Re: [ccp4bb] To bathe or not to bathe.

2007-11-27 Thread Richard Gillilan
Oops, sorry. The x axis of the previous plot is actually not  
resolution, but Q. My bad.


Richard


On Nov 27, 2007, at 11:26 AM, Richard Gillilan wrote:

Just a couple small images that may be of interest. The x scale is  
resolution in Angstroms, the y scale is intensity (arbitrary  
units). I don't recall if the plot is corrected for CCD pedestal  
values, so the difference is not quite as dramatic as x49, still  
very nice clean background if you are looking for a faint signal.


The CCD images below are equal exposures shown with equal contrast  
settings:


HeN2.tiff
HeN2imgs.tiff


Re: [ccp4bb] To bathe or not to bathe.

2007-11-27 Thread Nave, C (Colin)
Richard
I think the sharp spot is just small angle scattering (e.g. from domain
boundaries) resulting from the beam hitting one of the defining
apertures in your collimation system. If you ray trace the beam from the
last defining aperture, through the guard aperture then to the detector
you should get an area of high background where the guard aperture does
not cut this off and it is not covered by the beamstop. It then drops
off as the beamstop comes in to play appearing to give a sharp spike. In
fact, it is quite well set up with the beamstop not overly large.
The appearance of a very diffuse ring for the scatter from nitrogen is
due to the fact that the beamstop cuts off the air scatter, for the
parts of the beam nearer the beamstop.
Cheers
 Colin

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Richard Gillilan
Sent: 27 November 2007 16:40
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] To bathe or not to bathe.

Oops, sorry. The x axis of the previous plot is actually not resolution,
but Q. My bad.

Richard


On Nov 27, 2007, at 11:26 AM, Richard Gillilan wrote:

 Just a couple small images that may be of interest. The x scale is 
 resolution in Angstroms, the y scale is intensity (arbitrary units). I

 don't recall if the plot is corrected for CCD pedestal values, so the 
 difference is not quite as dramatic as x49, still very nice clean 
 background if you are looking for a faint signal.

 The CCD images below are equal exposures shown with equal contrast
 settings:

 HeN2.tiff
 HeN2imgs.tiff
DIVFONT size=1 color=grayThis e-mail and any attachments may contain 
confidential, copyright and or privileged material, and are for the use of the 
intended addressee only. If you are not the intended addressee or an authorised 
recipient of the addressee please notify us of receipt by returning the e-mail 
and do not use, copy, retain, distribute or disclose the information in or 
attached to the e-mail.
Any opinions expressed within this e-mail are those of the individual and not 
necessarily of Diamond Light Source Ltd. 
Diamond Light Source Ltd. cannot guarantee that this e-mail or any attachments 
are free from viruses and we cannot accept liability for any damage which you 
may sustain as a result of software viruses which may be transmitted in or with 
the message.
Diamond Light Source Limited (company no. 4375679). Registered in England and 
Wales with its registered office at Diamond House, Harwell Science and 
Innovation Campus, Didcot, Oxfordshire, OX11 0DE, United Kingdom
/FONT/DIV 


Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread Nave, C (Colin)
 
To bathers and non bathers

This is an interesting discussion with several relevant points. I agree
that, if small beams can pick up the best bits of the crystal that is a
very good reason for using them. The background arguments can be
relevant and having the beam size at the detector matched to the
detector resolution is a good idea to obtain the best signal above the
background for weak spots.

Several have raised the issue of radiation damage. The strategy which
Bob mentions can make sense, ensuring fresh parts of the crystal are
regularly brought in to the beam. I would have thought 5-10 micron beams
were rather too small if your crystal is several times bigger than this.
I think the microbeams can overcook the samples, particularly when
collecting large rotation ranges. The centre part of the crystal (if
centred!) is constantly in the beam in this case.

With a crystal of uniform quality, one should try and make use of the
entire volume if radiation damage is an issue. This could be by using
either using the method outlined by Bob or by bathing the whole crystal
in the beam. For large crystals, matching the beam at the crystal to the
crystal size and that at the detector to the detector resolution can
make a significant difference (see for example W. R. Wikoff, W.
Schildkamp and J. E. Johnson Acta Cryst. (2000). D56, 890-893   ,
http://scripts.iucr.org/cgi-bin/paper?S090744495941). 

We occasionally hear reports (citations needed!) of better data from
bending magnet beamlines and overcooking part of the crystal on an
undulator  may be one of the reasons. Undulator beamlines are great
especially for small crystals. However, the parallel nature of the beam
means that it is not always easy to get a big beam at the crystal and a
smaller one at the detector, though it can be done if the set up is
sufficiently flexible.

 Colin
-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Robert Sweet
Sent: 26 November 2007 00:28
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] To bathe or not to bathe.

Thanks, Ron,

Regarding the bathing question, these days the major source of error
we find in synchrotron-based data is crystal damage.  Several groups,
notably the two ID23s (one each of pairs of matched canted undulators at
ESRF and
APS) are producing small x-ray beams, on the order of 5-10 micron
diameters, and are saying that the principal use to which they're put is
to shoot a single larger crystal several times.  There's a certain
synergy here -- each new exposure is essentially a new and undamaged
tiny crystal, and the orientations of each succesive spot on the crystal
are essentially identical.  The orientations abut nearly precisely and
various data-reduction questions are simplified.  These days, where
x-tals tend to be smallish and synchrotron access is pretty available,
this seems to us to be an excellent strategy.

Regarding absorption corrections, here at the PXRR/NSLS a whole lot of
data are taken with wavelengths on the order of one Angstroem.  At these
wavelengths the absorption of a typical crystal in its mount, even one
that is, say 3M ammonium sulfate, is nearly negligible.  On the other
hand, if one tries to optimize S or I anomalous with 1.7 Angstroem
radiation, all bets are off.  Here I'm sure you would take excellent
data.

About empirical corrections: to employ Tony North's method in about 1969
I took what are formally psi scans (repeat mesurements on a
diffractometer of one reflection at increments of rotation about one
real-lattice vector) on the trypsin/soybean-trypsin-inhibitor complex.
I was using Wyckoff 
scans* to do this.  The crystal was an orthorhombic needle, and I was
rotating about the needle axis.  I observed that the rocking curve
varied from narrow to wide, with the extremes being separated by about
180 deg. 
I'd jammed the needle into a tapered capillary and it was slightly
bent(!), giving a focused beam in one direction (sharp) and an unfocused
one in the opposite (broad).  I'm told this happens these days with
frozen needles sticking out of the gob of vitreous mother liquor.

