Re: UL544 Leakage Limits Patient Equipment*

2000-03-29 Thread Robert Johnson

It may not be important whether you call it an applied part or not. You are not
making a direct electrical connection to the patient. In any case you should
meet the leakage requirement. This should be easy with properly insulated
earphones.
IEC 60950 also addresses this issue and throws in more considerations. For the
condition of telephone type headsets they are worried about transients from
telecom or power lines, and therefore have additional isolation requirements.
Clause 6.2.2 requires an impulse test of 2.5 KV or hipot of 1.5 KV (in Australia
this is increased to both 7KV impulse and 3KV hipot) applied to foil wrapped
around the headset.
Bob Johnson

Kenneth McCormick wrote:

 I also agree that the headphones are a patient connection.

 Hi,

 Its been a while since I looked at UL 544, but a few comments.

 1.  Headphones are an applied part (UL2601-1) or patient connection
 (UL544).  They come into direct contact with the patient.  You can argue
 that they are an ordinary patient connection (Clause 2.18 of UL544).  But,
 it is still a patient connection.  The limit is 50 uA for ordinary patient
 connection.

 2.  How long is this product going to be in the market?  UL 544 goes
 away on 1/1/2003 for new products and 1/1/2005 for all products.  You might
 be better off going to UL2601-1 now.  In UL2601-1 this type of patient
 connect is Type BF.  The patient leakage current limit is 100 uA in normal
 condition and 500 uA in single fault condition.

 3.  Call UL.  Maybe I missed some out in UL544.

 Ned Devine
 Entela, Inc.
 Program Manager III
 Phone 616 248 9671
 Fax  616 574 9752
 e-mail  ndev...@entela.com

 -Original Message-
 From: me...@aol.com [mailto:me...@aol.com]
 Sent: Friday, March 24, 2000 3:36 PM
 To: jjuh...@fiberoptions.com; m.r...@ieee.org;
 emc-p...@majordomo.ieee.org
 Subject: UL544 Leakage Limits Patient Equipment*

 Most esteemed colleagues, we are looking for your professional opinion on a
 UL 544 leakage limit (I think you will find this interesting):

 1. This particular product uses a UL 544 evaluated direct plug in power
 supply with outputs to the patient care equipment. In this case it is a
 diagnostic unit that sends an audible tone to headphones (audiometer). The
 plug in power supply Conditions of Acceptability indicate the outputs are
 not evaluated for patient leads (i.e. applied parts).

 2. Table 42.1 of UL 544 specifies leakage limits. patient connection
 footnote a references testing of patient leads (applied parts)
 connections. There is no written definition for patient leads or applied
 parts in UL544.

 As such NFPA 99 supplements UL 544 as it draws from the NEC and NFPA 99
 (referenced in UL 544):

 NFPA 99 defines the US definition of  Patient Lead = A deliberate
 electrical connection that can carry current between an appliance and
 patient. It is not intended to include adventitious or casual contacts such
 as a push button, bed surface, lamp, hand held appliance, etc.

 3. As the headphones of this audiometer are clearly not deliberate
 electrical connections we conclude these are not patient leads (applied
 parts) which would not fall under the limits for patient connection
 limits
 per 544. The applicable limits would be as defined under enclosure or
 chassis grounded  or double insulated Now be careful not to jump to

 a conclusion yet. You might say enclosure or chassis?, but if you
 examine this, you will find the footnotes reference UL 544's Enclosure
 definition:

 Enclosure =  That external portion of an appliance that serves to house or
 support component parts, or both. Enclosure of patient care equipment likely

 to be contacted by a patient include, for example, bedside monitors, bed
 frames, dental chairs, and examination stands.

 Our conclusion: Due to the US definition of patient leads (applied parts),

 the earphones of an audiometer (patient care equipment) are subjected to the

 leakage current limits for enclosure or chassis, and not the limits of
 patient connection. For this particular application, we conclude that
 based
 on the C of As, the output of the power supply has already been evaluated
 for
 enclosure or chassis leakage limits.

 Your Thoughts???

 Drew

 PS: If you care to look, CSA supports this position in that 50uA is related
 to cardiac tissue limits only. See Appendix A of CSA 22.2 125 (500uA).

 ---
 This message is from the IEEE EMC Society Product Safety
 Technical Committee emc-pstc discussion list.

