Dear Patrick,


Thanks a lot for the help and feedback, this is really appreciated.


For the second order term of the baseline, I believe one of the reason for it 
lies in our sideband suppression method. I'll now try to add that effect in the 
sensor_response variable and see if this impacts the baseline accordingly.


Thanks,
Eric


De : Patrick Eriksson 
Envoyé : mardi 22 juin 2021 20:39:56
À : Sauvageat, Eric (IAP); arts_users...@mailman.rrz.uni-hamburg.de
Cc : rita.edit.kaj...@ltu.se
Objet : Re: [arts-users] Tropospheric continuum retrieval in ARTS

Eric,

> (1998)). The values of the poly_order=0 are now quite small (mostly
> lower than 0.1K), however, I still only get positive values. Is that
> something that could be expected from the way the function
> "retrievalAddPolyfit" works maybe ?

No, there is nothing that promotes positive values. But you now get so
small values that nothing to worry about.
You get very consistent values for poly_order 1 and 2, despite that the
tropospheric opacity varies quite a bit. That should be an indication on
that receiver is stable and the baseline+tropospheric part of the
retrievals is working well.


> Would you have any tips on how to implement the retrieval of a single
> scaling parameter for a full water vapor profile ? I believe this is
> more or less what the CONTABS_DO parameter was implementing according to
> the first Qpack documentation ?

This you do by setting the retrieval grid to have a single point (e.g.
500 hPa) and set the corresponding covariance matrix to have size 1x1. A
suitable value for it is maybe 1 (meaning a 100% uncertainty at all
altitudes, fully correlated).

Bye,

Patrick
___
arts_users.mi mailing list
arts_users.mi@lists.uni-hamburg.de
https://mailman.rrz.uni-hamburg.de/mailman/listinfo/arts_users.mi


Re: [arts-users] Tropospheric continuum retrieval in ARTS

2021-06-22 Thread Patrick Eriksson

Eric,

(1998)). The values of the poly_order=0 are now quite small (mostly 
lower than 0.1K), however, I still only get positive values. Is that 
something that could be expected from the way the function 
"retrievalAddPolyfit" works maybe ?


No, there is nothing that promotes positive values. But you now get so 
small values that nothing to worry about.
You get very consistent values for poly_order 1 and 2, despite that the 
tropospheric opacity varies quite a bit. That should be an indication on 
that receiver is stable and the baseline+tropospheric part of the 
retrievals is working well.



Would you have any tips on how to implement the retrieval of a single 
scaling parameter for a full water vapor profile ? I believe this is 
more or less what the CONTABS_DO parameter was implementing according to 
the first Qpack documentation ?


This you do by setting the retrieval grid to have a single point (e.g. 
500 hPa) and set the corresponding covariance matrix to have size 1x1. A 
suitable value for it is maybe 1 (meaning a 100% uncertainty at all 
altitudes, fully correlated).


Bye,

Patrick
___
arts_users.mi mailing list
arts_users.mi@lists.uni-hamburg.de
https://mailman.rrz.uni-hamburg.de/mailman/listinfo/arts_users.mi


Re: [arts-users] Tropospheric continuum retrieval in ARTS

2021-06-11 Thread eric.sauvageat
Hi,


Thanks, I'll see if I can reduce that value by lowering the apriori cov for the 
polyfit order 0 then.


So the "single continuum fit" was really just setting a single altitude and 
retrieving a singe H2O VMR value at this altitude whereas the profile case was 
the same retrievals (species, apriori cov) but with an apriori profile instead 
of a single value. I'll check the rel unit then, I was not aware this could be 
used to retrieved a single scaling value on a profile.


Best regards,
Eric



De : Patrick Eriksson 
Envoyé : vendredi 11 juin 2021 10:23:58
À : Sauvageat, Eric (IAP); arts_users...@mailman.rrz.uni-hamburg.de
Objet : Re: [arts-users] Tropospheric continuum retrieval in ARTS

Eric,

It seems indeed that poly_order=0 is involved. I think you want this
value to be small. Especially as it seems to only be positive. This
behaviour would be OK if you thought that you have a calibration error
that is random, but just one-sided. Sounds not reasonable. I think more
realistic is to trust the calibration and let the tropospheric part take
care of this part. poly_order=0 and the tropospheric correction have a
very similar impact on the "baseline".