* Hal Wyckoff (the inventor of the true Inverse Beam Method**) reasoned
that there were three ways to integrate a reflection: one could rotate
the crystal through its rocking curve (the rotation method), one could
use a broad bandpass of wavelengths (the Laue method), or one could use
a widely convergent source (the Wyckoff scan).  To do this one increased
the take-off angle of the x-ray tube (Ron knows what this is, but those
of you who don't know should ask the oldest crystallographer nearby) so
that the angle subtended from the crystal would fill the rocking curve.
In principal the single largest measurement would equal the integrated
intensity.  To hedge ones bets, one did a five-point scan and summed the
top three.  This was how I could see the sharpness of the scans.

** Wyckoff also invented a diffractometer that had two opposing x-ray
generators

Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread Richard Gillilan




I just noticed this thread. I should make a few comments.

We regularly provide microbeam with and without Helium here at  
MacCHESS. Yes, there are cases in which microbeam can give you good  
diffraction on large crystals when a larger beam cannot. Just last  
week we had a user group collecting on column-shaped crystals about  
40 microns wide by 200-300 microns long. When exposed using a 150  
micron diameter beam, the diffraction patterns were a mess, producing  
multiple lattices. When they exposed the crystals using a 20 micron  
beam, they were able to find enough sweet spots with single lattice  
to give good results. I can only offer anecdotes at this time, but,  
based on user experiences so far, this is not uncommon at all.


I don't usually think of microbeam alone as a significant reduction  
in background except that one avoids hitting solvent (which is a  
major source of scatter!).  While the beam scatters less air due to  
its small size, microbeams also can have much increased flux which  
compensates for the smaller size ... so I don't think there will be  
less air scatter unless you are just using an aperture with no flux  
gain.


Regarding helium: we regularly use helium at 95K combined with a  
helium enclosure. The setup is awkward for manual sample mounting,  
but quite convenient for automounting. The main advantage here is  
that direct beam scattering with air is nearly eliminated.


How important is this effect? It depends upon how much free path  
there is between the end of the optic (slits, etc) and the beamstop.  
It also depends upon how small your sample is and how strongly it  
diffracts. In a carefully-controlled experiment using the identical  
crystal and orientation range with both nitrogen and helium, I saw a  
signal-to-noise improvement by over a factor of 3 in the 2.5 Angstrom  
range and lower. At higher resolution the benefit decreases, but  
still looks no worse than a factor of 2. This is with a 50 micron  
crystal illuminated with an 18 micron beam.


I am currently working on guidelines for when helium and microbeam  
are necessary (based on both simulations and explicit measurements).  
At the present time, my feeling is that crystals below 50 micron can  
certainly make the extra hassle worthwhile. It really depends upon  
how badly you want that extra resolution. In the case above, it  
pushed the resolution from above to below the 2.0 Angstrom mark based  
on I/SIG.


As for radiation damage, I do think it is important that a beam  
intensity profile be as flat at the top as possible (not a sharp  
hotspot) .. so defocusing a little may be useful. This is not usually  
a factor that regular users have any control over. I don't have any  
hard data on this.




Richard Gillilan
MacCHESS


Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread Juergen Bosch

Richard Gillilan wrote:




I am currently working on guidelines for when helium and microbeam  
are necessary (based on both simulations and explicit measurements).  
At the present time, my feeling is that crystals below 50 micron can  
certainly make the extra hassle worthwhile. It really depends upon  
how badly you want that extra resolution. In the case above, it  
pushed the resolution from above to below the 2.0 Angstrom mark based  
on I/SIG.



Hi Richard,

I/SigI based on which program ? Default modes or tweaked by expert ? I 
would give this particular dataset a chance to be processed by all 
available programs and then do the comparison, or actualy all the 
datasets you have with various setups. Should be a nice table comparing 
program X versus Y and Z with the given data. And if it's SeMet data 
that would even be better - runnig e.g. Shelx and demonstrating which 
setup leads to an interpretable electron density.


Juergen

--
Jürgen Bosch
University of Washington
Dept. of Biochemistry, K-426
1705 NE Pacific Street
Seattle, WA 98195
Box 357742
Phone:   +1-206-616-4510
FAX: +1-206-685-7002
Web: http://faculty.washington.edu/jbosch


Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread Nave, C (Colin)
 
Note the density of air is approximately 1000 times less than a protein
crystal. The total scatter for a beam going through a 50 micron thick
crystal will be similar to that from 50mm air. Most beamlines will have
a path length less than this but nevertheless the air scatter will be
significant with small crystals.
In principle, with smaller beams one can have smaller beamstops nearer
the sample thus reducing the path length through the air.

 Colin
-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Richard Gillilan
Sent: 26 November 2007 16:24
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] To bathe or not to bathe.



I just noticed this thread. I should make a few comments.

We regularly provide microbeam with and without Helium here at MacCHESS.
Yes, there are cases in which microbeam can give you good diffraction on
large crystals when a larger beam cannot. Just last week we had a user
group collecting on column-shaped crystals about 40 microns wide by
200-300 microns long. When exposed using a 150 micron diameter beam, the
diffraction patterns were a mess, producing multiple lattices. When they
exposed the crystals using a 20 micron beam, they were able to find
enough sweet spots with single lattice to give good results. I can
only offer anecdotes at this time, but, based on user experiences so
far, this is not uncommon at all.

I don't usually think of microbeam alone as a significant reduction in
background except that one avoids hitting solvent (which is a major
source of scatter!).  While the beam scatters less air due to its small
size, microbeams also can have much increased flux which compensates for
the smaller size ... so I don't think there will be less air scatter
unless you are just using an aperture with no flux gain.

Regarding helium: we regularly use helium at 95K combined with a helium
enclosure. The setup is awkward for manual sample mounting, but quite
convenient for automounting. The main advantage here is that direct beam
scattering with air is nearly eliminated.

How important is this effect? It depends upon how much free path there
is between the end of the optic (slits, etc) and the beamstop.  
It also depends upon how small your sample is and how strongly it
diffracts. In a carefully-controlled experiment using the identical
crystal and orientation range with both nitrogen and helium, I saw a
signal-to-noise improvement by over a factor of 3 in the 2.5 Angstrom
range and lower. At higher resolution the benefit decreases, but still
looks no worse than a factor of 2. This is with a 50 micron crystal
illuminated with an 18 micron beam.

