I know the the open load output of some instrument is 10Vpp and I think this 
is right because if  we want to connect this output on a 50Ohm line is correct 
to close the cable with the proper load impedance.
   I have found this level also on some trak System equipment.
   All my test are done using a 50 Ohm load in parallel to the HP and Trak 
outputs. I have tested two HP5065A and one HP5061B with the 1PPS option 
installed with the same spikes problem. The Trak , instead, work correctly, so 
I suspect some design problems on the HP dividers. I remember the same problem 
on the HP, and only HP, appear using an HP53132A TIC.
   To avoid any TAPR TICC input problem I am working on a small interface board 
with a 1M and 50Ohm input selection and a protection circuit for negative or 
non TTL signals.
   Luciano
   www.timeok.it


   Da "time-nuts" time-nuts-boun...@febo.com
   A "Discussion of precise time and frequency measurement" time-nuts@febo.com
   Cc
   Data Mon, 20 Mar 2017 09:48:24 -0700
   Oggetto Re: [time-nuts] ADEV query Timelab and TICC
   Luciano,

   This should not happen with the hp 5065A or 5061B frequency standards. I'm 
glad you worked around it by using a TAPR divider, but let's see if we can 
figure out the actual problem.

   One thing to know is that the 1PPS output level from the 5061 and 5065 is 
*HUGE*, even up to 10 volts. If you send this to most counters it will blow the 
inputs or cause other undesirable side-effects, like the bouncing and spikes 
that you speak of. So always check your output and input signal levels and 
waveforms using a 'scope. Do this in-circuit, with all cables attached. Use 
attenuators and termination as appropriate for your counter's input specs. Set 
your DC trigger level to best match the actual waveform seen by the counter 
(not the waveform sent by the frequency standard).

   Yes, the usual way you find out about this is that your ADEV measurements 
don't look right. The good news is that you can often tell within seconds that 
something is wrong. It's almost always a signal conditioning or trigger level 
issue, not a flaw in the instrument itself.

   The TAPR dividers tend not to have "this problem" because they output at 
wimpy TTL/CMOS levels.

   Older equipment can have powerful outputs. 10V into 50R is, what, 1/5th of 
an amp? Logarithmically, that puts a 5061A 1PPS closer to an automobile starter 
motor or heart defibrillator compared to modern logic gates.

   /tvb


   ----- Original Message -----
   From: "timeok" <tim...@timeok.it>
   To: <time-nuts@febo.com>
   Sent: Monday, March 20, 2017 12:56 AM
   Subject: Re: [time-nuts] ADEV query Timelab and TICC


   >
   > All,
   > the similar problem I have verified using the HP5065A and HP5061B 1PPS 
output, the dividers are pratically unusable for ADEV measurements. The 5/10MHz 
output of the same instruments using the TAPR divider are ok, so these dividers 
have some spike noise problems. It can be seen even using other TIC as The 
HP53132A.
   > Luciano
   > www.timeok.it
   >
   >
   > Da "time-nuts" time-nuts-boun...@febo.com
   > A "Discussion of precise time and frequency measurement" time-nuts@febo.com
   > Cc
   > Data Sun, 19 Mar 2017 20:03:29 -0700
   > Oggetto Re: [time-nuts] ADEV query Timelab and TICC
   > > I have sent a couple of files to Tom. They were taken simultaneously from
   > > an LTE Lite - one from the PPS and one from a PicDiv dividing the 10MHz 
to
   > > 1Hz. The glitches were on the PPS trace, but not on the PicDiv trace, so
   > > I'm fairly confident the TICC was working correctly.
   > >
   > > Orin.
   >
   > Hi Orin,
   >
   > Thanks for the raw data. It's very nice (2 hours 16 minutes = 8219 
points). Everything looks fine with the exception of 8 glitches. These are 
sometimes obvious jumps in phase, which cause massive spikes in frequency. Two 
plots attached.
   >
   > Almost every data point is within a few ns of each other. This is good. 
The standard deviation is a fraction of 1 ns. But once in a while there is a 
relatively massive phase jump. This is bad. Interestingly these 8 phase jumps 
all appear to be about 25 ns or a multiple of 25 ns in magnitude. The full list 
is (ns units):
   >
   > 24.575
   > 24.724
   > 24.831
   > 25.047
   > 25.087
   > 25.549
   > 25.589
   > 49.623
   >
   > 25 * N ns is not random. So I think this is not a Windows problem, not a 
USB problem, not a TimeLab problem, not a TICC problem either.
   >
   > It makes me wonder if this is a LTE-Lite problem. If Said or Keith from 
Jackson Labs is around -- is there anything on the LTE-Lite board that's close 
to 20 or 40 or 80 MHz? At this point I kind of trust Orin's data and I kind of 
trust the TICC. So when I see monster 25 ns phase jumps it makes me think 
there's a problem with the GSPDO board itself.
   >
   > (Please realize that only on time-nuts may we can use the words "monster" 
and "25 ns" in the same sentence; the rest of the world has larger problems)
   >
   > /tvb

   _______________________________________________
   time-nuts mailing list -- time-nuts@febo.com
   To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to