The resulting tropospheric attenuation depends to some extent on the
retrieved humidity profile, but that should be secondary. But you can of
course take a look at how the tropospheric humidity profile look in both
cases and compare.

By the way, how do you do the single column case? I suggest to use the
rel unit. So you scale the profile with a single relative value. Using
VMR for this is less realistic (here you would change the full profile
with the same VMR value, that is not realistic).

Regards,

Patrick



On 2021-06-11 10:02, eric.sauvag...@iap.unibe.ch wrote:
> Dear Patrick,
>
>
> Thanks for you answer.
>
>
> I did some further investigations on the value of the polyfit and as
> expected the value of the constant terms is indeed quite different for
> both options and very similar for 1st and 2nd degree terms (I appended
> some basic representative plots to give an idea).
>
>
> More specifically, it is smaller in the case of the "single continuum
> fit" which is the one giving a +10% ozone profile. It would mean
> that tropospheric attenuation is considered to be larger in the case of
> the single continuum fit compared to the other option. I think this is
> then consistent with the 10% higher ozone values obtained in this case,
> would you agree ?
>
>
> I remember having tried different constraint on the constant term but
> only making it larger so I should redo some test reducing it more. I
> believe one way to check if the tropospheric attenuation is correct is
> to compare the opacity value computed from ARTS to one computed manually
> on the spectra.
>
> I'll check that as well but thanks for your inputs already.
>
>
> Best regards,
>
> Eric
>
>
>
>
>
> 
> *De :* Patrick Eriksson 
> *Envoyé :* mercredi 9 juin 2021 23:39:33
> *À :* Sauvageat, Eric (IAP); arts_users...@mailman.rrz.uni-hamburg.de
> *Objet :* Re: [arts-users] Tropospheric continuum retrieval in ARTS
> Eric,
>
> I don't know about a fixed setup for dealing with the troposphere in
> observations of this type. As your observations at max gives one piece
> of information for the troposphere, I would say that the single
> retrieval point setup makes most sense. In any case it is plausible option.
>
> I assume that you fit the measurements in both cases.
>
> If I get it right, your main worry is that you get a 10% difference in
> the ozone profile between the options. This should likely originate in
> that the retrievals give you a troposphere having a 10% difference in
> transmission.
>
> My suggestion is then to look at the polyfit part. Is the fitted polyfit
> the same between the options? My guess is that it differs. And that
> gives room for retrieving a different tropospheric transmission.
>
> And it could be reasons to anyhow consider the polyfit part. I assume
> you have a good calibration and the uncertainty in the overall
> "baseline" level is due to the troposphere. Or expressed differently,
> you want to fit the overall baseline level by changing H2O in the
> troposphere, not by the polyfit. Or more exactly you want the first
> polyfit coefficient to be small, the polyfit should just take care of
> the "wiggling". To achieve this you should set the a priori uncertainty
> for the first (0-order) coefficient to be very small, to effectively
> enforce a low measurement response for the coefficient. With this, the
> retrieval will have to fully adjust the overall baseline level by the
> H2O profile, independen

Re: [arts-users] Tropospheric continuum retrieval in ARTS

2021-06-11 Thread eric.sauvageat
Dear Mathias,


Nice, I am happy to see that other people are working on this topic as well.

In terms of apriori, I am indeed doing the same kind of forcing on the ozone 
profile to avoid some high ozone VMR retrieved at low altitudes, problem that I 
got before using such a forcing.


May I ask what you are using as apriori profile and covariance matrix for the 
H2O profile ? I tested ECMWF and Fascod as apriori for the H2O and it did not 
produce significant changes, the thing that matters the most in that view seems 
to be the apriori cov value for H2O setup with constant std deviation around 
6e-4 or 6 ppm (I do all my retrievals in vmr units, both for O3 and H2O). Note 
that this value was mostly a guess but works (enable convergence...) for all my 
retrievals. This might change soon though depending on the current discussion..


Regarding the polyfit value, I am also not providing any specific apriori 
value, only a covariance matrix for each of the terms.


Best regards,

Eric



De : Mathias Milz 
Envoyé : jeudi 10 juin 2021 08:05:53
À : Sauvageat, Eric (IAP); arts_users.mi@lists.uni-hamburg.de
Cc : Rita Edit Kajtar; Uwe Raffalski
Objet : Re: Tropospheric continuum retrieval in ARTS


Dear Eric,



We are working with a similar setup (Just at another frequency) for the 
radiometers in Kiruna.