I am currently working on guidelines for when helium and microbeam are
necessary (based on both simulations and explicit measurements).  
At the present time, my feeling is that crystals below 50 micron can
certainly make the extra hassle worthwhile. It really depends upon how
badly you want that extra resolution. In the case above, it pushed the
resolution from above to below the 2.0 Angstrom mark based on I/SIG.

As for radiation damage, I do think it is important that a beam
intensity profile be as flat at the top as possible (not a sharp
hotspot) .. so defocusing a little may be useful. This is not usually a
factor that regular users have any control over. I don't have any hard
data on this.



Richard Gillilan
MacCHESS
DIVFONT size=1 color=grayThis e-mail and any attachments may contain 
confidential, copyright and or privileged material, and are for the use of the 
intended addressee only. If you are not the intended addressee or an authorised 
recipient of the addressee please notify us of receipt by returning the e-mail 
and do not use, copy, retain, distribute or disclose the information in or 
attached to the e-mail.
Any opinions expressed within this e-mail are those of the individual and not 
necessarily of Diamond Light Source Ltd. 
Diamond Light Source Ltd. cannot guarantee that this e-mail or any attachments 
are free from viruses and we cannot accept liability for any damage which you 
may sustain as a result of software viruses which may be transmitted in or with 
the message.
Diamond Light Source Limited (company no. 4375679). Registered in England and 
Wales with its registered office at Diamond House, Harwell Science and 
Innovation Campus, Didcot, Oxfordshire, OX11 0DE, United Kingdom
/FONT/DIV 


Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread Richard Gillilan
Hi Juergen, the original calculation was done with I/SIG's from  
scala. Yes, I am aware of the problems obtaining reliable and  
meaningful I/SIG with CCD data. I have gone through the exercise of  
trying to get agreement between scala and scalepack by optimizing  
error model parameters ... though not yet with this particular  
dataset. Keep in mind that these are nearly identical datasets  
(actually the same Bragg reflections), so it is a relative  
improvement figure and not absolute I/SIG that are relevant here.


I had not thought to go as far as density comparison. That's a good  
idea. Unfortunately the datasets were kept incomplete on purpose to  
reduce possible radiation damage effects.



Richard

On Nov 26, 2007, at 11:38 AM, Juergen Bosch wrote:


Richard Gillilan wrote:




I am currently working on guidelines for when helium and  
microbeam  are necessary (based on both simulations and explicit  
measurements).  At the present time, my feeling is that crystals  
below 50 micron can  certainly make the extra hassle worthwhile.  
It really depends upon  how badly you want that extra resolution.  
In the case above, it  pushed the resolution from above to below  
the 2.0 Angstrom mark based  on I/SIG.



Hi Richard,

I/SigI based on which program ? Default modes or tweaked by  
expert ? I would give this particular dataset a chance to be  
processed by all available programs and then do the comparison, or  
actualy all the datasets you have with various setups. Should be a  
nice table comparing program X versus Y and Z with the given data.  
And if it's SeMet data that would even be better - runnig e.g.  
Shelx and demonstrating which setup leads to an interpretable  
electron density.


Juergen

--
Jürgen Bosch
University of Washington
Dept. of Biochemistry, K-426
1705 NE Pacific Street
Seattle, WA 98195
Box 357742
Phone:   +1-206-616-4510
FAX: +1-206-685-7002
Web: http://faculty.washington.edu/jbosch





Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread Richard Gillilan
In our current helium box, there is a total of about 28 mm of beam  
exposed. 10 mm from the aperture of the optic and 18 mm from sample  
to beamstop. The 10 mm side working distance is very tight for hand  
mounting (little room for tongs) and falls just outside the shield  
stream for cryo. Could probably decrease the 28 mm some, but not  
below 10 mm I think.  I think our standard setup places the  
collimator ion chamber about 20 mm from sample ... so 30-50 mm may  
not be far off the mark for typical stations. Usually, it is the  
virus crystallographers who are fussy about getting  the smallest  
beamstop. Next time, I'll have to whip out a ruler and see what  
distance makes them happy.


It would be interesting to know typical distances for other  
facilities. Anyone else made these measurements?



Richard Gillilan
MacCHESS

Note the density of air is approximately 1000 times less than a  
protein

crystal. The total scatter for a beam going through a 50 micron thick
crystal will be similar to that from 50mm air. Most beamlines will  
have

a path length less than this but nevertheless the air scatter will be
significant with small crystals.
In principle, with smaller beams one can have smaller beamstops nearer
the sample thus reducing the path length through the air.

 Colin


Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread Jacob Keller
Concerning the bathing, I think there is at least this point, which I do not 
think has been mentioned, to wit:


Considering that radiation damage is now present, I would guess, in 
virtually all synchrotron-collected datasets, and considering that probably 
this information could be used in virtually all cases to help phase the data 
or at least be corrected for, it really makes sense to bathe the crystal 
completely, as this way all of the crystal is exposed evenly (assuming 
radiation damage is not direction-specific, and assuming the beam flux is 
evenly distributed across the beam cross-section). This way, the damage 
should be a simple function of time, not confounded with various parts of 
the crystal entering or exiting the beam. Therefore, if one wants either to 
use or correct for the radiation damage, it seems obvious to me that, 
concerning this aspect, bathing is superior.


Enjoying the debate,

Jacob Keller

***
Jacob Pearson Keller
Northwestern University
Medical Scientist Training Program
Dallos Laboratory
F. Searle 1-240
2240 Campus Drive
Evanston IL 60208
lab: 847.467.4049
cel: 773.608.9185
email: [EMAIL PROTECTED]
***

- Original Message - 
From: Richard Gillilan [EMAIL PROTECTED]

To: CCP4BB@JISCMAIL.AC.UK
Sent: Monday, November 26, 2007 12:02 PM
Subject: Re: [ccp4bb] To bathe or not to bathe.


In our current helium box, there is a total of about 28 mm of beam 
exposed. 10 mm from the aperture of the optic and 18 mm from sample  to 
beamstop. The 10 mm side working distance is very tight for hand  mounting 
(little room for tongs) and falls just outside the shield  stream for 
cryo. Could probably decrease the 28 mm some, but not  below 10 mm I 
think.  I think our standard setup places the  collimator ion chamber 
about 20 mm from sample ... so 30-50 mm may  not be far off the mark for 
typical stations. Usually, it is the  virus crystallographers who are 
fussy about getting  the smallest  beamstop. Next time, I'll have to whip 
out a ruler and see what  distance makes them happy.


