https://www.febo.com/pipermail/time-nuts/2017-March/104374.html
You bring up an interesting point. As the cesium is used up over the
years, it forms a cesium frost which deposits on everything inside the
tube. If it gets thick enough it conducts or can arc between frost
grains. A charged
It looks like that there is about 10% hysteresis on the cesium trip
off/on. That may not be enough to prevent cycling on and off. I may
not have made it clear but instability in the +3,500 voltage makes a
big difference in the threshold ion current required for activation.
If it fades it can
Hi
Some quick hints:
1) You need a way to digitize the phase input with adequate resolution. If you
have a 1 second period and want
1 ns, you need a way to digitize at a 1:1,000,000,000 sort of level. That’s in
the 30 bit range so a simple ADC
isn’t going to do it alone.
2) You need a way to
Hi
Slipped a factor of 1,000 on the DAC … sorry about that …
Bob
> On Mar 21, 2017, at 7:25 PM, James Peroulas wrote:
>
> Thanks for the hints and references everyone. I'll dig in and possibly come
> back with some more questions.
>
> BR,
> James
>
So it seems the threat of replacement and a bunch of plugging and
replugging into different computers has spurred my recalcitrant GPSDO
back into life and it is now producing output again... This leaves me
with the original problem of low sensitivity but at least I can solve
that problem by using
Hi
Ummm … e …. it’s a gas cell standard. I’d bet there is a pressure effect.
Bob
> On Mar 21, 2017, at 7:01 PM, Scott McGrath wrote:
>
> Noted
>
> However CSAC not subject to barometric effects as Rb units are
>
> Content by Scott
> Typos by Siri
>
>> On Mar 21,
scmcgr...@gmail.com said:
> However CSAC not subject to barometric effects as Rb units are
Does anybody tried to measure CSAC vs pressure?
--
These are my opinions. I hate spam.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe,
That's some very nice work, Donald. Looking back, I have junked one or two Cs
tubes that might have been usable if I'd thought through the problem of high
ion pump current as you and KB7APQ seem to have done.
Another good reason to raise the lockout threshold would be to cut down on the
Thanks for the hints and references everyone. I'll dig in and possibly come
back with some more questions.
BR,
James
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and
https://www.febo.com/pipermail/time-nuts/2017-March/104374.html
Paul:
I don't think that I made myself clear. This power supply is a rare
example of HP engineering incompetence. The transformer is EASILY
capable of 10 Watts out at 3500 V. Instead they made it where it
can't even put out 100
> https://goo.gl/photos/JZhBbFKFzkBAykti6
> Why would a GPS module produce jitter with a pattern like this?
Trent, I must admit I have not seen such a four hour "spike" before in
the NMEA latency, however a clue may be that the GPS ephemeris
(broadcast by the SV's) orbit description is valid only
Scott McGrath wrote:
> Or perhaps use the Symmetricom CSAC ...
Chris Albertson wrote:
> Get a weather balloon. Or there might already be an amateur group that
> launches these. Balloons can go much higher than your local mountains.
You'll both be interested to hear that CSAC+balloon was
Noted
However CSAC not subject to barometric effects as Rb units are
Content by Scott
Typos by Siri
> On Mar 21, 2017, at 4:18 PM, jimlux wrote:
>
>> On 3/21/17 12:51 PM, Scott McGrath wrote:
>> Or perhaps use the Symmetricom CSAC ...
>>
>> Relatively expensive but
HI
> On Mar 21, 2017, at 4:52 PM, Trent Piepho wrote:
>
> Thanks to all who responded. Yes, I know PPS is the way get a more
> accurate timestamps. That is the plan, but it takes more time to write
> FPGA programs. The surprise is not that there is considerable
On 3/21/17 1:40 PM, Michael Wouters wrote:
These are less stable than a rubidium eg tau=10e-11@1000s and monthly
ageing of 9e-10.
The price of these has gone up too- they're now about US5000.