For water vapour we use a rather loose constraint (large S_x/S_a value) to 
allow the fit of the profile for the H2O background (co-fit). Of course, H2O 
can be unrealistic as we just use it to fit the continuum/baseline



For O3 we use an altitude dependent constraint with a rather strong constraint 
below the tropopause (where we know the values are “negligible”) and above the 
sensitive area where we force the retrieval to the a priori knowledge. So we 
allow reasonable fitting only for the altitudes where the Jacobians show decent 
sensitivity. Note: We have good results with “rel” jacobians. “vmr”/absolute 
Jacobians caused problems.



We also co-fit a baseline  using “polyfit” of the lowest degree. However here 
we did not yet find a solution that we can start with a preselected value (e.g. 
an instrument-dependend baseline) so the a priori value is by default 0 and the 
a standard constrain constraint might cause difficulties with large large 
baselines due to high humidity or clouds with precipitation.



Maybe this helps.



Best regards,



Mathias









From: arts_users.mi-boun...@lists.uni-hamburg.de 
 on behalf of 
eric.sauvag...@iap.unibe.ch 
Date: Wednesday, 9 June 2021 at 16:22
To: arts_users.mi@lists.uni-hamburg.de 
Subject: [arts-users] Tropospheric continuum retrieval in ARTS

Dear ARTS community,



I am doing stratospheric O3 retrieval with a ground-based radiometer (f=142GHz) 
and am trying to deal with the absorption contribution of the troposphere 
directly in the OEM implemented in ARTS (avoiding tropospheric correction prior 
to the retrievals).



Up to now, I took inspiration from "qpack2_demo2.m" which suggests (if I 
understood it correctly) to implement a "H2O-PWR98, H2O" retrieval (main 
contributor of tropospheric opacity at these frequencies) on a lower atmosphere 
retrieval grid. This results in a water vapor profile retrieved together with 
my main ozone retrievals. Of course this profile has no good measurement 
response as my ozone radiometer is not designed to retrieve any H2O profile, 
but it seems that it provides the "right amount of opacity" needed to explain 
my spectrum. In addition, note that I am also performing a polyfit retrieval of 
degree 2 which is also fitting a constant term on my spectra which also 
probably contributes somehow to fit the global continuum absorption.



I found out recently, that such a continuum retrieval was implemented in QPack1 
(activated with "Q.CONTABS_DO") and from my understanding, it does not seem to 
retrieve any H2O profiles but only single values for the continuum (which 
somehow makes more sense to me). So I did try to provide a single grid 
retrieval point and single value for H2O cov matrix and it seems to work 
equally good as the retrieval including a full H2O profile (in the sense of 
convergence, correlation between both time series, ...) but it has a constant 
+10% VMR offset on my whole ozone profiles and I have no clue why.



Also, I have made different tests to check the impact of the selected species 
(continuum vs full absorption model defined with or without H2O) but it did 
only produce slight changes in the results. As well, the height of the H2O grid 
or its altitude resolution does not seem to have significant impact on the 
retrievals.



Sorry for this long email but I am really puzzled in what is the best way to 
deal with continuum absorption in ARTS and what I might be doing wrong. 
Therefore, any kind of feedback or help regarding this would be much 
appreciated. If needed, I can also provide examples plots of MR,

Re: [arts-users] Tropospheric continuum retrieval in ARTS

2021-06-10 Thread Mathias Milz
Dear Eric,

We are working with a similar setup (Just at another frequency) for the 
radiometers in Kiruna.

For water vapour we use a rather loose constraint (large S_x/S_a value) to 
allow the fit of the profile for the H2O background (co-fit). Of course, H2O 
can be unrealistic as we just use it to fit the continuum/baseline

For O3 we use an altitude dependent constraint with a rather strong constraint 
below the tropopause (where we know the values are “negligible”) and above the 
sensitive area where we force the retrieval to the a priori knowledge. So we 
allow reasonable fitting only for the altitudes where the Jacobians show decent 
sensitivity. Note: We have good results with “rel” jacobians. “vmr”/absolute 
Jacobians caused problems.

We also co-fit a baseline  using “polyfit” of the lowest degree. However here 
we did not yet find a solution that we can start with a preselected value (e.g. 
an instrument-dependend baseline) so the a priori value is by default 0 and the 
a standard constrain constraint might cause difficulties with large large 
baselines due to high humidity or clouds with precipitation.