It would be interesting to know typical distances for other  facilities. 
Anyone else made these measurements?



Richard Gillilan
MacCHESS


Note the density of air is approximately 1000 times less than a  protein
crystal. The total scatter for a beam going through a 50 micron thick
crystal will be similar to that from 50mm air. Most beamlines will  have
a path length less than this but nevertheless the air scatter will be
significant with small crystals.
In principle, with smaller beams one can have smaller beamstops nearer
the sample thus reducing the path length through the air.

 Colin




Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread Sanishvili, Ruslan
Folks,

As expected, discussion about to bathe or not to bathe are starting to
expand into other aspects of data collection and processing. Along the
way I saw reference to how we use our small beams on GM/CA beamlines in
sector 23 of APS. Without going into discussions of advantages and
disadvantages of various beam sizes, I'd like to clarify our usage of
small beams.
Our mini-beams (5-10 micron) are not a primary tool and are being used
when larger (100x25 um) beams have had problems. These problems include
Small crystals (less than 20-30 um)
Streaky spots (from any size crystal)
Very high mosaicity
Inconsistent quality (order) throughout the crystal

When larger crystal is more or less homogeneous in its quality, larger
beams are superior to be used and allow larger diffracting volume to
work for you. This is especially true for avoiding the radiation damage
as larger volumes allow equally lower flux to be used. We collect data
from larger crystals and 5-10 um beams only when one or more of the
above problems need to be resolved.

For those who prefer bathing, the recommendation still is to keep the
beam size closer to that of the sample. Air scatter, produced by the
(part of the) beam which is not contributing to the Bragg spots, is
still noticeable at most typical energies. 
Cheers,
N.


Ruslan Sanishvili (Nukri), Ph.D.

GM/CA-CAT, Bld. 436, D007
Biosciences Division, ANL
9700 S. Cass Ave.
Argonne, IL 60439

Tel: (630)252-0665
Fax: (630)252-0667
[EMAIL PROTECTED]


Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread bas
- Original Message -
From: Nave, C (Colin) [EMAIL PROTECTED]

 Several have raised the issue of radiation damage. The strategy which
 Bob mentions can make sense, ensuring fresh parts of the crystal are
 regularly brought in to the beam. I would have thought 5-10 micron beams
 were rather too small if your crystal is several times bigger than this.
 I think the microbeams can overcook the samples, particularly when
 collecting large rotation ranges. The centre part of the crystal (if
 centred!) is constantly in the beam in this case.


On the other hand, why perfectly centre the crystal? A set-up where a
narrow beam runs past rather than through the rotation axis will spare the
central voxel entirely; apart from a somewhat more often exposed doughnut
around the axis, only fresh parts of the crystal are illuminated that way.

Sebastiaan Werten.


Re: [ccp4bb] To bathe or not to bathe.

2007-11-26 Thread James Holton

Manjasetty, Babu wrote:

Hi there,
 
1. See attached for life-time of crystals (How long will my crystal lost?) at various beamlines in the world.
2. The citations for the beauty and quality of the datasets from bending magnet beamlines. Please read most of the papers from Dauter group. 
 
Cheers,
 
-Babu


Well, it's certainly interesting to see how far my little document has 
gone!  As one might infer from the URL in the header of the PDF Babu has 
just posted, I made this document for the radiation damage session at 
ACA this year.  So, I thought I should take this moment to claim 
responsibility for it.  If anyone out there feels like this document has 
errors or omissions, please do not flame Babu about it.  Send all 
complaints and corrections to [EMAIL PROTECTED]  Send them quickly too, 
I'm getting the impression that I should publish this document properly.




As for the original question that launched this thread: I generally 
advise my users to use a beam that is the same size as their crystal.  
Bathing is important, but after you are clean, it is a waste.


IMHO, The only time you want to use a beam smaller than your crystal is 
if there are parts of your crystal you want to avoid.  An example given 
already is a needle crystal that has been bent.  It is not hard to 
imagine a bent needle crystal as a pile of plate crystals stacked along 
a curved path.  Thus the bend results in a wider mosaic spread.  Using 
a smaller beam will make the mosaic spread smaller because plates 
nearer each other in the bent needle have more similar orientations.


The only time you want to use a beam larger than your crystal is if you 
think you can live with the extra background. An over-sized beam 
obviously simplifies the crystal-centering problem as well as the 
flicker noise induced if the crystal is vibrating.  However, in a 
properly done experiment (a centered crystal that doesn't shake), 
there is no real advantage to using an over-sized beam.  Since the 
strength of diffraction spots is proportional to the crystal volume 
(roughly the third power of the linear size) as your shoot smaller and 
smaller crystals, the background will eventually kill you.  I measured 
the background on ALS 8.3.1 on an absolute scale here:

http://bl831.als.lbl.gov/~jamesh/pickup/background_scatter.png
The rule of thumb (as Colin pointed out) is that 1 mm of air is about 
equal to 1 micron of sample.  This is because the air and the sample are 
basically made of the same stuff (nitrogen) and air is ~1/1000 the 
density of a protein crystal.  As you go to higher and higher angles, 
the form factor of water, protein or most any substance will approach 
that of a single atom of that substance.  Hence, the rule of thumb 
applies in the domain of high-res spots. 



WRT helium paths:
Scattering is proportional to the square of the number of electrons, so 
(at high res) He will produce (7/2)^2 ~1/12 the background from nitrogen 
(air).  When it comes to forward scattering, N2 has 14 electrons in it, 
not 7.  So, the difference in low-res scattering from N2 vs He is a 
factor of 49.  My colleague Scott Classen has verified this experimentally:

http://bl1231.als.lbl.gov/2006/05/helium_box.php
So I would say a He box is certainly worth it, as long as your sample 
(crystal and surrounding material) is on the order of ~1/1000th the air 
path (or smaller). 