 To cancel your subscription, send mail to:
  majord...@ieee.org
 with the single line:
  unsubscribe emc-pstc

 For help, send mail to the list administrators:
  Jim Bacher:  jim_bac...@mail.monarch.com
  Michael Garretson:pstc_ad...@garretson.org

 For policy questions, send mail to:
  Richard Nute:   ri...@ieee.org

 ---
 This message is from the IEEE EMC

RE: UL544 Leakage Limits Patient Equipment*

2000-03-27 Thread Kenneth McCormick


I also agree that the headphones are a patient connection.

Hi,

Its been a while since I looked at UL 544, but a few comments.

1.  Headphones are an applied part (UL2601-1) or patient connection
(UL544).  They come into direct contact with the patient.  You can argue
that they are an ordinary patient connection (Clause 2.18 of UL544).  But,
it is still a patient connection.  The limit is 50 uA for ordinary patient
connection.

2.  How long is this product going to be in the market?  UL 544 goes
away on 1/1/2003 for new products and 1/1/2005 for all products.  You might
be better off going to UL2601-1 now.  In UL2601-1 this type of patient
connect is Type BF.  The patient leakage current limit is 100 uA in normal
condition and 500 uA in single fault condition.

3.  Call UL.  Maybe I missed some out in UL544.



Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax  616 574 9752
e-mail  ndev...@entela.com



-Original Message-
From: me...@aol.com [mailto:me...@aol.com]
Sent: Friday, March 24, 2000 3:36 PM
To: jjuh...@fiberoptions.com; m.r...@ieee.org;
emc-p...@majordomo.ieee.org
Subject: UL544 Leakage Limits Patient Equipment*



Most esteemed colleagues, we are looking for your professional opinion on a
UL 544 leakage limit (I think you will find this interesting):

1. This particular product uses a UL 544 evaluated direct plug in power
supply with outputs to the patient care equipment. In this case it is a
diagnostic unit that sends an audible tone to headphones (audiometer). The
plug in power supply Conditions of Acceptability indicate the outputs are
not evaluated for patient leads (i.e. applied parts).

2. Table 42.1 of UL 544 specifies leakage limits. patient connection
footnote a references testing of patient leads (applied parts)
connections. There is no written definition for patient leads or applied
parts in UL544.

As such NFPA 99 supplements UL 544 as it draws from the NEC and NFPA 99
(referenced in UL 544):

NFPA 99 defines the US definition of  Patient Lead = A deliberate
electrical connection that can carry current between an appliance and
patient. It is not intended to include adventitious or casual contacts such
as a push button, bed surface, lamp, hand held appliance, etc.

3. As the headphones of this audiometer are clearly not deliberate
electrical connections we conclude these are not patient leads (applied
parts) which would not fall under the limits for patient connection
limits
per 544. The applicable limits would be as defined under enclosure or
chassis grounded  or double insulated Now be careful not to jump to

a conclusion yet. You might say enclosure or chassis?, but if you
examine this, you will find the footnotes reference UL 544's Enclosure
definition:

Enclosure =  That external portion of an appliance that serves to house or
support component parts, or both. Enclosure of patient care equipment likely

to be contacted by a patient include, for example, bedside monitors, bed
frames, dental chairs, and examination stands.

Our conclusion: Due to the US definition of patient leads (applied parts),

the earphones of an audiometer (patient care equipment) are subjected to the

leakage current limits for enclosure or chassis, and not the limits of
patient connection. For this particular application, we conclude that
based
on the C of As, the output of the power supply has already been evaluated
for
enclosure or chassis leakage limits.

Your Thoughts???

Drew

PS: If you care to look, CSA supports this position in that 50uA is related
to cardiac tissue limits only. See Appendix A of CSA 22.2 125 (500uA).

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
majord...@ieee.org
with the single line:
unsubscribe emc-pstc

For help, send mail to the list administrators:
Jim Bacher:  jim_bac...@mail.monarch.com
Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
Richard Nute:   ri...@ieee.org


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
majord...@ieee.org
with the single line:
unsubscribe emc-pstc

For help, send mail to the list administrators:
Jim Bacher:  jim_bac...@mail.monarch.com
Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
Richard Nute:   ri...@ieee.org

__
Get Your Private, Free Email at http://www.hotmail.com


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
majord...@ieee.org
with the single line

RE: UL544 Leakage Limits Patient Equipment

2000-03-27 Thread Dick Grobner

Good Day - First, why are you using UL544? Why not use UL2601, based on
IEC601-1 with US deviations? Beware, UL544  UL187 may be used until Jan 1
2003, after that use of UL2601 will be mandatory. Also, As of Jan 1 2005,
all Listed, Classified and Recognized medical and dental products, where
UL544 or UL187 was used to evaluate the product, must comply with UL2601-1
(direct quote from UL2601).

