > > > with > > > > the temperature measurements from both meters. The program works > fine > > on > > > > his Agilent and HP 3458As. > > > > > > > > On my two 3458As, it takes the temperature measurements just fine but > > > only > > > > takes a voltage measurement from the Agilent unit. The HP unit will > > not > > > > make any voltage measurements although I can see both the 3458As in > > talk > > > > states and, of course, it does sent the temperature measurements. > > > However, > > > > if I turn the HP unit off and then on again, the HP will start to > take > > > > voltage and temperature measurements and send them to the PC. > > > > > > > > Any ideas on what might be causing the strange behavior? > > > > > > > > Thanks, > > > > > > > > Randy Evans > > > > > > > ___ > > > volt-nuts mailing list -- volt-nuts@febo.com > > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > > > mailman/listinfo/volt-nuts > > > and follow the instructions there. > > > > > ___ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > > mailman/listinfo/volt-nuts > > and follow the instructions there. > > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] GPIB Issue with 3458A
Todd, I have version 1.01. Yes, please e-mail me your modified version. Maybe it will help. Thanks, Randy On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallefwrote: > Randy, > > Is there a version number at the top of the file? I had two versions (1.00 > and 1.01). I can't remember all that changed between the two since it has > been a while but I did find an additional timeout code that limited the > wait time for reading data to 1 minute. I am guessing I had a problem with > mine and added it. > > I can email it to you if you want to try it. If it works for you, I will > have Illya update it on xDevs. > > Todd > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans > wrote: > > > I am still unable to get my HP-3458A to work with the xdevs program > > "HP3458A_Measure_DMM_Noise.488" (https://xdevs.com/article/dmm_noise/). > > It > > works fine with my Agilent 3458A and with a friend's HP-3458A and his > > Agilent 3458A using the EZGPIB program. The SW versions are identical > with > > my friend's DMMs, the HP units are (8.2) and with the Agilent units are > > (9.2). However, both of my units work fine with the KE5FX GPIB Toolkit > > HP3458.EXE program to download the NVRAM CAL data. This would seem to > > indicate that my HP-3458 GPIB works at least partially. > > > > Does anyone have a EZGPIB program that I could try to run (that is > > different from the Xdev program)? It would be nice to see if I could get > > something to work other than the HP3458.EXE program. Or if any one has > > some ideas on what to look for. > > > > Thanks, > > > > Randy Evans > > > > On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans > > wrote: > > > > > I am hoping some of you experts might be able to assist with a strange > > > problem i am having trying to make automated measurements using EZGPIB > > with > > > an Agilent and an HP 3458A. I have a program given to me by Martin > > > Reynolds, which is a variant on the Xdevs noise measurement program. > > > Martin's program continuously reads the voltage measurements > sequentially > > > from the two meters and then stores then in a measurements file, along > > with > > > the temperature measurements from both meters. The program works fine > on > > > his Agilent and HP 3458As. > > > > > > On my two 3458As, it takes the temperature measurements just fine but > > only > > > takes a voltage measurement from the Agilent unit. The HP unit will > not > > > make any voltage measurements although I can see both the 3458As in > talk > > > states and, of course, it does sent the temperature measurements. > > However, > > > if I turn the HP unit off and then on again, the HP will start to take > > > voltage and temperature measurements and send them to the PC. > > > > > > Any ideas on what might be causing the strange behavior? > > > > > > Thanks, > > > > > > Randy Evans > > > > > ___ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > > mailman/listinfo/volt-nuts > > and follow the instructions there. > > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] GPIB Issue with 3458A
Randy, Is there a version number at the top of the file? I had two versions (1.00 and 1.01). I can't remember all that changed between the two since it has been a while but I did find an additional timeout code that limited the wait time for reading data to 1 minute. I am guessing I had a problem with mine and added it. I can email it to you if you want to try it. If it works for you, I will have Illya update it on xDevs. Todd On Sat, Dec 30, 2017 at 12:39 AM, Randy Evanswrote: > I am still unable to get my HP-3458A to work with the xdevs program > "HP3458A_Measure_DMM_Noise.488" (https://xdevs.com/article/dmm_noise/). > It > works fine with my Agilent 3458A and with a friend's HP-3458A and his > Agilent 3458A using the EZGPIB program. The SW versions are identical with > my friend's DMMs, the HP units are (8.2) and with the Agilent units are > (9.2). However, both of my units work fine with the KE5FX GPIB Toolkit > HP3458.EXE program to download the NVRAM CAL data. This would seem to > indicate that my HP-3458 GPIB works at least partially. > > Does anyone have a EZGPIB program that I could try to run (that is > different from the Xdev program)? It would be nice to see if I could get > something to work other than the HP3458.EXE program. Or if any one has > some ideas on what to look for. > > Thanks, > > Randy Evans > > On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans > wrote: > > > I am hoping some of you experts might be able to assist with a strange > > problem i am having trying to make automated measurements using EZGPIB > with > > an Agilent and an HP 3458A. I have a program given to me by Martin > > Reynolds, which is a variant on the Xdevs noise measurement program. > > Martin's program continuously reads the voltage measurements sequentially > > from the two meters and then stores then in a measurements file, along > with > > the temperature measurements from both meters. The program works fine on > > his Agilent and HP 3458As. > > > > On my two 3458As, it takes the temperature measurements just fine but > only > > takes a voltage measurement from the Agilent unit. The HP unit will not > > make any voltage measurements although I can see both the 3458As in talk > > states and, of course, it does sent the temperature measurements. > However, > > if I turn the HP unit off and then on again, the HP will start to take > > voltage and temperature measurements and send them to the PC. > > > > Any ideas on what might be causing the strange behavior? > > > > Thanks, > > > > Randy Evans > > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.