It is also important to bear in mind that not all of the background you 
see on a no-sample image is air scatter.  A significant contribution 
to background can come from the pinhole/slits, or even from stuff 
upstream of them.  The everything else line in the 
background_scatter.png graph above is generated by subtracting the 
calibrated air scatter from a no sample image.  Note the sharp spike 
near the beamstop.  I'm not sure what this is, but it is NOT air 
scatter.  It is also not x-ray fluorescence from the pinhole or 
beamstop.  I am still trying to figure out exactly what kind of physics 
produces these photons, but diffuse scatter from the pinhole (if there 
is such a thing) or SAXS from the mirrors may be culprits.  If anyone 
has any ideas about what could give you a sharp spike just around the 
beamstop, I'd like to hear it.


-James Holton
MAD Scientist


Re: [ccp4bb] To bathe or not to bathe.

2007-11-25 Thread Ronald E Stenkamp

Just a few comments on consider a crystal bathed in a uniform
beam.

I've not fully bought into the idea that it's OK to have the
beam smaller than the crystal.  I learned most of my crystallography
in a lab dedicated to precise structure determinations, and somewhere
along the line, I picked up the idea that it's better to remove 
systematic errors experimentally than to correct for them computationally.

(Maybe that had something to do with the computing power and programs
available at the time?)

Anyway, I thought the reason people went to smaller beams was that
it made it possible to resolve the spots on the film or detector. 
Isn't that the main reason for using small beams?


In practice, I guess the change in crystal volume actually diffracting
hasn't been a big issue and that frame-to-frame scaling deals with
the problem adequately.

I'm less convinced that frame-to-frame scaling can correct for 
absorption very well.  Due to our irregular-shaped protein crystals,

before the area detectors came along, we'd use an empirical correction
(one due to North comes to mind) based on rotation about the phi axis
of a four-circle goniostat.  It was clearly an approximation to the
more detailed calculations of path-lengths available for crystals with
well-defined faces not surrounded by drops of mother liquor and glass
capillaries.  Has anyone checked to see how frame-to-frame scaling
matches up with analytical determinations of absorption corrections?
It'd be interesting to determine the validity of the assumption that
absorption is simply a function of frame number.

Ron


Re: [ccp4bb] To bathe or not to bathe.

2007-11-25 Thread Jon Wright

It'd be interesting to determine the validity of the assumption that
absorption is simply a function of frame number.


... and direction. See, eg:
Acta Cryst. (1995). A51, 33-38[ doi:10.1107/S0108767394005726 ]
An empirical correction for absorption anisotropy
R. H. Blessing


Best,

Jon


Re: [ccp4bb] To bathe or not to bathe.

2007-11-25 Thread Robert Sweet

Thanks, Ron,

Regarding the bathing question, these days the major source of error we 
find in synchrotron-based data is crystal damage.  Several groups, notably 
the two ID23s (one each of pairs of matched canted undulators at ESRF and 
APS) are producing small x-ray beams, on the order of 5-10 micron 
diameters, and are saying that the principal use to which they're put is 
to shoot a single larger crystal several times.  There's a certain synergy 
here -- each new exposure is essentially a new and undamaged tiny crystal, 
and the orientations of each succesive spot on the crystal are essentially 
identical.  The orientations abut nearly precisely and various 
data-reduction questions are simplified.  These days, where x-tals tend to 
be smallish and synchrotron access is pretty available, this seems to us 
to be an excellent strategy.


Regarding absorption corrections, here at the PXRR/NSLS a whole lot of 
data are taken with wavelengths on the order of one Angstroem.  At these 
wavelengths the absorption of a typical crystal in its mount, even one 
that is, say 3M ammonium sulfate, is nearly negligible.  On the other 
hand, if one tries to optimize S or I anomalous with 1.7 Angstroem 
radiation, all bets are off.  Here I'm sure you would take excellent 
data.


About empirical corrections: to employ Tony North's method in about 1969 I 
took what are formally psi scans (repeat mesurements on a diffractometer 
of one reflection at increments of rotation about one real-lattice vector) 
on the trypsin/soybean-trypsin-inhibitor complex.  I was using Wyckoff 
scans* to do this.  The crystal was an orthorhombic needle, and I was 
rotating about the needle axis.  I observed that the rocking curve varied 
from narrow to wide, with the extremes being separated by about 180 deg. 
I'd jammed the needle into a tapered capillary and it was slightly 
bent(!), giving a focused beam in one direction (sharp) and an unfocused 
one in the opposite (broad).  I'm told this happens these days with frozen 
needles sticking out of the gob of vitreous mother liquor.


* Hal Wyckoff (the inventor of the true Inverse Beam Method**) reasoned 
that there were three ways to integrate a reflection: one could rotate 
the crystal through its rocking curve (the rotation method), one could use 
a broad bandpass of wavelengths (the Laue method), or one could use a 
widely convergent source (the Wyckoff scan).  To do this one increased the 
take-off angle of the x-ray tube (Ron knows what this is, but those of you 
who don't know should ask the oldest crystallographer nearby) so that the 
angle subtended from the crystal would fill the rocking curve.  In 
principal the single largest measurement would equal the integrated 
intensity.  To hedge ones bets, one did a five-point scan and summed the 
top three.  This was how I could see the sharpness of the scans.


** Wyckoff also invented a diffractometer that had two opposing x-ray 
generators, literally shooting at the crystal from opposite directions, 
and then two opposing detectors, each on its own 2-theta arm.  One 
detector would measure (h,k,l) at precisely the same time the other would 
measure (-h,-k,-l).  What we do now, to flip the rotation axis by 180 
degrees, seeing the (-h,-k,-l) reflections in a mirror image to the 
(h,k,l) ones, would better be termed the Friedel Flip.


Keep those cards and letters coming, folks.

Bob


On Sun, 25 Nov 2007, Ronald E Stenkamp wrote:


Just a few comments on consider a crystal bathed in a uniform
beam.

I've not fully bought into the idea that it's OK to have the
beam smaller than the crystal.  I learned most of my crystallography
in a lab dedicated to precise structure determinations, and somewhere
along the line, I picked up the idea that it's better to remove systematic 
errors experimentally than to correct for them computationally.

(Maybe that had something to do with the computing power and programs
available at the time?)

Anyway, I thought the reason people went to smaller beams was that
it made it possible to resolve the spots on the film or detector. Isn't that 
the main reason for using small beams?


In practice, I guess the change in crystal volume actually diffracting
hasn't been a big issue and that frame-to-frame scaling deals with
the problem adequately.