This may shed some light - definitions from UL2601 

Applied Part: Entirety of all parts of equipment including the patient leads
which come intentionally into contact with the patient to be examined or
treated. For some equipment, particular standards may consider parts in
contact with the operator as an applied part.

Enclosure: Exterior surface of equipment including:
- all accessible metal parts, knobs, grips and the like
- accessible shafts
- for the purpose of tests, metal foil, with specified dimensions, applied
in contact with parts of the exterior surface made of material with low
conductivity or made of insulating material (i.e. insulated / double
insulated enclosures)

Patient Circuit: Electrical circuit of which the patient forms a part. 

F-Type applied part: Applied part isolated from all other parts of the
equipment to such a degree that the patient leakage current allowable in
single fault condition is not exceeded when a voltage equal to 1.1 times the
highest rated mains voltage is applied between the applied part and earth.

Type B equipment: equipment providing a particular degree of protection
against electric shock, particularly regarding: allowable leakage current,
reliability of the protective earth connection (if present)

Type BF equipment: Type B equipment with an F-Type applied part.

Allowable leakage currents:
Enclosure Leakage: N.C. (normal condition) .1 milliamps, S.F.C. (single
fault condition) .5 milliamps
Same limits for Type B or BF equipment
Patient Leakage: Same as Enclosure Leakage for both Type B or BF equip.
(Taken from Table IV of UL2601, other leakage currents of concern are Earth
Leakage, Patient Aux Current, Patient Leakage Current with Mains Voltage
Applied)

With regards to your equipment, depending on your design, I would say it is
either a Type B or Type BF. I would lean towards the Type B. 
It would be VERY wise to confer with the engineer at your chosen NRTL
(domestic National Recognized Testing Lab). These are the people you need to
convince on the classification of your product and your chosen route to
compliance.

I hope this didn't confuse you any more, I know it can be and that's why I
do not hesitate to contact the engineers at UL from time to time. I have
found the ones I deal with helpful and very understanding and easy to work
with (this is not intended to promote UL, only my opinion of them).   

-Original Message-
From: me...@aol.com [mailto:me...@aol.com]
Sent: Friday, March 24, 2000 2:36 PM
To: jjuh...@fiberoptions.com; m.r...@ieee.org;
emc-p...@majordomo.ieee.org
Subject: UL544 Leakage Limits Patient Equipment*



Most esteemed colleagues, we are looking for your professional opinion on a 
UL 544 leakage limit (I think you will find this interesting):

1. This particular product uses a UL 544 evaluated direct plug in power 
supply with outputs to the patient care equipment. In this case it is a 
diagnostic unit that sends an audible tone to headphones (audiometer). The 
plug in power supply Conditions of Acceptability indicate the outputs are 
not evaluated for patient leads (i.e. applied parts).

2. Table 42.1 of UL 544 specifies leakage limits. patient connection 
footnote a references testing of patient leads (applied parts)  
connections. There is no written definition for patient leads or applied 
parts in UL544. 

As such NFPA 99 supplements UL 544 as it draws from the NEC and NFPA 99 
(referenced in UL 544):

NFPA 99 defines the US definition of  Patient Lead = A deliberate 
electrical connection that can carry current between an appliance and 
patient. It is not intended to include adventitious or casual contacts such 
as a push button, bed surface, lamp, hand held appliance, etc.

3. As the headphones of this audiometer are clearly not deliberate 
electrical connections we conclude these are not patient leads (applied 
parts) which would not fall under the limits for patient connection
limits 
per 544. The applicable limits would be as defined under enclosure or 
chassis grounded  or double insulated Now be careful not to jump to

a conclusion yet. You might say enclosure or chassis?, but if you 
examine this, you will find the footnotes reference UL 544's Enclosure 
definition:

Enclosure =  That external portion of an appliance that serves to house or 
support component parts, or both. Enclosure of patient care equipment likely

to be contacted by a patient include, for example, bedside monitors, bed 
frames, dental chairs, and examination stands.