Really? That's a big increase. I bought some last year (well, in
December 2015) and they were
Hi Guys,
Does anyone have a multi-day (at least 5 days) timelab file comparing the 1PPS
from a GPSDO to a Cesium standard, with a run of the mill rooftop antenna using
at least an HP 5370A quality TIC? I'm running a 7 day test like this, and I'd
like to have something to compare to. If you
These are less stable than a rubidium eg tau=10e-11@1000s and monthly
ageing of 9e-10.
The price of these has gone up too- they're now about US5000.
Cheers
Michael
On Wed, 22 Mar 2017 at 7:03 am, Scott McGrath wrote:
> Or perhaps use the Symmetricom CSAC ...
>
>
On 3/21/17 12:51 PM, Scott McGrath wrote:
Or perhaps use the Symmetricom CSAC ...
Relatively expensive but might work
The CSAC is 8E-12 AVAR at 1000 seconds, comparable to a Rb.
See also http://tycho.usno.navy.mil/ptti/2011papers/Paper27.pdf
which shows a bit better performance (3E-12 @
Or perhaps use the Symmetricom CSAC ...
Relatively expensive but might work
> On Mar 21, 2017, at 8:08 AM, Attila Kinali wrote:
>
> On Tue, 21 Mar 2017 13:38:51 +1100
> Hugh Blemings wrote:
>
>> This got me to wondering if a Rubidium based standard might
Hi Scott,
That's a nice project.
Combining quartz and pendulum like that is essentially how a GPSDO works. In
your case, instead of a 10 MHz oscillator you have a 1 MHz oscillator and
instead of 1PPS you have 1/2 PPS. Whether you use an analog loop or a digital
loop there are dozens of
Look at ICD-GPS-060.
10V into 50ohm, sub 50ns risetime and 20us pulse lenght is specified in figure
3-2.
-- Björn
Sent from my smartphone.
Original message From: Hal Murray
Date: 21/03/2017 19:18 (GMT+01:00) To: Tom Van Baak
Roland Best, "Phase Locked Loops". I see from Amazon that it is up the
6th edition. I think I have the 3rd edition.
https://www.amazon.com/Phase-Locked-Loops-Simulation-Applications/dp/0071493751/ref=sr_1_1?ie=UTF8=1490121310=8-1=best+phase+locked+loops
I took a PLL class at StorageTek a while
On Tue, 21 Mar 2017 11:19:22 -0400
paul swed wrote:
> Have to say a search for PLL book by Best generates lots of stuff. None of
> it about PLLs. Better hint please? I think Lyons may be a bit better.
t...@leapsecond.com said:
> The TAPR dividers tend not to have "this problem" because they output at
> wimpy TTL/CMOS levels.
Modern CMOS drivers have fast rise times. As long as the rise time is short
relative to the cable length, it gets doubled if the end of the cable is an
open circuit.
If you cannot get the ion current below 50ua or so after a week at 5kV
then you are out of luck.
Most likely you have resistive deposits on the ion pump insulator.
If you can get the tube to give a decent SN at those levels then you can
run with the alarm circuit bypassed.
I have run tubes a
Have to say a search for PLL book by Best generates lots of stuff. None of
it about PLLs. Better hint please? I think Lyons may be a bit better.
Regards
Paul
WB8TSL
On Tue, Mar 21, 2017 at 7:39 AM, Attila Kinali wrote:
> On Mon, 20 Mar 2017 21:32:28 -0700
> James Peroulas
Get a weather balloon. Or there might already be an amateur group that
launches these. Balloons can go much higher than your local mountains.
You al ill want to build an environmental chamber for the Rb clock. The
chamber is heated and pressurized.
Even for the maintain top experiment you will
it is tempting to use digital loop filter for PLL if a very low
bandwidth is required, but with the large time constant is a problem ;
the VCO will change it's frequency despite the tuning voltage remains
constant and that is not so simple to model.