I'm less convinced that frame-to-frame scaling can correct for absorption 
very well.  Due to our irregular-shaped protein crystals,

before the area detectors came along, we'd use an empirical correction
(one due to North comes to mind) based on rotation about the phi axis
of a four-circle goniostat.  It was clearly an approximation to the
more detailed calculations of path-lengths available for crystals with
well-defined faces not surrounded by drops of mother liquor and glass
capillaries.  Has anyone checked to see how frame-to-frame scaling
matches up with analytical determinations of absorption corrections?
It'd be interesting to determine the validity of the 

Re: [ccp4bb] To bathe or not to bathe.

2007-11-25 Thread Ethan A Merritt
On Sunday 25 November 2007 14:43, Ronald E Stenkamp wrote:

 Just a few comments on consider a crystal bathed in a uniform beam.

 Anyway, I thought the reason people went to smaller beams was that
 it made it possible to resolve the spots on the film or detector. 
 Isn't that the main reason for using small beams?

If you mean that the projected image of the crystal onto the detector
is smaller because of a smaller beam, I think could only be relevant in
the case of truly huge crystals.  On the other hand, as mentioned earlier
in this thread, there is a possibility that a small beam will illuminate
a sweet spot on the crystal with lower mosaicity.  In that case yes,
the smaller beam may make it possible to resolve spots that would
otherwise overlap due to high mosaicity.

I think that is the strongest argument being advanced recently for
the use of micro-beam apparatus.

The other argument is that a smaller beam will generate lower background
due to air-scatter.  So for weakly diffracting crystals you want a beam
that is no bigger than the crystal, as any part of the beam that doesn't
hit the crystal contributes to the background but not to the signal.  

This is true, but if we really took the air-scatter argument seriously we
would go back to the days of huge Helium-filled enclosures to get rid of
the air scatter.  Some beamlines currently do direct He outflow from the
collimator toward the crystal, which reduces air scatter by the
indident beam, but I have not seen many beamline helium box setups to
reduce also the air scatter from the diffracted beams.

 I'm less convinced that frame-to-frame scaling can correct for 
 absorption very well.  Due to our irregular-shaped protein crystals,
 before the area detectors came along, we'd use an empirical correction
 (one due to North comes to mind) based on rotation about the phi axis
 of a four-circle goniostat.

The current scaling algorithms for area detectors do more than generate
a frame-to-frame scale. Separate correction factors are routinely
calculated for different regions of the diffraction image.
These map back onto a set of approximately equal X-ray paths through
the crystal. Furthermore, the 3D profile fitting done by some processing
programs is a logical extension of those same empirical corrections that
we did back in the 70s.

 It'd be interesting to determine the validity of the assumption that
 absorption is simply a function of frame number.

I don't think any of the current generation of programs make that
assumption.  But maybe I'm giving them too much credit?

-- 
Ethan A Merritt
Biomolecular Structure Center
University of Washington, Seattle 98195-7742


Re: [ccp4bb] To bathe or not to bathe.

2007-11-25 Thread Thomas Earnest

This is true, but if we really took the air-scatter argument seriously we
would go back to the days of huge Helium-filled enclosures to get rid of
the air scatter.  Some beamlines currently do direct He outflow from the
collimator toward the crystal, which reduces air scatter by the
indident beam, but I have not seen many beamline helium box setups to
reduce also the air scatter from the diffracted beams.
  



Reducing air scatter between the collimator and beamstop makes the most 
significant reduction due to x-ray induced scattering background.
Simply thinking, calculate the intensity times path-length before the 
beamstop and compare to the scattered beam intensity times
the diffracted path length to estimate. This would suggest air-scatter 
from the diffracted beams (even totaled up) is small.


Absorption/attenuation of the diffracted beam is an issue that can be 
addressed by the He-box, and this air scatter should be balanced
against window(s) on the He-box that also absorb. My impression is that 
there are a few extreme cases where He box improves
data quality, but that this is rarer than the number of cases where it 
is used.it would be nice to have someone perform

a systematic study of this across a number of condition cases.

IMHO mini-beams (at least those which retain a small divergence) are 
critical for small crystals and rods where otherwise multiple
crystals would be needed, and seem to make a significant difference a 
number of cases, including mosiacity scanning as was

earlier mentioned.


- Thomas

Thomas Earnest, Ph.D.
Senior Scientist and Group Leader
Structural Proteomics Development Group
Physical Biosciences Division
MS64R0121
Lawrence Berkeley National Laboratory
Berkeley CA 94720

[EMAIL PROTECTED]
510 486 4603




Ethan A Merritt wrote:

On Sunday 25 November 2007 14:43, Ronald E Stenkamp wrote:

  

Just a few comments on consider a crystal bathed in a uniform beam.



  

Anyway, I thought the reason people went to smaller beams was that
it made it possible to resolve the spots on the film or detector. 
Isn't that the main reason for using small beams?



If you mean that the projected image of the crystal onto the detector
is smaller because of a smaller beam, I think could only be relevant in
the case of truly huge crystals.  On the other hand, as mentioned earlier
in this thread, there is a possibility that a small beam will illuminate
a sweet spot on the crystal with lower mosaicity.  In that case yes,
the smaller beam may make it possible to resolve spots that would
otherwise overlap due to high mosaicity.

I think that is the strongest argument being advanced recently for
the use of micro-beam apparatus.

The other argument is that a smaller beam will generate lower background
due to air-scatter.  So for weakly diffracting crystals you want a beam
that is no bigger than the crystal, as any part of the beam that doesn't
hit the crystal contributes to the background but not to the signal.  


This is true, but if we really took the air-scatter argument seriously we
would go back to the days of huge Helium-filled enclosures to get rid of
the air scatter.  Some beamlines currently do direct He outflow from the
collimator toward the crystal, which reduces air scatter by the
indident beam, but I have not seen many beamline helium box setups to
reduce also the air scatter from the diffracted beams.

  
I'm less convinced that frame-to-frame scaling can correct for 
absorption very well.  Due to our irregular-shaped protein crystals,

before the area detectors came along, we'd use an empirical correction
(one due to North comes to mind) based on rotation about the phi axis
of a four-circle goniostat.



The current scaling algorithms for area detectors do more than generate
a frame-to-frame scale. Separate correction factors are routinely
calculated for different regions of the diffraction image.
These map back onto a set of approximately equal X-ray paths through
the crystal. Furthermore, the 3D profile fitting done by some processing
programs is a logical extension of those same empirical corrections that
we did back in the 70s.

  

It'd be interesting to determine the validity of the assumption that
absorption is simply a function of frame number.