Maybe this helps.

Best regards,

Mathias




From: arts_users.mi-boun...@lists.uni-hamburg.de 
 on behalf of 
eric.sauvag...@iap.unibe.ch 
Date: Wednesday, 9 June 2021 at 16:22
To: arts_users.mi@lists.uni-hamburg.de 
Subject: [arts-users] Tropospheric continuum retrieval in ARTS

Dear ARTS community,



I am doing stratospheric O3 retrieval with a ground-based radiometer (f=142GHz) 
and am trying to deal with the absorption contribution of the troposphere 
directly in the OEM implemented in ARTS (avoiding tropospheric correction prior 
to the retrievals).



Up to now, I took inspiration from "qpack2_demo2.m" which suggests (if I 
understood it correctly) to implement a "H2O-PWR98, H2O" retrieval (main 
contributor of tropospheric opacity at these frequencies) on a lower atmosphere 
retrieval grid. This results in a water vapor profile retrieved together with 
my main ozone retrievals. Of course this profile has no good measurement 
response as my ozone radiometer is not designed to retrieve any H2O profile, 
but it seems that it provides the "right amount of opacity" needed to explain 
my spectrum. In addition, note that I am also performing a polyfit retrieval of 
degree 2 which is also fitting a constant term on my spectra which also 
probably contributes somehow to fit the global continuum absorption.


I found out recently, that such a continuum retrieval was implemented in QPack1 
(activated with "Q.CONTABS_DO") and from my understanding, it does not seem to 
retrieve any H2O profiles but only single values for the continuum (which 
somehow makes more sense to me). So I did try to provide a single grid 
retrieval point and single value for H2O cov matrix and it seems to work 
equally good as the retrieval including a full H2O profile (in the sense of 
convergence, correlation between both time series, ...) but it has a constant 
+10% VMR offset on my whole ozone profiles and I have no clue why.



Also, I have made different tests to check the impact of the selected species 
(continuum vs full absorption model defined with or without H2O) but it did 
only produce slight changes in the results. As well, the height of the H2O grid 
or its altitude resolution does not seem to have significant impact on the 
retrievals.



Sorry for this long email but I am really puzzled in what is the best way to 
deal with continuum absorption in ARTS and what I might be doing wrong. 
Therefore, any kind of feedback or help regarding this would be much 
appreciated. If needed, I can also provide examples plots of MR, AVK or 
profiles (not sure how it works for mailing list though).



Many thanks in advance,
Eric














___
arts_users.mi mailing list
arts_users.mi@lists.uni-hamburg.de
https://mailman.rrz.uni-hamburg.de/mailman/listinfo/arts_users.mi


Re: [arts-users] Tropospheric continuum retrieval in ARTS

2021-06-09 Thread Patrick Eriksson

Eric,

I don't know about a fixed setup for dealing with the troposphere in 
observations of this type. As your observations at max gives one piece 
of information for the troposphere, I would say that the single 
retrieval point setup makes most sense. In any case it is plausible option.


I assume that you fit the measurements in both cases.

If I get it right, your main worry is that you get a 10% difference in 
the ozone profile between the options. This should likely originate in 
that the retrievals give you a troposphere having a 10% difference in 
transmission.


My suggestion is then to look at the polyfit part. Is the fitted polyfit 
the same between the options? My guess is that it differs. And that 
gives room for retrieving a different tropospheric transmission.


And it could be reasons to anyhow consider the polyfit part. I assume 
you have a good calibration and the uncertainty in the overall 
"baseline" level is due to the troposphere. Or expressed differently, 
you want to fit the overall baseline level by changing H2O in the 
troposphere, not by the polyfit. Or more exactly you want the first 
polyfit coefficient to be small, the polyfit should just take care of 
the "wiggling". To achieve this you should set the a priori uncertainty 
for the first (0-order) coefficient to be very small, to effectively 
enforce a low measurement response for the coefficient. With this, the 
retrieval will have to fully adjust the overall baseline level by the 
H2O profile, independently on the grid and a priori uncertainty you use 
for H2O.


Bye,

Patrick





On 2021-06-09 16:21, eric.sauvag...@iap.unibe.ch wrote:

Dear ARTS community,


I am doing stratospheric O3 retrieval with a ground-based radiometer 
(f=142GHz) and am trying to deal with the absorption contribution of the 
troposphere directly in the OEM implemented in ARTS (avoiding 
tropospheric correction prior to the retrievals).