Our conclusion: Due to the US definition of patient leads (applied parts

RE: UL544 Leakage Limits Patient Equipment*

2000-03-27 Thread Ned Devine

Hi,

Its been a while since I looked at UL 544, but a few comments.

1.  Headphones are an applied part (UL2601-1) or patient connection
(UL544).  They come into direct contact with the patient.  You can argue
that they are an ordinary patient connection (Clause 2.18 of UL544).  But,
it is still a patient connection.  The limit is 50 uA for ordinary patient
connection.

2.  How long is this product going to be in the market?  UL 544 goes
away on 1/1/2003 for new products and 1/1/2005 for all products.  You might
be better off going to UL2601-1 now.  In UL2601-1 this type of patient
connect is Type BF.  The patient leakage current limit is 100 uA in normal
condition and 500 uA in single fault condition.

3.  Call UL.  Maybe I missed some out in UL544.  



Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax  616 574 9752
e-mail  ndev...@entela.com 



-Original Message-
From: me...@aol.com [mailto:me...@aol.com]
Sent: Friday, March 24, 2000 3:36 PM
To: jjuh...@fiberoptions.com; m.r...@ieee.org;
emc-p...@majordomo.ieee.org
Subject: UL544 Leakage Limits Patient Equipment*



Most esteemed colleagues, we are looking for your professional opinion on a 
UL 544 leakage limit (I think you will find this interesting):

1. This particular product uses a UL 544 evaluated direct plug in power 
supply with outputs to the patient care equipment. In this case it is a 
diagnostic unit that sends an audible tone to headphones (audiometer). The 
plug in power supply Conditions of Acceptability indicate the outputs are 
not evaluated for patient leads (i.e. applied parts).

2. Table 42.1 of UL 544 specifies leakage limits. patient connection 
footnote a references testing of patient leads (applied parts)  
connections. There is no written definition for patient leads or applied 
parts in UL544. 

As such NFPA 99 supplements UL 544 as it draws from the NEC and NFPA 99 
(referenced in UL 544):

NFPA 99 defines the US definition of  Patient Lead = A deliberate 
electrical connection that can carry current between an appliance and 
patient. It is not intended to include adventitious or casual contacts such 
as a push button, bed surface, lamp, hand held appliance, etc.

3. As the headphones of this audiometer are clearly not deliberate 
electrical connections we conclude these are not patient leads (applied 
parts) which would not fall under the limits for patient connection
limits 
per 544. The applicable limits would be as defined under enclosure or 
chassis grounded  or double insulated Now be careful not to jump to

a conclusion yet. You might say enclosure or chassis?, but if you 
examine this, you will find the footnotes reference UL 544's Enclosure 
definition:

Enclosure =  That external portion of an appliance that serves to house or 
support component parts, or both. Enclosure of patient care equipment likely

to be contacted by a patient include, for example, bedside monitors, bed 
frames, dental chairs, and examination stands.

Our conclusion: Due to the US definition of patient leads (applied parts),

the earphones of an audiometer (patient care equipment) are subjected to the

leakage current limits for enclosure or chassis, and not the limits of 
patient connection. For this particular application, we conclude that
based 
on the C of As, the output of the power supply has already been evaluated
for 
enclosure or chassis leakage limits.

Your Thoughts???

Drew

PS: If you care to look, CSA supports this position in that 50uA is related 
to cardiac tissue limits only. See Appendix A of CSA 22.2 125 (500uA). 

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



Re: UL544 Leakage Limits Patient Equipment*

2000-03-27 Thread Peter Merguerian

Yes, I also belive that your equipment has no patient leads and that the
enclosure leakage was performed as part of the power supply. If in doubt
regarding the enclosure leakage, best is to ask the power supply
manufacturer for a complete report.

Why are you using UL544 power supply? This standard is becoming obsolete
soon! There are many power supplies out there evaluated to UL2601-1 which
is harmonized with IEC 60601-1. 

Best Regards


At 15:35 24/03/2000 EST, me...@aol.com wrote:

Most esteemed colleagues, we are looking for your professional opinion on a 
UL 544 leakage limit (I think you will find this interesting):

1. This particular product uses a UL 544 evaluated direct plug in power 
supply with outputs to the patient care equipment. In this case it is a 
diagnostic unit that sends an audible tone to headphones (audiometer). The 
plug in power supply Conditions of Acceptability indicate the outputs are 
not evaluated for patient leads (i.e. applied parts).