I don't think any of the current generation of programs make that
assumption.  But maybe I'm giving them too much credit?

  


Re: [ccp4bb] To bathe or not to bathe.

2007-11-24 Thread harry powell

Hi

I'm not convinced that the first sentence here has much to do with  
the second (although both might be true).


The main reason was related to absorption. If you didn't completely  
bathe
the crystal in the xray beam, then the diffracting volume of the  
crystal
would be different during the data collection, and thus, scaling  
would be

inaccurate, especially when there was radiation damage.


Absorption can be a real problem when the path length through the  
crystal differs significantly, and is often not closely related to  
the diffracting volume - think of different paths through a flat  
plate or a needle. This is the main reason why old-fashioned  
crystallographers in days of yore used to grind their crystals into a  
sphere. Absorption problems are exacerbated when you have atoms  
present in your sample that absorb heavily (this should not be a  
surprise...).


As a former small-molecule crystallographer, I always made sure the  
crystal was bathed in the beam (so that the diffracting volume was  
the same), and usually tried to make sure the the whole crystal was  
in the central part of the beam (to try to make sure I was using the  
more uniform part of the beam). When I moved to macromolecular work,  
I found that most people seemed to prefer to get the whole beam going  
through the crystal, and not worry too much about bits of crystal  
hanging off outside the beam.


There are, of course, reasons for this, among which is that small  
molecule crystallographers are often spoilt for choice when it comes  
to picking out the right crystal, and protein crystallographers  
aren't (at least when I made the switch); they often needed all the  
crystals they could get in order to get a single dataset - this was  
in the days before cryo was standard, and room temperature data  
collection was de rigueur.


While practising small molecule crystallography, I still needed to  
apply absorption corrections to the data, especially where I had  
loads of strongly absorbing atoms (e.g. in third-row transition metal  
clusters), in order to both solve and refine the structures. The  
absorption corrections I used were based on different methods (psi- 
scans, analytical corrections based on the making precise  
measurements of the crystal itself, and the (ahem) wonderful Walker   
Stuart DIFABS (which could turn a pig's ear into a silk purse, in  
spite of what the purists might say about what it was actually doing  
to the data!)).


just my two ha'porth...


This was
especially true when you weren't sure that the crystal was well- 
centered
in the xray beam (in a cryostat, and therefore not visible). We  
typically
collected highly redundant data to help compensate for this. We  
also used
to correct for absorption by assigning Bragg indices to the crystal  
and

making precise measurements of crystal dimensions.

Scaling programs are now more extensive, and include options to  
calculate

a pseudo-absorption surface. In principle, if you have a beam that is
ALWAYS smaller than the crystal, then the same crystal volume is
illuminated by the xray beam, and will minimize scaling errors.

Bernie Santarsiero



On Fri, November 23, 2007 4:34 pm, Jim Pflugrath wrote:
It probably goes back to the days of using a single-counter  
diffractometer
where one didn't have multiple Bragg reflections on an image or  
film pack.

That is, each reflection was collected by itself.  Even in a small
molecule
crystal data collection nowadays, it would not hurt to have the  
crystal

completely bathed in the beam.

Also in the old days (let's say pre-cryo), there was plenty of  
radiation
damage going on even with a sealed-tube source.  We always  
corrected for

radiation damage by extrapolating back to zero dose in those days.

Jim

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Robert
Sweet
Sent: Friday, November 23, 2007 4:08 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] To bathe or not to bathe.

Jorge,

You said,


  I remember one former good (small molecule ?) crystallography book
with words a kind of this the crystals should be completely  
bathed by

the x-ray beam during the whole data collection  ...


The original motive for bathing the whole crystal was to assure  
that the

relative intensity of the data on each successive film pack was very
nearly constant.  This was possible (one might say necessary) in  
the old
days because the laboratory sources were very stable and the  
intensity was

low enough that there wasn't a lot of x-ray damage to the crystals.
There were a couple of other good reasons to pay attention to  
details like
this.  One was that methods for scaling images together were not  
quite as
good as now, and another was that film data were relatively very  
much less
accurate than what is achievable now with excellent detectors and  
brighter
sources.  To combat all of that, we tried to do everything  
possible to

make things better

Re: [ccp4bb] To bathe or not to bathe.

2007-11-24 Thread Juergen Bosch
One additional point to add not raised by Bob is that crystals are 
different. So you can shoot at one end of the crystal and say have a 
mosaicity of 0.2 degrees but somewhere else it might be 1.4 or even 
worse. In such cases e.g. rod like needles it pays off to have a smaller 
than crystal beam and walk over you crystal for the best spot to collect 
your dataset.


Jürgen

Robert Sweet wrote:


Jorge,

You said,

  I remember one former good (small molecule ?) crystallography book 
with words a kind of this the crystals should be completely bathed 
by the x-ray beam during the whole data collection  and also some 
other concerns about beam homogeneity in its cross section. How 
serious is this nowadays ? Can processing programs easily overcome, 
in a certain mounting, the fact that not all crystal orientations 
have the same number of unit cells exposed to x-rays ? What about 
inhomogeneities at the beam ? I understand that technical 
difficulties may lead you to exposed your crystal partially to the 
beam, etc..., but how hard should we care about this (how much effort 
to avoid this) ?



The original motive for bathing the whole crystal was to assure that 
the relative intensity of the data on each successive film pack was 
very nearly constant.  This was possible (one might say necessary) 
in the old days because the laboratory sources were very stable and 
the intensity was low enough that there wasn't a lot of x-ray damage 
to the crystals. There were a couple of other good reasons to pay 
attention to details like this.  One was that methods for scaling 
images together were not quite as good as now, and another was that 
film data were relatively very much less accurate than what is 
achievable now with excellent detectors and brighter sources.  To 
combat all of that, we tried to do everything possible to make things 
better.


These days scaling algorithms are good, the detectors are excellent, 
and very often it pays to employ a beam smaller than the x-tal.  This, 
the non-uniformity of many synchrotron beams, and the systematic 
damage to crystals that we observe now with synchrotron sources cause 
serious systematic errors.  We're forced to depend on good scaling and 
good detectors to get accurate measurements.  Making the measurements 
in many different crystal orientations (redundancy) helps to smooth 
out these systematic errors.


Nonetheless, it will always pay you to watch for EACH of these sources 
of error and to minimize them as best you can.