Up to now, I took inspiration from "qpack2_demo2.m" which suggests (if I 
understood it correctly) to implement a "H2O-PWR98, H2O" retrieval (main 
contributor of tropospheric opacity at these frequencies) on a lower 
atmosphere retrieval grid. This results in a water vapor profile 
retrieved together with my main ozone retrievals. Of course this profile 
has no good measurement response as my ozone radiometer is not designed 
to retrieve any H2O profile, but it seems that it provides the "right 
amount of opacity" needed to explain my spectrum.In addition, note that 
I am also performing a polyfit retrieval of degree 2 which is also 
fitting a constant term on my spectra which also probably contributes 
somehow to fit the global continuum absorption.



I found out recently, that such a continuum retrieval was implemented in 
QPack1 (activated with "Q.CONTABS_DO") and from my understanding, it 
does not seem to retrieve any H2O profiles but only single values for 
the continuum (which somehow makes more sense to me). So I did try to 
provide a single grid retrieval point and single value for H2O cov 
matrix and it seems to work equally good as the retrieval including a 
full H2O profile (in the sense of convergence, correlation between both 
time series, ...) but it has a constant +10% VMR offset on my whole 
ozone profiles and I have no clue why.



Also, I have made different tests to check the impact of the selected 
species (continuum vs full absorption model defined with or without H2O) 
but it did only produce slight changes in the results. As well, the 
height of the H2O grid or its altitude resolution does not seem to have 
significant impact on the retrievals.



Sorry for this long email but I am really puzzled in what is the best 
way to deal with continuum absorption in ARTS and what I might be doing 
wrong. Therefore, any kind of feedback or help regarding this would be 
much appreciated. If needed, I can also provide examples plots of MR, 
AVK or profiles (not sure how it works for mailing list though).



Many thanks in advance,
Eric










___
arts_users.mi mailing list
arts_users.mi@lists.uni-hamburg.de
https://mailman.rrz.uni-hamburg.de/mailman/listinfo/arts_users.mi


___
arts_users.mi mailing list
arts_users.mi@lists.uni-hamburg.de
https://mailman.rrz.uni-hamburg.de/mailman/listinfo/arts_users.mi


[arts-users] Tropospheric continuum retrieval in ARTS

2021-06-09 Thread eric.sauvageat
Dear ARTS community,


I am doing stratospheric O3 retrieval with a ground-based radiometer (f=142GHz) 
and am trying to deal with the absorption contribution of the troposphere 
directly in the OEM implemented in ARTS (avoiding tropospheric correction prior 
to the retrievals).


Up to now, I took inspiration from "qpack2_demo2.m" which suggests (if I 
understood it correctly) to implement a "H2O-PWR98, H2O" retrieval (main 
contributor of tropospheric opacity at these frequencies) on a lower atmosphere 
retrieval grid. This results in a water vapor profile retrieved together with 
my main ozone retrievals. Of course this profile has no good measurement 
response as my ozone radiometer is not designed to retrieve any H2O profile, 
but it seems that it provides the "right amount of opacity" needed to explain 
my spectrum. In addition, note that I am also performing a polyfit retrieval of 
degree 2 which is also fitting a constant term on my spectra which also 
probably contributes somehow to fit the global continuum absorption.


I found out recently, that such a continuum retrieval was implemented in QPack1 
(activated with "Q.CONTABS_DO") and from my understanding, it does not seem to 
retrieve any H2O profiles but only single values for the continuum (which 
somehow makes more sense to me). So I did try to provide a single grid 
retrieval point and single value for H2O cov matrix and it seems to work 
equally good as the retrieval including a full H2O profile (in the sense of 
convergence, correlation between both time series, ...) but it has a constant 
+10% VMR offset on my whole ozone profiles and I have no clue why.


Also, I have made different tests to check the impact of the selected species 
(continuum vs full absorption model defined with or without H2O) but it did 
only produce slight changes in the results. As well, the height of the H2O grid 
or its altitude resolution does not seem to have significant impact on the 
retrievals.


Sorry for this long email but I am really puzzled in what is the best way to 
deal with continuum absorption in ARTS and what I might be doing wrong. 
Therefore, any kind of feedback or help regarding this would be much 
appreciated. If needed, I can also provide examples plots of MR, AVK or 
profiles (not sure how it works for mailing list though).


Many thanks in advance,
Eric