2. Table 42.1 of UL 544 specifies leakage limits. patient connection 
footnote a references testing of patient leads (applied parts)  
connections. There is no written definition for patient leads or applied 
parts in UL544. 

As such NFPA 99 supplements UL 544 as it draws from the NEC and NFPA 99 
(referenced in UL 544):

NFPA 99 defines the US definition of  Patient Lead = A deliberate 
electrical connection that can carry current between an appliance and 
patient. It is not intended to include adventitious or casual contacts such 
as a push button, bed surface, lamp, hand held appliance, etc.

3. As the headphones of this audiometer are clearly not deliberate 
electrical connections we conclude these are not patient leads (applied 
parts) which would not fall under the limits for patient connection
limits 
per 544. The applicable limits would be as defined under enclosure or 
chassis grounded  or double insulated Now be careful not to jump to 
a conclusion yet. You might say enclosure or chassis?, but if you 
examine this, you will find the footnotes reference UL 544's Enclosure 
definition:

Enclosure =  That external portion of an appliance that serves to house or 
support component parts, or both. Enclosure of patient care equipment likely 
to be contacted by a patient include, for example, bedside monitors, bed 
frames, dental chairs, and examination stands.

Our conclusion: Due to the US definition of patient leads (applied parts), 
the earphones of an audiometer (patient care equipment) are subjected to the 
leakage current limits for enclosure or chassis, and not the limits of 
patient connection. For this particular application, we conclude that
based 
on the C of As, the output of the power supply has already been evaluated
for 
enclosure or chassis leakage limits.

Your Thoughts???

Drew

PS: If you care to look, CSA supports this position in that 50uA is related 
to cardiac tissue limits only. See Appendix A of CSA 22.2 125 (500uA). 

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



Peter Merguerian
Managing Director
Product Testing Division
I.T.L. (Product Testing) Ltd.
Hacharoshet 26, POB 211
Or Yehuda 60251, Israel

Tel: 972-3-5339022 Fax: 972-3-5339019
e-mail: pmerguer...@itl.co.il
website: http://www.itl.co.il 






---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



UL544 Leakage Limits Patient Equipment*

2000-03-24 Thread Metse

Most esteemed colleagues, we are looking for your professional opinion on a 
UL 544 leakage limit (I think you will find this interesting):

1. This particular product uses a UL 544 evaluated direct plug in power 
supply with outputs to the patient care equipment. In this case it is a 
diagnostic unit that sends an audible tone to headphones (audiometer). The 
plug in power supply Conditions of Acceptability indicate the outputs are 
not evaluated for patient leads (i.e. applied parts).

2. Table 42.1 of UL 544 specifies leakage limits. patient connection 
footnote a references testing of patient leads (applied parts)  
connections. There is no written definition for patient leads or applied 
parts in UL544. 

As such NFPA 99 supplements UL 544 as it draws from the NEC and NFPA 99 
(referenced in UL 544):

NFPA 99 defines the US definition of  Patient Lead = A deliberate 
electrical connection that can carry current between an appliance and 
patient. It is not intended to include adventitious or casual contacts such 
as a push button, bed surface, lamp, hand held appliance, etc.

3. As the headphones of this audiometer are clearly not deliberate 
electrical connections we conclude these are not patient leads (applied 
parts) which would not fall under the limits for patient connection limits 
per 544. The applicable limits would be as defined under enclosure or 
chassis grounded  or double insulated Now be careful not to jump to 
a conclusion yet. You might say enclosure or chassis?, but if you 
examine this, you will find the footnotes reference UL 544's Enclosure 
definition:

Enclosure =  That external portion of an appliance that serves to house or 
support component parts, or both. Enclosure of patient care equipment likely 
to be contacted by a patient include, for example, bedside monitors, bed 
frames, dental chairs, and examination stands.

Our conclusion: Due to the US definition of patient leads (applied parts), 
the earphones of an audiometer (patient care equipment) are subjected to the 
leakage current limits for enclosure or chassis, and not the limits of 
patient connection. For this particular application, we conclude that based 
on the C of As, the output of the power supply has already been evaluated for 
enclosure or chassis leakage limits.

Your Thoughts???

Drew

PS: If you care to look, CSA supports this position in that 50uA is related 
to cardiac tissue limits only. See Appendix A of CSA 22.2 125 (500uA). 

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org