Bob

=
Robert M. Sweet E-Dress: [EMAIL PROTECTED]
Group Leader, PXRR: Macromolecular   ^ (that's L
  Crystallography Research Resource at NSLSnot 1)
  http://px.nsls.bnl.gov/
Biology Dept
Brookhaven Nat'l Lab.   Phones:
Upton, NY  11973631 344 3401  (Office)
U.S.A.  631 344 2741  (Facsimile)
=




--
Jürgen Bosch
University of Washington
Dept. of Biochemistry, K-426
1705 NE Pacific Street
Seattle, WA 98195
Box 357742
Phone:   +1-206-616-4510
FAX: +1-206-685-7002
Web: http://faculty.washington.edu/jbosch


Re: [ccp4bb] To bathe or not to bathe.

2007-11-23 Thread Santarsiero, Bernard D.
The main reason was related to absorption. If you didn't completely bathe
the crystal in the xray beam, then the diffracting volume of the crystal
would be different during the data collection, and thus, scaling would be
inaccurate, especially when there was radiation damage. This was
especially true when you weren't sure that the crystal was well-centered
in the xray beam (in a cryostat, and therefore not visible). We typically
collected highly redundant data to help compensate for this. We also used
to correct for absorption by assigning Bragg indices to the crystal and
making precise measurements of crystal dimensions.

Scaling programs are now more extensive, and include options to calculate
a pseudo-absorption surface. In principle, if you have a beam that is
ALWAYS smaller than the crystal, then the same crystal volume is
illuminated by the xray beam, and will minimize scaling errors.

Bernie Santarsiero



On Fri, November 23, 2007 4:34 pm, Jim Pflugrath wrote:
 It probably goes back to the days of using a single-counter diffractometer
 where one didn't have multiple Bragg reflections on an image or film pack.
 That is, each reflection was collected by itself.  Even in a small
 molecule
 crystal data collection nowadays, it would not hurt to have the crystal
 completely bathed in the beam.

 Also in the old days (let's say pre-cryo), there was plenty of radiation
 damage going on even with a sealed-tube source.  We always corrected for
 radiation damage by extrapolating back to zero dose in those days.

 Jim

 -Original Message-
 From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
 Robert
 Sweet
 Sent: Friday, November 23, 2007 4:08 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: [ccp4bb] To bathe or not to bathe.

 Jorge,

 You said,

   I remember one former good (small molecule ?) crystallography book
 with words a kind of this the crystals should be completely bathed by
 the x-ray beam during the whole data collection  ...

 The original motive for bathing the whole crystal was to assure that the
 relative intensity of the data on each successive film pack was very
 nearly constant.  This was possible (one might say necessary) in the old
 days because the laboratory sources were very stable and the intensity was
 low enough that there wasn't a lot of x-ray damage to the crystals.
 There were a couple of other good reasons to pay attention to details like
 this.  One was that methods for scaling images together were not quite as
 good as now, and another was that film data were relatively very much less
 accurate than what is achievable now with excellent detectors and brighter
 sources.  To combat all of that, we tried to do everything possible to
 make things better.

 These days scaling algorithms are good, the detectors are excellent, and
 very often it pays to employ a beam smaller than the x-tal.  This, the
 non-uniformity of many synchrotron beams, and the systematic damage
 to crystals that we observe now with synchrotron sources cause serious
 systematic errors.  We're forced to depend on good scaling and good
 detectors to get accurate measurements.  Making the measurements in many
 different crystal orientations (redundancy) helps to smooth out these
 systematic errors.

 Nonetheless, it will always pay you to watch for EACH of these sources of
 error and to minimize them as best you can.

 Bob

 =
  Robert M. Sweet E-Dress: [EMAIL PROTECTED]
  Group Leader, PXRR: Macromolecular   ^ (that's L
Crystallography Research Resource at NSLSnot 1)
http://px.nsls.bnl.gov/
  Biology Dept
  Brookhaven Nat'l Lab.   Phones:
  Upton, NY  11973631 344 3401  (Office)
  U.S.A.  631 344 2741  (Facsimile)
 =



Re: [ccp4bb] To bathe or not to bathe.

2007-11-23 Thread Jim Pflugrath
It probably goes back to the days of using a single-counter diffractometer
where one didn't have multiple Bragg reflections on an image or film pack.
That is, each reflection was collected by itself.  Even in a small molecule
crystal data collection nowadays, it would not hurt to have the crystal
completely bathed in the beam.

Also in the old days (let's say pre-cryo), there was plenty of radiation
damage going on even with a sealed-tube source.  We always corrected for
radiation damage by extrapolating back to zero dose in those days.

Jim

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Robert
Sweet
Sent: Friday, November 23, 2007 4:08 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] To bathe or not to bathe.

Jorge,

You said,

   I remember one former good (small molecule ?) crystallography book 
 with words a kind of this the crystals should be completely bathed by 
 the x-ray beam during the whole data collection  ...

The original motive for bathing the whole crystal was to assure that the 
relative intensity of the data on each successive film pack was very 
nearly constant.  This was possible (one might say necessary) in the old 
days because the laboratory sources were very stable and the intensity was 
low enough that there wasn't a lot of x-ray damage to the crystals. 
There were a couple of other good reasons to pay attention to details like 
this.  One was that methods for scaling images together were not quite as 
good as now, and another was that film data were relatively very much less 
accurate than what is achievable now with excellent detectors and brighter 
sources.  To combat all of that, we tried to do everything possible to 
make things better.

These days scaling algorithms are good, the detectors are excellent, and 
very often it pays to employ a beam smaller than the x-tal.  This, the 
non-uniformity of many synchrotron beams, and the systematic damage 
to crystals that we observe now with synchrotron sources cause serious 
systematic errors.  We're forced to depend on good scaling and good 
detectors to get accurate measurements.  Making the measurements in many 
different crystal orientations (redundancy) helps to smooth out these 
systematic errors.

Nonetheless, it will always pay you to watch for EACH of these sources of 
error and to minimize them as best you can.

Bob

=
 Robert M. Sweet E-Dress: [EMAIL PROTECTED]
 Group Leader, PXRR: Macromolecular   ^ (that's L
   Crystallography Research Resource at NSLSnot 1)
   http://px.nsls.bnl.gov/
 Biology Dept
 Brookhaven Nat'l Lab.   Phones:
 Upton, NY  11973631 344 3401  (Office)
 U.S.A.  631 344 2741  (Facsimile)
=