73
KJ6UHN
Alex
On 3/21/2017 4:39 AM,
To get your loop to lock and keep phase noise down the loop filter would need a
bandwidth of .05 Hz or less. That would mean long lock times. Very long lock
times.
Engineering is the art of making what you want from what you can get at a
profit.
I like Polywell Fusion.
On Tuesday,
I know the the open load output of some instrument is 10Vpp and I think this
is right because if we want to connect this output on a 50Ohm line is correct
to close the cable with the proper load impedance.
I have found this level also on some trak System equipment.
All my test are
Hi
> On Mar 21, 2017, at 1:37 AM, Bruce Lane wrote:
>
> Fellow time-tickers,
>
> My thanks to Greg for getting back to me (and apologies for the spam
> filter bounce) but, apparently, MTI 240 OCXO's have become unobtanium.
>
> With this in mind: What
On Tue, 21 Mar 2017 13:38:51 +1100
Hugh Blemings wrote:
> This got me to wondering if a Rubidium based standard might do the trick
> - the Efratom SLCR-101s seem readily available for ~USD$200 mark.
As TvB wrote, a single one will not do the trick. You will need
a stability
H
> On Mar 21, 2017, at 4:58 AM, Tom Van Baak wrote:
>
> Hi Hugh,
>
>> If I do the math correctly that's about 14ns difference per 24h the
>> clocks are separated by that altitude. [1]
>
> That's correct. For your 1500m elevation gain, the gravitational redshift,
> the
Moin,
On Mon, 20 Mar 2017 21:07:03 -0400
David Scott Coburn wrote:
> I have built and tested a PLL circuit that will be used to generate a 1 MHz
> signal locked to a 0.5 HZ signal from a pendulum. (Details available upon
> request.)
[...]
> I tested this by feeding the
Hi
> On Mar 21, 2017, at 1:07 AM, Chris Albertson
> wrote:
>
> I built one of these using a PWM DAC also. The design was posted to this
> list so I can't take credit for the idea. But we used two PWM output
> pins. The PWM provides more voltage range than is
On Mon, 20 Mar 2017 21:32:28 -0700
James Peroulas wrote:
> I'm trying to understand how to design and analyze the loop filters in a
> digital PLL. Specifically, because of digital processing delays, the phase
> offset measured at time t will only produce a change on the VCXO
Andy, Bill, et al.
Attached is a GIF version of Scott's (unreadable?) histogram-utcday21613x.pdf
file.
/tvb
- Original Message -
From: "Andy"
To: "Discussion of precise time and frequency measurement"
Sent: Tuesday, March 21, 2017 12:23 AM
Hi Hugh,
> If I do the math correctly that's about 14ns difference per 24h the
> clocks are separated by that altitude. [1]
That's correct. For your 1500m elevation gain, the gravitational redshift, the
df/f frequency change, will be about 1.6e-13. To be able to measure with any
confidence
Second file successfully opened in Irfanview.
Three other PDF readers, including Adobe, could not open it.
Andy
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow
Fellow time-tickers,
My thanks to Greg for getting back to me (and apologies for the spam
filter bounce) but, apparently, MTI 240 OCXO's have become unobtanium.
With this in mind: What other OCXO's are compatible with the
Symmetricom TS2100? My understanding is whatever I choose
In message
, James Peroulas writes:
>I've found plenty of texts describing analog loop filters. Are there any
>recommendations for digital loop filter PLL design?
This may help:
I built one of these using a PWM DAC also. The design was posted to this
list so I can't take credit for the idea. But we used two PWM output
pins. The PWM provides more voltage range than is needed by the OXO's
EFC. To the output was scaled by a voltage divider. This also scaled down
thew
Wow. 350 nS. I had not been following this for the last few years. It
seems the ARM is a simply CPU with much more predictable interrupts timing.
One might ask way you'd need such good internal timing. What got me into
this years ago was scientific data acquisition. I wanted to time
43 matches